Re: DPO7254 SPC Fail

 

Silly question: Did you remember to unplug all input sources before starting SPC?

Raymond


Re: DPO7254 SPC Fail

 

The machine has plenty of clearance. I have tried running the SPC soon as I power it on and after it has been running for about 30-40 minutes. It keeps failing for the same code (** 1B failed -- no trigger found. **). I have reinstalled windows and software. What are my options.


Re: DPO7254 SPC Fail

 

Hi Paul,

Before you start the SPC Procedure the scope tells you the Purpose of this Procedure. That it failed after 5 minutes is a good sign. Maybe the internal temperature rises too much during this 5 minutes. Plese read the instructions about clearance. Thus not building it in between other instruments, obstructing airfliow etc. etc.

Success,

Egge Siert


Re: DPO7254 SPC Fail

 

Yes I ran an SPC, it runs for about 5 minutes then fails. I was able to get the faults from the console, it points to something about the trigger. Here is the fault:

SHOWTIME: SPC starting step : Trig gain/offset/hysteresis
** 1B failed -- no trigger found. **
18:04:11:125[2456] caltrig.cpp:calTrigGainOffsetAndHysteresis(123) error calibrating gain/hysteresis on channel 1
CAL: SPC returned an error in step Trig gain/offset/hysteresis

All the channels seem to work fine. I don't see any errors or offsets on any channel. The triggers all seem to work fine on all the channels as well. How important is it for the SPC to pass.


Re: DPO7254 SPC Fail

 

On Tue, Mar 19, 2024 at 06:15 AM, Paul Kalinchuk wrote:
DPO7254 that is failing SPC
DId you run the SPC?

Can be (it is?) a damaged acquisition board. (Was one of the channels over-volted? or cable static discharge damage?)
There is... at least... a resistor, a relay, a pre-amp in the channel input signal chain. (One...or all... might be damaged.)
The Signal Path Compensation signal could be routed by a relay too.

Do you see a measurement error on one or more of the channels??

You can substitute in a good acquisition board.
Or,try to trace the signal chain on the bad channel(s)... to find the faulty component(s)
Or live with the measurement error(s)?

AFAIR you can run some tests (and see the results) in the Instrument Diagnostics page, under the Utilities menu

Tek could (or could have) done component level repair to the acquisition board.?
Not sure if they are doing that now (DPO7000 series has been replaced by MSOs)
Another option is to find a parts unit with a good ack board.


Re: 7633 CRO

russell shaw
 

On 18/3/24 21:26, russell shaw wrote:
7633 CRO frame with service manual
I decided to keep the 7633.

Prefer analog over DSOs for doing most analog work.


DPO7254 SPC Fail

 

Hello, I purchased a DPO7254 that is failing SPC. The channels seem to all work in all modes. Both 1mOhm and 50Ohm and all voltage ranges on all channels seem to be working properly but SPC keeps failing. What are the common causes for this and how do I go about repairing it. This is my first Tektronix. The only other experience I have with oscilloscopes is a rigol that I¡¯ve owned for 10 years.?

If someone can put me in the right direction. Also, how do I retrieve error codes/logs?


7633 CRO

russell shaw
 

7633 CRO frame with service manual (no plugins, but might be available depending on demand). Hasn't be used for a long time, so would be advisable to do the 240V variac treatment (i ramp it up 20V per hour).

Also a 7490 service manual.

<>

Selling for a lowish amount and can be picked up at the EMDRC Car Boot Sale - Sunday 24 March 2024 10am-1pm EMDRC Clubrooms, 13a McCubbin St Burwood East
Melbourne, Australia

Also selling a vacuum chamber and couple of vacuum pumps among other things.
Good for re-impgregnating flyback transformers.

Chamber has glass door and a heater, so also useful for desiccating reels of SMD components.


7633 CRO

russell shaw
 

7633 CRO frame with service manual (no plugins, but might be available depending on demand). Hasn't be used for a long time, so would be advisable to do the 240V variac treatment (i ramp it up 20V per hour).

Also a 7490 service manual.

<>

Selling for a lowish amount and can be picked up at the EMDRC Car Boot Sale - Sunday 24 March 2024 10am-1pm EMDRC Clubrooms, 13a McCubbin St Burwood East
Melbourne, Australia

Also selling a vacuum chamber and couple of vacuum pumps among other things.
Good for re-impgregnating flyback transformers.


Re: Tek 7104, 7904, HP141T

russell shaw
 

On 15/3/24 12:12, George Livanes wrote:
Hi Russell,
I am interested in MOST? of this equipment particularly a tracking generator for my HP141T (and? a RF, Audio, plugin)
...

Hello George.L,

I am arranging to sell the gear as complete systems to two other enthusiast buyers.

These decisions are hard to make for much worthy interest.

Thankyou very much for your interest in keeping nice old machines running.

