My take is the chip is bad because it has a piece of finger stuck to it? Buy a new chip sans finger and the scope will run again. Rob
toggle quoted message
Show quoted text
-----Original Message----- From: TekScopes@... [mailto:TekScopes@...] On Behalf Of Alexandre Souza - Listas Sent: Saturday, July 07, 2012 11:40 AM To: TekScopes@... Subject: Re: [TekScopes] Re: tektronix 2445 help 3847K file just showing something useless? Wake up to life, Luciano! :oP Why can't you host the picture online or compress it better? :oP --- Enviado do meu Motorola PT550 Meu site: ----- Original Message ----- From: "Lucianopep Luciano" <lucianopep@...> To: <TekScopes@...> Sent: Saturday, July 07, 2012 1:34 PM Subject: Re: [TekScopes] Re: tektronix 2445 help [2 Attachments] ________________________________ Hello Tom photo 4193 pin 18 U800 pin 16 non have trace..and test point 59 pin 9 u 735 see Photo 4195 ..4196 Probe X1 set. hi ..luciano :-) ? Can you put the good scope on U800 pins 18 (tp60), 16 (tp63), and test point 59 (U735-9). Maybe take a quick picture of each. The one measurement that has me puzzled is U800 pin 11. It is over 1.5 volts low and is a power input pin. ----- Original Message ----- From: Lucianopep Luciano To: TekScopes@... Sent: Friday, July 06, 2012 6:32 PM Subject: Re: [TekScopes] Re: tektronix 2445 help
________________________________
Hello Tom What do you think? How about if I take once again the voltage all pins of U800 and you send a photo as you set my Telk 2445? The values ??that I have highlighted are those that determine the fault? Thanks Tom. Luciano :-)
? Here are the readings I get. I put them next to what you measured on your scope.
pin 1 = -2,62 V -2.187 pin 2= -1233mV +1.164 pin3= +14,52 V +10.21 pin4= +8,30V +38.7 - hor plate pin5= 0V 0 pin 6= + 78,7V +84.3 pin 7= +86,4V +86.3 pin 8= + 76,4 V +60.3 pin 9 =+73,7V +39.2 + hor plate pin10 = +15,01V +14.98 pin 11= +3,95V +5.68 This one puzzles me. The measurement I got is what I would expect. pin12= +4,29V +4.27
pin13=+7,90 mV +0.296 pin 14= +4,96V +4.96 pin 15= 0V 0 pin 16= -1685mV -1.683 pin17= +230mV -0.306 pin18= -57,4mV -0.007 pin19= +4,95V -4.98 pin20= +24,7mV -.003 pin21= -41,2mV -0.032 pin 22= -1367mV +0 .035 H pos centered, yours is ok here. pin 23= 0V 0 pin 24=+609mV -4.97 I need to review this one.
------------------------------------ Yahoo! Groups Links
|
?
The best price I have seen for the U800 chip is about $150 and they are
hard to find. There is also the issue of an in experienced person safely
removing a 24 pin dip in a multilayer board.
?
We are going to take the discussion offline so as to not clutter the
group.
?
Tom
?
?
?
toggle quoted message
Show quoted text
----- Original Message -----
Sent: Saturday, July 07, 2012 2:07
PM
Subject: RE: [TekScopes] Re: tektronix
2445 help
?
My take is the chip is bad because it has a piece of finger stuck to it?
Buy a new chip sans finger and the scope will run
again. Rob -----Original Message----- From: TekScopes@...
[mailto:TekScopes@...] On
Behalf Of Alexandre Souza - Listas Sent: Saturday, July 07, 2012 11:40
AM To: TekScopes@... Subject:
Re: [TekScopes] Re: tektronix 2445 help
3847K file just showing
something useless? Wake up to life, Luciano! :oP Why can't you host the
picture online or compress it better? :oP
--- Enviado do meu
Motorola PT550 Meu site:
-----
Original Message ----- From: "Lucianopep Luciano" <lucianopep@...> To:
<TekScopes@...> Sent:
Saturday, July 07, 2012 1:34 PM Subject: Re: [TekScopes] Re: tektronix 2445
help [2 Attachments]
________________________________ Hello Tom
photo 4193 pin 18 U800 pin 16 non have trace..and test point 59 pin 9 u
735 see Photo 4195 ..4196 Probe X1 set.
hi ..luciano
:-)
? Can you put the good scope on U800 pins 18 (tp60), 16 (tp63),
and test point 59 (U735-9). Maybe take a quick picture of each.
