2465 no traces, no OSD
Hi y'all, I just bought a 2467 on eBay. It came with option 9 - which I understand is the counter/timer? Unfortunately it doesn't seem to light up the CRT at all, no traces, no OSD.
On power-up, it does seem to go into diagnosis, as I have to hit one specific front-panel control before it'll respond to any others. After this, I can mess with the front panel controls, and e.g. the channel/trigger lights all toggle with their corresponding buttons. However, any time I change any setting, all the front panel LEDs flicker. Also, when I set channels 1/2 to the 50 Ohm input setting, they stick for about a second or so, but then I hear a relay toggle again, and the GND input lights up again.
Do those symptoms ring a bell with anyone here? I took it apart and pulled the PSU?to see whether there's anything obviously broken. It looks to me like the PSU has been re-capped, as there's a mix of colors on there. I tested all the rails, and they look about right, though I couldn't load the 5V with more than a single 7.5 ohm resistance at a time. When I loaded J303, I saw an undershoot for a second or two, then it seemed to drop into regulation at ~4.9V. The 7.5 ohm on J232 didn't quite seem to be enough load, though the voltage dropped a fair bit when I did it. The 10V reference was spot on, as were e.g. the 87 and the +-15 rails. The others were all high, but then they were unloaded... I guess next I'll measure the rails with the PSU in-place, check ripple etc? Are there any pictures here of 2467/2465A PSUs in original condition that I might use for comparison?
Are there other quick things I can look at to assess the damage/hopes of repair? I understand the 2467 has a "screen saver" that'll shut down the CRT on a timer? Presumably this is under digital control - so I'm hopeful that the CRT or control circuitry isn't beyond rescue?
I have a functional 2465 I was hoping to replace with this one, so I can have peek around. Siggi |
The very first thing I would do is check all voltages and ripple on J119.? There is a table of specifications for this in the service manual section for power supply trouble shooting, just before the flow chart sections 10 and 11.? If all J119 voltages and ripple are right I would use the working 2465 you have to check signals going to the CRT.? Especially check the unblanking signals, VZOUT and VGOUT, the prescribed waveforms are given in the service manual (waveforms 65 and 66 in the 2465 Service manual).??If there is activity closely resembling the prescribed activity in shape and, especially, amplitude the scope is trying to drive the CRT.? The problem is very likey the high voltage board (A9) or the CRT
itself. If you have a high voltage probe you can just measure the voltage on the CRT anode pole.? But don't do this if you?lack experience with high voltage measurement techniques as the voltage here is 14,000 volts or more. It sounds to me that the scope is trying to operate?and there are some good indications.? When there?are faults known to the scopes computer (board A5) it presents an error message on the CRT ("Test xxx fail? Push A/B trigger") and the scope then tries to operate.? But since your screen is blank?you cannot see this message.? If there were high voltage and a VZ signal you should at least see a bright spot on the CRT.? Since you cannot, either there is no high voltage, no VZ, or the CRT may not be lighting up?(no filament or just a?bad CRT). You can go to the Qservice web site and see pictures of every
subassembly for reference of how they should look.? Tektronix produced these scopes for several years and used different generations of capacitors.? So boards will look different from generation to generation but the overall layout will remain the same and there whould be uniformity within a particular build. From: Sigur?ur ?sgeirsson
To: TekScopes@... Sent: Sunday, January 26, 2014 11:38 AM Subject: [TekScopes] 2465 no traces, no OSD
?
Hi y'all,
I just bought a 2467 on eBay. It came with option 9 - which I understand is the counter/timer? Unfortunately it doesn't seem to light up the CRT at all, no traces, no OSD.
On power-up, it does seem to go into diagnosis, as I have to hit one specific front-panel control before it'll respond to any others. After this, I can mess with the front panel controls, and e.g. the channel/trigger lights all toggle with their corresponding buttons. However, any time I change any setting, all the front panel LEDs flicker. Also, when I set channels 1/2 to the 50 Ohm input setting, they stick for about a second or so, but then I hear a relay toggle again, and the GND input lights up again.