--
regards,
Russell Shaw, B.Eng, M.Eng(research)


Re: Tek 7104, 7904, HP141T

 

Hi Russell,
Forgot to mention that i am happy to buy items from? you (or send money to your favorite charity)...eg Have been after a HP8444a tracking generator for the Hp141 for years, to go with the pre selector, i just repaired (also would love to get ANY extender to suite the above gear)
Also, If a courier does not suit perhaps you could drop off at "pack and send who will pack it for you as well as send
Cheers
George


Re: Tek 7104, 7904, HP141T

 

Hi Russell,
I am interested in MOST? of this equipment particularly a tracking generator for my HP141T (and? a RF, Audio, plugin)
I am located in Sydney but i think i can get a courier to pick up from you without causing you too much trouble
I too am a collector with a lot of Tek and Hp gear
The 7104 would be nice (have a 7904 ) but perhaps too hard to get up to Sydney ??
I am very interested in Service Manuals and any extenders you have
Anything you send me will be used and? carefully restored so it will go to a good home
I'm retired these days and come from a Electronics Tech / Engineering background
Worked for many years with this era of Tek HP equipment and just love it
Hope to hear from you soon
Cheers
George
Phone 0403165332
george.livanes@...


Tek 7104, 7904, HP141T

russell shaw
 

Hi,

I've have a Tek 7104 with a 500MHz timebase plugin (7A29 ? it's around here somewhere) and a 7904.

The 7104 looks new and i have a thick service manual for it.
The 7904 is older looking and also has a manual.

I've got like 20+ vertical and timebase plugins.

I haven't switch them on for years but worked last time i used them.

Would be good to gradually ramp up the mains voltage to reform the caps using a variac.

I'm trying to make space, and so would be good for them to go to a suitable collector or someone that treats them well.

I can deliver to the Melbourne (Australia) and surrounding regions. I won't post them (unless someone convinces me otherwise).

I still do and think of analog stuff a lot, but the software and programming side of things (all aimed at analog electronic things eventually) means this gear won't be used for a long time.

Also got an HP141T spec-an system with all the plugins to cover from audio to 18GHz, and its 2 tracking generators (an 8443B 100kHz-110MHz and a 8444A-opt059 0.5-1500MHz). Got service manuals for all of them iirc. It all worked last time i used them. The caps would probably need reforming. This gear is a bit clunky to use compared to more modern CPU controlled gear, but is very repaireable. I have extender cards for servicing around somewhere too.

If there's too many takers, i would have to figure out the most worthy (probably someone that has the most gear, or that does the most impressive things).
If there's no one, i'll take them to hamfests.


Re: Converting TR502 tracking generator to TR503

 

The KO4BB website is not mine.? I just participated in a WIKI he operated in the past.? Apparently the TR502/TR503 conversion account (from about 15-years ago) was later saved to his website.? I found the link by searching for TR503 on the TEKWIKI websitesite.

Bruce, KG6OJI


Re: Converting TR502 tracking generator to TR503

 

Thanks Bruce,
I will use the 494 to align the TR502 as you suggested (just wanted to make sure it could be done)
I gather the Ko4bb site is yours ?
Essential site for all Test Gear minded? people..well done
George


Re: Converting TR502 tracking generator to TR503

 

Can't you just use the 494 to align the tracking generator?? I used a 492.

Bruce, KG6OJI


Re: Converting TR502 tracking generator to TR503

 

Hi.

In this day and age...

Take the 52.5MHz oscillator, divide by 21 to get 2.5MHz, then with a PLL lock a 55MHz VCO to that, via a divide by 22 counter.

It depends what you want, especially if you need the best noise floor you can, though what the original spec was (for example) the 7L13 etc and TR502/3 combination I know not.? Fixed frequency PLL's with a narrow loop bandwidth, can be quiet, especially if the VCO doesn't need much pulling to keep on frequency.

As to if it is cheaper than buying a "new" 55MHz crystal, who knows.? Depends I guess how you value your time.

Re bench space...? How can a 7603 take up more space than a 494? Curious, unless you have the TM5xx rack for the TG alongside it!

At Tek in the UK, when I was at Maidenhead, there were scope trolleys that could accommodate a TM504 below the scope (and that could be a 7704)? Very convenient.? The bench I had, had the utility TM504 on a shelf above it.

Just idle random thoughts..

73.

Dave B.

--
Sent from a MicroSoft free zone:


Re: Converting TR502 tracking generator to TR503

 

Just replacing the crystal with a programmable oscillator might not work.? A varactor tweaks the crystal oscillator frequency for tracking adjustment.? There are programmable oscillators like the SI571 which include a control voltage input, so something like that could work.
--John Gord


On Sun, Mar 3, 2024 at 05:28 PM, Sam Reaves wrote:
If you can get someone that has a NanoVNA H4 with the DSlord?firmware in it to sweep the crystal you have you should have all of the information you need to change to the 55MHz crystal.
?
There are some programmable?oscillator IC's?(is it SI Time?)
?
Or just order the programmable rock kit from QRP Labs and run it through a low pass filter:
?
?
73,
?
Sam
?
Sam Reaves
ARS W3OHM
Owner / Moderator of:
LeCroy Owners Group on Groups.io
Sencore Owners Group on Groups.io
Sprint Layout Group on Groups.io??
Pulsonix EDA Group on Groups.io
LPKF Owners Group on Groups.io
Electronics and Mechanical Hardware Design Engineering Manager
Staff Scientist Andritz Rolls Global Research Center (RETIRED)
?

