2247A Channel 1 issue


 

Hello everyone.

Initial test show an issue with channel 1.

Any direction on where to start?

Channel 2 displays inputs correctly from the FG.

Channel 1 displays 10 times the amplitude for the same signal.

Example a 1 volt sine wave on channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions at .5v / div.

The same signal on channel one reads as ~10v. When set to 5v / div it fills two div.

Thanks, KT


 

Some more checking shows channel 1 is not getting the code signal from the probe. Any idea on this circuit?

KT

--- In TekScopes@..., "lldskjfoi" <k.templin@...> wrote:

Hello everyone.

Initial test show an issue with channel 1.

Any direction on where to start?

Channel 2 displays inputs correctly from the FG.

Channel 1 displays 10 times the amplitude for the same signal.

Example a 1 volt sine wave on channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions at .5v / div.

The same signal on channel one reads as ~10v. When set to 5v / div it fills two div.

Thanks, KT


 

开云体育


I know it is probably silly to ask, but are you sure that the switch on one probe isn't set for the? x10 setting ?
PatM
============================================================
lldskjfoi wrote the following on 9/19/2012 9:59 PM:
?

Hello everyone.

Initial test show an issue with channel 1.

Any direction on where to start?

Channel 2 displays inputs correctly from the FG.

Channel 1 displays 10 times the amplitude for the same signal.

Example a 1 volt sine wave on channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions at .5v / div.

The same signal on channel one reads as ~10v. When set to 5v / div it fills two div.

Thanks, KT


.


 

They on 10x non swtichable.

KT

--- In TekScopes@..., Pat-W9ZO <malfam2@...> wrote:


I know it is probably silly to ask, but are you sure that the switch on
one probe isn't set for the x10 setting ?
PatM
============================================================
lldskjfoi wrote the following on 9/19/2012 9:59 PM:

Hello everyone.

Initial test show an issue with channel 1.

Any direction on where to start?

Channel 2 displays inputs correctly from the FG.

Channel 1 displays 10 times the amplitude for the same signal.

Example a 1 volt sine wave on channel 2 displays on the voltmeter
function correctly at ~1v and 2 divisions at .5v / div.

The same signal on channel one reads as ~10v. When set to 5v / div it
fills two div.

Thanks, KT

.


 

开云体育

Hi,

you should check, that there is no short on the "sensor" ring of the input which shows
the 10X signal. Normally the sensor, when open tells 1:1 probe and for probes
with 10:1 a.s.o. it is conncted via a resistor to ground !!!

Herbert


Am 20-09-2012 14:54, schrieb lldskjfoi:

?


They on 10x non swtichable.

KT
--- In TekScopes@..., Pat-W9ZO wrote:
>
>
> I know it is probably silly to ask, but are you sure that the switch on
> one probe isn't set for the x10 setting ?
> PatM
> ============================================================
> lldskjfoi wrote the following on 9/19/2012 9:59 PM:
> >
> > Hello everyone.
> >
> > Initial test show an issue with channel 1.
> >
> > Any direction on where to start?
> >
> > Channel 2 displays inputs correctly from the FG.
> >
> > Channel 1 displays 10 times the amplitude for the same signal.
> >
> > Example a 1 volt sine wave on channel 2 displays on the voltmeter
> > function correctly at ~1v and 2 divisions at .5v / div.
> >
> > The same signal on channel one reads as ~10v. When set to 5v / div it
> > fills two div.
> >
> > Thanks, KT
> >
> >
> .
>



Tom Jobe
 

开云体育

Maybe start by looking at the #1 "Vertical Inputs" schematic to see the simple circuit which starts at the junction of R101, C101 and R105?for channel 1.
That junction point is connected to the readout ring behind the BNC connector, so you could test for continuity and shorting to ground, etc.
If you compared both ends of? R101 and R105 (Chan 1)?with those of R102 and R106 (Chan 2) you might see if the readout voltage is getting that far.
If that looked okay, you could go to U506 and see if it arrives at pins 12 and 15 as you would expect.
With the scope on its back, you have excellent access to all of these resistors and U506 (16 pin DIP), with the blue outer cover?slid back the width of your hand.
From reading the service manual, it appears that the readout signal gets changed to a digital signal at U506, but is a simple voltage up to U506.
tom jobe...
?
?
?
?

----- Original Message -----
From: lldskjfoi
Sent: Wednesday, September 19, 2012 9:08 PM
Subject: [TekScopes] Re: 2247A Channel 1 issue

?

Some more checking shows channel 1 is not getting the code signal from the probe. Any idea on this circuit?

KT

--- In TekScopes@..., "lldskjfoi" wrote:
>
> Hello everyone.
>
> Initial test show an issue with channel 1.
>
> Any direction on where to start?
>
> Channel 2 displays inputs correctly from the FG.
>
> Channel 1 displays 10 times the amplitude for the same signal.
>
> Example a 1 volt sine wave on channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions at .5v / div.
>
> The same signal on channel one reads as ~10v. When set to 5v / div it fills two div.
>
> Thanks, KT
>


 

Thanks, thanks the info I was looking for. I started getting into the schematic last night I saw the path to U506. Will start checking it out.

Also notice just letting the scope run with the probe attached the attenuation factor jumps back and forth- correct / incorrect. Does not change with no probe attached so it is getting an intermitant signal.

