On Wed, 25 Jul 2018, Rick Boswell wrote:
Want to buy 495P v9.7 EPROMs which are installed on the A54 memory board.
P/Ns 160-8316-00 and 160-8315-00.
There are 9 EPROMs in v9.7 firmware in 494AP that should be the same as you
have. They are part numbers 160-8315-00 to 160-8323-00. First 2 are on that
memory board, rest is on GPIB board. I have all firmware images and can send
'em to you so you would be able to program your own chips.
There is another thing -- it is _NOT_ just those switches that tell the
firmware what instrument it runs on. There is also another MCU on the front
panel that runs its own firmware. Main instrument firmware asks the front
panel MCU about hardware version and that is taken into consideration when
deciding what hardware it runs on. For example, 494AP and 2756P have EXACTLY
SAME switches configuration and main firmware decides what it runs on based
on what front panel controller told. 494AP and 2756P are identical inside
just case and front panel is different. 2756P has full-blown 494AP chassis
inside sans front panel and its own front panel. Part of the case not
occupied by 494AP chassis is simply empty. If you remove that chassis and
install a 494AP front panel on it the thing will tell you it is 494AP and
will happily run as 494AP without flipping a single switch.
And BTW, did you replace _ALL_ electrolytic caps in the power supply except
those 2 big ones protruding through its cover? If you didn't start with this
first -- it is almost 100% probability that most (if not all) of those
capacitors are dead and your instrument just acts erratically.
There is absolutely no reason to even start doing anything on any part of
the instrument until you have the power supply refurbished.
You should also make sure the fan is working properly -- unlike some other
instruments in 49x it is CRITICAL. If it doesn't work your PS will overheat
in almost no time.
---
*
* KSI@home KOI8 Net < > The impossible we do immediately. *