Tektronix 585A Voltage Diagnoses Problem edited
Hello All,
I have a 585A scope that has me stumped.. the LV voltages are out of spec and I cannot find the problem. Here are the voltage readings; -150 = -182 -100 = -115 225 = 263 350 = 409 500 = 587 The -150v adjustment pot does not change the voltage value at all. All the tubes were pulled and tested for shorts and emission. Not being very good at diagnosis I shotgunned all the electrolytics and filter caps although most of the tested good in circuit I found a couple of bad ones. No improvement. I have the schematic and can upload that if it helps. Can anyone suggest a way I can isolate the circuit causing me grief? The scope is clean and intact otherwise. Thanks in advance. |
Roy Morgan
I no longer have my 585 or a manual, but I am pretty sure that all the other B+ voltages depend on the -150 supply. I suggest you focus on that one - you may find that all the others come to correct voltage if that one is right. I suggest you read the 585 manual section about the B+ supplies.
toggle quoted message
Show quoted text
¡°Typical OscilloscopeCircuitry¡± confirms this on pg 12-22. Fig 12-16 shows the supply for the 531A and 545A. The 585 supply is likely very similar. Your report below about the -150 volt adjustment pot having no effect is the smoking gun!! There are 4 capacitors .01 uF in the 531/545 supply shown. If ANY of them is a ¡°black beauty¡± cap, replace them ALL. Check for reasonable voltages on all 3 terminals of that adjustment pot. Note: the calibration of the entire scope depends on the -150 and the others to be right. I assume you can rely on your voltmeter to be right. Roy sends. On Aug 23, 2019, at 6:43 PM, randolphbeebe@... wrote: |
Chuck Harris
Everything is referenced to the -150V supply. It must
toggle quoted message
Show quoted text
be right, or the others will behave truly weird... like you are seeing. Start there. Also, the 585A has DC filaments. If it doesn't have a plugin, it switches off the -100V supply that is used to power the filaments on the plugin... Or, it has an internal dummy load... can't remember right now. -Chuck Harris randolphbeebe@... wrote: Hello All, |
Chuck Harris
Hi Randolph,
toggle quoted message
Show quoted text
There is no reason to expect the plugin would pull the voltages off their mark. Get the -150V right and everything else should fall into line. -Chuck Harris randolphbeebe@... wrote: Hi Chuck, |
Hi Randolph,
You checked all the tubes, but also the voltage reference tube V609? Should measure about 85 V. The grid voltages of V 624 should be approximately equal, well within 1 V or so. The voltage at the grids should be what you expect from the divider R615/616/617 with "your" -150V. If the fault is elsewhere then likely you will find grid pin 7 far too negative and pin 6 (anode) far too positive, almost at "your" +100V. Then continue with V634. Is the screen voltage about normal (not very negative)? And so on (for now). Albert |
Albert,
I checked the voltages and here is what I measured after a 10 minute warm up; Tube Pin Voltage V609 7 -107.4 V624 6 +1.5 7 -107.6 2 -104.0 V634 6 -38.0 7 and 2 -87.0 1 -87.1 4 -2.4 V627 9 +69.0 V637 same V647 same As you mentioned in you previous reply I found that pins 6 and 7 of V624 vary widely. I changed that 12AX7 with no change. Thanks Again, Randy |
Albert,
I am not sure if this is related but I seem to get a short between Ground and the positive terminals of C790-791-792 ( All have been replaced now) of the regulated heater supply (Upper Left portion of the LV Power Supply Schematic). It appears to be isolated between the Emitter of Q797 and Diode 793 and Diode 792 somewhere. Is this normal? Thanks |