Hello everyone.
Initial test show an issue with channel 1.
Any direction on where to start?
Channel 2 displays inputs correctly from the FG.
Channel 1 displays 10 times the amplitude for the same signal.
Example a 1 volt sine wave on channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions at .5v / div.
The same signal on channel one reads as ~10v. When set to 5v / div it fills two div.
Thanks, KT
|
Some more checking shows channel 1 is not getting the code signal from the probe. Any idea on this circuit?
KT
toggle quoted message
Show quoted text
--- In TekScopes@..., "lldskjfoi" <k.templin@...> wrote: Hello everyone.
Initial test show an issue with channel 1.
Any direction on where to start?
Channel 2 displays inputs correctly from the FG.
Channel 1 displays 10 times the amplitude for the same signal.
Example a 1 volt sine wave on channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions at .5v / div.
The same signal on channel one reads as ~10v. When set to 5v / div it fills two div.
Thanks, KT
|
I know it is probably silly to ask, but are you sure that the
switch on one probe isn't set for the? x10 setting ?
PatM
============================================================
lldskjfoi wrote the following on 9/19/2012 9:59 PM:
?
Hello everyone.
Initial test show an issue with channel 1.
Any direction on where to start?
Channel 2 displays inputs correctly from the FG.
Channel 1 displays 10 times the amplitude for the same
signal.
Example a 1 volt sine wave on channel 2 displays on the
voltmeter function correctly at ~1v and 2 divisions at .5v
/ div.
The same signal on channel one reads as ~10v. When set to
5v / div it fills two div.
Thanks, KT
.
|
They on 10x non swtichable.
KT
toggle quoted message
Show quoted text
--- In TekScopes@..., Pat-W9ZO <malfam2@...> wrote:
I know it is probably silly to ask, but are you sure that the switch on one probe isn't set for the x10 setting ? PatM ============================================================ lldskjfoi wrote the following on 9/19/2012 9:59 PM:
Hello everyone.
Initial test show an issue with channel 1.
Any direction on where to start?
Channel 2 displays inputs correctly from the FG.
Channel 1 displays 10 times the amplitude for the same signal.
Example a 1 volt sine wave on channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions at .5v / div.
The same signal on channel one reads as ~10v. When set to 5v / div it fills two div.
Thanks, KT
.
|
Hi,
you should check, that there is no short on the "sensor" ring of the
input which shows
the 10X signal. Normally the sensor, when open tells 1:1 probe and
for probes
with 10:1 a.s.o. it is conncted via a resistor to ground !!!
Herbert
Am 20-09-2012 14:54, schrieb lldskjfoi:
toggle quoted message
Show quoted text
?
They on 10x non swtichable.
KT
--- In TekScopes@...,
Pat-W9ZO wrote:
>
>
> I know it is probably silly to ask, but are you sure
that the switch on
> one probe isn't set for the x10 setting ?
> PatM
>
============================================================
> lldskjfoi wrote the following on 9/19/2012 9:59 PM:
> >
> > Hello everyone.
> >
> > Initial test show an issue with channel 1.
> >
> > Any direction on where to start?
> >
> > Channel 2 displays inputs correctly from the FG.
> >
> > Channel 1 displays 10 times the amplitude for
the same signal.
> >
> > Example a 1 volt sine wave on channel 2 displays
on the voltmeter
> > function correctly at ~1v and 2 divisions at .5v
/ div.
> >
> > The same signal on channel one reads as ~10v.
When set to 5v / div it
> > fills two div.
> >
> > Thanks, KT
> >
> >
> .
>
|
Maybe start by looking at the #1 "Vertical Inputs" schematic
to see the simple circuit which starts at the junction of R101, C101 and
R105?for channel 1.
That junction point is connected to the readout ring behind
the BNC connector, so you could test for continuity and shorting to ground,
etc.
If you compared both ends of? R101 and R105 (Chan
1)?with those of R102 and R106 (Chan 2) you might see if the readout
voltage is getting that far.
If that looked okay, you could go to U506 and see if it
arrives at pins 12 and 15 as you would expect.
With the scope on its back, you have excellent access to all
of these resistors and U506 (16 pin DIP), with the blue outer cover?slid
back the width of your hand.
From reading the service manual, it appears that the readout
signal gets changed to a digital signal at U506, but is a simple voltage up to
U506.
tom jobe...
?
?
?
?
toggle quoted message
Show quoted text
----- Original Message -----
Sent: Wednesday, September 19, 2012 9:08
PM
Subject: [TekScopes] Re: 2247A Channel 1
issue
?
Some more checking shows channel 1 is not getting the code signal from the
probe. Any idea on this circuit?
KT
--- In TekScopes@...,
"lldskjfoi" wrote: > > Hello
everyone. > > Initial test show an issue with channel 1. >
> Any direction on where to start? > > Channel 2 displays
inputs correctly from the FG. > > Channel 1 displays 10 times the
amplitude for the same signal. > > Example a 1 volt sine wave on
channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions
at .5v / div. > > The same signal on channel one reads as ~10v.
When set to 5v / div it fills two div. > > Thanks,
KT >
|
Thanks, thanks the info I was looking for. I started getting into the schematic last night I saw the path to U506. Will start checking it out.
Also notice just letting the scope run with the probe attached the attenuation factor jumps back and forth- correct / incorrect. Does not change with no probe attached so it is getting an intermitant signal.
KT
toggle quoted message
Show quoted text
--- In TekScopes@..., "Tom Jobe" <tomjobe@...> wrote: Maybe start by looking at the #1 "Vertical Inputs" schematic to see the simple circuit which starts at the junction of R101, C101 and R105 for channel 1. That junction point is connected to the readout ring behind the BNC connector, so you could test for continuity and shorting to ground, etc. If you compared both ends of R101 and R105 (Chan 1) with those of R102 and R106 (Chan 2) you might see if the readout voltage is getting that far. If that looked okay, you could go to U506 and see if it arrives at pins 12 and 15 as you would expect. With the scope on its back, you have excellent access to all of these resistors and U506 (16 pin DIP), with the blue outer cover slid back the width of your hand. From reading the service manual, it appears that the readout signal gets changed to a digital signal at U506, but is a simple voltage up to U506. tom jobe...
----- Original Message ----- From: lldskjfoi To: TekScopes@... Sent: Wednesday, September 19, 2012 9:08 PM Subject: [TekScopes] Re: 2247A Channel 1 issue
Some more checking shows channel 1 is not getting the code signal from the probe. Any idea on this circuit?
KT
--- In TekScopes@..., "lldskjfoi" <k.templin@> wrote: > > Hello everyone. > > Initial test show an issue with channel 1. > > Any direction on where to start? > > Channel 2 displays inputs correctly from the FG. > > Channel 1 displays 10 times the amplitude for the same signal. > > Example a 1 volt sine wave on channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions at .5v / div. > > The same signal on channel one reads as ~10v. When set to 5v / div it fills two div. > > Thanks, KT >
|
For the record, I slid the cover back on a very dead 2246 and
checked pin 12 and pin 15 of U506, they both had +2.49 volts with no probes
attached.
This scope does not have any lights on the front, and none of
the buttons work, so I could not see?the reading I would get at ground, or
with a 1X or?a 10X probe.
tom jobe...
?
?
?
?
?
toggle quoted message
Show quoted text
----- Original Message -----
Sent: Thursday, September 20, 2012 8:28
AM
Subject: [TekScopes] Re: 2247A Channel 1
issue
?
Thanks, thanks the info I was looking for. I started getting into the
schematic last night I saw the path to U506. Will start checking it
out.
Also notice just letting the scope run with the probe attached the
attenuation factor jumps back and forth- correct / incorrect. Does not change
with no probe attached so it is getting an intermitant
signal.
KT
--- In TekScopes@..., "Tom
Jobe" wrote: > > Maybe start by looking at the
#1 "Vertical Inputs" schematic to see the simple circuit which starts at the
junction of R101, C101 and R105 for channel 1. > That junction point is
connected to the readout ring behind the BNC connector, so you could test for
continuity and shorting to ground, etc. > If you compared both ends of
R101 and R105 (Chan 1) with those of R102 and R106 (Chan 2) you might see if
the readout voltage is getting that far. > If that looked okay, you
could go to U506 and see if it arrives at pins 12 and 15 as you would
expect. > With the scope on its back, you have excellent access to all
of these resistors and U506 (16 pin DIP), with the blue outer cover slid back
the width of your hand. > From reading the service manual, it appears
that the readout signal gets changed to a digital signal at U506, but is a
simple voltage up to U506. > tom jobe... > > >
> > ----- Original Message ----- > From: lldskjfoi
> To: TekScopes@...