Do those symptoms ring a bell with anyone here? I took it apart and pulled the PSU?to see whether there's anything obviously broken. It looks to me like the PSU has been re-capped, as there's a mix of colors on there. I tested all the rails, and they look about right, though I couldn't load the 5V with more than a single 7.5 ohm resistance at a time. When I loaded J303, I saw an undershoot for a second or two, then it seemed to drop into regulation at ~4.9V. The 7.5 ohm on J232 didn't quite seem to be enough load, though the voltage dropped a fair bit when I did it. The 10V reference was spot on, as were e.g. the 87 and the +-15 rails. The others were all high, but then they were unloaded... I guess next I'll measure the rails with the PSU in-place, check ripple etc? Are there any pictures here of 2467/2465A PSUs in original condition that I might use for comparison?
Are there other quick things I can look at to assess the damage/hopes of repair? I understand the 2467 has a "screen saver" that'll shut down the CRT on a timer? Presumably this is under digital control - so I'm hopeful that the CRT or control circuitry isn't beyond rescue?
I have a functional 2465 I was hoping to replace with this one, so I can have peek around. Siggi |
Bob Koller
If the power supply seems OK, just turn up the CRT Bias pot and see if it lights up. A common failure is the Z-Axis hybrid, which, if bad, will cause no display. If a display appears, you may be able to see which self test is failing, and take it from there. On Sunday, January 26, 2014 10:10 AM, machine guy wrote:
?
The very first thing I would do is check all voltages and ripple on J119.? There is a table of specifications for this in the service manual section for power supply trouble shooting, just before the flow chart sections 10 and 11.? If all J119 voltages and ripple are right I would use the working 2465 you have to check signals going to the CRT.? Especially check the unblanking signals, VZOUT and VGOUT, the prescribed waveforms are given in the service manual (waveforms 65 and 66 in the 2465 Service manual).??If there is activity closely resembling the prescribed activity in shape and, especially, amplitude the scope is trying to drive the CRT.? The problem is very likey the high voltage board (A9) or the
CRT
itself. If you have a high voltage probe you can just measure the voltage on the CRT anode pole.? But don't do this if you?lack experience with high voltage measurement techniques as the voltage here is 14,000 volts or more. It sounds to me that the scope is trying to operate?and there are some good indications.? When there?are faults known to the scopes computer (board A5) it presents an error message on the CRT ("Test xxx fail? Push A/B trigger") and the scope then tries to operate.? But since your screen is blank?you cannot see this message.? If there were high voltage and a VZ signal you should at least see a bright spot on the CRT.? Since you cannot, either there is no high voltage, no VZ, or the CRT may not be lighting up?(no filament or just a?bad CRT). You can go to the
Qservice web site and see pictures of every
subassembly for reference of how they should look.? Tektronix produced these scopes for several years and used different generations of capacitors.? So boards will look different from generation to generation but the overall layout will remain the same and there whould be uniformity within a particular build. From: Sigur?ur ?sgeirsson
To: TekScopes@... Sent: Sunday, January 26, 2014 11:38 AM Subject: [TekScopes] 2465 no traces, no OSD
?
Hi y'all,
I just bought a 2467 on eBay. It came with option 9 - which I understand is the counter/timer? Unfortunately it doesn't seem to light up the CRT at all, no traces, no OSD.
On power-up, it does seem to go into diagnosis, as I have to hit one specific front-panel control before it'll respond to any others. After this, I can mess with the front panel controls, and e.g. the channel/trigger lights all toggle with their corresponding buttons. However, any time I change any setting, all the front panel LEDs flicker. Also, when I set channels 1/2 to the 50 Ohm input setting, they stick for about a second or so, but then I hear a relay toggle again, and the GND input lights up again.
