2440 cal errors


 

Well I guess you get what you pay for. At $50 for a 2440 should have been a clue. So far the first 2440 I turned on has cal errors
4000 FPP fail
7000 CCD fail
8000 PA fail
all others passed

Maybe I should go through an external calibrate first?
Bob


Bob Albert
 

Am I still on the list for one of those 'scopes?
Bob818 894-2887

On Monday, August 19, 2019, 10:46:17 AM PDT, Robert Simpson via Groups.Io <go_boating_fast@...> wrote:

Well I guess you get what you pay for. At $50 for a 2440 should have been a clue. So far the first 2440 I turned on has cal errors
4000 FPP? fail
7000 CCD fail
8000 PA? ? fail
all others passed

Maybe I should go through an external calibrate first?
Bob


 

Hi,

Yes, definitely. The FPP error is 99% because of the nvram battery failed
an lost the calibration. All other fails are likely a result of it. You can
verify it by going into the extended diagnostics and check the detailed
result of the FPP fail.
Then you'll need to replace the battery (older models) or the nvram on
newer models (or hack them for external battery).
But first you shoul try self cal and ext cal, without powering down the
scope and see if the errors go away. They should. If not, you should still
try the full calibration as desvribed in the adjustments section of the
service man.

Mandatory warning: don't run the scope without cover without external
cooling, you risk damaging the CCDmodules.

Szabolcs

Robert Simpson via Groups.Io <go_boating_fast@...> ezt ¨ªrta
(id?pont: 2019. aug. 19., H¨¦t 19:46):

Well I guess you get what you pay for. At $50 for a 2440 should have been
a clue. So far the first 2440 I turned on has cal errors
4000 FPP fail
7000 CCD fail
8000 PA fail
all others passed

Maybe I should go through an external calibrate first?
Bob




 

OK,
Thanks, Bob


Bob Albert
 

So replacing the battery might solve most of the problems...

On Tuesday, August 20, 2019, 08:19:55 AM PDT, Robert Simpson via Groups.Io <go_boating_fast@...> wrote:

OK,
Thanks, Bob


 

Does anybody know around which serial # the 2440 switched from using a backup battery to using NVRAM? I'm looking at purchasing one on eBay that has an FPP diagnostic error - serial #B013530 (I think).

Thanks!
David Bono