I have an ancient 475 scope which works fine with an external trigger, but will not trigger on any of the internal sources. Should I start by reviewing the recent 468 triggering issue thread to get me going? If not, can anyone recommend a thread I should start with? Of course, I'd be delighted to receive help on this thread.
Stan
|
The 475 trigger pickoff circuit is significantly different than the 468 circuit. From your description, something is wrong with or around U370 on schematic 3. Take a close look at R403 which if open would affect triggering of both channels. X-Y mode and the channel 2 vertical signal out on the rear of the oscilloscope also depend on biasing from R403 so checking those functions may provide more information. On Fri, 25 Jan 2013 20:46:14 -0000, "stan_katz" <stan_katz@...> wrote: I have an ancient 475 scope which works fine with an external trigger, but will not trigger on any of the internal sources. Should I start by reviewing the recent 468 triggering issue thread to get me going? If not, can anyone recommend a thread I should start with? Of course, I'd be delighted to receive help on this thread.
Stan
|
On 01/25/2013 03:46 PM, stan_katz
wrote:
?
I have an ancient 475 scope which works fine with an
external trigger, but will not trigger on any of the
internal sources. Should I start by reviewing the recent
468 triggering issue thread to get me going? If not, can
anyone recommend a thread I should start with? Of course,
I'd be delighted to receive help on this thread.
Stan
Probably not.
Going from what I had on 465Bs the triggering problem is often that
the tunnel diodes have changed in value.
There are many! threads about this in the archives.
It may be worth your time to try swapping the TDs from the B sweep
section
with the A section to see if things change.
I had this problem come up when I was hunting for a process control
network
problem while working in factory in a different country.
I swapped the TDs and the scope has been near perfect ever since.
Go figure.
|
If you do work on the tunnel diodes, be sure to use a heat sink right up
against the diode body. Ge tunnel diodes are easily ruined with heat.
?
However, I think David's comments are more likely the issue. That is
because of the external trigger working.
?
Do you have the correct service manual for your 475?
?
Regards,
tom
?
toggle quoted message
Show quoted text
----- Original Message -----
Sent: Friday, January 25, 2013 5:09
PM
Subject: Re: [TekScopes] 475 Triggering
Issue -- Will recent 468 Triggering Issue Thread Help Me?
?
On 01/25/2013 03:46 PM, stan_katz wrote:
?
I have an ancient 475 scope which works fine with an external trigger,
but will not trigger on any of the internal sources. Should I start by
reviewing the recent 468 triggering issue thread to get me going? If not,
can anyone recommend a thread I should start with? Of course, I'd be
delighted to receive help on this
thread.
Stan
Probably not. Going from
what I had on 465Bs the triggering problem is often that the tunnel diodes
have changed in value. There are many! threads about this in the
archives. It may be worth your time to try swapping the TDs from the B
sweep section with the A section to see if things change. I had this
problem come up when I was hunting for a process control network problem
while working in factory in a different country. I swapped the TDs and the
scope has been near perfect ever since. Go figure.
|
Hello Tom,
I will gladly avoid touching the tunnel diodes, and will start with David's suggestion. The only service manuals I have are from BAMA, which are the Army TM--- model manuals. Will these do, or can you point me in a better direction? BTW, Bert's post is a warning to me that the 475 has potentially unobtanium parts. Right now, Tucker has Tek tunnel diode kits for sale. Can you, or Bert, post the proper kit part number for my scope? I should buy a kit now if it's one of those listed at Tucker. Tom, I noticed my thanks for your input on my 3325 repair effort got cut. Let me thank you here.
Stan
toggle quoted message
Show quoted text
--- In TekScopes@..., "Tom Miller" wrote: If you do work on the tunnel diodes, be sure to use a heat sink right up against the diode body. Ge tunnel diodes are easily ruined with heat.
However, I think David's comments are more likely the issue. That is because of the external trigger working.
Do you have the correct service manual for your 475?
Regards, tom
----- Original Message ----- From: Bert Haskins To: TekScopes@... Sent: Friday, January 25, 2013 5:09 PM Subject: Re: [TekScopes] 475 Triggering Issue -- Will recent 468 Triggering Issue Thread Help Me?