Do those symptoms ring a bell with anyone here? I took it apart and pulled the PSU?to see whether there's anything obviously broken. It looks to me like the PSU has been re-capped, as there's a mix of colors on there. I tested all the rails, and they look about right, though I couldn't load the 5V with more than a single 7.5 ohm resistance at a time. When I loaded J303, I saw an undershoot for a second or two, then it seemed to drop into regulation at ~4.9V. The 7.5 ohm on J232 didn't quite seem to be enough load, though the voltage dropped a fair bit when I did it. The 10V reference was spot on, as were e.g. the 87 and the +-15 rails. The others were all high, but then they were unloaded... I guess next I'll measure the rails with the PSU in-place, check ripple etc? Are there any pictures here of 2467/2465A PSUs in original condition that I might use for comparison?
Are there other quick things I can look at to assess the damage/hopes of repair? I understand the 2467 has a "screen saver" that'll shut down the CRT on a timer? Presumably this is under digital control - so I'm hopeful that the CRT or control circuitry isn't beyond rescue?
I have a functional 2465 I was hoping to replace with this one, so I can have peek around. Siggi |
--- In TekScopes@..., machine guy <machineguy59@...> wrote:
The rails all look good, nicely inside specs and nary a ripple in sight. If all J119 voltages and ripple are right I would use the working 2465 you have to check signals going to the CRT.?? Especially check the unblanking signals, VZOUT and VGOUT, the prescribed waveforms are given in the service manual (waveforms 65 and 66 in the 2465 Service manual).????If there is activity closely resembling the prescribed activity in shape and, especially, amplitude the scope is trying to drive the CRT.?? The problem is very likey the high voltage board (A9) or the CRT itself.Ugh, I see I mistitled my post - the scope's a 2467, though I suspect it's much the same at the level I'm grousing? I had some trouble orienting myself between the scaled-down printout of the schematic the poor lighting where I'm trying to do this. There are cutouts in the main PCB where it feeds off to the CRT - is it the aftermost one of those that's VZOUT/VGOUT? The signal on there didn't look altogether alien, though the levels didn't look right - I'll have to look again in the light of day... If you have a high voltage probe you can just measure the voltage on the CRT anode pole.?? But don't do this if you??lack experience with high voltage measurement techniques as the voltage here is 14,000 volts or more.Thanks - it starting to look like a terminal failure to me (for someone of my limited skill and knowledge) - alas :(. You can go to the Qservice web site and see pictures of every subassembly for reference of how they should look.?? Tektronix produced these scopes for several years and used different generations of capacitors.?? So boards will look different from generation to generation but the overall layout will remain the same and there whould be uniformity within a particular build. |
--- In TekScopes@..., Bob Koller <testtech@...> wrote:
Sadly no joy on cranking the bias pot up. Thanks for the hint, though.
|
--- In TekScopes@..., "tinkerer138" <siggi@...> wrote:
I take that back - turns out I was messing with the MCP bias which isn't too helpful. When I cranked up the CRT bias I got an error message display - very exciting - thanks! "TEST 05 FAIL 42". Apparently this is a line trigger self test. Given that I'm getting a picture, the CRT and the high-voltage supply are presumably OK. Can I also assume the horizontal/vertical/Z machinery is OK? Looking at pictures of original PSUs, I'm now thinking this PSU is pristine - e.g. hasn't been re-capped. Perhaps I'll take another hard look at the rails, it'd be great if it were that simple. I seem to recall a ~10mV spiky "beat" on one of the 5V rails presumably the analog - I'll have to look closer tonight. The other 5V rail was pretty fuzzy, but only around 5-10mV. I did see it take a little bit of a wiggle when I mess with the front panel controls, though I'm hard put to quantify the amount. Also, between the front panel LED flicker and the 50 Ohm input setting reverting, I'm thinking this might be (or at least manifest as) a control issue. While the LED shift registers are non-latched, I don't think I ought to be seeing noticable flicker on every LED each time they update? The 50Ohm termination resistors are monitored for overheating, and presumably the toggle-back means the CPU's "seeing" them warm. So if the rails are really checking out, I may start looking back from the thermistors toward the CPU. Any other ideas? |