> Sent: Wednesday, September 19, 2012 9:08 PM > Subject:
[TekScopes] Re: 2247A Channel 1 issue > > > > Some
more checking shows channel 1 is not getting the code signal from the probe.
Any idea on this circuit? > > KT > > --- In TekScopes@...,
"lldskjfoi" wrote: > > > > Hello
everyone. > > > > Initial test show an issue with channel
1. > > > > Any direction on where to start? > >
> > Channel 2 displays inputs correctly from the FG. > >
> > Channel 1 displays 10 times the amplitude for the same
signal. > > > > Example a 1 volt sine wave on channel 2
displays on the voltmeter function correctly at ~1v and 2 divisions at .5v /
div. > > > > The same signal on channel one reads as ~10v.
When set to 5v / div it fills two div. > > > > Thanks,
KT > > >
|
Thanks again for the responce. I think we are getting there.
Using cahannel 2 as the "working reference"
pin 12 (channel 2) on u506 goes from the +2.5 to +1.4 when a 10x probe is connected.
pin (15 channel 1) on U506 goes from +.9 to +.6 with the probe connected.
Checking R102 for channel 2 it has +2.5
R101, (channel 1), shows +2.5 in but only +1.8 out.
What are your thoughts? Is it simply a bad resistor or would one of the cap, C111 cause the voltage drop?
Thanks again, KT
toggle quoted message
Show quoted text
--- In TekScopes@..., "Tom Jobe" <tomjobe@...> wrote: For the record, I slid the cover back on a very dead 2246 and checked pin 12 and pin 15 of U506, they both had +2.49 volts with no probes attached. This scope does not have any lights on the front, and none of the buttons work, so I could not see the reading I would get at ground, or with a 1X or a 10X probe. tom jobe...
----- Original Message ----- From: lldskjfoi To: TekScopes@... Sent: Thursday, September 20, 2012 8:28 AM Subject: [TekScopes] Re: 2247A Channel 1 issue
Thanks, thanks the info I was looking for. I started getting into the schematic last night I saw the path to U506. Will start checking it out.
Also notice just letting the scope run with the probe attached the attenuation factor jumps back and forth- correct / incorrect. Does not change with no probe attached so it is getting an intermitant signal.
KT
--- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > Maybe start by looking at the #1 "Vertical Inputs" schematic to see the simple circuit which starts at the junction of R101, C101 and R105 for channel 1. > That junction point is connected to the readout ring behind the BNC connector, so you could test for continuity and shorting to ground, etc. > If you compared both ends of R101 and R105 (Chan 1) with those of R102 and R106 (Chan 2) you might see if the readout voltage is getting that far. > If that looked okay, you could go to U506 and see if it arrives at pins 12 and 15 as you would expect. > With the scope on its back, you have excellent access to all of these resistors and U506 (16 pin DIP), with the blue outer cover slid back the width of your hand. > From reading the service manual, it appears that the readout signal gets changed to a digital signal at U506, but is a simple voltage up to U506. > tom jobe... > > > > > ----- Original Message ----- > From: lldskjfoi > To: TekScopes@... > Sent: Wednesday, September 19, 2012 9:08 PM > Subject: [TekScopes] Re: 2247A Channel 1 issue > > > > Some more checking shows channel 1 is not getting the code signal from the probe. Any idea on this circuit? > > KT > > --- In TekScopes@..., "lldskjfoi" <k.templin@> wrote: > > > > Hello everyone. > > > > Initial test show an issue with channel 1. > > > > Any direction on where to start? > > > > Channel 2 displays inputs correctly from the FG. > > > > Channel 1 displays 10 times the amplitude for the same signal. > > > > Example a 1 volt sine wave on channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions at .5v / div. > > > > The same signal on channel one reads as ~10v. When set to 5v / div it fills two div. > > > > Thanks, KT > > >
|
I just checked R101 and R102 on a bare 2246 mainboard and they
both measured 8.2k ohms in circuit.
If the resistance is correct on your scope, then something is
drawing enough current to drop 0.7 volts across R101.
Nothing in that simple circuit looks like a suspect to me, no
nasty 10uF to 47uF tantalums, etc.
tom jobe...
?
?
?
toggle quoted message
Show quoted text
----- Original Message -----
Sent: Thursday, September 20, 2012 1:20
PM
Subject: [TekScopes] Re: 2247A Channel 1
issue
?
Thanks again for the responce. I think we are getting
there.
Using cahannel 2 as the "working reference"
pin 12
(channel 2) on u506 goes from the +2.5 to +1.4 when a 10x probe is
connected.
pin (15 channel 1) on U506 goes from +.9 to +.6 with the
probe connected.
Checking R102 for channel 2 it has +2.5
R101,
(channel 1), shows +2.5 in but only +1.8 out.
What are your thoughts?
Is it simply a bad resistor or would one of the cap, C111 cause the voltage
drop?
Thanks again, KT --- In TekScopes@..., "Tom
Jobe" wrote: > > For the record, I slid the
cover back on a very dead 2246 and checked pin 12 and pin 15 of U506, they
both had +2.49 volts with no probes attached. > This scope does not have
any lights on the front, and none of the buttons work, so I could not see the
reading I would get at ground, or with a 1X or a 10X probe. > tom
jobe... > > > > > > ----- Original
Message ----- > From: lldskjfoi > To: TekScopes@...
> Sent: Thursday, September 20, 2012 8:28 AM > Subject:
[TekScopes] Re: 2247A Channel 1 issue > > > >
Thanks, thanks the info I was looking for. I started getting into the
schematic last night I saw the path to U506. Will start checking it
out. > > Also notice just letting the scope run with the probe
attached the attenuation factor jumps back and forth- correct / incorrect.
Does not change with no probe attached so it is getting an intermitant
signal. > > KT > > --- In TekScopes@..., "Tom
Jobe" wrote: > > > > Maybe start by looking
at the #1 "Vertical Inputs" schematic to see the simple circuit which starts
at the junction of R101, C101 and R105 for channel 1. > > That
junction point is connected to the readout ring behind the BNC connector, so
you could test for continuity and shorting to ground, etc. > > If you
compared both ends of R101 and R105 (Chan 1) with those of R102 and R106 (Chan
2) you might see if the readout voltage is getting that far. > > If
that looked okay, you could go to U506 and see if it arrives at pins 12 and 15
as you would expect. > > With the scope on its back, you have
excellent access to all of these resistors and U506 (16 pin DIP), with the
blue outer cover slid back the width of your hand. > > From reading
the service manual, it appears that the readout signal gets changed to a
digital signal at U506, but is a simple voltage up to U506. > > tom
jobe... > > > > > > > > > >
----- Original Message ----- > > From: lldskjfoi > > To:
TekScopes@...
> > Sent: Wednesday, September 19, 2012 9:08 PM > >
Subject: [TekScopes] Re: 2247A Channel 1 issue > > > >
> > > > Some more checking shows channel 1 is not getting
the code signal from the probe. Any idea on this circuit? > >
> > KT > > > > --- In TekScopes@...,
"lldskjfoi" wrote: > > > > > >
Hello everyone. > > > > > > Initial test show an
issue with channel 1. > > > > > > Any direction on
where to start? > > > > > > Channel 2 displays inputs
correctly from the FG. > > > > > > Channel 1 displays
10 times the amplitude for the same signal. > > > > >
> Example a 1 volt sine wave on channel 2 displays on the voltmeter
function correctly at ~1v and 2 divisions at .5v / div. > > >
> > > The same signal on channel one reads as ~10v. When set to
5v / div it fills two div. > > > > > > Thanks,
KT > > > > > >
|
I get 8.2 and 8.2 as well.
KT
toggle quoted message
Show quoted text
--- In TekScopes@..., "Tom Jobe" <tomjobe@...> wrote: I just checked R101 and R102 on a bare 2246 mainboard and they both measured 8.2k ohms in circuit. If the resistance is correct on your scope, then something is drawing enough current to drop 0.7 volts across R101. Nothing in that simple circuit looks like a suspect to me, no nasty 10uF to 47uF tantalums, etc. tom jobe...
