开云体育

Date

Re: Anybody have schematics or link for schematics of HP 3488A switch control unit?

 

Yes and thanks guys, the 3488A is a very robust instrument - virtually indestructible.? Never was an I/O card ever sent back for warranty repair or for out of specification replacement.? Those Aromat relays were well qualified and life tested with plenty of margin.? Nothing in the specs was ever cut close -- everything I speced was for 3-sigma uncertainty and guaranteed to work in a class B environment.? Never did a card ever fail final test for being out of spec.---? the design of the circuits will work with components at the extreme ends of the selection bin of typical manufacturing variations.

Not sure how many main frames were finally produced but probably in the 50K range.? They were very popular for use in manufacturing test systems.? We were told the big users like Motorola would describe their test systems by how many 3488a frames would be required.? We talked about making an expander to give users more I/O slots but told by the users it was not necessary because the units were cheap enough as is and they would just purchase multiple units if needing more slots so an expander was never seriously considered.? It is indeed amusing the industry rated test systems by how many 3488As would be required.

It has now been replaced by the 3499 and 34970A data acquisition unit.
George? ? ? ? ? ? ? ? April 19. 2019? 11:33pm


Re: My $25 "In Poor Shape" 3468A has arrived

 

Alex,
Those Vishay resistors from China are the real deal.? They are pulls and affordable.? That is if they are from the company I have purchased them from previously.? ?A year ago I purchased qty.75? ?10K and 40K S102C resistors from a Chinese source and they we original Vishay.? If you want to purchase state side look to Digi-Key.? The? S102 series is the way to go.? Resistors you are looking at with 10ppm/C are not much better than the Dale resistor I designed into the 3468/78 meters.




Your comment:???the service manual seems to be very critical about which mode to select for calibration and recommends to to do so in the ohms mode that the instrument will be used (...the most). So it sounds like one would always have to make a small compromise, and having one of each basically sounded like being able to eliminate that compromise. Or not really.

My current comments supersede the manual for I wrote those sections of the manual -? LOL

Theoretically the offset between 2-W and 4-W cals is as I pointed out in my last post is 41mOhm when shorting the inputs but more realistically 60-90mOhm with the presence of thermal offsets and contact resistances when ZERO caling the instruments with the shorts applied to the inputs.? It makes no sense to cal in 2-W ohm with a short at the input.? You want to cal in 2-W mode with your test leads attached and shorted at the probe tips.

You will not find a source for the RED Devil resistors for they were never marketed to the industry.? They are proprietary components that were made at HP Components in Loveland Colorado for use as attenuator and gain setting resistors in the HP instruments.? This was before Vishay came along with a better product.? The RED Devils are precision wire wound resistors with low TC characteristics and tolerances.? You will find them in all the early high performance multimeters from the 60s and 70s.? Even the 3456A has some of them.


ANSWER:
And on the other hand there are no 190M ohm resistors available at all, the closest standard value seems to be 187M ohm. Closest possible exact match for a series configuration would be to use four 47.5M ohm, but they are 100ppm rated, so not sure it would be useful long term beyond just finding out if the ranges would fall in compliance after connecting this resistor combination in parallel with R201.

The value is not critical for testing as long as the shunt resistor in somewhere near 200 Meg:? 175 - 225 Meg would do the trick.

I will try to get the first installment of the HP3468A/78A design series out this weekend.
George




Re: Service Manual for HP 3708A

 

Hi Gerald, I sent you an e-mail from my private address, to get news on the manual, but I didn't get an answer. I don't know if they arrived.

Dario IU3ARP


Re: My $25 "In Poor Shape" 3468A has arrived

 

Comparison readings between 3478A and 3468A
-----------------------------------------------------------------

Used my little resistor standard box I build around a resistor array 1k - 10M 0.1% and a couple of extra separate 0.1% resistors for the 10 and 100 ohm positions. All readings taken in 4W ohms mode and with the chinese kelvin cables, at 28 C ambient temp.

