开云体育

Date

Re: HP 8007B Pulse Generator - Broken Attenuator Paddle - Fixed

 

开云体育

Tam,
In the spirit of OpenSCAD and open source, of course. What did you have in mind? You’re welcome to anything on the web page. Let me know if you need something more specific.

BTW, did you see my Elektor article “Simple Analog ESR Meter”? I printed the case with OpenSCAD.?

George?

image


On Aug 21, 2022, at 12:24 PM, Tam Hanna <tamhan@...> wrote:

?Hello,

thank you all so much. It was my honour that the OpenSCAD book was helpful.


Would you be willing to share a picture or two with me for Instagram?


Tam

With best regards
Tam HANNA

Enjoy electronics? Join 21k4 other followers by visiting the Crazy Electronics Lab at https://www.instagram.com/tam.hanna/

On 2022. 08. 21. 0:44, Paul Amaranth wrote:
That's great! ?Nice work, and thanks for publishing the openscad source.

Is this the same lever used in the 8640b?

On Sat, Aug 20, 2022 at 01:07:47PM -0700, Glydeck via groups.io wrote:
Gang,

Just shy of a year I can happily report that my HP 8007B is now back in service.? The big break through was acquiring a 3D printer.? A friend of mine had purchased a Creality Ender 3 during the pandemic, and after printing his fill of Dungeon and Dragon characters asked me if I wanted it. Of course I said yes. Armed with the Printer, Wilko’s sample lever and Tam’s book I was able to print the needed part. I’ve updated and will continue to update the details on this web page:

http://glydeck.com/hp_switch.html

Thanks again to Wilko and Tam for the help. Here is a photo of the 8007B back in action.








Re: HP 8007B Pulse Generator - Broken Attenuator Paddle - Fixed

 

Hello,

thank you all so much. It was my honour that the OpenSCAD book was helpful.


Would you be willing to share a picture or two with me for Instagram?


Tam

With best regards
Tam HANNA

Enjoy electronics? Join 21k4 other followers by visiting the Crazy Electronics Lab at

On 2022. 08. 21. 0:44, Paul Amaranth wrote:
That's great! Nice work, and thanks for publishing the openscad source.

Is this the same lever used in the 8640b?

On Sat, Aug 20, 2022 at 01:07:47PM -0700, Glydeck via groups.io wrote:
Gang,

Just shy of a year I can happily report that my HP 8007B is now back in service.? The big break through was acquiring a 3D printer.? A friend of mine had purchased a Creality Ender 3 during the pandemic, and after printing his fill of Dungeon and Dragon characters asked me if I wanted it. Of course I said yes. Armed with the Printer, Wilko’s sample lever and Tam’s book I was able to print the needed part. I’ve updated and will continue to update the details on this web page:



Thanks again to Wilko and Tam for the help. Here is a photo of the 8007B back in action.


Re: HP 8007B Pulse Generator - Broken Attenuator Paddle - Fixed

 

Wilco,
Agreed. I still need to update the dimensioned drawing with the measurements I took from the part you sent.

Paul,
Yes, it seems that the 8640B uses the same part. A subtle difference is that the lever on the 8640B has a thin black line.

George

On Aug 21, 2022, at 6:39 AM, Wilko Bulte <wkb@...> wrote:

?Nice to see that the part that I saved from the dumpster has passed on its genetic makeup ?

Wilko





Has anyone got a Boonton 160-A Q-meter manual?

 

I'm looking for a manual specifically for the above instrument. I've found manuls around that cover the 160-A anf 260-A in the one manual, and also 160-A and 170-A in the same manual, but I can't help but feel there must have been one specific for the 160-A.

Dr David Kirkby Ph.D
Email: drkirkby@... Web:
Kirkby Microwave Ltd (Tel 01621-680100 / +44 1621-680100)
Stokes Hall Lodge, Burnham Rd, Chelmsford, Essex, CM3 6DT.




Re: HP6268B repairability

 

Well a beast for sure. Can you lift it??Humor aside. Its a classical?linear supply design. If your comfortable with regulators and pass transistors then its very reasonable. The manuals online so take a look.
Though the manual I ran into was a poor copy and difficult to read. Maybe Artek has a better copy.
Will firmly believe the various transistors can be replaced with a wide range of surplus transistors.
Good luck.
Regards
Paul


Re: HP 8007B Pulse Generator - Broken Attenuator Paddle - Fixed

 

Nice to see that the part that I saved from the dumpster has passed on its genetic makeup ?

Wilko


Re: HP6268B repairability

 

I have repaired at least 4 of those beasts and one of them is currently holding my workbench in place in case of an earthquake...;-)

The most trouble-prone area IMO is the SCR-based pre-regulator. Just about every one of the ones I played with had an issue in that area. As I recall - and it's been about 15 years since I worked on one - the back end regulation is all analog. Other than the sheer mass of the unit, they were not all that hard to work on.