----- Original Message ----- From: lldskjfoi To: TekScopes@... Sent: Thursday, September 20, 2012 1:20 PM Subject: [TekScopes] Re: 2247A Channel 1 issue
Thanks again for the responce. I think we are getting there.
Using cahannel 2 as the "working reference"
pin 12 (channel 2) on u506 goes from the +2.5 to +1.4 when a 10x probe is connected.
pin (15 channel 1) on U506 goes from +.9 to +.6 with the probe connected.
Checking R102 for channel 2 it has +2.5
R101, (channel 1), shows +2.5 in but only +1.8 out.
What are your thoughts? Is it simply a bad resistor or would one of the cap, C111 cause the voltage drop?
Thanks again, KT --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > For the record, I slid the cover back on a very dead 2246 and checked pin 12 and pin 15 of U506, they both had +2.49 volts with no probes attached. > This scope does not have any lights on the front, and none of the buttons work, so I could not see the reading I would get at ground, or with a 1X or a 10X probe. > tom jobe... > > > > > > ----- Original Message ----- > From: lldskjfoi > To: TekScopes@... > Sent: Thursday, September 20, 2012 8:28 AM > Subject: [TekScopes] Re: 2247A Channel 1 issue > > > > Thanks, thanks the info I was looking for. I started getting into the schematic last night I saw the path to U506. Will start checking it out. > > Also notice just letting the scope run with the probe attached the attenuation factor jumps back and forth- correct / incorrect. Does not change with no probe attached so it is getting an intermitant signal. > > KT > > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > > > Maybe start by looking at the #1 "Vertical Inputs" schematic to see the simple circuit which starts at the junction of R101, C101 and R105 for channel 1. > > That junction point is connected to the readout ring behind the BNC connector, so you could test for continuity and shorting to ground, etc. > > If you compared both ends of R101 and R105 (Chan 1) with those of R102 and R106 (Chan 2) you might see if the readout voltage is getting that far. > > If that looked okay, you could go to U506 and see if it arrives at pins 12 and 15 as you would expect. > > With the scope on its back, you have excellent access to all of these resistors and U506 (16 pin DIP), with the blue outer cover slid back the width of your hand. > > From reading the service manual, it appears that the readout signal gets changed to a digital signal at U506, but is a simple voltage up to U506. > > tom jobe... > > > > > > > > > > ----- Original Message ----- > > From: lldskjfoi > > To: TekScopes@... > > Sent: Wednesday, September 19, 2012 9:08 PM > > Subject: [TekScopes] Re: 2247A Channel 1 issue > > > > > > > > Some more checking shows channel 1 is not getting the code signal from the probe. Any idea on this circuit? > > > > KT > > > > --- In TekScopes@..., "lldskjfoi" <k.templin@> wrote: > > > > > > Hello everyone. > > > > > > Initial test show an issue with channel 1. > > > > > > Any direction on where to start? > > > > > > Channel 2 displays inputs correctly from the FG. > > > > > > Channel 1 displays 10 times the amplitude for the same signal. > > > > > > Example a 1 volt sine wave on channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions at .5v / div. > > > > > > The same signal on channel one reads as ~10v. When set to 5v / div it fills two div. > > > > > > Thanks, KT > > > > > >
|
What are the resistance readings across R105 and
R106?
That might help narrow the search down.
tom jobe...
?
?
?
toggle quoted message
Show quoted text
----- Original Message -----
Sent: Thursday, September 20, 2012 5:19
PM
Subject: [TekScopes] Re: 2247A Channel 1
issue
?
I get 8.2 and 8.2 as well.
KT --- In TekScopes@..., "Tom
Jobe" wrote: > > I just checked R101 and R102
on a bare 2246 mainboard and they both measured 8.2k ohms in circuit. >
If the resistance is correct on your scope, then something is drawing enough
current to drop 0.7 volts across R101. > Nothing in that simple circuit
looks like a suspect to me, no nasty 10uF to 47uF tantalums, etc. > tom
jobe... > > > > ----- Original Message -----
> From: lldskjfoi > To: TekScopes@...
> Sent: Thursday, September 20, 2012 1:20 PM > Subject:
[TekScopes] Re: 2247A Channel 1 issue > > > >
> Thanks again for the responce. I think we are getting there. >
> Using cahannel 2 as the "working reference" > > pin 12
(channel 2) on u506 goes from the +2.5 to +1.4 when a 10x probe is
connected. > > pin (15 channel 1) on U506 goes from +.9 to +.6
with the probe connected. > > Checking R102 for channel 2 it has
+2.5 > > R101, (channel 1), shows +2.5 in but only +1.8
out. > > What are your thoughts? Is it simply a bad resistor or
would one of the cap, C111 cause the voltage drop? > > Thanks
again, KT > --- In TekScopes@..., "Tom
Jobe" wrote: > > > > For the record, I slid
the cover back on a very dead 2246 and checked pin 12 and pin 15 of U506, they
both had +2.49 volts with no probes attached. > > This scope does not
have any lights on the front, and none of the buttons work, so I could not see
the reading I would get at ground, or with a 1X or a 10X probe. > >
tom jobe... > > > > > > > > >
> > > ----- Original Message ----- > > From: lldskjfoi
> > To: TekScopes@...
> > Sent: Thursday, September 20, 2012 8:28 AM > > Subject:
[TekScopes] Re: 2247A Channel 1 issue > > > > > >
> > Thanks, thanks the info I was looking for. I started getting
into the schematic last night I saw the path to U506. Will start checking it
out. > > > > Also notice just letting the scope run with
the probe attached the attenuation factor jumps back and forth- correct /
incorrect. Does not change with no probe attached so it is getting an
intermitant signal. > > > > KT > > > >
--- In TekScopes@..., "Tom
Jobe" wrote: > > > > > > Maybe start
by looking at the #1 "Vertical Inputs" schematic to see the simple circuit
which starts at the junction of R101, C101 and R105 for channel 1. >
> > That junction point is connected to the readout ring behind the BNC
connector, so you could test for continuity and shorting to ground,
etc. > > > If you compared both ends of R101 and R105 (Chan 1)
with those of R102 and R106 (Chan 2) you might see if the readout voltage is
getting that far. > > > If that looked okay, you could go to U506
and see if it arrives at pins 12 and 15 as you would expect. > > >
With the scope on its back, you have excellent access to all of these
resistors and U506 (16 pin DIP), with the blue outer cover slid back the width
of your hand. > > > From reading the service manual, it appears
that the readout signal gets changed to a digital signal at U506, but is a
simple voltage up to U506. > > > tom jobe... > > >
> > > > > > > > > > > >
----- Original Message ----- > > > From: lldskjfoi > >
> To: TekScopes@...
> > > Sent: Wednesday, September 19, 2012 9:08 PM > >
> Subject: [TekScopes] Re: 2247A Channel 1 issue > > > >
> > > > > > > > Some more checking shows
channel 1 is not getting the code signal from the probe. Any idea on this
circuit? > > > > > > KT > > > >
> > --- In TekScopes@...,
"lldskjfoi" wrote: > > > > > > >
> Hello everyone. > > > > > > > > Initial
test show an issue with channel 1. > > > > > > >
> Any direction on where to start? > > > > > >
> > Channel 2 displays inputs correctly from the FG. > > >
> > > > > Channel 1 displays 10 times the amplitude for the
same signal. > > > > > > > > Example a 1 volt
sine wave on channel 2 displays on the voltmeter function correctly at ~1v and
2 divisions at .5v / div. > > > > > > > > The
same signal on channel one reads as ~10v. When set to 5v / div it fills two
div. > > > > > > > > Thanks, KT > >
> > > > > > > >
|
10k and 10k. What do you thin? the only thing left are the caps or the U506 itself.
KT
toggle quoted message
Show quoted text
--- In TekScopes@..., "Tom Jobe" <tomjobe@...> wrote: What are the resistance readings across R105 and R106? That might help narrow the search down. tom jobe...
----- Original Message ----- From: lldskjfoi To: TekScopes@... Sent: Thursday, September 20, 2012 5:19 PM Subject: [TekScopes] Re: 2247A Channel 1 issue
I get 8.2 and 8.2 as well.