Resistor?????? 3478A??? ??????? 3468A
--------------------------------------------------
10 ohm??? ?? ? 10.013x??? ???? 010.030
100 ohm??? ?? 100.037??? ??? 100.028-9
1.0001K??? ??? 0.99997K????? 0.99959K
10.01K??? ????? 10.0099K????? 10.0070K
101.01K??? ??? 101.010K????? 100.968K
1M??? ???????????? 1.00093-4M? 1.00065-6M?
10M??? ?? ? ? ? ? 9.97xxM??? ???? 09.99xxM

The "x" denotes digit(s) with unstable continuously fluctuating reading.

Results from the Red Devil Resistors
-------------------------------------------------
20K?????????????? 19.9721K?????? 19.9662K
98.975K??????? 99.022K???????? 98.980k

Far from perfect, but not that bad.
Wish it would be as easy to get a stable multi-voltage standard to be able to check all those VDC ranges as well.


DSO-X 3034A power supply pins

 

I am looking for information to trouble shoot scope that just quit?

there is no help or drawing in service manual?


Re: Anybody have schematics or link for schematics of HP 3488A switch control unit?

 

开云体育

Hi Tim,

?

??????????????? I used two 3488A units for several years, this was a very reliable instrument and very useful. You and your colleagues did an great job with this 3488A. Congratulations !

?

Regards

Yves

De?: [email protected] <[email protected]> De la part de ghnatiuk@...
贰苍惫辞测é?: 19 avril 2019 09:44
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] Anybody have schematics or link for schematics of HP 3488A switch control unit?

?

Tim,
Tell me what it is you are looking for and I can help you.? I was the analog design engineer on the HP3488A.
I designed the 444XX series of I/O cards and have a large supply of all the parts for the mainframe and 444XX series of cards.
That project began in 1981 and completed in 1983.? It was one of the highest selling products the HP Loveland Division ever produced.? ?Tons of them out there these days for very little money.??

I see there is interest in the power supply.? It was not an OEM supply.? The power supply was designed in-house by one of the other engineers on the team by the name of Tom Rodine.? There were 6 of us all together on this project:
Project manager
Analog Design I/O and main frame interface
Firmware / software design
Digital design
Mechanical design
Power supply design

George? Hnatiuk? ?April 19, 2019? 8:32am


Re: Anybody have schematics or link for schematics of HP 3488A switch control unit?

 

开云体育

Brilliant George,

Sounds good and if you have a small amount of time during the visit we can potentially meet up. I am only in Stevenage/Luton area (approx 30 miles north of London... simples.

Understood about the back connectors as any unit taken from a rack either had the rear panel plate removed or the cable cut...

Useful to know about the 3421A... i didn't know that.
Regards
Nigel


Sent from Samsung by banging a nail into a piece of wood.

-------- Original message --------
From: ghnatiuk@...
Date:19/04/2019 19:01 (GMT+00:00)
Subject: Re: [HP-Agilent-Keysight-equipment] Anybody have schematics or link for schematics of HP 3488A switch control unit?

Nigel,
I will see what I can do.? The same connector blocks were used on the HP3421A portable data acquisition unit so look out for those.? Unfortunately when you purchase cards, the connector blocks and housing are missing.? Best to look for completed units with cards stuffed into them that have the external cables cut at the blocks.

There is a chance I will be in the UK within the next 20 days on business.? If that goes thru and I find spares I will bring them along.? Will be flying into Heathrow.

Some people earlier were looking for floppy drives for the 9122/9133 series of disc drives.? If there is a need in the UK, I could drag some of those along with me.? ?Let me know and we can work out a price for the units.

George


Re: Anybody have schematics or link for schematics of HP 3488A switch control unit?

 

Nigel,
I will see what I can do.? The same connector blocks were used on the HP3421A portable data acquisition unit so look out for those.? Unfortunately when you purchase cards, the connector blocks and housing are missing.? Best to look for completed units with cards stuffed into them that have the external cables cut at the blocks.

There is a chance I will be in the UK within the next 20 days on business.? If that goes thru and I find spares I will bring them along.? Will be flying into Heathrow.

Some people earlier were looking for floppy drives for the 9122/9133 series of disc drives.? If there is a need in the UK, I could drag some of those along with me.? ?Let me know and we can work out a price for the units.

