More on my Tektronix 545B
JOSE V. GAVILA (EB5AGV/EC5AAU)
Hello!
I have been cleaning the 545B/1A1 and now it shines :-) Then I have connected it and I have got a curious trace; I have put a picture of it at: Center line is Channel 2, connected to GND. The other lines should be a calibrator signal on Channel 1. As you can see, the trace goes back and is distorted. It is synchronized with LINE, so it seems as a power supply filtering trouble. But as I have no manual nor schematics, I can't work on this easily. Do you have any hint for this trouble?. Also, if somebody could scan the schematics (or at least the power supply section) of the 545B it would sure help a lot! Regards, JOSE ---------------------------------------------------------------------- 73 EB5AGV / EC5AAU - JOSE V. GAVILA La Canyada - Valencia (SPAIN) EB5AGV Vintage Radio Site: European Boatanchors List: |
Don Black
Hi Jose,
toggle quoted message
Show quoted text
It looks like the power supply isn't regulating properly and has ripple on it. I'd check the supply lines for ripple if you've got another CRO. Probably either voltage set too high, weak electrolytics or regulator tubes. Sorry, I don't have the schematic. Don Black. "JOSE V. GAVILA (EB5AGV/EC5AAU)" wrote: Hello! |
Stan or Patricia Griffiths
Hi Jose,
toggle quoted message
Show quoted text
I would agree that it looks like a power supply problem. You should be able to find the test points for the power supplies on a ceramic strip on the upper chassis toward the rear of the scope. The voltages will be marked on the chassis (-150, +100, +225, etc.) They should each be within 2% of nominal and should have only about 5-10 mv of AC ripple on them. The calibrator and both sweep circuits should be off when checking ripple and you should be able to vary the line voltage + or - 10% of nominal and still have well regulating power supplies. In the US, we consider nominal line voltage to be 115VAC and vary it from 105 to 125 VAC while watching the ripple and voltage on each power supply with another scope. Both should remain stable. My guess is that you have 230 VAC and you should vary the mains voltage from 210 to 250 VAC while watching the voltage and ripple on each of the 5 power supplies (-150, +100, +225, +350, and +500 volts). Morris Odell mentioned a pull switch in the vertical plugin compartment. This switch is found in the 547, 546, and 544 but not the 545B and 543B. This means you MUST have a plugin installed in the 545B for the power supplies to regulate properly. You also mentioned that the power switch burned up. Yes, I have seen this many times. They just develop high contact resistance over time and you need to replace it. There is nothing special about this switch except that it should be rated for 10 amps. With regard to the slow starting of your scope (sometimes NO starting), if you had a manual, you would see that the heater voltage applied to the delay relay (looks like a vacuum tube) is applied through a couple of normally closed contacts on the mechanical relay that eventually closes and allows the scope to function. If those contacts develop high resistance, the turn-on time will vary irratically. You should carefully burnish all of the contacts on the mechanical relay. Yes, it is normal for some of the internal neons to turn off and on as the scope warms up and their operation will change again when the mechanical relay closes. I have read the other comments and this still sounds like a power supply problem in the 545B to me . . . Stan w7ni@... JOSE V. GAVILA (EB5AGV/EC5AAU) wrote: Hello! |
Hi,
I followed this discussion for a while. I have a 1A1 manual series no. 20000 up with FETs in the input stages. The circuit between the CH1 input and CH1 output is only supplied by the -150V, +100V and +225V. I think the "bump" you see see is supposed to come from one of these 3 voltages. The -150V is the reference voltage for the whole scope and also for the other supply voltages. Its one of the first things you need to check and adjust during a calibration and once it is adjusted you must never touch it again, or you need to run your calibration again. The voltages are supplied through the following pins of the interconnecting plug: +225V pin11, -150V pin9, and the +100V runs through the filament of V252 and V261 in the timebase B and drives a current of about 150mA into pin15 of the interconnecting plug. This current flows through a voltage divider in the plug in. The voltage divider supplies +39, +11 and +5V to the plug in. The voltage at pin 15 is about 75V. If you need a scan of the 1A1 CH1 input circuit or the 545B power supply, let me know. I can do that on the weekend. By the way there is 545B manual on Ebay in Germany. I think its worth the try to see whether the seller ships it to Spain. If you can't get it shipped to Spain, let me know. Leo --- In TekScopes@y..., Stan or Patricia Griffiths <w7ni@e...> wrote: Hi Jose,should be able to find the test points for the power supplies on a ceramic strip onthe upper chassis toward the rear of the scope. The voltages will be markedon the chassis (-150, +100, +225, etc.) They should each be within 2% ofnominal and should have only about 5-10 mv of AC ripple on them. Thecalibrator and both sweep circuits should be off when checking ripple and you should beable to vary the line voltage + or - 10% of nominal and still have wellregulating power supplies. In the US, we consider nominal line voltage to be115VAC and vary it from 105 to 125 VAC while watching the ripple and voltageon each power supply with another scope. Both should remain stable.voltage from 210 to 250 VAC while watching the voltage and ripple on each of the5 power supplies (-150, +100, +225, +350, and +500 volts).compartment. This switch is found in the 547, 546, and 544 but not the 545B and543B. This means you MUST have a plugin installed in the 545B for the power suppliesto regulate properly.seen this many times. They just develop high contact resistance over time and youneed to replace it. There is nothing special about this switch except thatit should be rated for 10 amps.starting), if you had a manual, you would see that the heater voltage applied to thedelay relay (looks like a vacuum tube) is applied through a couple of normallyclosed contacts on the mechanical relay that eventually closes and allowsthe scope to function. If those contacts develop high resistance, the turn-ontime will vary irratically. You should carefully burnish all of the contactson the mechanical relay.as the scope warms up and their operation will change again when themechanical relay closes.supply problem in the 545B to me . . .put a should be apicture of it at: synchronizedcalibrator signal on Channel 1. as I havewith LINE, so it seems as a power supply filtering trouble. But have any hintno manual nor schematics, I can't work on this easily. Do you (or at leastfor this trouble?. Also, if somebody could scan the schematics ----the power supply section) of the 545B it would sure help a lot! 73 EB5AGV / EC5AAU - JOSE V. GAVILA
|
to navigate to use esc to dismiss