This is my first group post, so don't hesitate to correct my citing style. I'm using yahoo's online web post function. Regarding your problem: I think I had similar problems (flickering, 50 Ohm Overload) with my 2440 as I bought it. There was a 0.1 Ohm overcurrent sensing resistor in the 5V supply which changed its value over time (0.11 Ohm @ 5% tolerance). It sits between two heatsinks, that may have caused the massive change in value. It looked also burned. Good luck! |
--- In TekScopes@..., <schneidertim541@...> wrote:
Me too... Regarding your problem: I think I had similar problems (flickering, 50 Ohm Overload) with my 2440 as I bought it. There was a 0.1 Ohm overcurrent sensing resistor in the 5V supply which changed its value over time (0.11 Ohm @ 5% tolerance). It sits between two heatsinks, that may have caused the massive change in value. It looked also burned.That's very interesting, thanks. I see one of those on the +5V line for the analog supply schematic. |
GREAT NEWS. Before starting a recap I suggest you find the source of these problems.? Otherwise, any mistake made during recap will make it necessary to find two problems at one time.? That is much harder than two problems at two times. The fact that you are getting a picture says much of the scope is working (Horiz, Vert, Z, HV).? But it doesn't say they are completely OK, just that your priorities can/should be other features. Faulty power supplies can cause a variety of misleading symptoms but I don't think that is what is happening here.? Double check the voltages on J119 and if they are reasonably close move on for now.? Tek gives specs for an instrument that does
laboratory quality measurements so small amounts of ripple are detracting.? But you aren't looking for a small error.? I suggest you find that first unless voltages and ripple are far out of spec and/or changing during operation. It sounds like you may have a control board (A5) issue.? Since the case is open you can easily do a visual check of the A5 board.? Look for leaking or bulging caps and check the date on the memory chip and its battery.? The 246x scopes (except for the 2465 and 2445) used a battery backed ram to save memory contents.? These batteries are known to fail after many years and must be replaced before all memory is lost.? There is a procedure to do this replacement.? Its necessary to provide temporary power to the memory chip while you remove the battery and connect a new one.? Even a momentary loss of power to the chip will lose all contents and a full calibration
will be necessary.? A failing A5 or its memory could explain all the problems you are seeing.? I suggest you?focus on that for now.? From: tinkerer138
To: TekScopes@... Sent: Tuesday, January 28, 2014 7:40 AM Subject: [TekScopes] Re: 2465 no traces, no OSD
?
--- In TekScopes@..., "tinkerer138" wrote:
> > --- In TekScopes@..., Bob Koller wrote: > > > > If the power supply seems OK, just turn up the CRT Bias pot and see if it lights up. A common failure is the Z-Axis hybrid, which, if bad, will cause no display. If a display appears, you may be able to see which self test is failing, and take it from there. > > > > Sadly no joy on cranking the bias pot up. Thanks for the hint, though. > I take that back - turns out I was messing with the MCP bias which isn't too helpful. When I cranked up the CRT bias I got an error message display - very exciting - thanks! "TEST 05 FAIL 42". Apparently this is a line trigger self test. Given that I'm getting a picture, the CRT and the high-voltage supply are presumably OK. Can I also assume the horizontal/vertical/Z machinery is OK? Looking at pictures of original PSUs, I'm now thinking this PSU is pristine - e.g. hasn't been re-capped. Perhaps I'll take another hard look at the rails, it'd be great if it were that simple. I seem to recall a ~10mV spiky "beat" on one of the 5V rails presumably the analog - I'll have to look closer tonight. The other 5V rail was pretty fuzzy, but only around 5-10mV. I did see it take a little bit of a wiggle when I mess with the front panel controls, though I'm hard put to quantify the amount. Also, between the front panel LED flicker and the 50 Ohm input setting reverting, I'm thinking this might be (or at least manifest as) a control issue. While the LED shift registers are non-latched, I don't think I ought to be seeing noticable flicker on every LED each time they update? The 50Ohm termination resistors are monitored for overheating, and presumably the toggle-back means the CPU's "seeing" them warm. So if the rails are really checking out, I may start looking back from the thermistors toward the CPU. Any other ideas? |
--- In TekScopes@..., machine guy <machineguy59@...> wrote:
