Hi Reed,

?

Is the NTE538 or a ECG538 ?a “X6” multiplier? Is the datasheet available?

?

Marcus

?

De: TekScopes@... [mailto:TekScopes@...] Em nome de Reed Dickinson
Enviada em: ter?a-feira, 7 de fevereiro de 2012 18:56
Para: TekScopes@...
Cc: Laura Tighe
Assunto: [TekScopes] 475 dead voltage multiplier (was: 475 high voltage circuit very

?

?

I have encountered at least a dozen bad HV multipliers in various 4XX
series scopes. In all cases, except the 485, I have used the ECG538 or
the NTE538 replacement available through Mouser for $26. It is
necessary to cut the HV lead from the 538 and splice the original Tek
connector on the cable. I use a thick wall plastic tube with the ID the
same as the largest wire. Make the solder joint free of sharp points to
avoid arcing, cover the joint with silicon RTV and slide the plastic
tube over the joint. Connect the LOW lead on the 538 to ground. Do not
use the D or FOCUS leads but insulate both from ground. Connect the IN
terminal on the 538 to the flyback transformer. After the RTV cures
mount the 538 where the original multiplier was by drilling two holes in
the metal bottom and securing with two screws. It is necessary to trim
the metal HV cage cover to get the cable through.

Reed Dickinson


 

开云体育

Hi:
The 4XX series of scopes, except for the 485, use a X3 multiplier.? The 485 uses an X6.? I have an old ECG catalog for a reference.? ECG was bought out by NTE.? There should be a datasheet? for the NTE538 on the internet but I have not checked this out.

Reed ?????????????????????????????

On 2/7/2012 1:10 PM, Marcus - PP5MS wrote:

?

Hi Reed,

?

Is the NTE538 or a ECG538 ?a “X6” multiplier? Is the datasheet available?

?

Marcus

?

De: TekScopes@... [mailto:TekScopes@...] Em nome de Reed Dickinson
Enviada em: ter?a-feira, 7 de fevereiro de 2012 18:56
Para: TekScopes@...
Cc: Laura Tighe
Assunto: [TekScopes] 475 dead voltage multiplier (was: 475 high voltage circuit very

?

?

I have encountered at least a dozen bad HV multipliers in various 4XX
series scopes. In all cases, except the 485, I have used the ECG538 or
the NTE538 replacement available through Mouser for $26. It is
necessary to cut the HV lead from the 538 and splice the original Tek
connector on the cable. I use a thick wall plastic tube with the ID the
same as the largest wire. Make the solder joint free of sharp points to
avoid arcing, cover the joint with silicon RTV and slide the plastic
tube over the joint. Connect the LOW lead on the 538 to ground. Do not
use the D or FOCUS leads but insulate both from ground. Connect the IN
terminal on the 538 to the flyback transformer. After the RTV cures
mount the 538 where the original multiplier was by drilling two holes in
the metal bottom and securing with two screws. It is necessary to trim
the metal HV cage cover to get the cable through.

Reed Dickinson


 

开云体育

Hi Reed,
?
The 465 and 475 use a X6 multiplier. See the X-ray, courtesy of a group member,?of a 475 CW multiplier in the photos. Six diodes and six capacitors clearly visible.
?
Regards,
Tom

----- Original Message -----
Sent: Tuesday, February 07, 2012 4:55 PM
Subject: Re: RES: [TekScopes] 475 dead voltage multiplier (was: 475 high voltage circuit very

?

Hi:
The 4XX series of scopes, except for the 485, use a X3 multiplier.? The 485 uses an X6.? I have an old ECG catalog for a reference.? ECG was bought out by NTE.? There should be a datasheet? for the NTE538 on the internet but I have not checked this out.

Reed ?????????????????????????????

On 2/7/2012 1:10 PM, Marcus - PP5MS wrote:

?

Hi Reed,

Is the NTE538 or a ECG538 ?a “X6” multiplier? Is the datasheet available?

Marcus

De: TekScopes@... [mailto:TekScopes@...] Em nome de Reed Dickinson
Enviada em: ter?a-feira, 7 de fevereiro de 2012 18:56
Para: TekScopes@...
Cc: Laura Tighe
Assunto: [TekScopes] 475 dead voltage multiplier (was: 475 high voltage circuit very

?

I have encountered at least a dozen bad HV multipliers in various 4XX
series scopes. In all cases, except the 485, I have used the ECG538 or
the NTE538 replacement available through Mouser for $26. It is
necessary to cut the HV lead from the 538 and splice the original Tek
connector on the cable. I use a thick wall plastic tube with the ID the
same as the largest wire. Make the solder joint free of sharp points to
avoid arcing, cover the joint with silicon RTV and slide the plastic
tube over the joint. Connect the LOW lead on the 538 to ground. Do not
use the D or FOCUS leads but insulate both from ground. Connect the IN
terminal on the 538 to the flyback transformer. After the RTV cures
mount the 538 where the original multiplier was by drilling two holes in
the metal bottom and securing with two screws. It is necessary to trim
the metal HV cage cover to get the cable through.

Reed Dickinson


 

Apparently there never has been a data sheet for the NTE538 or the precursor ECG538. The net revealed nothing but robotic responses with no data sheets. Mouser has only one left and zero on order. The NTE website reports the remaining stock:



I guess I'll take a hit and buy one to see what I can do with it. Or maybe not if I'm going to build one for my 455.

Arden

Is the NTE538 or a ECG538 a "X6" multiplier? Is the datasheet available?