The
one measurement that has me puzzled is U800 pin 11. It is over 1.5 volts
low and is a power input pin.
----- Original Message -----
>From: Lucianopep Luciano >To: TekScopes@... >Sent:
Friday, July 06, 2012 6:32 PM >Subject: Re: [TekScopes] Re: tektronix
2445
help > > > > > > > >________________________________ > >Hello
Tom What do you think? How about if I take once again the voltage >all
pins of U800 >and you send a photo as you set my Telk 2445? >The
values ??that I have highlighted are those that determine the
fault? >Thanks Tom. Luciano :-) > > >? >Here are
the readings I get. I put them next to what you measured on your
>scope. > > >pin 1 = -2,62 V -2.187 >pin 2=
-1233mV +1.164 >pin3= +14,52 V +10.21 >pin4= +8,30V +38.7 - hor
plate >pin5= 0V 0 >pin 6= + 78,7V +84.3 >pin 7= +86,4V
+86.3 >pin 8= + 76,4 V +60.3 >pin 9 =+73,7V +39.2 + hor
plate >pin10 = +15,01V +14.98 >pin 11= +3,95V +5.68 This one
puzzles me. The measurement I got is what >I would
expect. >pin12= +4,29V
+4.27 > > > > > > >pin13=+7,90 mV
+0.296 >pin 14= +4,96V +4.96 >pin 15= 0V 0 >pin 16= -1685mV
-1.683 >pin17= +230mV -0.306 >pin18= -57,4mV -0.007 >pin19=
+4,95V -4.98 >pin20= +24,7mV -.003 >pin21= -41,2mV
-0.032 >pin 22= -1367mV +0 .035 H pos centered, yours is ok
here. >pin 23= 0V 0 >pin 24=+609mV -4.97 I need to review this
one. > > > >>> > >
------------------------------------
Yahoo!
Groups Links
|
?
Hi Tom,
How could?the discussion of this Tektronix 2445 repair be
considered "clutter" to a Tektronix Scope Group?
I for one would like to read about what you guys are doing to
fix this 2445.
Already you have both listed the U800 pin voltages found on
your scopes, and you were working your way though the discrepancies between your
voltage lists.
I seems a shame to take this discussion off?group because
of one negative comment.
U800 problems are chronic on 24x5 scopes, so having the chance
to see how knowledgeable people (such as yourself) attack the problem of
identifying the actual problem is just too good to miss.
Please reconsider?taking this 2445
repair?discussion?off group!
tom jobe...
?
?
?
?
toggle quoted message
Show quoted text
----- Original Message -----
Sent: Saturday, July 07, 2012 11:18
AM
Subject: Re: [TekScopes] Re: tektronix
2445 help
?
?
The best price I have seen for the U800 chip is about $150 and they are
hard to find. There is also the issue of an in experienced person safely
removing a 24 pin dip in a multilayer board.
?
We are going to take the discussion offline so as to not clutter the
group.
?
Tom
?
?
?
----- Original Message -----
Sent: Saturday, July 07, 2012 2:07
PM
Subject: RE: [TekScopes] Re: tektronix
2445 help
?
My take is the chip is bad because it has a piece of finger stuck to it?
Buy a new chip sans finger and the scope will run
again. Rob -----Original Message----- From: TekScopes@...
[mailto:TekScopes@...] On
Behalf Of Alexandre Souza - Listas Sent: Saturday, July 07, 2012 11:40
AM To: TekScopes@... Subject:
Re: [TekScopes] Re: tektronix 2445 help
3847K file just showing
something useless? Wake up to life, Luciano! :oP Why can't you host the
picture online or compress it better? :oP
--- Enviado do meu
Motorola PT550 Meu site:
-----
Original Message ----- From: "Lucianopep Luciano" <lucianopep@...> To:
<TekScopes@...> Sent:
Saturday, July 07, 2012 1:34 PM Subject: Re: [TekScopes] Re: tektronix
2445 help [2 Attachments]
________________________________ Hello
Tom photo 4193 pin 18 U800 pin 16 non have trace..and test point 59 pin
9 u 735 see Photo 4195 ..4196 Probe X1 set.
hi ..luciano
:-)
? Can you put the good scope on U800 pins 18 (tp60), 16
(tp63), and test point 59 (U735-9). Maybe take a quick picture of
each.