KT --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > I just checked R101 and R102 on a bare 2246 mainboard and they both measured 8.2k ohms in circuit. > If the resistance is correct on your scope, then something is drawing enough current to drop 0.7 volts across R101. > Nothing in that simple circuit looks like a suspect to me, no nasty 10uF to 47uF tantalums, etc. > tom jobe... > > > > ----- Original Message ----- > From: lldskjfoi > To: TekScopes@... > Sent: Thursday, September 20, 2012 1:20 PM > Subject: [TekScopes] Re: 2247A Channel 1 issue > > > > > Thanks again for the responce. I think we are getting there. > > Using cahannel 2 as the "working reference" > > pin 12 (channel 2) on u506 goes from the +2.5 to +1.4 when a 10x probe is connected. > > pin (15 channel 1) on U506 goes from +.9 to +.6 with the probe connected. > > Checking R102 for channel 2 it has +2.5 > > R101, (channel 1), shows +2.5 in but only +1.8 out. > > What are your thoughts? Is it simply a bad resistor or would one of the cap, C111 cause the voltage drop? > > Thanks again, KT > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > > > For the record, I slid the cover back on a very dead 2246 and checked pin 12 and pin 15 of U506, they both had +2.49 volts with no probes attached. > > This scope does not have any lights on the front, and none of the buttons work, so I could not see the reading I would get at ground, or with a 1X or a 10X probe. > > tom jobe... > > > > > > > > > > > > ----- Original Message ----- > > From: lldskjfoi > > To: TekScopes@... > > Sent: Thursday, September 20, 2012 8:28 AM > > Subject: [TekScopes] Re: 2247A Channel 1 issue > > > > > > > > Thanks, thanks the info I was looking for. I started getting into the schematic last night I saw the path to U506. Will start checking it out. > > > > Also notice just letting the scope run with the probe attached the attenuation factor jumps back and forth- correct / incorrect. Does not change with no probe attached so it is getting an intermitant signal. > > > > KT > > > > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > > > > > Maybe start by looking at the #1 "Vertical Inputs" schematic to see the simple circuit which starts at the junction of R101, C101 and R105 for channel 1. > > > That junction point is connected to the readout ring behind the BNC connector, so you could test for continuity and shorting to ground, etc. > > > If you compared both ends of R101 and R105 (Chan 1) with those of R102 and R106 (Chan 2) you might see if the readout voltage is getting that far. > > > If that looked okay, you could go to U506 and see if it arrives at pins 12 and 15 as you would expect. > > > With the scope on its back, you have excellent access to all of these resistors and U506 (16 pin DIP), with the blue outer cover slid back the width of your hand. > > > From reading the service manual, it appears that the readout signal gets changed to a digital signal at U506, but is a simple voltage up to U506. > > > tom jobe... > > > > > > > > > > > > > > > ----- Original Message ----- > > > From: lldskjfoi > > > To: TekScopes@... > > > Sent: Wednesday, September 19, 2012 9:08 PM > > > Subject: [TekScopes] Re: 2247A Channel 1 issue > > > > > > > > > > > > Some more checking shows channel 1 is not getting the code signal from the probe. Any idea on this circuit? > > > > > > KT > > > > > > --- In TekScopes@..., "lldskjfoi" <k.templin@> wrote: > > > > > > > > Hello everyone. > > > > > > > > Initial test show an issue with channel 1. > > > > > > > > Any direction on where to start? > > > > > > > > Channel 2 displays inputs correctly from the FG. > > > > > > > > Channel 1 displays 10 times the amplitude for the same signal. > > > > > > > > Example a 1 volt sine wave on channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions at .5v / div. > > > > > > > > The same signal on channel one reads as ~10v. When set to 5v / div it fills two div. > > > > > > > > Thanks, KT > > > > > > > > > >
|
What is the voltages?on either side of?R105 and
R106?
It seems like C111 and U506 are the only possibilities if the
voltage leaving R105 more or less agrees with what you see arriving at U506 pin
15.
tom jobe...
?
?
?
toggle quoted message
Show quoted text
----- Original Message -----
Sent: Thursday, September 20, 2012 6:04
PM
Subject: [TekScopes] Re: 2247A Channel 1
issue
?
10k and 10k. What do you thin? the only thing left are the caps or the
U506 itself.
KT --- In TekScopes@..., "Tom
Jobe" wrote: > > What are the resistance
readings across R105 and R106? > That might help narrow the search
down. > tom jobe... > > > > ----- Original
Message ----- > From: lldskjfoi > To: TekScopes@... >
Sent: Thursday, September 20, 2012 5:19 PM > Subject: [TekScopes] Re:
2247A Channel 1 issue > > > > > I get 8.2
and 8.2 as well. > > KT > --- In TekScopes@..., "Tom
Jobe" wrote: > > > > I just checked R101
and R102 on a bare 2246 mainboard and they both measured 8.2k ohms in
circuit. > > If the resistance is correct on your scope, then
something is drawing enough current to drop 0.7 volts across R101. >
> Nothing in that simple circuit looks like a suspect to me, no nasty 10uF
to 47uF tantalums, etc. > > tom jobe... > > > >
> > > > ----- Original Message ----- > > From:
lldskjfoi > > To: TekScopes@...
> > Sent: Thursday, September 20, 2012 1:20 PM > > Subject:
[TekScopes] Re: 2247A Channel 1 issue > > > > > >
> > > > Thanks again for the responce. I think we are
getting there. > > > > Using cahannel 2 as the "working
reference" > > > > pin 12 (channel 2) on u506 goes from the
+2.5 to +1.4 when a 10x probe is connected. > > > > pin (15
channel 1) on U506 goes from +.9 to +.6 with the probe connected. > >
> > Checking R102 for channel 2 it has +2.5 > > >
> R101, (channel 1), shows +2.5 in but only +1.8 out. > > >
> What are your thoughts? Is it simply a bad resistor or would one of the
cap, C111 cause the voltage drop? > > > > Thanks again,
KT > > --- In TekScopes@..., "Tom
Jobe" wrote: > > > > > > For the
record, I slid the cover back on a very dead 2246 and checked pin 12 and pin
15 of U506, they both had +2.49 volts with no probes attached. > >
> This scope does not have any lights on the front, and none of the buttons
work, so I could not see the reading I would get at ground, or with a 1X or a
10X probe. > > > tom jobe... > > > > > >
> > > > > > > > > > > >
----- Original Message ----- > > > From: lldskjfoi > >
> To: TekScopes@...
> > > Sent: Thursday, September 20, 2012 8:28 AM > >
> Subject: [TekScopes] Re: 2247A Channel 1 issue > > > >
> > > > > > > > Thanks, thanks the info I was
looking for. I started getting into the schematic last night I saw the path to
U506. Will start checking it out. > > > > > > Also
notice just letting the scope run with the probe attached the attenuation
factor jumps back and forth- correct / incorrect. Does not change with no
probe attached so it is getting an intermitant signal. > > >
> > > KT > > > > > > --- In TekScopes@..., "Tom
Jobe" wrote: > > > > > > > >
Maybe start by looking at the #1 "Vertical Inputs" schematic to see the simple
circuit which starts at the junction of R101, C101 and R105 for channel
1. > > > > That junction point is connected to the readout ring
behind the BNC connector, so you could test for continuity and shorting to
ground, etc. > > > > If you compared both ends of R101 and R105
(Chan 1) with those of R102 and R106 (Chan 2) you might see if the readout
voltage is getting that far. > > > > If that looked okay, you
could go to U506 and see if it arrives at pins 12 and 15 as you would
expect. > > > > With the scope on its back, you have excellent
access to all of these resistors and U506 (16 pin DIP), with the blue outer
cover slid back the width of your hand. > > > > From reading
the service manual, it appears that the readout signal gets changed to a
digital signal at U506, but is a simple voltage up to U506. > > >
> tom jobe... > > > > > > > > > >
> > > > > > > > > > ----- Original
Message ----- > > > > From: lldskjfoi > > > >
To: TekScopes@...
> > > > Sent: Wednesday, September 19, 2012 9:08 PM >
> > > Subject: [TekScopes] Re: 2247A Channel 1 issue > >
> > > > > > > > > > > > >
> Some more checking shows channel 1 is not getting the code signal from
the probe. Any idea on this circuit? > > > > > > >
> KT > > > > > > > > --- In TekScopes@...,
"lldskjfoi" wrote: > > > > > > >
> > > Hello everyone. > > > > > > > >
> > Initial test show an issue with channel 1. > > > >
> > > > > > Any direction on where to start? >
> > > > > > > > > Channel 2 displays inputs
correctly from the FG. > > > > > > > > >
> Channel 1 displays 10 times the amplitude for the same signal. >
> > > > > > > > > Example a 1 volt sine wave on
channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions
at .5v / div. > > > > > > > > > > The
same signal on channel one reads as ~10v. When set to 5v / div it fills two
div. > > > > > > > > > > Thanks,
KT > > > > > > > > > > >
> > > >
|
I have 2.5 at both ends of R106.
