And the output capacitor I used was 100 uF/25 volts, with
measured 100 kHz ESR of 0.26 ohms.
Steve
On 9/16/2022 10:00 PM, Steve Ratzlaff
via groups.io wrote:
toggle quoted message
Show quoted text
As noted, first test was at -40 dBm. I raised the level just
now to max, -17 dBm--very little change from before--the output
impedance decreased about 0.03 ohms for each frequency.
Steve
On 9/16/2022 8:55 PM, Tom Lee wrote:
Those numbers have a believable trend, but the magnitudes seem
too high. I don't know what the parasitic emitter resistance is
of a 2N5109, but I would be surprised if it were bigger than an
ohm. The output capacitor's ESR adds to that, of course, but a
good cap shouldn't have several ohms of ESR. So, if you include
realistic parasitics, the output resistance should be an ohm or
less. The milliohms number I gave earlier is the ideal value you
would get in the limit of no parasitics.
How large an excitation are you using to probe the output Z?
Tom
--
Prof. Thomas H. Lee
Allen Ctr., Rm. 205
350 Jane Stanford Way
Stanford University
Stanford, CA 94305-4070
On 9/16/2022 19:48, Steve Ratzlaff
wrote:
I breadboarded the circuit just now using 2N5109's and
swept the output Z with the DG8SAQ VNWA, 10 kHz-30 MHz. The
circuit draws 117 mA; the output transistor draws 84 mA. The
output impedance is mostly flat, gradually rising after
about 10 MHz.
50 kHz 4.23 ohms
1 MHz 4.42 ohms
10 MHz 5.27 ohms
20 MHz 6.80 ohms
30 MHz 8.77 ohms
Steve AA7U
On 9/16/2022 5:44 PM, Mikek
wrote:
Someone sent me a private messaged and corrected me on the
output impedance of the 50 to 1 transformer,
it is 0.001¦¸ or 1 milli¦¸. So, impedance ratio of 2500 x
-.001 = 2.5¦¸ primary impedance. I'm not sure I know how to
figure the output impedance of the
power amp in the HP4342A, I would have thought 220¦¸.
I'm posting the schematic of the impedance converter the
drives the injection transformer primary and asking,
What is the output impedance of the impedance converter?
????????????????? Thanks, Mikek