I've begun to go through the calibration process on the Tek 570 I picked up last summer. It's an early model, S/N 169. It works well but I wanted to give it a once over since I have no idea how long ago it was last calibrated.
For some reason the floating supplies, +400v and -300v are both reading high; +446v and -350v
The manual says the must be within 5% of the proper voltage, so these are out of spec. Im performing my tests with line voltage set to 117vac per the manual. I confirmed that the transformer is properly wired for 117v.
I have the front panel controls set per the manual. I tested the 6ax5 and 6x4 rectifiers, both sections strong on both. The 80uf electrolytics were already replaced with modern parts by the previous owner.
Any ideas? You can see in my pic (assuming I can figure out how to attach it) that the secondaries on the transformer are reading a bit high for both sections in question - but not too bad. (the 325-0-325 for the +400v 6ax5 is reading 342-0-342 and the 250-0-250 for the -300v 6x4 is reading 265-0-265)
I checked the diagram for the unregulated transformer at VintageTek, p/n 120-0069-00 to confirm that it was indeed made for 117v. Anyways, lowering to 115v doesn't bring it into spec - have to go all the way down below 110v.
I plan to start looking downstream and see if something's not pulling enough current, but figured I'd get this post started now.
Thanks,
Phil