1.9 and 1 at R105.
toggle quoted message
Show quoted text
--- In TekScopes@..., "Tom Jobe" <tomjobe@...> wrote: What is the voltages on either side of R105 and R106? It seems like C111 and U506 are the only possibilities if the voltage leaving R105 more or less agrees with what you see arriving at U506 pin 15. tom jobe...
----- Original Message ----- From: lldskjfoi To: TekScopes@... Sent: Thursday, September 20, 2012 6:04 PM Subject: [TekScopes] Re: 2247A Channel 1 issue
10k and 10k. What do you thin? the only thing left are the caps or the U506 itself.
KT --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > What are the resistance readings across R105 and R106? > That might help narrow the search down. > tom jobe... > > > > ----- Original Message ----- > From: lldskjfoi > To: TekScopes@... > Sent: Thursday, September 20, 2012 5:19 PM > Subject: [TekScopes] Re: 2247A Channel 1 issue > > > > > I get 8.2 and 8.2 as well. > > KT > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > > > I just checked R101 and R102 on a bare 2246 mainboard and they both measured 8.2k ohms in circuit. > > If the resistance is correct on your scope, then something is drawing enough current to drop 0.7 volts across R101. > > Nothing in that simple circuit looks like a suspect to me, no nasty 10uF to 47uF tantalums, etc. > > tom jobe... > > > > > > > > ----- Original Message ----- > > From: lldskjfoi > > To: TekScopes@... > > Sent: Thursday, September 20, 2012 1:20 PM > > Subject: [TekScopes] Re: 2247A Channel 1 issue > > > > > > > > > > Thanks again for the responce. I think we are getting there. > > > > Using cahannel 2 as the "working reference" > > > > pin 12 (channel 2) on u506 goes from the +2.5 to +1.4 when a 10x probe is connected. > > > > pin (15 channel 1) on U506 goes from +.9 to +.6 with the probe connected. > > > > Checking R102 for channel 2 it has +2.5 > > > > R101, (channel 1), shows +2.5 in but only +1.8 out. > > > > What are your thoughts? Is it simply a bad resistor or would one of the cap, C111 cause the voltage drop? > > > > Thanks again, KT > > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > > > > > For the record, I slid the cover back on a very dead 2246 and checked pin 12 and pin 15 of U506, they both had +2.49 volts with no probes attached. > > > This scope does not have any lights on the front, and none of the buttons work, so I could not see the reading I would get at ground, or with a 1X or a 10X probe. > > > tom jobe... > > > > > > > > > > > > > > > > > > ----- Original Message ----- > > > From: lldskjfoi > > > To: TekScopes@... > > > Sent: Thursday, September 20, 2012 8:28 AM > > > Subject: [TekScopes] Re: 2247A Channel 1 issue > > > > > > > > > > > > Thanks, thanks the info I was looking for. I started getting into the schematic last night I saw the path to U506. Will start checking it out. > > > > > > Also notice just letting the scope run with the probe attached the attenuation factor jumps back and forth- correct / incorrect. Does not change with no probe attached so it is getting an intermitant signal. > > > > > > KT > > > > > > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > > > > > > > Maybe start by looking at the #1 "Vertical Inputs" schematic to see the simple circuit which starts at the junction of R101, C101 and R105 for channel 1. > > > > That junction point is connected to the readout ring behind the BNC connector, so you could test for continuity and shorting to ground, etc. > > > > If you compared both ends of R101 and R105 (Chan 1) with those of R102 and R106 (Chan 2) you might see if the readout voltage is getting that far. > > > > If that looked okay, you could go to U506 and see if it arrives at pins 12 and 15 as you would expect. > > > > With the scope on its back, you have excellent access to all of these resistors and U506 (16 pin DIP), with the blue outer cover slid back the width of your hand. > > > > From reading the service manual, it appears that the readout signal gets changed to a digital signal at U506, but is a simple voltage up to U506. > > > > tom jobe... > > > > > > > > > > > > > > > > > > > > ----- Original Message ----- > > > > From: lldskjfoi > > > > To: TekScopes@... > > > > Sent: Wednesday, September 19, 2012 9:08 PM > > > > Subject: [TekScopes] Re: 2247A Channel 1 issue > > > > > > > > > > > > > > > > Some more checking shows channel 1 is not getting the code signal from the probe. Any idea on this circuit? > > > > > > > > KT > > > > > > > > --- In TekScopes@..., "lldskjfoi" <k.templin@> wrote: > > > > > > > > > > Hello everyone. > > > > > > > > > > Initial test show an issue with channel 1. > > > > > > > > > > Any direction on where to start? > > > > > > > > > > Channel 2 displays inputs correctly from the FG. > > > > > > > > > > Channel 1 displays 10 times the amplitude for the same signal. > > > > > > > > > > Example a 1 volt sine wave on channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions at .5v / div. > > > > > > > > > > The same signal on channel one reads as ~10v. When set to 5v / div it fills two div. > > > > > > > > > > Thanks, KT > > > > > > > > > > > > > > >
|
Well, unless I missed something you are down to C111 or U506
as being the culprit.
I would carefully lift one leg of C111.
It is near U506 pin 14, slightly further from the front of the
scope.
It might also be orange in color.
tom jobe...
?
toggle quoted message
Show quoted text
----- Original Message -----
Sent: Thursday, September 20, 2012 6:27
PM
Subject: [TekScopes] Re: 2247A Channel 1
issue
?
I have 2.5 at both ends of R106.
1.9 and 1 at R105.
---
In TekScopes@...,
"Tom Jobe" wrote: > > What is the voltages on
either side of R105 and R106? > It seems like C111 and U506 are the only
possibilities if the voltage leaving R105 more or less agrees with what you
see arriving at U506 pin 15. > tom jobe... > > >
> ----- Original Message ----- > From: lldskjfoi > To: TekScopes@...
> Sent: Thursday, September 20, 2012 6:04 PM > Subject:
[TekScopes] Re: 2247A Channel 1 issue > > > >
> 10k and 10k. What do you thin? the only thing left are the caps or
the U506 itself. > > KT > --- In TekScopes@..., "Tom
Jobe" wrote: > > > > What are the
resistance readings across R105 and R106? > > That might help narrow
the search down. > > tom jobe... > > > > >
> > > ----- Original Message ----- > > From: lldskjfoi
> > To: TekScopes@...
> > Sent: Thursday, September 20, 2012 5:19 PM > > Subject:
[TekScopes] Re: 2247A Channel 1 issue > > > > > >
> > > > I get 8.2 and 8.2 as well. > > >
> KT > > --- In TekScopes@..., "Tom
Jobe" wrote: > > > > > > I just
checked R101 and R102 on a bare 2246 mainboard and they both measured 8.2k
ohms in circuit. > > > If the resistance is correct on your scope,
then something is drawing enough current to drop 0.7 volts across
R101. > > > Nothing in that simple circuit looks like a suspect to
me, no nasty 10uF to 47uF tantalums, etc. > > > tom
jobe... > > > > > > > > > > >
> ----- Original Message ----- > > > From: lldskjfoi >
> > To: TekScopes@...
> > > Sent: Thursday, September 20, 2012 1:20 PM > >
> Subject: [TekScopes] Re: 2247A Channel 1 issue > > > >
> > > > > > > > > > > Thanks
again for the responce. I think we are getting there. > > >
> > > Using cahannel 2 as the "working reference" > >
> > > > pin 12 (channel 2) on u506 goes from the +2.5 to +1.4
when a 10x probe is connected. > > > > > > pin (15
channel 1) on U506 goes from +.9 to +.6 with the probe connected. > >
> > > > Checking R102 for channel 2 it has +2.5 > >
> > > > R101, (channel 1), shows +2.5 in but only +1.8
out. > > > > > > What are your thoughts? Is it simply
a bad resistor or would one of the cap, C111 cause the voltage drop? >
> > > > > Thanks again, KT > > > --- In TekScopes@..., "Tom
Jobe" <tomjobe@> wrote: > > > > > > > >
For the record, I slid the cover back on a very dead 2246 and checked pin 12
and pin 15 of U506, they both had +2.49 volts with no probes attached. >
> > > This scope does not have any lights on the front, and none of
the buttons work, so I could not see the reading I would get at ground, or
with a 1X or a 10X probe. > > > > tom jobe... > > >
> > > > > > > > > > > > >
> > > > > > > > ----- Original Message -----
> > > > From: lldskjfoi > > > > To: TekScopes@...
