Tek 422 HV power supply issues


 

So I'm looking at this 422 and the trace doesn't get visible until the
intensity control is around 3 o'clock. Blanking is adjusted, so
that's not the problem

This doesn't seem right, so I dug out the HV probe and took some
measurements.

The Accelerating potential is spot on at +4900.

But all of the negative voltages are just a tad off. The high voltage test point
at G1 on the schematic (labeled -1420 or -1400 depending on which side of the
line you're looking at) never gets below -1396. (At midrange it's -1390)

With the intensity at midrange, the cathode is -1360, not -1375 as spec'd on
the schematic.

Right now at the end of the intensity control, I get a range from -1375 to -1390,
so it looks like 10 or 15V can make a difference.

I hate working on HV supplies, but there's not a lot of parts in this one.
Anyone have any experience with this one? I'd appreciate some pointers.

Paul
--
Paul Amaranth, GCIH | Rochester MI, USA
Aurora Group, Inc. | Security, Systems & Software
paul@... | Unix & Windows


 

Hi Paul --
Check the resistors in the series string where the focus & intensity pots
live. Your voltages aren't that far off the mark. Per the schematic,
'normal' should be a bias of around -45 V. If your difference is less
and the intensity is still weak, you may be looking at a crt.
-ls-

Paul Amaranth <paul@...> wrote:

So I'm looking at this 422 and the trace doesn't get visible until the
intensity control is around 3 o'clock. Blanking is adjusted, so
that's not the problem

This doesn't seem right, so I dug out the HV probe and took some
measurements.

The Accelerating potential is spot on at +4900.

But all of the negative voltages are just a tad off. The high voltage test point
at G1 on the schematic (labeled -1420 or -1400 depending on which side of the
line you're looking at) never gets below -1396. (At midrange it's -1390)

With the intensity at midrange, the cathode is -1360, not -1375 as spec'd on
the schematic.

Right now at the end of the intensity control, I get a range from -1375 to -1390,
so it looks like 10 or 15V can make a difference.

I hate working on HV supplies, but there's not a lot of parts in this one.
Anyone have any experience with this one? I'd appreciate some pointers.

Paul
--
Paul Amaranth, GCIH | Rochester MI, USA
Aurora Group, Inc. | Security, Systems & Software
paul@... | Unix & Windows



------------------------------------

Yahoo! Groups Links



 

Resistors looked more or less OK, that was one of the first things
I checked. The 2 2.2M in series had drifted down to 4.2M, while the
1M focus was up to 1.2M, so more or less a wash.

More measurements, here's the bias range:

Intensity: Min Midrange High
G1 bias -95 -42 -6

I don't have any experience with this measurment, is this reasonable?
Is this pointing to low emision on the CRT?

On Wed, Feb 20, 2013 at 09:54:01PM -0500, larrys@... wrote:
Hi Paul --
Check the resistors in the series string where the focus & intensity pots
live. Your voltages aren't that far off the mark. Per the schematic,
'normal' should be a bias of around -45 V. If your difference is less
and the intensity is still weak, you may be looking at a crt.
-ls-

Paul Amaranth <paul@...> wrote:
So I'm looking at this 422 and the trace doesn't get visible until the
intensity control is around 3 o'clock. Blanking is adjusted, so
that's not the problem
--
Paul Amaranth, GCIH | Rochester MI, USA
Aurora Group, Inc. | Security, Systems & Software
paul@... | Unix & Windows


 

A little more investigation showed the blanking levels did not
match the waveform diagrams; putting a meter on the +55 showed
+66, so now it's on to the LV ps.

Anyone have a schematic or link to a 422 AC Only PS? I have
a 016-0072-00 The online manuals I've found so far only seem to have
the AC/DC supply.

And I thought this was just going to be a quick clean up and
toss on the shelf.

Paul
--
Paul Amaranth, GCIH | Rochester MI, USA
Aurora Group, Inc. | Security, Systems & Software
paul@... | Unix & Windows


 

Paul,

I uploaded "Tektronix 422 Oscilloscope AC Power Supply Schematic" to the ko4bb.com site. Check there under the Manuals&#92;Tektronix directory. (It might get moved into the Manuals&#92;Tektronix&#92;422 directory.)

--John Gord

--- In TekScopes@..., Paul Amaranth <paul@...> wrote:

A little more investigation showed the blanking levels did not
match the waveform diagrams; putting a meter on the +55 showed
+66, so now it's on to the LV ps.

Anyone have a schematic or link to a 422 AC Only PS? I have
a 016-0072-00 The online manuals I've found so far only seem to have
the AC/DC supply.

And I thought this was just going to be a quick clean up and
toss on the shelf.

Paul
--