Keyboard Shortcuts
ctrl + shift + ? :
Show all keyboard shortcuts
ctrl + g :
Navigate to a group
ctrl + shift + f :
Find
ctrl + / :
Quick actions
esc to dismiss
Likes
Search
7603 horizontal performance
I have just refurbished and calibrated my 7603.
The frame itself showed 2.5ns rise time and bandwidth was 140MHz. With 7A26 system rise time is 3ns and bandwidth seems to extend to 130MHz. Small dip around 100MHz? Now I have 4 channel >100MHz ¡±old school¡± scope in my lab. But I have also 7B87 and using high sweep speeds (<5ns/div) horizontal linearity is poor. Has anybody tried to improve the horizontal amplifier to have a bit better performance? Regards, Jouko |
Hello there: Very fine to yave these mainframes.
I have the similar 7904, with 7B92A HOR plugins, HOR LIN is fine, however at the fastest sweep rates first CM of sweep start has some slight aberrations. I think that is normal. IF the fast rate sweeps of you HOR are nonlin, either the 7B87 OR the mainframe HOR 7603 amps may be the culprits. More likely it is the plugin. Did you HOR cal the 7603? RE 7B87, try in another TEK 7000 mainframe? Was it calibrated? Scope photos of the nonlinearity on the fast rate as well as linearity on slow rates, (with calibrator or function gen signal) will get you fast get advice. Bon Courage, Jon |
On Mon, Mar 8, 2021 at 05:50 PM, Jouko Koskinen wrote:
The faster rates of the 7B87 are too fast for the 7603, by far: Fastest calibrated hor. sweep is 5ns/div. 7B87 goes down to 1 ns/div (with 10x multiplier). Raymond |
The fastest sweep the 7603 is "specified for" is 5ns/div which is obtainable with a 7B53A set the highest sweep (.05us/div) with the x10 multiplier engaged.
toggle quoted message
Show quoted text
I'm amazed you find the 7B85 or 7B87 work well with the 7603 as I thought the delay line in the 7603 was too short for these. FWIW if you are looking for detail in a sweep at faster than 5ns/div then you also need a (much) faster vertical amp. David -----Original Message-----
From: [email protected] <[email protected]> On Behalf Of Jouko Koskinen Sent: 08 March 2021 16:50 To: [email protected] Subject: [TekScopes] 7603 horizontal performance Has anybody tried to improve the horizontal amplifier to have a bit better performance?Regards, Jouko |
On Mon, Mar 8, 2021 at 11:18 PM, David C. Partridge wrote:
I guess that depends on what you call "work well". You just won't see (part of) the triggering edge but otherwise, the image is fine. Apart from it possibly being a nice exercise, I don't see the merit in trying to improve Tek's 7603 design. 7704's and 7904's are very affordable these days. Yes, they're a bit wider. Raymond |
There are faster frames frames available.
What I like in this 7603 is the large screen. Therefore I thought about this exercise. This CRT is without distributed deflection. Seems to be that all the faster tubes use distributed approach at least in vertical. I assume this is limiting the vertical response. Am I right? BR, Jouko |
On Tue, Mar 9, 2021 at 12:19 PM, Jean-Paul wrote:
I guess that the screens of the 2467 and 7104 are smaller not so much because of BW requirements but because the angle of incidence of the electron beam varies more across a larger screen, affecting the "transparency" of the microchannels and thereby the MCP's gain across the screen. The 2445B and 2465B, both with 400 MHz BW, have a larger screen size, as do the 400MHz and 500 MHz instruments in the 7000 series (7844, 7854, 7904(A)). Smaller screen size allows higher brightness (writing speed), witness models like the 7904 option 4, with its smaller (4 x 5 cm, instead of 8 x 10 cm) graticule. 24 kV total acceleration potential helps as that model... Analog storage CRT's often have a smaller screen as well (7613/23/33 vs 7603!). BTW: Models like the 2430/32/40, with their modest BW requirements, use the same complex CRT as the 2445 and 2465. I guess that's to do with CRT availability... Raymond |
On Tue, Mar 9, 2021 at 09:09 AM, Jouko Koskinen wrote:
The 7603 works well with a 7S14 sampling plugin. 1 GHz BW, two 50 Ohm channels only but a P6202 FET probe can provide Hi-Z input. The non-A (so, plain P6202, not P6202A) has its own power supply. High BW, low input capacity, easy-to-operate sampling plugin, 350 ps rise time, and all that in a 7603, you can't lose! Raymond |
Hi , if my memory serves me correctly the P6202 FET probe is only 500MHz so you will likely need a P6201 and an 1101 psu for maximum B/W
toggle quoted message
Show quoted text
Brian On Tuesday, 9 March 2021, 13:51:28 GMT, Jouko Koskinen <jouko.koskinen@...> wrote:
Hi Raymond, I already have 7S14. Have to look into FET probes. Thread "Sampling with 7S14". BR, Jouko |
On Tue, Mar 9, 2021 at 02:55 PM, Brian wrote:
Absolutely, Brian. For full BW you need (something even faster than) the P6201. I was thinking about the convenience of having an "integrated" power supply, as opposed to needing to hunt for an 1101. I should have mentioned that but thanks for bringing it up. Raymond Raymond |
Hi Raymond?I cant afford better than the P6201 and a home made 1101 substitute , if I need better I have a P6056 and P6057 resistive probes but of course no Hi-Z input?