> > > > Sent: Thursday, September 20, 2012 8:28 AM >
> > > Subject: [TekScopes] Re: 2247A Channel 1 issue > >
> > > > > > > > > > > > >
> Thanks, thanks the info I was looking for. I started getting into the
schematic last night I saw the path to U506. Will start checking it
out. > > > > > > > > Also notice just letting
the scope run with the probe attached the attenuation factor jumps back and
forth- correct / incorrect. Does not change with no probe attached so it is
getting an intermitant signal. > > > > > > > >
KT > > > > > > > > --- In TekScopes@..., "Tom
Jobe" wrote: > > > > > > > >
> > Maybe start by looking at the #1 "Vertical Inputs" schematic to see
the simple circuit which starts at the junction of R101, C101 and R105 for
channel 1. > > > > > That junction point is connected to the
readout ring behind the BNC connector, so you could test for continuity and
shorting to ground, etc. > > > > > If you compared both ends
of R101 and R105 (Chan 1) with those of R102 and R106 (Chan 2) you might see
if the readout voltage is getting that far. > > > > > If
that looked okay, you could go to U506 and see if it arrives at pins 12 and 15
as you would expect. > > > > > With the scope on its back,
you have excellent access to all of these resistors and U506 (16 pin DIP),
with the blue outer cover slid back the width of your hand. > > >
> > From reading the service manual, it appears that the readout signal
gets changed to a digital signal at U506, but is a simple voltage up to
U506. > > > > > tom jobe... > > > > >
> > > > > > > > > > > > >
> > > > > > > ----- Original Message ----- >
> > > > From: lldskjfoi > > > > > To: TekScopes@...
> > > > > Sent: Wednesday, September 19, 2012 9:08
PM > > > > > Subject: [TekScopes] Re: 2247A Channel 1
issue > > > > > > > > > > > >
> > > > > > > > Some more checking shows channel 1
is not getting the code signal from the probe. Any idea on this
circuit? > > > > > > > > > > KT >
> > > > > > > > > --- In TekScopes@...,
"lldskjfoi" <k.templin@> wrote: > > > > > > >
> > > > > Hello everyone. > > > > > >
> > > > > > Initial test show an issue with channel
1. > > > > > > > > > > > > Any
direction on where to start? > > > > > > > >
> > > > Channel 2 displays inputs correctly from the FG. >
> > > > > > > > > > > Channel 1 displays
10 times the amplitude for the same signal. > > > > > >
> > > > > > Example a 1 volt sine wave on channel 2
displays on the voltmeter function correctly at ~1v and 2 divisions at .5v /
div. > > > > > > > > > > > > The
same signal on channel one reads as ~10v. When set to 5v / div it fills two
div. > > > > > > > > > > > >
Thanks, KT > > > > > > > > > >
> > > > > > > > >
> >
|
KT,
I money is on U506 has a bad input at pin 15. Lots of current's flowing into it, but it shouldn't be. Probably got zapped with a big dose of static electricity at the ring.
Jimmy
toggle quoted message
Show quoted text
--- In TekScopes@..., "lldskjfoi" <k.templin@...> wrote:
I have 2.5 at both ends of R106.
1.9 and 1 at R105.
--- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
What is the voltages on either side of R105 and R106? It seems like C111 and U506 are the only possibilities if the voltage leaving R105 more or less agrees with what you see arriving at U506 pin 15. tom jobe...
----- Original Message ----- From: lldskjfoi To: TekScopes@... Sent: Thursday, September 20, 2012 6:04 PM Subject: [TekScopes] Re: 2247A Channel 1 issue
10k and 10k. What do you thin? the only thing left are the caps or the U506 itself.
KT --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > What are the resistance readings across R105 and R106? > That might help narrow the search down. > tom jobe... > > > > ----- Original Message ----- > From: lldskjfoi > To: TekScopes@... > Sent: Thursday, September 20, 2012 5:19 PM > Subject: [TekScopes] Re: 2247A Channel 1 issue > > > > > I get 8.2 and 8.2 as well. > > KT > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > > > I just checked R101 and R102 on a bare 2246 mainboard and they both measured 8.2k ohms in circuit. > > If the resistance is correct on your scope, then something is drawing enough current to drop 0.7 volts across R101. > > Nothing in that simple circuit looks like a suspect to me, no nasty 10uF to 47uF tantalums, etc. > > tom jobe... > > > > > > > > ----- Original Message ----- > > From: lldskjfoi > > To: TekScopes@... > > Sent: Thursday, September 20, 2012 1:20 PM > > Subject: [TekScopes] Re: 2247A Channel 1 issue > > > > > > > > > > Thanks again for the responce. I think we are getting there. > > > > Using cahannel 2 as the "working reference" > > > > pin 12 (channel 2) on u506 goes from the +2.5 to +1.4 when a 10x probe is connected. > > > > pin (15 channel 1) on U506 goes from +.9 to +.6 with the probe connected. > > > > Checking R102 for channel 2 it has +2.5 > > > > R101, (channel 1), shows +2.5 in but only +1.8 out. > > > > What are your thoughts? Is it simply a bad resistor or would one of the cap, C111 cause the voltage drop? > > > > Thanks again, KT > > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > > > > > For the record, I slid the cover back on a very dead 2246 and checked pin 12 and pin 15 of U506, they both had +2.49 volts with no probes attached. > > > This scope does not have any lights on the front, and none of the buttons work, so I could not see the reading I would get at ground, or with a 1X or a 10X probe. > > > tom jobe... > > > > > > > > > > > > > > > > > > ----- Original Message ----- > > > From: lldskjfoi > > > To: TekScopes@... > > > Sent: Thursday, September 20, 2012 8:28 AM > > > Subject: [TekScopes] Re: 2247A Channel 1 issue > > > > > > > > > > > > Thanks, thanks the info I was looking for. I started getting into the schematic last night I saw the path to U506. Will start checking it out. > > > > > > Also notice just letting the scope run with the probe attached the attenuation factor jumps back and forth- correct / incorrect. Does not change with no probe attached so it is getting an intermitant signal. > > > > > > KT > > > > > > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > > > > > > > Maybe start by looking at the #1 "Vertical Inputs" schematic to see the simple circuit which starts at the junction of R101, C101 and R105 for channel 1. > > > > That junction point is connected to the readout ring behind the BNC connector, so you could test for continuity and shorting to ground, etc. > > > > If you compared both ends of R101 and R105 (Chan 1) with those of R102 and R106 (Chan 2) you might see if the readout voltage is getting that far. > > > > If that looked okay, you could go to U506 and see if it arrives at pins 12 and 15 as you would expect. > > > > With the scope on its back, you have excellent access to all of these resistors and U506 (16 pin DIP), with the blue outer cover slid back the width of your hand. > > > > From reading the service manual, it appears that the readout signal gets changed to a digital signal at U506, but is a simple voltage up to U506. > > > > tom jobe... > > > > > > > > > > > > > > > > > > > > ----- Original Message ----- > > > > From: lldskjfoi > > > > To: TekScopes@... > > > > Sent: Wednesday, September 19, 2012 9:08 PM > > > > Subject: [TekScopes] Re: 2247A Channel 1 issue > > > > > > > > > > > > > > > > Some more checking shows channel 1 is not getting the code signal from the probe. Any idea on this circuit? > > > > > > > > KT > > > > > > > > --- In TekScopes@..., "lldskjfoi" <k.templin@> wrote: > > > > > > > > > > Hello everyone. > > > > > > > > > > Initial test show an issue with channel 1. > > > > > > > > > > Any direction on where to start? > > > > > > > > > > Channel 2 displays inputs correctly from the FG. > > > > > > > > > > Channel 1 displays 10 times the amplitude for the same signal. > > > > > > > > > > Example a 1 volt sine wave on channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions at .5v / div. > > > > > > > > > > The same signal on channel one reads as ~10v. When set to 5v / div it fills two div. > > > > > > > > > > Thanks, KT > > > > > > > > > > > > > > >
|
I think so too.
I have a "parts" 2246 that has the same chip.
I will swap out. It will be next week before I can get into the A10's.
I will let you know.
