7A15A issue or limitation??
Hi,
I have been playing around with my 7A15A's, in different 7603 Mainframes, and have a 'jittering' or 'flickering' of the 4V Calibration signal. The 'jittering' is a movement, always to the left, of the trace along the horizontal axis and return to the correct position. I can define this issue or limitation as follows:- Independent of Mainframe - 2 x 7603's used Independent of Timebases - 7B53 and 7B92A used Only occurs in TRIG SOURCE/VERT MODE with VERT MODE/ALT - both traces are stable in all other Modes Only one of the traces becomes unstable Only occurs when traces from each Vert Amp are 'un-overlapped' - when the two traces are overlapped, both traces are stable The Manual states that in TRIG SOURCE/VERT MODE .... "each sweep is triggered by the signal being displayed on the CRT" This suggests to me that when I dive into the Circuit Descriptions, that I should be looking for some 'switching mechanism' which is swapping the Triggering mechanism between the signals from each Vert Amp. If this 'switching mechanism' is faulty or receiving a weak signal, then this maybe causing poor triggering, hence the 'jittering'. However, this 'switching' between Vert Amp signals would occur in the Timebase or Mainframe. But that doesn't make sense because the 'jittering' occurs in both 7A15A's, independently (ie. 7A15A + 7A26 + 7B92A etc), and in 2 different Mainframes, and with different Timebases. So .... where have I gone wrong here???? ..... or is this a limitation of the 7A15A??? Any guidance is appreciated. -- Cheers, Ian, Melbourne, Australia |
Chuck Harris
Simple explanation, I think:
toggle quoted message
Show quoted text
"Vertical mode" triggering is only valid when there is a single vertical signal displayed on the screen, or you are in ALT (alternate) mode. It gets more complicated if I try to tell you why: The only modes where the traces on the screen are exactly what they appear is when there is only a single channel displayed on the screen. Whenever there is more than one channel on the screen, there is some form of eye trickery going on. This is because there is only a single beam in the CRT to draw on the screen. A couple of scope models have multiple beams, but they are the exception, rather than the rule... In ALT (alternate) mode, it may appear to the eye that there are two signals on the screen at the same time, but in reality, the two signals are being shown singly, alternating: A, B, A, B... If the sweep time is fast enough, the eye merges the alternating images so they appear to the brain to be simultaneous. At the slower sweeps, the eye can easily see the traces alternate. In CHOP (chopped) mode, it may appear to the eye that there are two signals on the screen at the same time, but in reality, the two signals are chopped up into alternating tidbits.... a tidbit of A, a tidbit of B, a tidbit of A, a tidbit of B...all the while the sweep marches on, making a sort of square wave with the tops being the tidbits of one channel, and the bottoms being the tidbits of the other channel. Because the chopping of the A and B signals is faster than the sweep speed, and with no time relationship to the sweep speed, or the signals, the eye sees the chopped up signal to have merged into what looks like simultaneous display. But the trigger isn't a human eye, so it isn't fooled by this trickery. When the trigger mode is in "Vertical", the trigger is given the exact signal the plugin is trying to display on the screen: Single channel mode, gives the single channel. Alternate mode, gives an alternation of Channel A and Channel B. Chop mode gives a composite signal of a little bit of A, B, and their vertical positions... A real dog's breakfast. "Vertical mode" triggering is only valid when there is a single vertical signal displayed on the screen, or you in ALT (alternate) mode. -Chuck Harris tinkera123 wrote: Hi, |
When the trigger mode is in "Vertical", the trigger is given*** Okay ... I understood ALT, CHOP, ADD etc .... individually ... but what I didn't grasp is the 'implication' of them collectively ... If I can re-state what you said ... the selection of "VERT MODE/ALT and TRIG SOURCE/VERT MODE" is the only mode that requires two sweeps (one for trace A and one for trace B) and 2 different trigger events. All the other modes only require one sweep and one trigger event. Doing some further experimenting with this today, I am finding that the 'jitter' is substantially less when I use higher bandwidth probes, and no co-ax, no bnc t-connector etc, to connect the Calibrator signal to the 7A15A's. So ... somehow .... maybe this second trigger event is being compromised by my low BW equipment?? |
to navigate to use esc to dismiss