toggle quoted message
Show quoted text
Brian On Tuesday, 9 March 2021, 14:02:22 GMT, Raymond Domp Frank <hewpatek@...> wrote:
On Tue, Mar? 9, 2021 at 02:55 PM, Brian wrote: Absolutely, Brian. For full BW you need (something even faster than) the P6201. I was thinking about the convenience of having an "integrated" power supply, as opposed to needing to hunt for an 1101. I should have mentioned that but thanks for bringing it up. Raymond Raymond |
On Tue, Mar 9, 2021 at 03:01 PM, Raymond Domp Frank wrote:
Further to this, AFAIK the OP wasn't looking for 1 GHz BW as such, but looking for a significantly faster time base and vertical BW. Even the P6201 "only" achieves 900 MHz "on its own". Talking about convenience: I often prefer using a 7A29 over a 7A19, even at around 500 MHz, because of the former's variable attenuation knob - and resettable overvoltage protection... Raymond |
On Tue, Mar 9, 2021 at 03:05 PM, Brian wrote:
Hi Brian, I love the P6201 and actually prefer it over the P6202 because of its handling convenience. I have the luxury of several 'scopes with probe power sockets. Still, I have modified two P6202's (the ones with permanently connected power supply) by cutting the power wire somewhere near the middle and inserting a Lemo male/female plug set like Tek used for probe power. That way, I have a stand-alone probe power supply to use with a P6201, should a probe power socket be absent, in the process converting the P6202's to P6202A's (?)... Raymond |
On Tue, Mar 9, 2021 at 03:11 PM, Jean-Paul wrote:
I have P6056/6057's and FET probes - and a P6230. Guess which one I use most. Raymond |
I like the P6201 also , I'm lucky enough to have 4 complete sets . I have not used them with a 7S14 though only the resistive probes . I have a 7S14 acquired long before the P6201's but the sampler has not been used that much for a while as I have since purchased 2 x 7104's along with some 7A29's and 7A19's . I also like the 7A29 over the 7A19 as its much easier to reset the input protection than keep replacing fuses that are really hard to find .
toggle quoted message
Show quoted text
Brian On Tuesday, 9 March 2021, 14:23:43 GMT, Raymond Domp Frank <hewpatek@...> wrote:
On Tue, Mar? 9, 2021 at 03:05 PM, Brian wrote: Hi Brian, I love the P6201 and actually prefer it over the P6202 because of its handling convenience. I have the luxury of several 'scopes with probe power sockets. Still, I have modified two P6202's (the ones with permanently connected power supply) by cutting the power wire somewhere near the middle and inserting a Lemo male/female plug set like Tek used for probe power. That way, I have a stand-alone probe power supply to use with a P6201, should a probe power socket be absent, in the process converting the P6202's to P6202A's (?)... Raymond |
Hi Raymond?Would I be anywhere near right if I suggested the P6230 -- is that the one providing DC offset . If I am wrong then I would gp for the P6056/6057?
toggle quoted message
Show quoted text
Brian On Tuesday, 9 March 2021, 14:28:40 GMT, Raymond Domp Frank <hewpatek@...> wrote:
On Tue, Mar? 9, 2021 at 03:11 PM, Jean-Paul wrote: I have P6056/6057's and FET probes - and a P6230. Guess which one I use most. Raymond |
to navigate to use esc to dismiss