Thanks for help, KT
toggle quoted message
Show quoted text
--- In TekScopes@..., "jtjewell83" <jtjewell83@...> wrote: KT,
I money is on U506 has a bad input at pin 15. Lots of current's flowing into it, but it shouldn't be. Probably got zapped with a big dose of static electricity at the ring.
Jimmy
--- In TekScopes@..., "lldskjfoi" <k.templin@> wrote:
I have 2.5 at both ends of R106.
1.9 and 1 at R105.
--- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
What is the voltages on either side of R105 and R106? It seems like C111 and U506 are the only possibilities if the voltage leaving R105 more or less agrees with what you see arriving at U506 pin 15. tom jobe...
----- Original Message ----- From: lldskjfoi To: TekScopes@... Sent: Thursday, September 20, 2012 6:04 PM Subject: [TekScopes] Re: 2247A Channel 1 issue
10k and 10k. What do you thin? the only thing left are the caps or the U506 itself.
KT --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > What are the resistance readings across R105 and R106? > That might help narrow the search down. > tom jobe... > > > > ----- Original Message ----- > From: lldskjfoi > To: TekScopes@... > Sent: Thursday, September 20, 2012 5:19 PM > Subject: [TekScopes] Re: 2247A Channel 1 issue > > > > > I get 8.2 and 8.2 as well. > > KT > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > > > I just checked R101 and R102 on a bare 2246 mainboard and they both measured 8.2k ohms in circuit. > > If the resistance is correct on your scope, then something is drawing enough current to drop 0.7 volts across R101. > > Nothing in that simple circuit looks like a suspect to me, no nasty 10uF to 47uF tantalums, etc. > > tom jobe... > > > > > > > > ----- Original Message ----- > > From: lldskjfoi > > To: TekScopes@... > > Sent: Thursday, September 20, 2012 1:20 PM > > Subject: [TekScopes] Re: 2247A Channel 1 issue > > > > > > > > > > Thanks again for the responce. I think we are getting there. > > > > Using cahannel 2 as the "working reference" > > > > pin 12 (channel 2) on u506 goes from the +2.5 to +1.4 when a 10x probe is connected. > > > > pin (15 channel 1) on U506 goes from +.9 to +.6 with the probe connected. > > > > Checking R102 for channel 2 it has +2.5 > > > > R101, (channel 1), shows +2.5 in but only +1.8 out. > > > > What are your thoughts? Is it simply a bad resistor or would one of the cap, C111 cause the voltage drop? > > > > Thanks again, KT > > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > > > > > For the record, I slid the cover back on a very dead 2246 and checked pin 12 and pin 15 of U506, they both had +2.49 volts with no probes attached. > > > This scope does not have any lights on the front, and none of the buttons work, so I could not see the reading I would get at ground, or with a 1X or a 10X probe. > > > tom jobe... > > > > > > > > > > > > > > > > > > ----- Original Message ----- > > > From: lldskjfoi > > > To: TekScopes@... > > > Sent: Thursday, September 20, 2012 8:28 AM > > > Subject: [TekScopes] Re: 2247A Channel 1 issue > > > > > > > > > > > > Thanks, thanks the info I was looking for. I started getting into the schematic last night I saw the path to U506. Will start checking it out. > > > > > > Also notice just letting the scope run with the probe attached the attenuation factor jumps back and forth- correct / incorrect. Does not change with no probe attached so it is getting an intermitant signal. > > > > > > KT > > > > > > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > > > > > > > Maybe start by looking at the #1 "Vertical Inputs" schematic to see the simple circuit which starts at the junction of R101, C101 and R105 for channel 1. > > > > That junction point is connected to the readout ring behind the BNC connector, so you could test for continuity and shorting to ground, etc. > > > > If you compared both ends of R101 and R105 (Chan 1) with those of R102 and R106 (Chan 2) you might see if the readout voltage is getting that far. > > > > If that looked okay, you could go to U506 and see if it arrives at pins 12 and 15 as you would expect. > > > > With the scope on its back, you have excellent access to all of these resistors and U506 (16 pin DIP), with the blue outer cover slid back the width of your hand. > > > > From reading the service manual, it appears that the readout signal gets changed to a digital signal at U506, but is a simple voltage up to U506. > > > > tom jobe... > > > > > > > > > > > > > > > > > > > > ----- Original Message ----- > > > > From: lldskjfoi > > > > To: TekScopes@... > > > > Sent: Wednesday, September 19, 2012 9:08 PM > > > > Subject: [TekScopes] Re: 2247A Channel 1 issue > > > > > > > > > > > > > > > > Some more checking shows channel 1 is not getting the code signal from the probe. Any idea on this circuit? > > > > > > > > KT > > > > > > > > --- In TekScopes@..., "lldskjfoi" <k.templin@> wrote: > > > > > > > > > > Hello everyone. > > > > > > > > > > Initial test show an issue with channel 1. > > > > > > > > > > Any direction on where to start? > > > > > > > > > > Channel 2 displays inputs correctly from the FG. > > > > > > > > > > Channel 1 displays 10 times the amplitude for the same signal. > > > > > > > > > > Example a 1 volt sine wave on channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions at .5v / div. > > > > > > > > > > The same signal on channel one reads as ~10v. When set to 5v / div it fills two div. > > > > > > > > > > Thanks, KT > > > > > > > > > > > > > > >
|
The 4051 CMOS multiplexor is one of those very common "jelly bean" ICs with many alternate sources and even improved versions. On Fri, 21 Sep 2012 01:51:39 -0000, "lldskjfoi" <k.templin@...> wrote: I think so too.
I have a "parts" 2246 that has the same chip.
I will swap out. It will be next week before I can get into the A10's.
I will let you know.
Thanks for help, KT
--- In TekScopes@..., "jtjewell83" <jtjewell83@...> wrote:
KT,
I money is on U506 has a bad input at pin 15. Lots of current's flowing into it, but it shouldn't be. Probably got zapped with a big dose of static electricity at the ring.
Jimmy
--- In TekScopes@..., "lldskjfoi" <k.templin@> wrote:
I have 2.5 at both ends of R106.
1.9 and 1 at R105.
--- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
What is the voltages on either side of R105 and R106? It seems like C111 and U506 are the only possibilities if the voltage leaving R105 more or less agrees with what you see arriving at U506 pin 15. tom jobe...
----- Original Message ----- From: lldskjfoi To: TekScopes@... Sent: Thursday, September 20, 2012 6:04 PM Subject: [TekScopes] Re: 2247A Channel 1 issue
10k and 10k. What do you thin? the only thing left are the caps or the U506 itself.
