Marco,

?

Our calibration constants are really similar, which thickens the mystery.

With these results, we will have no choice but to measure at different locations and compare our measurements.

Let's take 2 easy points, AUX IF OUTPUT and AUX VIDEO OUTPUT on the rear panel.

?

There should be no problem with that, but check if AMP CORR is OFF.

CAL

MORE

MORE

MORE (4/4)

AMP CORR

AMP CORR OFF

--------------

- PRESET

- Put the 300 MHz REF on the 8591E connector

- SPAN 1 MHz

- ZERO SPAN

?

The measurement on another Spectrum (or tinySA) at 21.4MHz on the AUX IF OUTPUT connector gives me about -51dBm, with a voltmeter on the AUX VIDEO OUTPUT output I have about - 300mV.

?

If these two tests are OK, then we will have to measure near the ADC. I will wait for your results.

?

?

Here comparison between the calibration¡¯s constants.

?

?

??

?

?

Test AUX IF and VIDEO output

Join [email protected] to automatically receive all group messages.