:) Faulty power supplies can cause a variety of misleading symptoms but I don't think that is what is happening here.?? Double check the voltages on J119 and if they are reasonably close move on for now.?? Tek gives specs for an instrument that does laboratory quality measurements so small amounts of ripple are detracting.?? But you aren't looking for a small error.?? I suggest you find that first unless voltages and ripple are far out of spec and/or changing during operation.I checked this out again, and it's well inside spec. The ripple is "ugly" though, e.g. there are sharp spikes at ~20/40kHz. Assuming I get it working, I'll be planning to do the recap. One surprising thing I found on a closer look; after I acknowledge the initial fault, something starts injecting about 40mV of a ~5MHz clock signal onto my 10V reference. Thankfully this doesn't propagate to the other rails though. Interesting, thanks. I'm not seeing anything obvious to my eye, though I should probably take a harder look. A failing A5 or its memory could explain all the problems you are seeing.?? I suggest you??focus on that for now.??Thanks, this looks promising (not to mention fun:). One thing I forgot to mention is that the scope doesn't trigger on the calibration signal, although the signal itself is fine viewed on my other scope. Staring at the circuit diagram, I figured many of the symptoms could be explained by a DAC or analog MUX failure. It looks as though the DAC/MUX is to the center of e.g. reading all analog signals, setting trigger levels, etc. Working through the A5 test points, I find the CPU clock and the interrupt signals are just fine, as expected. The DAC-derived reference(?) voltages at TP2420 and TP2421 are way off, however, so I've something to look into there. I have to admit that I can't for the life of me figure out where the -1.25V reference comes from, though. It appears to be under CPU control through the DAC/MUX, but it's also used to derive the +1.36V reference, which in turn is the centerpoint of the DAC's I->V conversion resistors (I'm a software guy, this is all alien territory to me). Anyways, I guess working forward from the DAC's reference is something to do :). |
> I checked this out again, and it's well inside spec. The ripple is "ugly" though, e.g. there are sharp spikes at ~20/40kHz. Assuming I get it working, I'll be planning to do the recap. That is the normal sawtooth of a switching power supply.? As long as its low in amplitude you can ignore it and come back later. > One surprising thing I found on a closer look; after I acknowledge the initial fault, something starts injecting about 40mV of a ~5MHz clock signal onto my 10V reference. Thankfully this doesn't propagate to the other rails though. That sounds like the microprocessor clock.? 40 mv is small enough to ignore for now but
I don't think it should be there.? It may be a clue to the erratic behavior you see on the controls. >Thanks, this looks promising (not to mention fun:). Are we having fun yet?? lol.? I wont be much help to you with the control board.? I know the generalities but my scope is a 2465 and I haven't actually worked on a 2467.? I hope someone with more experience here will jump in with ideas.? But the Tektronix Service Manual is very good on these scopes.? There?are Diagnostic Routines you can command from the front panel.??The service manual tells how to activate them and gives details of what tests are appropriate for your series of scope.? For my scope you activate the Diagnostic Routine menu by holding deltaT and delta V buttons then pressing Slope button on the front panel.? I would run them and see how it does.? Again, we are
seeking easy information and others on this forum may have suggestions on which of these tests are important here. >One thing I forgot to mention is that the scope doesn't trigger on the calibration signal, although the signal itself is fine viewed on my other scope. Faulty trigger will have to be tracked down and repaired? But that should not cause the erratic panel operation you described before.? I would put that on hold until you get?stable operation.? One possibility is that trigger logic is controlled by the same computer that controls the panel.? Your trigger problems may clear up once the panel is stable.? The computer board (A5) sends controls that eventually reach?the trigger hybrid (U500) so you can trace these command lines with your test scope.? But, as I said,?check the processor first. >I have to admit that I can't for the life of me figure out where the -1.25V reference comes from? Sorry, I cant help you here since I have a different processor board and a different DAC configuration.? But in my 2465 scope the DAC voltage reference is derived by voltage divider from the +10 V reference from the Low Voltage Power Supply.? I will bet yours is similar.? Tek used a simple?resistor divider to set the voltages for the DAC input.? It has a 47 uf cap to filter is at the DAC input. > (I'm a software guy, this is all alien territory to me).? I did Hardware, then software, when they found out I failed at both they made me a manager. From: tinkerer138
To: TekScopes@... Sent: Wednesday, January 29, 2014 8:42 AM Subject: [TekScopes] Re: 2465 no traces, no OSD
?