Have fun.

73,
geo - n4ua


Re: HP6268B repairability

 

Looking at the services manual, they are the standard phase control on the AC input for pre-regulation and linear output regulation. I have worked on a couple of supplies using this architecture through the years and the pre-regulator is not much fun to work on, but the output end is pretty straight forward. I don't remember it being a big problem, but it's been a while.

BobH


Re: HP 8591E phase noise question

 

Thank you John. Looks like I expected too much.
--

Regards,

Slawek/SP9BSL


Re: HP 8007B Pulse Generator - Broken Attenuator Paddle - Fixed

 

Paul,

It might be. I have a 8640B and can check.

George

On Aug 20, 2022, at 3:44 PM, Paul Amaranth <paul@...> wrote:

?That's great! Nice work, and thanks for publishing the openscad source.

Is this the same lever used in the 8640b?

On Sat, Aug 20, 2022 at 01:07:47PM -0700, Glydeck via groups.io wrote:
Gang,

Just shy of a year I can happily report that my HP 8007B is now back in service. The big break through was acquiring a 3D printer. A friend of mine had purchased a Creality Ender 3 during the pandemic, and after printing his fill of Dungeon and Dragon characters asked me if I wanted it. Of course I said yes. Armed with the Printer, Wilko’s sample lever and Tam’s book I was able to print the needed part. I’ve updated and will continue to update the details on this web page:



Thanks again to Wilko and Tam for the help. Here is a photo of the 8007B back in action.


--
Paul Amaranth, GCIH | Manchester MI, USA
Aurora Group of Michigan, LLC | Security, Systems & Software
paul@... | Unix/Linux - We don't do windows





Re: HP 8007B Pulse Generator - Broken Attenuator Paddle - Fixed

 

That's great! Nice work, and thanks for publishing the openscad source.

Is this the same lever used in the 8640b?

On Sat, Aug 20, 2022 at 01:07:47PM -0700, Glydeck via groups.io wrote:
Gang,

Just shy of a year I can happily report that my HP 8007B is now back in service.? The big break through was acquiring a 3D printer.? A friend of mine had purchased a Creality Ender 3 during the pandemic, and after printing his fill of Dungeon and Dragon characters asked me if I wanted it. Of course I said yes. Armed with the Printer, Wilko’s sample lever and Tam’s book I was able to print the needed part. I’ve updated and will continue to update the details on this web page:



Thanks again to Wilko and Tam for the help. Here is a photo of the 8007B back in action.
--
Paul Amaranth, GCIH | Manchester MI, USA
Aurora Group of Michigan, LLC | Security, Systems & Software
paul@... | Unix/Linux - We don't do windows


Re: Noise head ENR calibration - how do keysight do it? A question that was put to Keysight.

 

开云体育

Hi David:

Yes.? The HP 4395A when in Spectrum Analyzer mode (actually all modes) uses a DSP IF with I & Q detection which can do RMW, rather than peak detection that's used in analog spectrum analyzers.
This allows displaying noise directly in /root hertz.? The display icon is "/Hz" since there's no square root symbol in the character generator.

-- 
Have Fun,

Brooke Clarke

axioms:
1. The extent to which you can fix or improve something will be limited by how well you understand how it works.
2. Everybody, with no exceptions, holds false beliefs.


Re: HP 8007B Pulse Generator - Broken Attenuator Paddle - Fixed

 

Gang,

Just shy of a year I can happily report that my HP 8007B is now back in service.? The big break through was acquiring a 3D printer.? A friend of mine had purchased a Creality Ender 3 during the pandemic, and after printing his fill of Dungeon and Dragon characters asked me if I wanted it. Of course I said yes. Armed with the Printer, Wilko’s sample lever and Tam’s book I was able to print the needed part. I’ve updated and will continue to update the details on this web page:
?


Thanks again to Wilko and Tam for the help. Here is a photo of the 8007B back in action.


Re: looking for 5315A counter handle

 

On Saturday (08/20/2022 at 08:13AM -0700), Alex wrote:
You are right the HP3468A DMM shares the exact same handle/bail stand. Been looking for one of these handles myself for a good while now, as mine came broken with the DMM and missing the whole horizontal portion. I only ever saw one brand new handle being offered on ebay about 2-3 years ago, and I always regret not having taken it back then. Never seen any other one, even used, come up ever again since.
Yes. Sorry to hear it and now I am overwhelmed with guilt as I have
found exactly one of these handles from a 3468A DMM that a friend has,
which was brutalized internally in a prior life and not worth fixing
electrically. So, I get to have the handle.

When I am not using the handle, I will keep it in my safe, along with
my unobtainium and my pure liquid latinum.

