Re: E5052A Power On Test (3.3V Bus Supply)


 

Great!? That's the best kind of repair; when you learn something in addition to getting it working again.

Jim Ford
Laguna Hills, California, USA

On Saturday, March 1, 2025 at 05:43:25 AM CST, jmr via groups.io <jmrhzu@...> wrote:


It's all OK now, I took it apart and fixed it :) :) :)
?
On the DSP/ADC module A3 the 5V rail is now 4.97V (was 4.70V typical) and the 3V3 rail is now 3.24V (was 2.97V and very drifty).
?
I disassembled the main firmware and worked out how it does the power on testing of the power supplies and this gave me the pass/fail limits for all the supplies.?
I was able to trace through the power supply distribution and see where it was losing voltage in a poor connection. A thermal camera helped a lot here. It's now all fixed and fine again!
?
I also found out a few things about how it manages the internal 10 MHz reference and how to adjust the high stability 10 MHz OCXO. So that is now calibrated.
?
I am so pleased, I thought it was slowly dying on me. There's no way I could afford to send it to Keysight for repair.
?
I should be able to fix the other E5052A as well as it has a similar issue with a sagging 5V and a sagging 3V3 supply.
?
?
?
?