On Sun, Mar 3, 2024 at 6:51?PM George Livanes <gmt1@...> wrote:

Hi Bruce,
Yes I saw your message, on ko4bb Thanks
My 7603 scope is temporarily sick and waiting my attention so my 7L13 system is out of action
Not sure if I will bother with this system anymore as the 7L13 is just too basic a specy (especially since I don't have the associated Tek frequency counter)...and the 7603 takes up too much real estate on my bench ....
So I was hoping to be able to convert the B.P.Filter without using the 7L13 (using a suitable VNA) so I can use my 494 with the modified Tracking generator

The Xtal is a worry..I think I can get it made locally but I suspect they will need more information

than it is a series resonant (I think I read 10ppm stability somewhere on the forum)
George

?

?


Re: Converting TR502 tracking generator to TR503

 

There is also in Belgium:



Sam Reaves
ARS W3OHM



On Sun, Mar 3, 2024 at 11:32?PM Sam Reaves via <sam.reaves=[email protected]> wrote:
I have heard that Bomar Crystals is back in business making crystals for ham use:


Sam Reaves
ARS W3OHM




On Sun, Mar 3, 2024 at 8:28?PM Sam Reaves via <sam.reaves=[email protected]> wrote:
If you can get someone that has a NanoVNA H4 with the DSlord?firmware in it to sweep the crystal you have you should have all of the information you need to change to the 55MHz crystal.

There are some programmable?oscillator IC's?(is it SI Time?)

Or just order the programmable rock kit from QRP Labs and run it through a low pass filter:



73,

Sam

Sam Reaves
ARS W3OHM
Owner / Moderator of:
LeCroy Owners Group on Groups.io
Sencore Owners Group on Groups.io
Sprint Layout Group on Groups.io??
Pulsonix EDA Group on Groups.io
LPKF Owners Group on Groups.io
Electronics and Mechanical Hardware Design Engineering Manager
Staff Scientist Andritz Rolls Global Research Center (RETIRED)



On Sun, Mar 3, 2024 at 6:51?PM George Livanes <gmt1@...> wrote:

Hi Bruce,
Yes I saw your message, on ko4bb Thanks
My 7603 scope is temporarily sick and waiting my attention so my 7L13 system is out of action
Not sure if I will bother with this system anymore as the 7L13 is just too basic a specy (especially since I don't have the associated Tek frequency counter)...and the 7603 takes up too much real estate on my bench ....
So I was hoping to be able to convert the B.P.Filter without using the 7L13 (using a suitable VNA) so I can use my 494 with the modified Tracking generator

The Xtal is a worry..I think I can get it made locally but I suspect they will need more information

than it is a series resonant (I think I read 10ppm stability somewhere on the forum)
George


Re: Converting TR502 tracking generator to TR503

 

I have heard that Bomar Crystals is back in business making crystals for ham use:


Sam Reaves
ARS W3OHM




On Sun, Mar 3, 2024 at 8:28?PM Sam Reaves via <sam.reaves=[email protected]> wrote:
If you can get someone that has a NanoVNA H4 with the DSlord?firmware in it to sweep the crystal you have you should have all of the information you need to change to the 55MHz crystal.

There are some programmable?oscillator IC's?(is it SI Time?)

Or just order the programmable rock kit from QRP Labs and run it through a low pass filter:



73,

Sam

Sam Reaves
ARS W3OHM
Owner / Moderator of:
LeCroy Owners Group on Groups.io
Sencore Owners Group on Groups.io
Sprint Layout Group on Groups.io??
Pulsonix EDA Group on Groups.io
LPKF Owners Group on Groups.io
Electronics and Mechanical Hardware Design Engineering Manager
Staff Scientist Andritz Rolls Global Research Center (RETIRED)



On Sun, Mar 3, 2024 at 6:51?PM George Livanes <gmt1@...> wrote:

Hi Bruce,
Yes I saw your message, on ko4bb Thanks
My 7603 scope is temporarily sick and waiting my attention so my 7L13 system is out of action
Not sure if I will bother with this system anymore as the 7L13 is just too basic a specy (especially since I don't have the associated Tek frequency counter)...and the 7603 takes up too much real estate on my bench ....
So I was hoping to be able to convert the B.P.Filter without using the 7L13 (using a suitable VNA) so I can use my 494 with the modified Tracking generator

The Xtal is a worry..I think I can get it made locally but I suspect they will need more information

than it is a series resonant (I think I read 10ppm stability somewhere on the forum)
George