On 01/25/2013 03:46 PM, stan_katz wrote:
I have an ancient 475 scope which works fine with an external trigger, but will not trigger on any of the internal sources. Should I start by reviewing the recent 468 triggering issue thread to get me going? If not, can anyone recommend a thread I should start with? Of course, I'd be delighted to receive help on this thread.
Stan
Probably not. Going from what I had on 465Bs the triggering problem is often that the tunnel diodes have changed in value. There are many! threads about this in the archives. It may be worth your time to try swapping the TDs from the B sweep section with the A section to see if things change. I had this problem come up when I was hunting for a process control network problem while working in factory in a different country. I swapped the TDs and the scope has been near perfect ever since. Go figure.
|
I used the same military manual from Bama. I had two reasons not to suggest that the tunnel diodes are the problem: 1. The common failure point for the symptoms does not implicate the tunnel diode part of the trigger circuit. External triggering and I would assume line triggering both work and neither of them go through U370 which could cause all of the problems. 2. The tunnel diodes are fragile and difficult to replace so fiddling with them if they are not definitely the problem is asking for trouble. :) On Sat, 26 Jan 2013 23:56:53 -0000, "stan_katz" <stan_katz@...> wrote: Hello Tom,
I will gladly avoid touching the tunnel diodes, and will start with David's suggestion. The only service manuals I have are from BAMA, which are the Army TM--- model manuals. Will these do, or can you point me in a better direction? BTW, Bert's post is a warning to me that the 475 has potentially unobtanium parts. Right now, Tucker has Tek tunnel diode kits for sale. Can you, or Bert, post the proper kit part number for my scope? I should buy a kit now if it's one of those listed at Tucker. Tom, I noticed my thanks for your input on my 3325 repair effort got cut. Let me thank you here.
Stan
--- In TekScopes@..., "Tom Miller" wrote:
If you do work on the tunnel diodes, be sure to use a heat sink right up against the diode body. Ge tunnel diodes are easily ruined with heat.
However, I think David's comments are more likely the issue. That is because of the external trigger working.
Do you have the correct service manual for your 475?
Regards, tom
----- Original Message ----- From: Bert Haskins To: TekScopes@... Sent: Friday, January 25, 2013 5:09 PM Subject: Re: [TekScopes] 475 Triggering Issue -- Will recent 468 Triggering Issue Thread Help Me?
On 01/25/2013 03:46 PM, stan_katz wrote:
I have an ancient 475 scope which works fine with an external trigger, but will not trigger on any of the internal sources. Should I start by reviewing the recent 468 triggering issue thread to get me going? If not, can anyone recommend a thread I should start with? Of course, I'd be delighted to receive help on this thread.
Stan
Probably not. Going from what I had on 465Bs the triggering problem is often that the tunnel diodes have changed in value. There are many! threads about this in the archives. It may be worth your time to try swapping the TDs from the B sweep section with the A section to see if things change. I had this problem come up when I was hunting for a process control network problem while working in factory in a different country. I swapped the TDs and the scope has been near perfect ever since. Go figure.
|
Hi Stan,
?
It's up to you and your wallet as to picking up a spare set of tunnel
diodes. For me, it seems that parts 475 scopes are available at good prices.
Since there are four TDs in each scope, your chances for getting a good one are
reasonable. The main problem is many times?you?wind up with?a
parts unit that can be restored.
?
Email manuals@... (Dave) with
your serial number and ask for a quote. He will be sure you get the correct
manual for your serial number range.
?
Glad you found the problem with the 3325 and that it was a simple fix. Good
job.
?
Best regards,
Tom
?
?
toggle quoted message
Show quoted text
----- Original Message -----
Sent: Saturday, January 26, 2013 6:56
PM
Subject: [TekScopes] Re: 475 Triggering
Issue -- Will recent 468 Triggering Issue Thread Help Me?
?