George


Re: My $25 "In Poor Shape" 3468A has arrived

 

Hi George,

Very nice explanations. I will follow your advice and calibrate the 3478 as well on 4W. Its just that I was a bit less trusting about its calibration than the 3468, so my reasoning went to use 4W on what I perceived as being the better of the two. But as you point out the '78 having the the lower ohms range, it definitively would make sense to use it for low value 4W measurements. Although I have to say that with the zero cal on the 3468 and the chinese kelvin cable I was for the first time able to read and verify the value of some 0.01 resistors I had purchased some time ago to use as shunts in a power supply project but never used them. They are more like a U shaped piece of wire, and so far all of them. And same goes for basically any other low value resistor in my parts stock that I threw at the 68, it measured up quite well within the expected results.

I really need to compare resistor readings between the 68 and 78, which I did once, but then also this time write down the values for comparison and post them here just for the record. I will recalibrate the 78 in 4W and do a comparison of readings.

My reasoning to have one DVM on 2W and the other on 4W stems from the fact that the service manual seems to be very critical about which mode to select for calibration and recommends to to do so in the ohms mode that the instrument will be used (...the most). So it sounds like one would always have to make a small compromise, and having one of each basically sounded like being able to eliminate that compromise. Or not really.

For the time being and following your suggestion, I will not attempt to calibrate the 30k/300k ohms ranges by using the Red Devils. No need for absolute accuracy.
In regards to your suggestion about R201 did some research at Mouser, my preferred supplier:

To substitute R201 outright there are no 40.0K series S102 resistors available for that exact value. Closest I could find is Vishay PTF series 40K 1/8W 0.05% 10PPM/C at $3.38 each, 58 currently in stock. Would that be still an improvement over the existing resistor?

https://www.mouser.com/ProductDetail/Vishay-Dale/PTF5640K000AYEK?qs=sGAEpiMZZMsPqMdJzcrNwo3vE%252BHgilrQScfmxiejoUM%3D

Second best options would be another Vishay PTF series 40K 1/2W 0.1% 10ppm/C. One does cost $4.99 and the other only $0.89. Frankly, with a side by side comparison, can't seem to find any differences between these? two other than the selling price (both prices are for qty one unit purchase).

71-PTF5640K000BYEK
https://www.mouser.com/ProductDetail/?qs=41svs21nsHBzvOZxn1wkhQ%3D%3D

71-PTF5640K000BYEB


And there is actually a Chinese seller on ebay offering used Vishay S102K at $5, 40K 0.01% which as pictured are obvious pulls. But given ebay's history with remarked fake parts mainly from China I doubt this could be trusted to be the real deal, and even if it where given aging and accumulated drift, etc would probably preclude this as an option in this case. If this where the only option, figure probably better off staying with my current R201.
https://www.ebay.com/itm/122674485353

And on the other hand there are no 190M ohm resistors available at all, the closest standard value seems to be 187M ohm. Closest possible exact match for a series configuration would be to use four 47.5M ohm, but they are 100ppm rated, so not sure it would be useful long term beyond just finding out if the ranges would fall in compliance after connecting this resistor combination in parallel with R201.

As many others, looking forward to your video, and when the time comes hoping for some more of those Red Devil resistors that you might want to sell for use as standards. BTW did a G search on them and closest results is for Ohmite "Brown Devil" resistors, a whole different beast. Basically zip on the Red ones.

Thanks for all the Hi-Q information.
Alex


Re: Existing HP 70K, HP 7000 thread on EEVBlog, was .... Keypad for 70820A module,

 

For HP70K, ?HP7000 users:

There is also a thread for HP70K technical & repair issues, ?on the EEVBlog with photos, schematics & interleaved text.



?


Re: Anybody have schematics or link for schematics of HP 3488A switch control unit?

 

Seconded! I have two, and love 'em. They are wonderful units, damn
handy. Nice work George and team.

-Dave

On 4/19/19 12:47 PM, nigel adams via Groups.Io wrote:
Hi George,

Congratulations on a nicely executed design. I love it... simple to use
and straightforward to fix.

As i said,i have two chassis with various switch or i/o cards in each.

What i don't have for 3 of the relay cards is the rear plate and
connector plus its associated cable clamp. Do you have any of these?
Mainly required for the 44470A relay cards. If you have any to spare and
can ship to UK for not too expensive the i could be interested.

Great bits of kit though, well done.
Regards
Nigel


Sent from Samsung by banging a nail into a piece of wood.