--- In TekScopes@..., machine guy wrote:
> > GREAT NEWS. :) > Faulty power supplies can cause a variety of misleading symptoms but I don't think that is what is happening here.?? Double check the voltages on J119 and if they are reasonably close move on for now.?? Tek gives specs for an instrument that does laboratory quality measurements so small amounts of ripple are detracting.?? But you aren't looking for a small error.?? I suggest you find that first unless voltages and ripple are far out of spec and/or changing during operation. I checked this out again, and it's well inside spec. The ripple is "ugly" though, e.g. there are sharp spikes at ~20/40kHz. Assuming I get it working, I'll be planning to do the recap. One surprising thing I found on a closer look; after I acknowledge the initial fault, something starts injecting about 40mV of a ~5MHz clock signal onto my 10V reference. Thankfully this doesn't propagate to the other rails though. > > It sounds like you may have a control board (A5) issue.?? Since the case is open you can easily do a visual check of the A5 board.?? Look for leaking or bulging caps and check the date on the memory chip and its battery.?? The 246x scopes (except for the 2465 and 2445) used a battery backed ram to save memory contents.?? These batteries are known to fail after many years and must be replaced before all memory is lost.?? There is a procedure to do this replacement.?? Its necessary to provide temporary power to the memory chip while you remove the battery and connect a new one.?? Even a momentary loss of power to the chip will lose all contents and a full calibration will be necessary.?? > Interesting, thanks. I'm not seeing anything obvious to my eye, though I should probably take a harder look. > A failing A5 or its memory could explain all the problems you are seeing.?? I suggest you??focus on that for now.?? > Thanks, this looks promising (not to mention fun:). One thing I forgot to mention is that the scope doesn't trigger on the calibration signal, although the signal itself is fine viewed on my other scope. Staring at the circuit diagram, I figured many of the symptoms could be explained by a DAC or analog MUX failure. It looks as though the DAC/MUX is to the center of e.g. reading all analog signals, setting trigger levels, etc. Working through the A5 test points, I find the CPU clock and the interrupt signals are just fine, as expected. The DAC-derived reference(?) voltages at TP2420 and TP2421 are way off, however, so I've something to look into there. I have to admit that I can't for the life of me figure out where the -1.25V reference comes from, though. It appears to be under CPU control through the DAC/MUX, but it's also used to derive the +1.36V reference, which in turn is the centerpoint of the DAC's I->V conversion resistors (I'm a software guy, this is all alien territory to me). Anyways, I guess working forward from the DAC's reference is something to do :). |
--- In TekScopes@..., machine guy <machineguy59@...> wrote:
Hell yes!Thanks, this looks promising (not to mention fun:).Are we having fun yet??? lol.?? I wont be much help to you with the control board.No worries, you've set me on the right path, and I think I may be able to sort out at least what I've discovered so far. I'd hope noone here minds if I keep a "diary" of what I find as I go along. I'm hoping this might help someone else later, as I've been helped by others lore and repair logs. Last night I found that VR2420 was shorted, and was pulling the -1.25 reference to ground at ~10Ohm. This is a 3V zener that's presumably there to clamp the reference to +.7- -3V. It looks like whatever took out the zener also burned U2420D (1/4 TL074), as when I removed the zener, the rail went to -5V, irrespective the drive on the input. So, whatever happened there may have been a fairly violent event. I'm guessing downstream from the -1.25V rail from A5 I may find more trouble. I'll have to buy some parts and replace the op-amp, see what happens next. |
>Last night I found that VR2420 was shorted You are definitely on the right path.? The flashing LED and variations when you select a control points to the control board.? The DAC reference is a key factor in the control logic.? But it puzzles my why it would fail in such fashion.? I would look around for physical damage by a clumsy or unskilled technician.? No matter, you know what to do and it may be the major fault here. From: tinkerer138
To: TekScopes@... Sent: Thursday, January 30, 2014 8:15 AM Subject: [TekScopes] Re: 2465 no traces, no OSD
?