Hello Tom,
I will gladly avoid touching the tunnel diodes, and will
start with David's suggestion. The only service manuals I have are from BAMA,
which are the Army TM--- model manuals. Will these do, or can you point me in
a better direction? BTW, Bert's post is a warning to me that the 475 has
potentially unobtanium parts. Right now, Tucker has Tek tunnel diode kits for
sale. Can you, or Bert, post the proper kit part number for my scope? I should
buy a kit now if it's one of those listed at Tucker. Tom, I noticed my thanks
for your input on my 3325 repair effort got cut. Let me thank you
here.
Stan
--- In TekScopes@..., "Tom
Miller" wrote: > > If you do work on the tunnel diodes, be sure to
use a heat sink right up against the diode body. Ge tunnel diodes are easily
ruined with heat. > > However, I think David's comments are more
likely the issue. That is because of the external trigger working. >
> Do you have the correct service manual for your 475? > >
Regards, > tom > > ----- Original Message ----- >
From: Bert Haskins > To: TekScopes@...
> Sent: Friday, January 25, 2013 5:09 PM > Subject: Re:
[TekScopes] 475 Triggering Issue -- Will recent 468 Triggering Issue Thread
Help Me? > > > > > On 01/25/2013 03:46 PM,
stan_katz wrote: > > > I have an ancient 475 scope which
works fine with an external trigger, but will not trigger on any of the
internal sources. Should I start by reviewing the recent 468 triggering issue
thread to get me going? If not, can anyone recommend a thread I should start
with? Of course, I'd be delighted to receive help on this thread. >
> Stan > > > > Probably not. > Going
from what I had on 465Bs the triggering problem is often that > the
tunnel diodes have changed in value. > There are many! threads about
this in the archives. > It may be worth your time to try swapping the
TDs from the B sweep section > with the A section to see if things
change. > I had this problem come up when I was hunting for a process
control network > problem while working in factory in a different
country. > I swapped the TDs and the scope has been near perfect ever
since. > Go figure. >
|
--- In TekScopes@..., "stan_katz" wrote: I have an ancient 475 scope which works fine with an external trigger, but will not trigger on any of the internal sources. Should I start by reviewing the recent 468 triggering issue thread to get me going? If not, can anyone recommend a thread I should start with? Of course, I'd be delighted to receive help on this thread.
Stan
I think the TD is the problem. I got a similar problem in my 3B4 plugin. Try to detach one terminal of the TD CR566 (using adequate protection against heating, like a pliers) and check if the behavior is the same. If the TD is defective (i.e. open), the trigger still works, but only in presence of sufficiently steep and large signals, like square waves and/or external trigger pulses (which are generally square waves or pulses). With sinewaves it's likely that trigger doesn't work, but with squares do. The same test may apply to CR556 (the arming TD). You can also check for the presence of a pulse signal on either bases of Q556 and Q566. If they're present, all the circuitry preceding Q556/566 are well-working. Max
|
Amazing!?
Tunnel diodes in a 561A!
Tek must have started using tunnel diodes shortly after they came out in 1960. Leave it to Tek to go with bleeding edge esoteric. Since I'm a rather low budget experimenter, the only other scope I have is a Telequiment D51 "student scope" with a 4mhz bandwidth. Is this scope adequate to view the pulses at Q556/Q566?
Thanks, Stan
toggle quoted message
Show quoted text
--- In TekScopes@..., "Max Mazza" wrote:
--- In TekScopes@..., "stan_katz" wrote:
I have an ancient 475 scope which works fine with an external trigger, but will not trigger on any of the internal sources. Should I start by reviewing the recent 468 triggering issue thread to get me going? If not, can anyone recommend a thread I should start with? Of course, I'd be delighted to receive help on this thread.
Stan
I think the TD is the problem. I got a similar problem in my 3B4 plugin.
Try to detach one terminal of the TD CR566 (using adequate protection against heating, like a pliers) and check if the behavior is the same.
If the TD is defective (i.e. open), the trigger still works, but only in presence of sufficiently steep and large signals, like square waves and/or external trigger pulses (which are generally square waves or pulses). With sinewaves it's likely that trigger doesn't work, but with squares do. The same test may apply to CR556 (the arming TD).
