2465 Ch 1 probe not switching to 10x
On Thu, 15 Nov 2018 14:32:56 -0800, you wrote:
Can anyone point me in the right direction to fix this? The scope recognizes the 10x probe when connected to the other three channels.Look at the ring on the probe. It's likely connected to some circuitry that forms part of a voltage divider. The other part of the divider is the resistor in the probe itself. Try taking a number of resistors and run them from the ring to ground. If you get no readout change then the mesure/compare circuit isn't working, or the feedback from the circuit to the readout circuitry isn't happy. Harvey
|
Good point.
toggle quoted message
Show quoted text
-----Original Message-----
From: Harvey White Sent: Friday, November 16, 2018 12:34 PM To: [email protected] Subject: Re: [TekScopes] 2465 Ch 1 probe not switching to 10x On Thu, 15 Nov 2018 14:32:56 -0800, you wrote: Can anyone point me in the right direction to fix this? The scope recognizes the 10x probe when connected to the other three channels.Look at the ring on the probe. It's likely connected to some circuitry that forms part of a voltage divider. The other part of the divider is the resistor in the probe itself. Try taking a number of resistors and run them from the ring to ground. If you get no readout change then the mesure/compare circuit isn't working, or the feedback from the circuit to the readout circuitry isn't happy. Harvey
-- Jack |
Each of the four probe rings is connected to -1.25V by a 10k resistor so with the scope powered off you should be able to measure the resistance between each pair of probe rings at not more than 20k. If the pairs involving Ch1 give different results to the other pairs (ie Ch1-Ch2 is different from Ch2-Ch3 etc) then there is probably a break in the wiring of the Ch1 probe ring. There is a three way ribbon cable emerging from the Ch1 attenuator, if you unplug this you should get zero ohms from the probe ring to pin 3, the orange wire.
The next check is resistance from the Ch1 probe ring to pin 1 of the analogue multiplexer chip U2601. This should read 10k, if not measure from pin 1 to either side of R2704. If all the resistance checks prove OK then as a last resort measure the voltage on pin1 of U2601 and see if it changes when you plug the 10X probe to channel 1. To compare with a working channel measure pin 5 of U2601 while plugging the probe into Ch2. If you don't find anything wrong by now you probably have a dead U2601 since everything after this is common to all four probe channels. Roger |
Ok, it appears that the bnc jack on the channel 1 attenuators is bad. The connection to the probe ring was loose and intermittent. The bnc jacks alone are not available for any reasonable price, so I picked up another attenuator.
Btw, the component numbers you referred to were for the 2465b. My 2465 has the same circuit with different component numbers. Thanks again for the pointers. Greg |
Hey Greg,
toggle quoted message
Show quoted text
I replaced the BNCs on my 2467 a couple of years ago. See message /g/TekScopes/message/115057 for the kind gentleman who sourced replacement BNCs for me at a very reasonable price. Maybe he has more of those. I'd think the precise attenuation ratios of your attenuators get baked into the calibration, so you'll likely stay closer by replacing the BNCs than the entire attenuator. If your scope hasn't been calibrated in the last couple of decades or so, it may not make for a hill of beans, though :). Siggi On Fri, Nov 16, 2018 at 5:14 PM Greg <gcdco@...> wrote:
Ok, it appears that the bnc jack on the channel 1 attenuators is bad. The |
If the replacement attenuator works acceptably, I¡¯ll use it. If not, I¡¯ll pull the bnc from the new one and use my existing attenuator. Appreciate the suggestion re the possible source, though.
When I got this scope, for free, about two months ago, the ch 1 bnc was covered with duct tape and the statement that channel 1 didn¡¯t work. Since then, I have visually inspected all of the caps, blown out the dust, verified the lvps voltages as correct and determined that the channel 1 issue seems to be cured by chilling U600. I suspect an intermittent terminal on U600 and will clean those contacts. All channels now work correctly, and the calibration seems pretty good still. Just working on the probe sensor ring and planning a future re capping right now. This forum has been a great resource for me. Thanks again for the offer re the bnc. Greg |