Chris
--
Chris Elmquist


HP6268B repairability

 

I am being (remotely) offered a HP6268B for a somewhat decent price but its said to be in "unkown/untested" condition. Pysically and cosmetically it looks clean only minor usual blemishes. Wanted to ask if these power supplies are complex beasts that are difficult to repair, and/or the parts might not be always available. Also how reliable are they after all these years, are there any typical faults or certain details one needs to address or look out for? Assume this power supply uses switching technology as opposed to analog regulation, which in itself would certainly not make it any easier to work on one of these. I already have the service manual but haven't dug into it much yet, wanted to hear what others think of it before investing additional time into this.

Has anyone serviced or repaired one of these or a similar model that could give me his opinion? Thank you.


Re: looking for 5315A counter handle

 

You are right the HP3468A DMM shares the exact same handle/bail stand. Been looking for one of these handles myself for a good while now, as mine came broken with the DMM and missing the whole horizontal portion. I only ever saw one brand new handle being offered on ebay about 2-3 years ago, and I always regret not having taken it back then. Never seen any other one, even used, come up ever again since.


Re: Add option 002 to 3245A?...

 

On Sat, 20 Aug 2022 at 05:08, Radu Bogdan Dicher <vondicher@...> wrote:
Dave,
No connection with the 4284/4285s that I am aware of, but for the 3245A, Xdevs has a good set of documents for this (including CLIP, which is very complete and good res, and includes option 002), high res pics, and a very thorough overview of the unit.
Radu.?

If the circuit diagram is clear, and the components are available, it should be possible to make the board if you really want the option. If it's a highly desirable option, you might even be able to sell the boards, or if you feel in a good mood, make the files available so others can get a PCB from any PCB fabrication place.



Re: Noise head ENR calibration - how do keysight do it? A question that was put to Keysight.

 

On Fri, Aug 19, 2022 at 01:33 PM, <n8zmTWH@...> wrote:

?

?Why would you ignore the vector error correction? An accurate measure of the DUT gain and other s-parameters is essential to the proper calculation of the NF using the cold source method. Keysight claims that the uncertainty using the PNA-X’s cold source technique cuts the uncertainty compared to the Y-factor method to less than ?, mostly because the DUT gain is known more accurately.

?

Tom, N8ZM

I wrote “ignoring the vector error correction, you have not gained anything” because I was trying to keep it simple and say that the error correction doesn’t change the need to have something calibrated.


Re: Noise head ENR calibration - how do keysight do it? A question that was put to Keysight.

 

开云体育

Thanks, Charles -- I deeply appreciate your finding those references/that reference. I note that the author himself disavows the method in the comments section after his mwrf article. As he notes, there's nothing wrong with the derivation per se, but the devil, as always, is in the details.

-- Cheers,
Tom
-- 
Prof. Thomas H. Lee
Allen Ctr., Rm. 205
350 Jane Stanford Way
Stanford University
Stanford, CA 94305-4070
On 8/19/2022 19:28, Charles Edmonds wrote:

Hi,

the article was in Microwaves & RF in April 2013. There is some discussion about its accuracy but I'm uncertain about the actual validity of the suggestion. The article can be found in a few places:




regards

Charles Edmonds VK3CLE
Mob: 0418309921
charles.edmonds@...


Re: Noise head ENR calibration - how do keysight do it? A question that was put to Keysight.

 

开云体育

Hi,

the article was in Microwaves & RF in April 2013. There is some discussion about its accuracy but I'm uncertain about the actual validity of the suggestion. The article can be found in a few places:

https://www.mwrf.com/technologies/test-measurement-analyzers/article/21845669/measure-noise-without-a-calibrated-source

https://www.researchgate.net/publication/295615411_Measure_Noise_Without_A_Calibrated_Source


regards

Charles Edmonds VK3CLE
Mob: 0418309921
charles.edmonds@...




------ Original Message ------
From: "Conrad, PA5Y" <g0ruz@...>
To: "[email protected]" <[email protected]>
Sent: Saturday, 20 Aug, 2022 At 7:01 AM
Subject: Re: [HP-Agilent-Keysight-equipment] Noise head ENR calibration - how do keysight do it? A question that was put to Keysight.

Having re-read the David Stockton presentation again, that along with the fact that these days we have high IP3 MMICs with very low noise figures and good return loss, it makes sense to look more ?closely at these devices. Tuning for minimum on the bench is much less risky and negative noise figures should be a thing of the past.

?

I can see that my winter will be a little less boring.

?

Interesting discussion, thanks everyone. The expensive box that Keysight sell will be rather like the R&S FSWP, nothing more than a dream.

?

Regards

?

Conrad PA5Y

?

From: [email protected] <[email protected]> On Behalf Of n8zmTWH via groups.io
Sent: 19 August 2022 21:14
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] Noise head ENR calibration - how do keysight do it? A question that was put to Keysight.