You can also check for the presence of a pulse signal on either bases of Q556 and Q566. If they're present, all the circuitry preceding Q556/566 are well-working.
Max
|
I am not exactly sure when Tektronix started using tunnel diodes but the 545 lacks them while the 547 has them which explains the later's outstanding triggering bandwidth. It may be that they had the only engineers able to understand how to use them for trigger circuits. You will not need a very fast oscilloscope to see if the tunnel diode circuit is operating correctly. Yours should work fine. On Sun, 27 Jan 2013 03:51:35 -0000, "stan_katz" <stan_katz@...> wrote: Amazing!?
Tunnel diodes in a 561A!
Tek must have started using tunnel diodes shortly after they came out in 1960. Leave it to Tek to go with bleeding edge esoteric. Since I'm a rather low budget experimenter, the only other scope I have is a Telequiment D51 "student scope" with a 4mhz bandwidth. Is this scope adequate to view the pulses at Q556/Q566?
Thanks, Stan
--- In TekScopes@..., "Max Mazza" wrote:
--- In TekScopes@..., "stan_katz" wrote:
I have an ancient 475 scope which works fine with an external trigger, but will not trigger on any of the internal sources. Should I start by reviewing the recent 468 triggering issue thread to get me going? If not, can anyone recommend a thread I should start with? Of course, I'd be delighted to receive help on this thread.
Stan
I think the TD is the problem. I got a similar problem in my 3B4 plugin.
Try to detach one terminal of the TD CR566 (using adequate protection against heating, like a pliers) and check if the behavior is the same.
If the TD is defective (i.e. open), the trigger still works, but only in presence of sufficiently steep and large signals, like square waves and/or external trigger pulses (which are generally square waves or pulses). With sinewaves it's likely that trigger doesn't work, but with squares do. The same test may apply to CR556 (the arming TD).
You can also check for the presence of a pulse signal on either bases of Q556 and Q566. If they're present, all the circuitry preceding Q556/566 are well-working.
Max
|
Hi David,
I sampled rear panel ch2 vertical signal per your suggestion. It's coming out at only 1/20th the signal true amplitude. So maybe R403 should be examined? However, a confounding symptom, occurs when I try 1v to 20v sine waves at 60hz with line triggering. I couldn't get a lock. no lock at any other freq. for line source. Another bad sign, which has presented itself since I started exercising all controls through full range...neither of the v/div controls will advance past 5v/div...no 10/20/50v positions possible on either channel. Is this op. error, or what? Ch 2 v/div control skirt illuminator is not lit. Just burned out, or is illuminator driven by other circuitry?
Stan
toggle quoted message
Show quoted text
--- In TekScopes@..., David wrote: I used the same military manual from Bama.
I had two reasons not to suggest that the tunnel diodes are the problem:
1. The common failure point for the symptoms does not implicate the tunnel diode part of the trigger circuit. External triggering and I would assume line triggering both work and neither of them go through U370 which could cause all of the problems.
2. The tunnel diodes are fragile and difficult to replace so fiddling with them if they are not definitely the problem is asking for trouble. :)
On Sat, 26 Jan 2013 23:56:53 -0000, "stan_katz" wrote:
Hello Tom,
I will gladly avoid touching the tunnel diodes, and will start with David's suggestion. The only service manuals I have are from BAMA, which are the Army TM--- model manuals. Will these do, or can you point me in a better direction? BTW, Bert's post is a warning to me that the 475 has potentially unobtanium parts. Right now, Tucker has Tek tunnel diode kits for sale. Can you, or Bert, post the proper kit part number for my scope? I should buy a kit now if it's one of those listed at Tucker. Tom, I noticed my thanks for your input on my 3325 repair effort got cut. Let me thank you here.
Stan
--- In TekScopes@..., "Tom Miller" wrote:
If you do work on the tunnel diodes, be sure to use a heat sink right up against the diode body. Ge tunnel diodes are easily ruined with heat.
However, I think David's comments are more likely the issue. That is because of the external trigger working.
Do you have the correct service manual for your 475?
