Re: Very poor thru (barrel) supplied with NanoVNAs


 

On Thu, Oct 10, 2019 at 02:55 PM, <qrp.ddc@...> wrote:


On Thu, Oct 10, 2019 at 03:17 AM, Kurt Poulsen wrote:


Wrong, the cables are part of the NanoVNA hardware and cancelled out during
calibration.
That's Wrong. You cannot cancel bad cable effect with calibration. You can
only reduce it for a little.

For example, here is calibration+measurement performed with rg174 32 cm (which
come with NanoVNA) and with rg405 10 cm. See attachment.

The test cases are following:
1) perform calibration with rg174
2) measure rg174

1) perform calibration with rg405
2) measure rg405

If it were possible to eliminate bad cable effect (as you mentioned), both
measurement will be exactly the same. But they are different, and rg174 are
much worse.
You should calibrate at the end of the cable and the difference in the picture measurements you show is caused by the cable length, not the cable quality.