tds744a tube crt affichage
oudot damien
Bonjour sur mon tds744a quand je l'allume, sur l'cran il apparait des bandes horizontal plus lumineuse et de couleur qui d¨¦filent verticalement.
Apr¨¨s plusieurs minutes elles disparaissent. Pouvez vous m'aider?? ou touver les sh¨¦ma de la partie tube crt et alimentation?? Cordialement Christophe |
Hello,
I am not sure and hopefully others have more detailed experience with helpful feedback. However, this may help though if you're not aware of... pg. 241 is the Troubleshooting section of the Service Manual: There are some maintenance figures and instruction starting on pg. 185. Detailed block diagrams starting on pg. 268. That manual is found on this page: Meilleures salutations |
On Tue, Feb 26, 2019 at 4:01 AM oudot damien <lcristoftrain@...>
wrote: Bonjour sur mon tds744a quand je l'allume, sur l'cran il apparait desI had occasional image "tearing" - where vertical lines on the CRT wiggled to one side occasionally I re-capped the CRT driver board, which helped for a while. None of the electrolytic capacitors had leaked, but some of them were getting marginal on ESR and capacity as I recall. The actual problem, however, turned out to be a poor connection to the horizontal yoke. See this message /g/TekScopes/message/133472 for a description - apparently others have had the same problem. |