Regards, tom
----- Original Message ----- From: Bert Haskins To: TekScopes@... Sent: Friday, January 25, 2013 5:09 PM Subject: Re: [TekScopes] 475 Triggering Issue -- Will recent 468 Triggering Issue Thread Help Me?
On 01/25/2013 03:46 PM, stan_katz wrote:
I have an ancient 475 scope which works fine with an external trigger, but will not trigger on any of the internal sources. Should I start by reviewing the recent 468 triggering issue thread to get me going? If not, can anyone recommend a thread I should start with? Of course, I'd be delighted to receive help on this thread.
Stan
Probably not. Going from what I had on 465Bs the triggering problem is often that the tunnel diodes have changed in value. There are many! threads about this in the archives. It may be worth your time to try swapping the TDs from the B sweep section with the A section to see if things change. I had this problem come up when I was hunting for a process control network problem while working in factory in a different country. I swapped the TDs and the scope has been near perfect ever since. Go figure.
|
The extra three markings on the Volts/Div knob are there for use with a x10 probe, which enables a different lamp behind the Volts/div skirt, so the 5V/Div setting is mechanically the same as the 50V/div setting. The change of lamp is only made with a probe with the correct BNC with probe range pin which connects a resistor from the outer ring of the BNC socket to ground to indicate the type of probe.
Dave
toggle quoted message
Show quoted text
-----Original Message----- From: TekScopes@... [mailto:TekScopes@...] On Behalf Of stan_katz Sent: 27 January 2013 19:58 To: TekScopes@... Subject: [TekScopes] Re: 475 Triggering Issue -- Will recent 468 Triggering Issue Thread Help Me?
Hi David,
I sampled rear panel ch2 vertical signal per your suggestion. It's coming out at only 1/20th the signal true amplitude. So maybe R403 should be examined? However, a confounding symptom, occurs when I try 1v to 20v sine waves at 60hz with line triggering. I couldn't get a lock. no lock at any other freq. for line source. Another bad sign, which has presented itself since I started exercising all controls through full range...neither of the v/div controls will advance past 5v/div...no 10/20/50v positions possible on either channel. Is this op. error, or what? Ch 2 v/div control skirt illuminator is not lit. Just burned out, or is illuminator driven by other circuitry?
Stan
|
I've uncovered what I think is a significant symptom. If I inject a 1 volt square wave, at 1khz or above, set v/div to 20mv/div, and play with the position control, I can get the trigger source lamp to light. If I try the same stunt with a sine wave, I need at least 500khz to get the scope to trigger. This symptom seems to play into what Max was suggesting by lifting one lead of a tunnel diode, and then determining if I can get trigger with high slew rate sources. Of course, with the scope setup to the above test conditions, it's useless.
Stan
toggle quoted message
Show quoted text
--- In TekScopes@..., "stan_katz" wrote: Hi David,
I sampled rear panel ch2 vertical signal per your suggestion. It's coming out at only 1/20th the signal true amplitude. So maybe R403 should be examined? However, a confounding symptom, occurs when I try 1v to 20v sine waves at 60hz with line triggering. I couldn't get a lock. no lock at any other freq. for line source. Another bad sign, which has presented itself since I started exercising all controls through full range...neither of the v/div controls will advance past 5v/div...no 10/20/50v positions possible on either channel. Is this op. error, or what? Ch 2 v/div control skirt illuminator is not lit. Just burned out, or is illuminator driven by other circuitry?
Stan
--- In TekScopes@..., David wrote:
I used the same military manual from Bama.
I had two reasons not to suggest that the tunnel diodes are the problem:
1. The common failure point for the symptoms does not implicate the tunnel diode part of the trigger circuit. External triggering and I would assume line triggering both work and neither of them go through U370 which could cause all of the problems.
2. The tunnel diodes are fragile and difficult to replace so fiddling with them if they are not definitely the problem is asking for trouble. :)
On Sat, 26 Jan 2013 23:56:53 -0000, "stan_katz" wrote:
Hello Tom,
I will gladly avoid touching the tunnel diodes, and will start with David's suggestion. The only service manuals I have are from BAMA, which are the Army TM--- model manuals. Will these do, or can you point me in a better direction? BTW, Bert's post is a warning to me that the 475 has potentially unobtanium parts. Right now, Tucker has Tek tunnel diode kits for sale. Can you, or Bert, post the proper kit part number for my scope? I should buy a kit now if it's one of those listed at Tucker. Tom, I noticed my thanks for your input on my 3325 repair effort got cut. Let me thank you here.