The one measurement that has me puzzled is U800 pin 11. It is
over 1.5 volts low and is a power input pin.
----- Original
Message ----- >From: Lucianopep Luciano >To: TekScopes@... >Sent:
Friday, July 06, 2012 6:32 PM >Subject: Re: [TekScopes] Re: tektronix
2445
help > > > > > > > >________________________________ > >Hello
Tom What do you think? How about if I take once again the voltage
>all pins of U800 >and you send a photo as you set my Telk
2445? >The values ??that I have highlighted are those that determine
the fault? >Thanks Tom. Luciano
:-) > > >? >Here are the readings I get. I put them
next to what you measured on your >scope. > > >pin
1 = -2,62 V -2.187 >pin 2= -1233mV +1.164 >pin3= +14,52 V
+10.21 >pin4= +8,30V +38.7 - hor plate >pin5= 0V 0 >pin 6=
+ 78,7V +84.3 >pin 7= +86,4V +86.3 >pin 8= + 76,4 V
+60.3 >pin 9 =+73,7V +39.2 + hor plate >pin10 = +15,01V
+14.98 >pin 11= +3,95V +5.68 This one puzzles me. The measurement I
got is what >I would expect. >pin12= +4,29V
+4.27 > > > > > > >pin13=+7,90 mV
+0.296 >pin 14= +4,96V +4.96 >pin 15= 0V 0 >pin 16=
-1685mV -1.683 >pin17= +230mV -0.306 >pin18= -57,4mV
-0.007 >pin19= +4,95V -4.98 >pin20= +24,7mV -.003 >pin21=
-41,2mV -0.032 >pin 22= -1367mV +0 .035 H pos centered, yours is ok
here. >pin 23= 0V 0 >pin 24=+609mV -4.97 I need to review this
one. > > > >>> > >
------------------------------------
Yahoo!
Groups Links
|
How could?the discussion of this Tektronix 2445 repair be considered "clutter" to a Tektronix Scope Group?No way, the discussion of this repair is a fantastic source of information and there isn?t problem at all to continue as an open source. I seems a shame to take this discussion off?group because of one negative comment. I agree, just ignore the negative comments and go on. Regards, Carlos.
Em 07/07/2012 20:57, Tom Jobe < tomjobe@... > escreveu:
?
Hi Tom,
How could?the discussion of this Tektronix 2445 repair be considered "clutter" to a Tektronix Scope Group?
I for one would like to read about what you guys are doing to fix this 2445.
Already you have both listed the U800 pin voltages found on your scopes, and you were working your way though the discrepancies between your voltage lists.
I seems a shame to take this discussion off?group because of one negative comment.
U800 problems are chronic on 24x5 scopes, so having the chance to see how knowledgeable people (such as yourself) attack the problem of identifying the actual problem is just too good to miss.
Please reconsider?taking this 2445 repair?discussion?off group!
tom jobe...
toggle quoted message
Show quoted text
----- Original Message -----
From:
To:
Sent: Saturday, July 07, 2012 11:18 AM
Subject: Re: [TekScopes] Re: tektronix 2445 help
?
The best price I have seen for the U800 chip is about $150 and they are hard to find. There is also the issue of an in experienced person safely removing a 24 pin dip in a multilayer board.
We are going to take the discussion offline so as to not clutter the group.
