TD744A Probe cal storage


 

Every time I power up my 'scope it complains that two (and only two)
probes have changed and "do you want to use the existing cal data or
re-cal".

If I choose to use the existing cal data it works fine, or if I
recalibrate the probe it works fine too but I still get challenged the
next time I power up.

Do I need to "enable write access" to the nvram when I re-cal a probe ??

Or is the nvram failing ??

The 'scope is working fine in all other respects.

Best regards,
Dave


 

I believe this is normal. It will do this anytime you connect a probe that
has readout scaling or power up with any connected. My 754C does the same
thing.

Dave Casey

On Sat, Jun 23, 2018 at 3:49 AM, David Slipper <softfoot@...> wrote:

Every time I power up my 'scope it complains that two (and only two)
probes have changed and "do you want to use the existing cal data or
re-cal".

If I choose to use the existing cal data it works fine, or if I
recalibrate the probe it works fine too but I still get challenged the
next time I power up.

Do I need to "enable write access" to the nvram when I re-cal a probe ??

Or is the nvram failing ??

The 'scope is working fine in all other respects.

Best regards,
Dave





 

The thing is - I have 4 identical probes on this 4 channel 'scope - but
only 2 channels do this.

Regards,
Dave

On 23/06/2018 10:17, Dave Casey wrote:
I believe this is normal. It will do this anytime you connect a probe that
has readout scaling or power up with any connected. My 754C does the same
thing.

Dave Casey

On Sat, Jun 23, 2018 at 3:49 AM, David Slipper <softfoot@...> wrote:

Every time I power up my 'scope it complains that two (and only two)
probes have changed and "do you want to use the existing cal data or
re-cal".

If I choose to use the existing cal data it works fine, or if I
recalibrate the probe it works fine too but I still get challenged the
next time I power up.

Do I need to "enable write access" to the nvram when I re-cal a probe ??

Or is the nvram failing ??

The 'scope is working fine in all other respects.

Best regards,
Dave





.


 

David, try switching the two probes on the channels that do this with the two that don't so you can check whether the problem is with the scope channels or with the probes.
--Eric
Sent from my Samsung Galaxy S8.

-------- Original message --------From: David Slipper <softfoot@...> Date: 6/23/18 7:25 AM (GMT-06:00) To: [email protected] Subject: Re: [TekScopes] TD744A Probe cal storage

The thing is - I have 4 identical probes on this 4 channel 'scope - but
only 2 channels do this.

Regards,
Dave



On 23/06/2018 10:17, Dave Casey wrote:
I believe this is normal. It will do this anytime you connect a probe that
has readout scaling or power up with any connected. My 754C does the same
thing.

Dave Casey

On Sat, Jun 23, 2018 at 3:49 AM, David Slipper <softfoot@...> wrote:

Every time I power up my 'scope it complains that two (and only two)
probes have changed and "do you want to use the existing cal data or
re-cal".

If I choose to use the existing cal data it works fine, or if I
recalibrate the probe it works fine too but I still get challenged the
next time I power up.

Do I need to "enable write access" to the nvram when I re-cal a probe ??

Or is the nvram failing ??

The 'scope is working fine in all other respects.

Best regards,
Dave





.