Stan
--- In TekScopes@..., "Tom Miller" wrote:
If you do work on the tunnel diodes, be sure to use a heat sink right up against the diode body. Ge tunnel diodes are easily ruined with heat.
However, I think David's comments are more likely the issue. That is because of the external trigger working.
Do you have the correct service manual for your 475?
Regards, tom
----- Original Message ----- From: Bert Haskins To: TekScopes@... Sent: Friday, January 25, 2013 5:09 PM Subject: Re: [TekScopes] 475 Triggering Issue -- Will recent 468 Triggering Issue Thread Help Me?
On 01/25/2013 03:46 PM, stan_katz wrote:
I have an ancient 475 scope which works fine with an external trigger, but will not trigger on any of the internal sources. Should I start by reviewing the recent 468 triggering issue thread to get me going? If not, can anyone recommend a thread I should start with? Of course, I'd be delighted to receive help on this thread.
Stan
Probably not. Going from what I had on 465Bs the triggering problem is often that the tunnel diodes have changed in value. There are many! threads about this in the archives. It may be worth your time to try swapping the TDs from the B sweep section with the A section to see if things change. I had this problem come up when I was hunting for a process control network problem while working in factory in a different country. I swapped the TDs and the scope has been near perfect ever since. Go figure.
|
--- In TekScopes@..., "stan_katz" wrote: I've uncovered what I think is a significant symptom. If I inject a 1 volt square wave, at 1khz or above, set v/div to 20mv/div, and play with the position control, I can get the trigger source lamp to light. If I try the same stunt with a sine wave, I need at least 500khz to get the scope to trigger. This symptom seems to play into what Max was suggesting by lifting one lead of a tunnel diode, and then determining if I can get trigger with high slew rate sources. Of course, with the scope setup to the above test conditions, it's useless.
Stan
This was exactly the same behavior I noted when debugging my 3B4 timebase unit. With 1 KHz square wave of adequate amplitude, the triggers works. With a sinewave, it must be 100 KHz or above to trigger, with difficult. If disconnecting the TD the behavior is exactly the same, assume that 99% the TD is open. To get a definitive certainty, you should check if some signal is present on the anode of TD (with TD in place). If the amplitude swing during the TD's commutation is around 200-300mV pk-pk, the TD is OK. If a few mV, the TD is shorted. If more than 500mV, the TD is open. Do that check before you detach one TD's lead and then after. Compare the results. Let us know. Max
|
The rear output signal it's not directly correlated to the input signal amplitude, but to the amplitude in major divisions on the screen. Tek stated that output voltage is approximately 20mV/division into 1 Megaohm load or 10mV/division into 50 ohm load. So I think that you've observed is absolutely correct.
Max
toggle quoted message
Show quoted text
--- In TekScopes@..., "stan_katz" wrote: Hi David,
I sampled rear panel ch2 vertical signal per your suggestion. It's coming out at only 1/20th the signal true amplitude. So maybe R403 should be examined?
|
Line triggering triggers the sweep from the AC power line so the sweep will synchronize with AC power supplies. It will not synchronize with other low frequency signals unless they are also synchronized with the AC power line. The 10/20/50 volts/division attenuator settings depend on using a x10 probe which attenuates the input signal by 10 times. Whether the x1 or x10 attenuator skirt illumination occurs depends on the readout connection from the probe used. A normal BNC connection or probe without that feature will always illuminated the x1 indicator on the skirt. Tektronix has made a very nice PDF copy of the 475 service manual available here although you need to register with them: The KO4BB manual is excellent also: On Sun, 27 Jan 2013 19:57:34 -0000, "Stan_katz" <stan_katz@...> wrote: Hi David,
I sampled rear panel ch2 vertical signal per your suggestion. It's coming out at only 1/20th the signal true amplitude. So maybe R403 should be examined? However, a confounding symptom, occurs when I try 1v to 20v sine waves at 60hz with line triggering. I couldn't get a lock. no lock at any other freq. for line source. Another bad sign, which has presented itself since I started exercising all controls through full range...neither of the v/div controls will advance past 5v/div...no 10/20/50v positions possible on either channel. Is this op. error, or what? Ch 2 v/div control skirt illuminator is not lit. Just burned out, or is illuminator driven by other circuitry?