KT --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > What are the resistance readings across R105 and R106? > That might help narrow the search down. > tom jobe... > > > > ----- Original Message ----- > From: lldskjfoi > To: TekScopes@... > Sent: Thursday, September 20, 2012 5:19 PM > Subject: [TekScopes] Re: 2247A Channel 1 issue > > > > > I get 8.2 and 8.2 as well. > > KT > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > > > I just checked R101 and R102 on a bare 2246 mainboard and they both measured 8.2k ohms in circuit. > > If the resistance is correct on your scope, then something is drawing enough current to drop 0.7 volts across R101. > > Nothing in that simple circuit looks like a suspect to me, no nasty 10uF to 47uF tantalums, etc. > > tom jobe... > > > > > > > > ----- Original Message ----- > > From: lldskjfoi > > To: TekScopes@... > > Sent: Thursday, September 20, 2012 1:20 PM > > Subject: [TekScopes] Re: 2247A Channel 1 issue > > > > > > > > > > Thanks again for the responce. I think we are getting there. > > > > Using cahannel 2 as the "working reference" > > > > pin 12 (channel 2) on u506 goes from the +2.5 to +1.4 when a 10x probe is connected. > > > > pin (15 channel 1) on U506 goes from +.9 to +.6 with the probe connected. > > > > Checking R102 for channel 2 it has +2.5 > > > > R101, (channel 1), shows +2.5 in but only +1.8 out. > > > > What are your thoughts? Is it simply a bad resistor or would one of the cap, C111 cause the voltage drop? > > > > Thanks again, KT > > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > > > > > For the record, I slid the cover back on a very dead 2246 and checked pin 12 and pin 15 of U506, they both had +2.49 volts with no probes attached. > > > This scope does not have any lights on the front, and none of the buttons work, so I could not see the reading I would get at ground, or with a 1X or a 10X probe. > > > tom jobe... > > > > > > > > > > > > > > > > > > ----- Original Message ----- > > > From: lldskjfoi > > > To: TekScopes@... > > > Sent: Thursday, September 20, 2012 8:28 AM > > > Subject: [TekScopes] Re: 2247A Channel 1 issue > > > > > > > > > > > > Thanks, thanks the info I was looking for. I started getting into the schematic last night I saw the path to U506. Will start checking it out. > > > > > > Also notice just letting the scope run with the probe attached the attenuation factor jumps back and forth- correct / incorrect. Does not change with no probe attached so it is getting an intermitant signal. > > > > > > KT > > > > > > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > > > > > > > Maybe start by looking at the #1 "Vertical Inputs" schematic to see the simple circuit which starts at the junction of R101, C101 and R105 for channel 1. > > > > That junction point is connected to the readout ring behind the BNC connector, so you could test for continuity and shorting to ground, etc. > > > > If you compared both ends of R101 and R105 (Chan 1) with those of R102 and R106 (Chan 2) you might see if the readout voltage is getting that far. > > > > If that looked okay, you could go to U506 and see if it arrives at pins 12 and 15 as you would expect. > > > > With the scope on its back, you have excellent access to all of these resistors and U506 (16 pin DIP), with the blue outer cover slid back the width of your hand. > > > > From reading the service manual, it appears that the readout signal gets changed to a digital signal at U506, but is a simple voltage up to U506. > > > > tom jobe... > > > > > > > > > > > > > > > > > > > > ----- Original Message ----- > > > > From: lldskjfoi > > > > To: TekScopes@... > > > > Sent: Wednesday, September 19, 2012 9:08 PM > > > > Subject: [TekScopes] Re: 2247A Channel 1 issue > > > > > > > > > > > > > > > > Some more checking shows channel 1 is not getting the code signal from the probe. Any idea on this circuit? > > > > > > > > KT > > > > > > > > --- In TekScopes@..., "lldskjfoi" <k.templin@> wrote: > > > > > > > > > > Hello everyone. > > > > > > > > > > Initial test show an issue with channel 1. > > > > > > > > > > Any direction on where to start? > > > > > > > > > > Channel 2 displays inputs correctly from the FG. > > > > > > > > > > Channel 1 displays 10 times the amplitude for the same signal. > > > > > > > > > > Example a 1 volt sine wave on channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions at .5v / div. > > > > > > > > > > The same signal on channel one reads as ~10v. When set to 5v / div it fills two div. > > > > > > > > > > Thanks, KT > > > > > > > > > > > > > > >
|
The culprit was C111. Sniped one end and bingo.
Amazing what a 10 cent part can do.
Thanks Tom I really appreciate your help. You start overthinking the problem and going off in a wrong direction.
It's nice to have those "outside eyes" helping.
Snipping the cap was a great idea.
KT
toggle quoted message
Show quoted text
--- In TekScopes@..., "Tom Jobe" <tomjobe@...> wrote: Well, unless I missed something you are down to C111 or U506 as being the culprit. I would carefully lift one leg of C111. It is near U506 pin 14, slightly further from the front of the scope. It might also be orange in color. tom jobe...
----- Original Message ----- From: lldskjfoi To: TekScopes@... Sent: Thursday, September 20, 2012 6:27 PM Subject: [TekScopes] Re: 2247A Channel 1 issue
I have 2.5 at both ends of R106.
1.9 and 1 at R105.
--- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > What is the voltages on either side of R105 and R106? > It seems like C111 and U506 are the only possibilities if the voltage leaving R105 more or less agrees with what you see arriving at U506 pin 15. > tom jobe... > > > > ----- Original Message ----- > From: lldskjfoi > To: TekScopes@... > Sent: Thursday, September 20, 2012 6:04 PM > Subject: [TekScopes] Re: 2247A Channel 1 issue > > > > > 10k and 10k. What do you thin? the only thing left are the caps or the U506 itself. > > KT > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > > > What are the resistance readings across R105 and R106? > > That might help narrow the search down. > > tom jobe... > > > > > > > > ----- Original Message ----- > > From: lldskjfoi > > To: TekScopes@... > > Sent: Thursday, September 20, 2012 5:19 PM > > Subject: [TekScopes] Re: 2247A Channel 1 issue > > > > > > > > > > I get 8.2 and 8.2 as well. > > > > KT > > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > > > > > I just checked R101 and R102 on a bare 2246 mainboard and they both measured 8.2k ohms in circuit. > > > If the resistance is correct on your scope, then something is drawing enough current to drop 0.7 volts across R101. > > > Nothing in that simple circuit looks like a suspect to me, no nasty 10uF to 47uF tantalums, etc. > > > tom jobe... > > > > > > > > > > > > ----- Original Message ----- > > > From: lldskjfoi > > > To: TekScopes@... > > > Sent: Thursday, September 20, 2012 1:20 PM > > > Subject: [TekScopes] Re: 2247A Channel 1 issue > > > > > > > > > > > > > > > Thanks again for the responce. I think we are getting there. > > > > > > Using cahannel 2 as the "working reference" > > > > > > pin 12 (channel 2) on u506 goes from the +2.5 to +1.4 when a 10x probe is connected. > > > > > > pin (15 channel 1) on U506 goes from +.9 to +.6 with the probe connected. > > > > > > Checking R102 for channel 2 it has +2.5 > > > > > > R101, (channel 1), shows +2.5 in but only +1.8 out. > > > > > > What are your thoughts? Is it simply a bad resistor or would one of the cap, C111 cause the voltage drop? > > > > > > Thanks again, KT > > > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > > > > > > > For the record, I slid the cover back on a very dead 2246 and checked pin 12 and pin 15 of U506, they both had +2.49 volts with no probes attached. > > > > This scope does not have any lights on the front, and none of the buttons work, so I could not see the reading I would get at ground, or with a 1X or a 10X probe. > > > > tom jobe... > > > > > > > > > > > > > > > > > > > > > > > > ----- Original Message ----- > > > > From: lldskjfoi > > > > To: TekScopes@... > > > > Sent: Thursday, September 20, 2012 8:28 AM > > > > Subject: [TekScopes] Re: 2247A Channel 1 issue > > > > > > > > > > > > > > > > Thanks, thanks the info I was looking for. I started getting into the schematic last night I saw the path to U506. Will start checking it out. > > > > > > > > Also notice just letting the scope run with the probe attached the attenuation factor jumps back and forth- correct / incorrect. Does not change with no probe attached so it is getting an intermitant signal. > > > > > > > > KT > > > > > > > > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > > > > > > > > > Maybe start by looking at the #1 "Vertical Inputs" schematic to see the simple circuit which starts at the junction of R101, C101 and R105 for channel 1. > > > > > That junction point is connected to the readout ring behind the BNC connector, so you could test for continuity and shorting to ground, etc. > > > > > If you compared both ends of R101 and R105 (Chan 1) with those of R102 and R106 (Chan 2) you might see if the readout voltage is getting that far. > > > > > If that looked okay, you could go to U506 and see if it arrives at pins 12 and 15 as you would expect. > > > > > With the scope on its back, you have excellent access to all of these resistors and U506 (16 pin DIP), with the blue outer cover slid back the width of your hand. > > > > > From reading the service manual, it appears that the readout signal gets changed to a digital signal at U506, but is a simple voltage up to U506. > > > > > tom jobe... > > > > > > > > > > > > > > > > > > > > > > > > > ----- Original Message ----- > > > > > From: lldskjfoi > > > > > To: TekScopes@... > > > > > Sent: Wednesday, September 19, 2012 9:08 PM > > > > > Subject: [TekScopes] Re: 2247A Channel 1 issue > > > > > > > > > > > > > > > > > > > > Some more checking shows channel 1 is not getting the code signal from the probe. Any idea on this circuit? > > > > > > > > > > KT > > > > > > > > > > --- In TekScopes@..., "lldskjfoi" <k.templin@> wrote: > > > > > > > > > > > > Hello everyone. > > > > > > > > > > > > Initial test show an issue with channel 1. > > > > > > > > > > > > Any direction on where to start? > > > > > > > > > > > > Channel 2 displays inputs correctly from the FG. > > > > > > > > > > > > Channel 1 displays 10 times the amplitude for the same signal. > > > > > > > > > > > > Example a 1 volt sine wave on channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions at .5v / div. > > > > > > > > > > > > The same signal on channel one reads as ~10v. When set to 5v / div it fills two div. > > > > > > > > > > > > Thanks, KT > > > > > > > > > > > > > > > > > > > > >
|