It is true that adding a 1Meg paralleled with a 20 pF to the scope input will halve the capacitance, & double the DC resistance. Both of these results will reduce the loading on the DUT, which is the precise reason that 10X probes are used !
The scope's input capacitance is determined by wiring, relays, rotary attenuator switches, etc.
All these values are fixed & cannot be made to disappear.