Stan
|
Yes, Tek engineers started to experiment and using TD as soon as they were introduced. Tek 544, 546 and 547 already they had TD in their circuitry. I've gained expertise working and studying circuits in my TEK 3B4, 3T77, 3T77A plugins and 465 also. TDs are very rare and hard to find, but the models used in 465 and 475 scopes (respectively 152-0125-00/01 and 152-0140-01) still can be found with some patience. I've got two of these, Tek original, from a Greek surplus dealer for a reasonable price, about 40$ plus shipping, two years ago. One of these replaced the broken TD in my 3B4. The other is jealously guarded in my drawer with other Tek original spares, since the TD model 152-0125-00 fits equally well in #b4 and 465....
The only, very UNOBTAINABLE TD is the 1N3129, used in 3T77 and 3T77A sampling sweep units. My 3T77 had one TD dead. After 3 years of fruitless search, the past year I finally decided to make the jump: I bought some russian TD, GaAs military, from an ukrainian dealer, carefully choosing and comparing their specs against the original Teks ones.
Then I changed the polarization network, to adapt the GaAs diode (Ge diodes has Vp-Vv of about 300-500mV, GaAs can be 800 mV or more, for the same Ip, so the polarization network must be modified accordingly)
After some hours of calculations, graph comparisons, measurements of TD curves on my curve tracer (a curve tracer I made especially for TD measurements), I changed some resistors in the 3T77 unit, put the russian GaAs TD in place of the original dead 1N3129 and voila! My 3T77 returned to live, with the same specs of the original: it triggers with sine wave signals up to 600/700Mhz and a mere 20 mV amplitude. With external trigger signal it works up to 1Ghz with no problems, in conjunction with my 3S1 sampling unit.
For the chronicle, two years ago I found ONE 1N3129 offered to the miserable price of $ 150 (I refused, damn... was gold or platinum? Or maybe, diamond?), while the russian diodes I've paid about $ 10 each
Max
toggle quoted message
Show quoted text
--- In TekScopes@..., "stan_katz" wrote: Amazing!?
Tunnel diodes in a 561A!
Tek must have started using tunnel diodes shortly after they came out in 1960. Leave it to Tek to go with bleeding edge esoteric. Since I'm a rather low budget experimenter, the only other scope I have is a Telequiment D51 "student scope" with a 4mhz bandwidth. Is this scope adequate to view the pulses at Q556/Q566?
Thanks, Stan
|
For the chronicle, I investigated why the TD on my 3T77 unit, purchased used on ebay two years ago as "not working" was dead.
The 3T77 does not trigger at all, and that because simply for the fact that one TD was dead. Incidentally, the TD was the one that its signal was sent ALSO to the TRIGGER OUT BNC on the front panel via a simple pulse transformer. Very useful but... imagine if you, by mistake, send a "full-bodied, fat" trigger signal (5,10 V or so) TO the scope putting it in the WRONG connector (TRIGGER OUT instead on TRIGGER IN)? The TD surely reacts in a very negative manner! That was the event made by the previous owner, that destroyed the 1N3129 TD, IMO.
Max
|
Max,
Remember, these measurements are made with a student scope that has its own problems and can't measure below 100mv/div. So with 10x probe its hard to read. Amplitude and pulse widths given are very approximate.
