Re: 59.5keV spectrum cleanup


 

That run ended, reset the auto Threshold control, restarted, will report back over next 4 hours, already looks better tho.

Geo


From: GEOelectronics@...
To: [email protected]
Sent: Friday, October 30, 2020 4:58:38 PM
Subject: Re: [XRF] 59.5keV spectrum cleanup

Click on LINEAR view, looks pretty good to me, the L X-rays are the subject, but you are right I didn't do the threshold routine. Do you think 25.6 is too slow? I'm afraid if something with high count rates comes along it will not be the best, maybe 19 was a better compromise ?

Geo

----- Original Message -----
From: Dude <dfemer@...>
To: [email protected]
Sent: Fri, 30 Oct 2020 17:18:57 -0400 (EDT)
Subject: Re: [XRF] 59.5keV spectrum cleanup

It
looks like the low energy electronic noise is a problem below 1.8 keV. You¡¯ll
need to try resetting the threshold filters or try a higher PT as noise counts
are 10 times your signal

Dud

?

?

From: [email protected]
[mailto:[email protected]] On Behalf Of GEOelectronics@...
Sent: Friday, October 30, 2020 2:04 PM
To: [email protected]
Subject: Re: [XRF] 59.5keV spectrum cleanup

?

This is real time an still counting but it accumulated
11,250 secs now. .mca below.