--- In TekScopes@..., machine guy wrote:
> >Thanks, this looks promising (not to mention fun:). > Are we having fun yet??? lol.?? Hell yes! > I wont be much help to you with the control board. No worries, you've set me on the right path, and I think I may be able to sort out at least what I've discovered so far. I'd hope noone here minds if I keep a "diary" of what I find as I go along. I'm hoping this might help someone else later, as I've been helped by others lore and repair logs. Last night I found that VR2420 was shorted, and was pulling the -1.25 reference to ground at ~10Ohm. This is a 3V zener that's presumably there to clamp the reference to +.7- -3V. It looks like whatever took out the zener also burned U2420D (1/4 TL074), as when I removed the zener, the rail went to -5V, irrespective the drive on the input. So, whatever happened there may have been a fairly violent event. I'm guessing downstream from the -1.25V rail from A5 I may find more trouble. I'll have to buy some parts and replace the op-amp, see what happens next. |
Last night I found that VR2420 was shorted, and was pulling the -1.25 reference to ground at ~10Ohm. This is a 3V zener that's presumably there to clamp the reference to +.7- -3V.PROGRESS! After replacing VR2420 and U2420, I have OSD, though I'm still getting "TEST 05 FAIL 42" on startup. The 50Ohm input setting revert is also fixed. This makes sense as the -1.25V is used for the overload sense. In hindsight, it was probably not necessary to replace U2420, as it seems the -1.25 does not drop into spec while the A5 board is isolated. Presumably it's set by measuring it against something else on startup. The -1.25 voltage is pretty near the mark (-1.251 according to my Fluke), and the +1.36 measures at +1.37, which I hope is near enough. If I'm reading this right, any deviation here can be (presymably has been) trimmed at the DAC reference trimmer. As a bonus problem the beam find switch is flaky, so I spent a good 20 minutes trying to bring the compressed OSD into focus by playing with the adjustment trimmers before I caught on :/. I'm also getting traces, but only sporadically - even on auto, and she's glitchy as hell. The "C" in the "DIAGNSTIC" flickers, and spurious pixels appear here and there on the display with the same kind of frequency. When I manage to get an A/B trace of the calibration signal, has weird fading, which likely means I need to re-tune the Z response. I've seen talk of corrosion on the hybrids causing this sort of symptom, and there's certainly plenty of corrosion on the BNC connectors, pots and switches. Looks like the poor thing was warehoused somewhere terrible, probably for years. I looked at the operation timer, and she has ~3600 hours on her in ~650 power cycles. This doesn't sound like a whole lot of time, and there's certainly no visible burn-in on the CRT. I guess the next thing I'll look at is to pop one of the hybrids, probably whichever one is responsible for triggering. Is there a recommended product/procedure for de-glitching switches, pots and hybrids? |
I'm also getting traces, but only sporadically - even on auto, and she's glitchy as hell. The "C" in the "DIAGNSTIC" flickers, and spurious pixels appear here and there on the display with the same kind of frequency.After pulling the trigger/sweep/horizontal/Z hybrids (U700, U600, U900, U950), cleaning the contacts and sockets with isopropyl alcohol, the glitching in the triggering and in the sweeps is gone. I now have four glorious, glitch free traces with attenuation, A/B sweeps, the works. Looks like all the important parts of her are viable :). I only get "normal" triggering though, no auto sweeps and no auto trigger level, which is presumably related to the 05/42 error. The first hybrid I treated I had to re-do, as I didn't wipe the contacts quite vigorously enough, or perhaps it was the wriggling during re-seating it that did it. I'll have to take a look at the readout board and see whether I can fix the (slight remaining) OSD glitching by cleaning contacts or the like. Next up is to calibrate the ADC to see whether that'll fix the 05-42 error. After that I'm going to re-calibrate the CRT, as I've clearly messed it up while I was trying to focus the OSD on beamfinder :/. |