Re: 59.5keV spectrum cleanup


 

Hi Charles, close but no Am involved. The exciter is natural U-238, we are seeing the decay of the daughter Pa-234m (also known as UX2). We are essentially seeing only the birth of the?U-234daughter from? Pa-234m, just that part of the natural U-238 decays series.
When depleted U has been chemically separated from its daughters you wind up with only U-238, the quickly the ingrowth of the first 3 daughters Th-234>Pa-234m>U234.? U-234 has a 240,000 year half-life, so lower daughters will eventually grow back in, but in geological time scales.?

The disc is actually marked "UX2 beta, and is source of Pa-234m beta particles, but we're looking deeper than that.

A picture of that decay series and its STOP sign is in an earlier post.

Geo

----- Original Message -----
From: Charles David Young <charlesdavidyoung@...>
To: XRF <[email protected]>
Sent: Fri, 30 Oct 2020 22:27:34 -0400 (EDT)
Subject: Re: [XRF] 59.5keV spectrum cleanup

I am not sure exactly what I am looking at here but I like the sharpness of the peaks and the low noise.? Is this a scan with an Am241 exciter?? See attached.

Charles

----- Original Message -----
From: Charles David Young <charlesdavidyoung@...>
To: XRF <[email protected]>
Sent: Fri, 30 Oct 2020 22:27:34 -0400 (EDT)
Subject: Re: [XRF] 59.5keV spectrum cleanup

I am not sure exactly what I am looking at here but I like the sharpness of the peaks and the low noise.? Is this a scan with an Am241 exciter?? See attached.

Charles

On Fri, Oct 30, 2020 at 5:36 PM <GEOelectronics@...> wrote:

25.6 like last run, but auto threshold adjust applied. 8500s update .mca below. Didn't help?

"noise" at low end, did noticeably improve definition of peaks above 3 keV.

3.3 yes it is Ka1 of Potassium. But let's think about element K. for a moment, a portion of it is K-40 and that has something like 10% decay path to Ar gas. K-40 is loaded with 3.19 2.98 and 3.19 keVs. Then add the stable Ar in the air which is really easily ionized, you might be getting a huge dose of Ar XRF because of? the radioactive source.

Also the "noise" at the low end- I'm wondering if it has more to do with the X-Ray photon's photopeak being scattered? by O2, and other air molecules.?

Geo

? K at 3.3. noise at 4.5?? Cr at 5.42 and U Ll at 11.64






Join [email protected] to automatically receive all group messages.