-------- Original message --------
From: ghnatiuk@...
Date:19/04/2019 14:44 (GMT+00:00)
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] Anybody have schematics or
link for schematics of HP 3488A switch control unit?

Tim,
Tell me what it is you are looking for and I can help you.? I was the
analog design engineer on the HP3488A.
I designed the 444XX series of I/O cards and have a large supply of all
the parts for the mainframe and 444XX series of cards.
That project began in 1981 and completed in 1983.? It was one of the
highest selling products the HP Loveland Division ever produced.? ?Tons
of them out there these days for very little money.??

I see there is interest in the power supply.? It was not an OEM supply.?
The power supply was designed in-house by one of the other engineers on
the team by the name of Tom Rodine.? There were 6 of us all together on
this project:
Project manager
Analog Design I/O and main frame interface
Firmware / software design
Digital design
Mechanical design
Power supply design

George? Hnatiuk? ?April 19, 2019? 8:32am
--
Dave McGuire, AK4HZ
New Kensington, PA


Re: HP 53310A POST errors

 

开云体育

Hello,

now I am awake again.


IMHO; the next test we must do is to check if the CPU card is farting out a video waveform. If you see the "marching ants pattern" on the LEDs of the card, it is very likely that at least the CPU is alive.


Tam

On 2019. 04. 19. 7:17, Jack2015 wrote:
After power on,all 8 leds were on for less than one second,then 6,then 4,then all leds were off (completed power on self test)
after a little while the leds flashed one after one in sequence.(each time only one led was on).

I kept front panel faced up and powered it on for couple minutes,then?powered it off and on again
while holding one button,for couple times,but nothing changed.

Jack
-- 
With best regards
Tam HANNA 

Enjoy electronics? Join 15k7 other followers by visiting the Crazy Electronics Lab at  


Re: Anybody have schematics or link for schematics of HP 3488A switch control unit?

 

开云体育

Hi George,

Congratulations on a nicely executed design. I love it... simple to use and straightforward to fix.

As i said,i have two chassis with various switch or i/o cards in each.

What i don't have for 3 of the relay cards is the rear plate and connector plus its associated cable clamp. Do you have any of these? Mainly required for the 44470A relay cards. If you have any to spare and can ship to UK for not too expensive the i could be interested.

Great bits of kit though, well done.
Regards
Nigel


Sent from Samsung by banging a nail into a piece of wood.

-------- Original message --------
From: ghnatiuk@...
Date:19/04/2019 14:44 (GMT+00:00)
Subject: Re: [HP-Agilent-Keysight-equipment] Anybody have schematics or link for schematics of HP 3488A switch control unit?

Tim,
Tell me what it is you are looking for and I can help you.? I was the analog design engineer on the HP3488A.
I designed the 444XX series of I/O cards and have a large supply of all the parts for the mainframe and 444XX series of cards.
That project began in 1981 and completed in 1983.? It was one of the highest selling products the HP Loveland Division ever produced.? ?Tons of them out there these days for very little money.??

I see there is interest in the power supply.? It was not an OEM supply.? The power supply was designed in-house by one of the other engineers on the team by the name of Tom Rodine.? There were 6 of us all together on this project:
Project manager
Analog Design I/O and main frame interface
Firmware / software design
Digital design
Mechanical design
Power supply design

George? Hnatiuk? ?April 19, 2019? 8:32am


Re: hp8753B_Phase_Lock_Lost

 

开云体育

Are you sure it is the YTO?


Peter

On Apr 19, 2019, at 12:11 PM, Armin Roesch <armin.roesch@...> wrote:

Newly this error appears.
Does somebody has a reliable source for spare YIG.
In may vna is a 5086-7718 mounted.
Armin
HB9MFL


hp8753B_Phase_Lock_Lost

 

Newly this error appears.
Does somebody has a reliable source for spare YIG.
In may vna is a 5086-7718 mounted.
Armin
HB9MFL


Re: Anybody have schematics or link for schematics of HP 3488A switch control unit?

 

Tim,
Tell me what it is you are looking for and I can help you.? I was the analog design engineer on the HP3488A.
I designed the 444XX series of I/O cards and have a large supply of all the parts for the mainframe and 444XX series of cards.
That project began in 1981 and completed in 1983.? It was one of the highest selling products the HP Loveland Division ever produced.? ?Tons of them out there these days for very little money.??