Tom
----- Original Message -----
From:
To:
Sent: Saturday, July 07, 2012 2:07 PM
Subject: RE: [TekScopes] Re: tektronix 2445 help
My take is the chip is bad because it has a piece of finger stuck to it? Buy a new chip sans finger and the scope will run again. Rob -----Original Message----- From: [mailto:] On Behalf Of Alexandre Souza - Listas Sent: Saturday, July 07, 2012 11:40 AM To: Subject: Re: [TekScopes] Re: tektronix 2445 help
3847K file just showing something useless? Wake up to life, Luciano! :oP Why can't you host the picture online or compress it better? :oP
--- Enviado do meu Motorola PT550 Meu site:
----- Original Message ----- From: "Lucianopep Luciano" <lucianopep@...> To: <TekScopes@...> Sent: Saturday, July 07, 2012 1:34 PM Subject: Re: [TekScopes] Re: tektronix 2445 help [2 Attachments]
________________________________ Hello Tom photo 4193 pin 18 U800 pin 16 non have trace..and test point 59 pin 9 u 735 see Photo 4195 ..4196 Probe X1 set.
hi ..luciano :-)
? Can you put the good scope on U800 pins 18 (tp60), 16 (tp63), and test point 59 (U735-9). Maybe take a quick picture of each.
The one measurement that has me puzzled is U800 pin 11. It is over 1.5 volts low and is a power input pin.
----- Original Message --
--- >From: Lucianopep Luciano >To: TekScopes@... >Sent: Friday, July 06, 2012 6:32 PM >Subject: Re: [TekScopes] Re: tektronix 2445 help > > > > > > > >________________________________ > >Hello Tom What do you think? How about if I take once again the voltage >all pins of U800 >and you send a photo as you set my Telk 2445? >The values ??that I have highlighted are those that determine the fault? >Thanks Tom. Luciano :-) > > >? >Here are the readings I get. I put them next to what you measured on your >scope. > > >pin 1 = -2,62 V -2.187 >pin 2= -1233mV +1.164 >pin3= +14,52 V +10.21 >pin4= +8,30V +38.7 - h
or plate >pin5= 0V 0 >pin 6= + 78,7V +84.3 >pin 7= +86,4V +86.3 >pin 8= + 76,4 V +60.3 >pin 9 =+73,7V +39.2 + hor plate >pin10 = +15,01V +14.98 >pin 11= +3,95V +5.68 This one puzzles me. The measurement I got is what >I would expect. >pin12= +4,29V +4.27 > > > > > > >pin13=+7,90 mV +0.296 >pin 14= +4,96V +4.96 >pin 15= 0V 0 >pin 16= -1685mV -1.683 >pin17= +230mV -0.306 >pin18= -57,4mV -0.007 >pin19= +4,95V -4.98 >pin20= +24,7mV -.003 >pin21= -41,2mV -0.032 >pin 22= -1367mV +0 .035 H pos centered, yours is ok here. >pin 23= 0V 0 >pin 24=+609mV -4.97 I need to review this one. > > > >>> > >
------------------------------------
Yahoo!
Groups Links
?
?
|
I think a summary at the end would be much more helpful than following all the blind alleys.
toggle quoted message
Show quoted text
On Sat, Jul 7, 2012 at 5:57 PM, Tom Jobe <tomjobe@...> wrote:
?
Hi Tom,
How could?the discussion of this Tektronix 2445 repair be
considered "clutter" to a Tektronix Scope Group?
|
I believe the outcome is the OP found a good horizontal sweep signal being fed to the input of U800 (pin 18). Since the output (pins 4 & 9) is not correct, it is reasonable to assume that U800 has failed, especially since U800 is a known problem area.
The low voltage measured by the OP on pin 11 indicates the current flow to that pin is greater than normal (pin 11 receives voltage via a 330 ohm resistor connected to the +15V power supply.) I suppose that also is indicative of a failed U800.
Thanks,
Patrick Wong AK6C
toggle quoted message
Show quoted text
--- In TekScopes@..., David DiGiacomo <daviddigiacomo@...> wrote: I think a summary at the end would be much more helpful than following all the blind alleys.
On Sat, Jul 7, 2012 at 5:57 PM, Tom Jobe <tomjobe@...> wrote:
Hi Tom, How could the discussion of this Tektronix 2445 repair be considered "clutter" to a Tektronix Scope Group?
|
Hi Patrick,
?
Seeing the waveform on the input along with the DC component of 60 mV on
pin 18 leads me to believe the chip is bad. I didn't want him to go through
changing the chip unless I was sure that was the problem. It's no trivial task
for a beginner to remove the main board just to get access to U800.
