¿ªÔÆÌåÓý

ctrl + shift + ? for shortcuts
© 2025 Groups.io

7904 trace dimming


 

Hello All,


I am noticing that as my 7904 is warming, the trace dims. It does this about 3-5 mins after power up and does so all at once. In other words, it is not gradual dimming. I need to increase the intensity by one mark on the knob scale in order to return the trace to similar intensity as before it dimmed. It does not dim beyond that throughout the day after this initial "dimming."


Has anyone else experienced this?


Just tinkering in the lab.............


Cheers,


David


 

Hi David,

I have seen that on other 7000 scopes and even on some of the small TM500
scopes (SC501, SC502, SC503, SC504, and MR501). It is subtle. It was the
subject of some comments on the forum many, many years ago.

I will ask a friend of mine who was in the CRT group if he knows what causes
it.

Dennis Tillman W7PF

-----Original Message-----
From: TekScopes@... [mailto:TekScopes@...]
Sent: Friday, December 19, 2014 8:55 AM
To: TekScopes@...
Subject: [TekScopes] 7904 trace dimming

Hello All,

I am noticing that as my 7904 is warming, the trace dims. It does this
about 3-5 mins after power up and does so all at once. In other words, it
is not gradual dimming. I need to increase the intensity by one mark on the
knob scale in order to return the trace to similar intensity as before it
dimmed. It does not dim beyond that throughout the day after this initial
"dimming."

Has anyone else experienced this?
Just tinkering in the lab.............

Cheers,
David

------------------------------------
Posted by: d.garrido@...
------------------------------------


 

Many of mine change in brightness as they warm up but it is gradual and stops
after a few minutes.

If it happens all at once, then I would suspect a high voltage part in the
z-axis DC restorer like a diode or capacitor has marginal leakage and may be
close to failing. I think it would be easier to just do a general replacement
of those high voltage parts than trying to find which one is causing the
problem.

Before doing that I would test with a different timebase and do the z-axis
calibration procedures. If you want to make a measurement, check the low
voltage DC level at the output of the z-axis amplifier with a voltmeter and see
if it shifts when the brightness changes all at once.

I would not use freeze spray on the z-axis circuits because of condensation and
high voltage.

Another thing I might consider after the above tests is removing the z-axis
board and cleaning it thoroughly but if I did, then I would change the high
voltage diodes and capacitors anyway.

On 19 Dec 2014 08:55:18 -0800, you wrote:

Hello All,

I am noticing that as my 7904 is warming, the trace dims. It does this about 3-5 mins after power up and does so all at once. In other words, it is not gradual dimming. I need to increase the intensity by one mark on the knob scale in order to return the trace to similar intensity as before it dimmed. It does not dim beyond that throughout the day after this initial "dimming."

Has anyone else experienced this?

Just tinkering in the lab.............

Cheers,

David


 

A simpler fault would be an unstable trimpot somewhere in the Z-axis amplifier itself. Such was the case in my 7854. Intensity changed suddenly after warming up and often also after switching between display modes (analog or stored waveform). I too would first of all perform a Z-axis calibration and readjust the settings back and forth even when they seem on spec.

Albert


---In TekScopes@..., <davidwhess@...> wrote :

Many of mine change in brightness as they warm up but it is gradual and stops
after a few minutes.

If it happens all at once, then I would suspect a high voltage part in the
z-axis DC restorer like a diode or capacitor has marginal leakage and may be
close to failing. I think it would be easier to just do a general replacement
of those high voltage parts than trying to find which one is causing the
problem.

Before doing that I would test with a different timebase and do the z-axis
calibration procedures. If you want to make a measurement, check the low
voltage DC level at the output of the z-axis amplifier with a voltmeter and see
if it shifts when the brightness changes all at once.
----


 

Hi David,

Here is what the head of Tek's Laboratories (he also designed the vertical
deflection plates in the 7104 CRT) had to say:

"Dennis: I turned on both my 7904 and SC503. In a ten minute span I did
not see any change in the trace brightness. So I don't have any evidence
that there is such a drop -- at least on these two scopes. There was
nothing that we ever recommended from the CRT group that would have led to
such a design feature. Maybe it's caused by a noisy intensity control."

Dennis Tillman W7PF

-----Original Message-----
From: TekScopes@... [mailto:TekScopes@...]
Sent: Friday, December 19, 2014 9:56 AM
To: TekScopes@...
Subject: RE: [TekScopes] 7904 trace dimming

Hi David,

I have seen that on other 7000 scopes and even on some of the small TM500
scopes (SC501, SC502, SC503, SC504, and MR501). It is subtle. It was the
subject of some comments on the forum many, many years ago.

I will ask a friend of mine who was in the CRT group if he knows what causes
it.

Dennis Tillman W7PF

-----Original Message-----
From: TekScopes@... [mailto:TekScopes@...]
Sent: Friday, December 19, 2014 8:55 AM
To: TekScopes@...
Subject: [TekScopes] 7904 trace dimming

Hello All,

I am noticing that as my 7904 is warming, the trace dims. It does this
about 3-5 mins after power up and does so all at once. In other words, it
is not gradual dimming. I need to increase the intensity by one mark on the
knob scale in order to return the trace to similar intensity as before it
dimmed. It does not dim beyond that throughout the day after this initial
"dimming."

Has anyone else experienced this?
Just tinkering in the lab.............

Cheers,
David

------------------------------------
Posted by: d.garrido@...
------------------------------------