I see there is interest in the power supply.? It was not an OEM supply.? The power supply was designed in-house by one of the other engineers on the team by the name of Tom Rodine.? There were 6 of us all together on this project:
Project manager
Analog Design I/O and main frame interface
Firmware / software design
Digital design
Mechanical design
Power supply design

George? Hnatiuk? ?April 19, 2019? 8:32am


Re: Anybody have schematics or link for schematics of HP 3488A switch control unit?

 

开云体育

George,

Nice work. I still have a few, one of my sons used a couple for a High School project with a 3456A to plot the discharge curves of several different types of cells.?

These are a great bargain right now.?

Peter

On Apr 19, 2019, at 9:32 AM, "ghnatiuk@..." <ghnatiuk@...> wrote:

Tim,
Tell me what it is you are looking for and I can help you.? I was the analog design engineer on the HP3488A.
I designed the 444XX series of I/O cards and have a large supply of all the parts for the mainframe and 444XX series of cards.
That project began in 1981 and completed in 1983.? It was one of the highest selling products the HP Loveland Division ever produced.? ?Tons of them out there these days for very little money.??
\
George? Hnatiuk? ?April 19, 2019? 8:32am


Re: Anybody have schematics or link for schematics of HP 3488A switch control unit?

 

开云体育

The psu is a module which is likely to have been a 3rd party module so schematics?
may not be available.

Regards
Nigel

Sent from Samsung by banging a nail into a piece of wood.

-------- Original message --------
From: Tim hughes
Date:19/04/2019 03:28 (GMT+00:00)
To: HP Agilent Equipment
Subject: [HP-Agilent-Keysight-equipment] Anybody have schematics or link for schematics of HP 3488A switch control unit?

Keysight website only has operator manuals and similarly BAMA has operator/programming manuals like 03488-90002 or 3 etc.

The manual numbered? 03488-90009 is supposed to be the repair manual. But I could not find a free download.??
From the title it sounds more like a block level repair manual. But it does have some sort of schematics.?

I am only really interested in schematics particularly the power supplies.

Any help appreciated.
Tim Hughes


Re: Anybody have schematics or link for schematics of HP 3488A switch control unit?

 

开云体育

I have two of these units and downloaded the manuals direct from the Keysight site.

This includes the schematics for the main unit and the various switch cards.

These are useful bits of kit
Good luck with yours.
Regards
Nigel


Sent from Samsung by banging a nail into a piece of wood.

-------- Original message --------
From: Jos Raven
Date:19/04/2019 10:20 (GMT+00:00)
Subject: Re: [HP-Agilent-Keysight-equipment] Anybody have schematics or link for schematics of HP 3488A switch control unit?

Hi Tim,
I have this document ; HP-3488A-90012 3488A Service manual
I downloaded that from the Agilent website .
I do not know if it is still available there.

Look for it and if you do not find it I am happy to upload it to KO4BB.

Regards,
Jos Raven PA0AMX


Re: Anybody have schematics or link for schematics of HP 3488A switch control unit?

Tim hughes
 

A number of now have people sent me copies of 90012nservice manual,? thanks for? all replies on and off list!

The manual is actually on the Agilent keysight website, but it does no come up if you do an official search on the sight. It appears they took the link down, but it is still on the site and the link can be found via a general google seach.

This was from Nigel GM8PZR :
>>>>>>
Keysight does have the service manual, the one you need is 03488-90012, but for some reason it doesn't seem to show in their document list for the product, although a google search on "HP3488A service manual" does find it.
Another link :
https://elektrotanya.com/hp_agilent-technologies_3488a_switch_control_unit.pdf/download.htm
<<<<<

Thanks again.
Tim


On Friday, April 19, 2019, 2:20:23 AM PDT, Jos Raven <jos.raven@...> wrote:


Hi Tim,
I have this document ; HP-3488A-90012 3488A Service manual
I downloaded that from the Agilent website .
I do not know if it is still available there.

Look for it and if you do not find it I am happy to upload it to KO4BB.

Regards,
Jos Raven PA0AMX