?
There are other failure modes that can cause the same symptoms.
?
Of course you or I could have arrived at that diagnosis within minutes but
we are dealing with a language barrier as well as a time delay of a
day?between replies.
?
Rob has found two sources for U800 from $117 to $135 so that is very
helpful. Now to find out if he can get some experienced help to change the
chip.
?
?
Regards,
Tom
?
?
?
?
toggle quoted message
Show quoted text
----- Original Message -----
Sent: Saturday, July 07, 2012 11:00
PM
Subject: [TekScopes] Re: tektronix 2445
help
?
I believe the outcome is the OP found a good horizontal sweep signal being
fed to the input of U800 (pin 18). Since the output (pins 4 & 9) is not
correct, it is reasonable to assume that U800 has failed, especially since
U800 is a known problem area.
The low voltage measured by the OP on
pin 11 indicates the current flow to that pin is greater than normal (pin 11
receives voltage via a 330 ohm resistor connected to the +15V power supply.) I
suppose that also is indicative of a failed
U800.
Thanks,
Patrick Wong AK6C
--- In TekScopes@..., David
DiGiacomo wrote: > > I think a summary
at the end would be much more helpful than following all the blind
alleys. > > On Sat, Jul 7, 2012 at 5:57 PM, Tom Jobe
wrote: > > > Hi Tom, > > How
could the discussion of this Tektronix 2445 repair be considered "clutter" to
a Tektronix Scope Group?
|
I question whether replacing U800 on a 2445 is worth the labor and the cost of the device. I suppose if it's a beloved family heirloom would make it about the only reason to keep it working.
toggle quoted message
Show quoted text
--- In TekScopes@..., "Tom Miller" <tmiller11147@...> wrote: Hi Patrick,
Seeing the waveform on the input along with the DC component of 60 mV on pin 18 leads me to believe the chip is bad. I didn't want him to go through changing the chip unless I was sure that was the problem. It's no trivial task for a beginner to remove the main board just to get access to U800.
There are other failure modes that can cause the same symptoms.
Of course you or I could have arrived at that diagnosis within minutes but we are dealing with a language barrier as well as a time delay of a day between replies.
Rob has found two sources for U800 from $117 to $135 so that is very helpful. Now to find out if he can get some experienced help to change the chip.
Regards, Tom
----- Original Message ----- From: Patrick Wong To: TekScopes@... Sent: Saturday, July 07, 2012 11:00 PM Subject: [TekScopes] Re: tektronix 2445 help
I believe the outcome is the OP found a good horizontal sweep signal being fed to the input of U800 (pin 18). Since the output (pins 4 & 9) is not correct, it is reasonable to assume that U800 has failed, especially since U800 is a known problem area.
The low voltage measured by the OP on pin 11 indicates the current flow to that pin is greater than normal (pin 11 receives voltage via a 330 ohm resistor connected to the +15V power supply.) I suppose that also is indicative of a failed U800.
Thanks,
Patrick Wong AK6C
--- In TekScopes@..., David DiGiacomo <daviddigiacomo@> wrote: > > I think a summary at the end would be much more helpful than following all the blind alleys. > > On Sat, Jul 7, 2012 at 5:57 PM, Tom Jobe <tomjobe@> wrote: > > > Hi Tom, > > How could the discussion of this Tektronix 2445 repair be considered "clutter" to a Tektronix Scope Group?
|
Well, he could most likely part out the good hybrids, attenuaters, power
supply & fan, the A5 board, HV power supply, and the CRT for enough to buy a
working 2465 and lunch. Especially in Italy. Or just keep it for spares.
?
:)
?
Tom
?
?
?
toggle quoted message
Show quoted text
----- Original Message -----
Sent: Sunday, July 08, 2012 12:57
AM
Subject: [TekScopes] Re: tektronix 2445
help
?
I question whether replacing U800 on a 2445 is worth the labor and the cost
of the device. I suppose if it's a beloved family heirloom would make it about
the only reason to keep it working.