?

Hi Conrad…

?

Actually, even the expensive box that Keysight sells to do the cal boils down to just comparing the unknown source to one that is traceable,? and simply adding the ?difference to the known ENR. So you are on the right track. As you know, the problem at low NF’s is quite simply that the magnitude of the accumulated uncertainties can become equal to or greater than the value of the NF.

?

You moonbounce guys also already know that an impedance mismatch between the antenna and the preamp input can really offset the benefits of a preamp tuned to perfection ?on the bench. Stockton hinted at this in his paper.

?

Tom, N8ZM

?

From: [email protected] <[email protected]> On Behalf Of Conrad, PA5Y
Sent: Friday, August 19, 2022 12:50 PM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] Noise head ENR calibration - how do keysight do it? A question that was put to Keysight.

?

Hello Tom, yes, I already read all this material (again) as well as the SM5BSZ stuff. As usual Leif’s pages are well worth a read. I will just compare with the calibrated Noise Com NC-346A that I have. It is good enough. The stuff that ?I do is sub-0.4dB which as you know is very tricky. Above 432 system analysis with celestial sources seems a whole lot more reliable. I am fortunate to live in a quiet location.

?

Anyway, it was good to refresh my memory. All brought about due to the acquisition of an HP-8790B and to help a friend with his home brew noise source.

?

Regards


Conrad PA5Y

?

From: [email protected] <[email protected]> On Behalf Of n8zmTWH via groups.io
Sent: 19 August 2022 17:21
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] Noise head ENR calibration - how do keysight do it? A question that was put to Keysight.

?

This link will take you to a paper by David Stockton, GM4ZNX, the architect of the HP 8973 series Noise Figure analyzers, and and Ian White GM3SEK . In it, they ?discuss the errors in NF measurements, including ENR calibration.

?

?

The title of the paper to look for is:

?

GM4ZNX-GM3SEK_Noise-Figure-Measurment_EME2012-Long.pdf

?

There is also a procedure for NF measurement using a network analyzer that does not require knowing the ENR, as the VNA determines the DUT gain, thus eliminating ?the need for the Y-factor. It can be a bit more tedious, and there is, as always, a devil in the details, but it is very accurate. Dr. Dunsmore covers it in his microwave measurements handbook.

?

?

Tom, N8ZM

?

From: [email protected] <[email protected]> On Behalf Of Tom Lee
Sent: Friday, August 19, 2022 7:17 AM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] Noise head ENR calibration - how do keysight do it? A question that was put to Keysight.

?

Hi Charles,

Can you point us to where you saw this claim? Like David Kirby, I'm skeptical, but I'd like to see the details of the argument. It is true that, since noise factor is dimensionless, a purely ratiometric measurement process is in principle possible. But details ?matter.

-- Cheers,
Tom

-- 
Prof. Thomas H. Lee
Allen Ctr., Rm. 205
350 Jane Stanford Way
Stanford University
Stanford, CA 94305-4070

On 8/19/2022 01:33, charles.edmonds@... wrote:

Hi, about 18 months ago I put the below on a Keysight forum expecting some helpful replies. Sadly I have not had a single reply as yet. (The concept of doing the measurement without the ENR ?being known is to compare derivatives of things not absolute values. This is the point where my mathematical cabilities implode).

Measuring the ENR of a Noise Source

Hi, I have read that it is possible to accurately measure the noise figure of a device without having a calibrated ?noise source. This is done by including an extra amplifier and an attenuator in the process. If the amplifier gain and the attenuator loss are very accurately known then knowing the exact ENR of the noise source is not necessary. The?device noise figure can?be?then ?determined as accurately as doing the normal Y-factor method with a calibrated noise source.

?

If this is correct then surely the next step would be to do the normal Y-factor process and change the ENR of the ?un-calibrated noise source in the calculation until the result given is the same as in the above procedure. So now the ENR of the un-calibrated noise source is known.

?

So how accurate could this result be? Could it be claimed that the un-calibrated noise source is now calibrated or ?only measured?

?

Would it be possible to create a spreadsheet to calculate all this using only the noise marker readings from something ?like an old 859xE or 856xE spectrum analyzer? (As an aside I don't think this would work using an E440xB spectrum analyzer would it?)?

?

My thought is that a hobbyist could buy 3 of the small amplifier modules seen on auction sites, the best being the ?ones that use the SBB5089 MMIC, as it have the right gain and a fairly flat frequency response. Also needed would be a termination and 2 attenuators. On one of the amplifiers the termination would be placed on the input and one of the attenuators placed on ?the output. This would be the un-calibrated noise source. The other preamplifiers would be the one normally used on the front end of the spectrum analyzer for a Y-factor method? measurement and the last is the DUT to be measured.

?

? ? ? ?