Re: 8594E CAL: SPAN SENS FAIL


 

On Tue, Apr 30, 2024 at 10:17 AM, Scott Townley wrote:
I have an 8594E spectrum analyzer that gives CAL: SPAN SENS FAIL when self-calibrating.
I went through the internal service-diagnostic routines and?
passes:? 2V reference detector, ground reference detector
fails:? main coil driver, FM coil driver, sweep ramp, 10V reference detector, -10V reference detector

The service guide points to the A7 Analog Interface for all of these failures.? So...simple enough, I just need to swap A7 boards.
And I did, and nothing has changed...exact same fails.

So either I got an A7 with the exact same failures as my original, or something else is actually wrong.
Any suggestions would be greatly appreciated!
===
Are all supply voltages correct?

+10V/-10V diagnostics are very basic tests, it is easier to debug them which hopefully shows the root cause. +10V reference starts from U25 pin6 and its buffered version shows up at emitter of Q150 also J-22 on A7 board. If you don't see +10V at that location, either +15V/-15V supplies are missing or you have a board with a faulty reference section like the original one (unlikely).

Assuming you see +10V reference at emitter of Q150, this voltage goes through test muxes U15A->U17->right side of R17 (J1-31) with 1/10 attenuation and shows up as 1V during +10V diagnostic routine. If the muxes don't pick the right input there could be an issue with address/data bus in the connector of A7.

If you see +1V (attenuated 10Vref) at J1-31 during +10V test, and -1V at -10V test next possibility is ANA_TEST input of U64 in processor board is damaged (or ANA_TEST trace is damaged, worth checking the connectivity). 2V reference is passing, which is another input to the same mux so rest of the chain looks fine. If you have another processor board (08590-60365 ? ) you could swap and check. You can also look at output of U64 (or U45 with x1.68 gain) in 10V and 2V reference routines.?

I don't know if the test voltages stay static of pulsed during diagnostic tests, if DVM doesn't show expected voltages at mux outputs you may need an oscilloscope.
Ozan
?