I'm still working up the courage to pull out a TD lead. Anyway, with both TDs in circuit:
2v p-p square wave in .5v/div ... auto ... trigger source lamp extinguished CR566 -- nothing CR556 -- approx. -500mv nice commutation approx. 300us pulse width
2v p-p square wave in 20mv/div...auto...trigger source lamp lit CR566...crummy,noisy elongated commutation pulses approx -300mv approx 400us width. CR556 -- approx. -500mv nice commutation approx. 300us pulse width
Isn't this experiment sufficient to prove CR566 is shorted? Should I still pull a lead. If so. On CR566 or CR556? And what/where do I measure? I'm confused about the post pull procedure.
stan
toggle quoted message
Show quoted text
--- In TekScopes@..., "Max Mazza" wrote:
--- In TekScopes@..., "stan_katz" wrote:
I've uncovered what I think is a significant symptom. If I inject a 1 volt square wave, at 1khz or above, set v/div to 20mv/div, and play with the position control, I can get the trigger source lamp to light. If I try the same stunt with a sine wave, I need at least 500khz to get the scope to trigger. This symptom seems to play into what Max was suggesting by lifting one lead of a tunnel diode, and then determining if I can get trigger with high slew rate sources. Of course, with the scope setup to the above test conditions, it's useless.
Stan This was exactly the same behavior I noted when debugging my 3B4 timebase unit. With 1 KHz square wave of adequate amplitude, the triggers works. With a sinewave, it must be 100 KHz or above to trigger, with difficult. If disconnecting the TD the behavior is exactly the same, assume that 99% the TD is open.
To get a definitive certainty, you should check if some signal is present on the anode of TD (with TD in place). If the amplitude swing during the TD's commutation is around 200-300mV pk-pk, the TD is OK. If a few mV, the TD is shorted. If more than 500mV, the TD is open. Do that check before you detach one TD's lead and then after. Compare the results. Let us know.
Max
|
I think you would be better off using a triangle or sine wave as a test signal so you can distinguish the TD changing state and the test signal. I noticed that the earlier oscilloscopes which use the same trigger circuit like the 465 and 464 have more detailed schematics which show voltage levels and waveforms. They also show the details of U520 and U720 (they are paraphase amplifiers used to selectively invert the trigger signal which selects the slope) although that is not really needed. On Mon, 28 Jan 2013 04:45:35 -0000, "stan_katz" <stan_katz@...> wrote: Max,
Remember, these measurements are made with a student scope that has its own problems and can't measure below 100mv/div. So with 10x probe its hard to read. Amplitude and pulse widths given are very approximate.
I'm still working up the courage to pull out a TD lead. Anyway, with both TDs in circuit:
2v p-p square wave in .5v/div ... auto ... trigger source lamp extinguished CR566 -- nothing CR556 -- approx. -500mv nice commutation approx. 300us pulse width
2v p-p square wave in 20mv/div...auto...trigger source lamp lit CR566...crummy,noisy elongated commutation pulses approx -300mv approx 400us width. CR556 -- approx. -500mv nice commutation approx. 300us pulse width
Isn't this experiment sufficient to prove CR566 is shorted? Should I still pull a lead. If so. On CR566 or CR556? And what/where do I measure? I'm confused about the post pull procedure.
stan
--- In TekScopes@..., "Max Mazza" wrote:
--- In TekScopes@..., "stan_katz" wrote:
I've uncovered what I think is a significant symptom. If I inject a 1 volt square wave, at 1khz or above, set v/div to 20mv/div, and play with the position control, I can get the trigger source lamp to light. If I try the same stunt with a sine wave, I need at least 500khz to get the scope to trigger. This symptom seems to play into what Max was suggesting by lifting one lead of a tunnel diode, and then determining if I can get trigger with high slew rate sources. Of course, with the scope setup to the above test conditions, it's useless.
Stan This was exactly the same behavior I noted when debugging my 3B4 timebase unit. With 1 KHz square wave of adequate amplitude, the triggers works. With a sinewave, it must be 100 KHz or above to trigger, with difficult. If disconnecting the TD the behavior is exactly the same, assume that 99% the TD is open.
To get a definitive certainty, you should check if some signal is present on the anode of TD (with TD in place). If the amplitude swing during the TD's commutation is around 200-300mV pk-pk, the TD is OK. If a few mV, the TD is shorted. If more than 500mV, the TD is open. Do that check before you detach one TD's lead and then after. Compare the results. Let us know.
Max
|