--- In TekScopes@..., "Tom
Miller" wrote: > > Hi Patrick, >
> Seeing the waveform on the input along with the DC component of 60 mV
on pin 18 leads me to believe the chip is bad. I didn't want him to go through
changing the chip unless I was sure that was the problem. It's no trivial task
for a beginner to remove the main board just to get access to U800. >
> There are other failure modes that can cause the same symptoms.
> > Of course you or I could have arrived at that diagnosis
within minutes but we are dealing with a language barrier as well as a time
delay of a day between replies. > > Rob has found two sources for
U800 from $117 to $135 so that is very helpful. Now to find out if he can get
some experienced help to change the chip. > > >
Regards, > Tom > > > > > -----
Original Message ----- > From: Patrick Wong > To: TekScopes@...
> Sent: Saturday, July 07, 2012 11:00 PM > Subject: [TekScopes]
Re: tektronix 2445 help > > > > I believe the
outcome is the OP found a good horizontal sweep signal being fed to the input
of U800 (pin 18). Since the output (pins 4 & 9) is not correct, it is
reasonable to assume that U800 has failed, especially since U800 is a known
problem area. > > The low voltage measured by the OP on pin 11
indicates the current flow to that pin is greater than normal (pin 11 receives
voltage via a 330 ohm resistor connected to the +15V power supply.) I suppose
that also is indicative of a failed U800. > > Thanks, >
> Patrick Wong AK6C > > --- In TekScopes@..., David
DiGiacomo wrote: > > > > I think a
summary at the end would be much more helpful than following all the blind
alleys. > > > > On Sat, Jul 7, 2012 at 5:57 PM, Tom Jobe
wrote: > > > > > Hi Tom, > >
> How could the discussion of this Tektronix 2445 repair be considered
"clutter" to a Tektronix Scope Group? >
|
Tom,
Now you're talking!
toggle quoted message
Show quoted text
--- In TekScopes@..., "Tom Miller" <tmiller11147@...> wrote: Well, he could most likely part out the good hybrids, attenuaters, power supply & fan, the A5 board, HV power supply, and the CRT for enough to buy a working 2465 and lunch. Especially in Italy. Or just keep it for spares.
:)
Tom
----- Original Message ----- From: victor_j_silva To: TekScopes@... Sent: Sunday, July 08, 2012 12:57 AM Subject: [TekScopes] Re: tektronix 2445 help
I question whether replacing U800 on a 2445 is worth the labor and the cost of the device. I suppose if it's a beloved family heirloom would make it about the only reason to keep it working.
--- In TekScopes@..., "Tom Miller" <tmiller11147@> wrote: > > Hi Patrick, > > Seeing the waveform on the input along with the DC component of 60 mV on pin 18 leads me to believe the chip is bad. I didn't want him to go through changing the chip unless I was sure that was the problem. It's no trivial task for a beginner to remove the main board just to get access to U800. > > There are other failure modes that can cause the same symptoms. > > Of course you or I could have arrived at that diagnosis within minutes but we are dealing with a language barrier as well as a time delay of a day between replies. > > Rob has found two sources for U800 from $117 to $135 so that is very helpful. Now to find out if he can get some experienced help to change the chip. > > > Regards, > Tom > > > > > ----- Original Message ----- > From: Patrick Wong > To: TekScopes@... > Sent: Saturday, July 07, 2012 11:00 PM > Subject: [TekScopes] Re: tektronix 2445 help > > > > I believe the outcome is the OP found a good horizontal sweep signal being fed to the input of U800 (pin 18). Since the output (pins 4 & 9) is not correct, it is reasonable to assume that U800 has failed, especially since U800 is a known problem area. > > The low voltage measured by the OP on pin 11 indicates the current flow to that pin is greater than normal (pin 11 receives voltage via a 330 ohm resistor connected to the +15V power supply.) I suppose that also is indicative of a failed U800. > > Thanks, > > Patrick Wong AK6C > > --- In TekScopes@..., David DiGiacomo <daviddigiacomo@> wrote: > > > > I think a summary at the end would be much more helpful than following all the blind alleys. > > > > On Sat, Jul 7, 2012 at 5:57 PM, Tom Jobe <tomjobe@> wrote: > > > > > Hi Tom, > > > How could the discussion of this Tektronix 2445 repair be considered "clutter" to a Tektronix Scope Group? >
|