KT

--- In TekScopes@..., "Tom Jobe" <tomjobe@...> wrote:

Maybe start by looking at the #1 "Vertical Inputs" schematic to see the simple circuit which starts at the junction of R101, C101 and R105 for channel 1.
That junction point is connected to the readout ring behind the BNC connector, so you could test for continuity and shorting to ground, etc.
If you compared both ends of R101 and R105 (Chan 1) with those of R102 and R106 (Chan 2) you might see if the readout voltage is getting that far.
If that looked okay, you could go to U506 and see if it arrives at pins 12 and 15 as you would expect.
With the scope on its back, you have excellent access to all of these resistors and U506 (16 pin DIP), with the blue outer cover slid back the width of your hand.
From reading the service manual, it appears that the readout signal gets changed to a digital signal at U506, but is a simple voltage up to U506.
tom jobe...




----- Original Message -----
From: lldskjfoi
To: TekScopes@...
Sent: Wednesday, September 19, 2012 9:08 PM
Subject: [TekScopes] Re: 2247A Channel 1 issue



Some more checking shows channel 1 is not getting the code signal from the probe. Any idea on this circuit?

KT

--- In TekScopes@..., "lldskjfoi" <k.templin@> wrote:
>
> Hello everyone.
>
> Initial test show an issue with channel 1.
>
> Any direction on where to start?
>
> Channel 2 displays inputs correctly from the FG.
>
> Channel 1 displays 10 times the amplitude for the same signal.
>
> Example a 1 volt sine wave on channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions at .5v / div.
>
> The same signal on channel one reads as ~10v. When set to 5v / div it fills two div.
>
> Thanks, KT
>


Tom Jobe
 

开云体育

For the record, I slid the cover back on a very dead 2246 and checked pin 12 and pin 15 of U506, they both had +2.49 volts with no probes attached.
This scope does not have any lights on the front, and none of the buttons work, so I could not see?the reading I would get at ground, or with a 1X or?a 10X probe.
tom jobe...
?
?
?
?
?

----- Original Message -----
From: lldskjfoi
Sent: Thursday, September 20, 2012 8:28 AM
Subject: [TekScopes] Re: 2247A Channel 1 issue

?

Thanks, thanks the info I was looking for. I started getting into the schematic last night I saw the path to U506. Will start checking it out.

Also notice just letting the scope run with the probe attached the attenuation factor jumps back and forth- correct / incorrect. Does not change with no probe attached so it is getting an intermitant signal.

KT

--- In TekScopes@..., "Tom Jobe" wrote:
>
> Maybe start by looking at the #1 "Vertical Inputs" schematic to see the simple circuit which starts at the junction of R101, C101 and R105 for channel 1.
> That junction point is connected to the readout ring behind the BNC connector, so you could test for continuity and shorting to ground, etc.
> If you compared both ends of R101 and R105 (Chan 1) with those of R102 and R106 (Chan 2) you might see if the readout voltage is getting that far.
> If that looked okay, you could go to U506 and see if it arrives at pins 12 and 15 as you would expect.
> With the scope on its back, you have excellent access to all of these resistors and U506 (16 pin DIP), with the blue outer cover slid back the width of your hand.
> From reading the service manual, it appears that the readout signal gets changed to a digital signal at U506, but is a simple voltage up to U506.
> tom jobe...
>
>
>
>
> ----- Original Message -----
> From: lldskjfoi
> To: TekScopes@...
> Sent: Wednesday, September 19, 2012 9:08 PM
> Subject: [TekScopes] Re: 2247A Channel 1 issue
>
>
>
> Some more checking shows channel 1 is not getting the code signal from the probe. Any idea on this circuit?
>
> KT
>
> --- In TekScopes@..., "lldskjfoi" wrote:
> >
> > Hello everyone.
> >
> > Initial test show an issue with channel 1.
> >
> > Any direction on where to start?
> >
> > Channel 2 displays inputs correctly from the FG.
> >
> > Channel 1 displays 10 times the amplitude for the same signal.
> >
> > Example a 1 volt sine wave on channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions at .5v / div.
> >
> > The same signal on channel one reads as ~10v. When set to 5v / div it fills two div.
> >
> > Thanks, KT
> >
>


 

Thanks again for the responce. I think we are getting there.

Using cahannel 2 as the "working reference"

pin 12 (channel 2) on u506 goes from the +2.5 to +1.4 when a 10x probe is connected.

pin (15 channel 1) on U506 goes from +.9 to +.6 with the probe connected.

Checking R102 for channel 2 it has +2.5

R101, (channel 1), shows +2.5 in but only +1.8 out.

What are your thoughts? Is it simply a bad resistor or would one of the cap, C111 cause the voltage drop?

Thanks again, KT

--- In TekScopes@..., "Tom Jobe" <tomjobe@...> wrote:

For the record, I slid the cover back on a very dead 2246 and checked pin 12 and pin 15 of U506, they both had +2.49 volts with no probes attached.
This scope does not have any lights on the front, and none of the buttons work, so I could not see the reading I would get at ground, or with a 1X or a 10X probe.
tom jobe...





----- Original Message -----
From: lldskjfoi
To: TekScopes@...
Sent: Thursday, September 20, 2012 8:28 AM
Subject: [TekScopes] Re: 2247A Channel 1 issue



Thanks, thanks the info I was looking for. I started getting into the schematic last night I saw the path to U506. Will start checking it out.

Also notice just letting the scope run with the probe attached the attenuation factor jumps back and forth- correct / incorrect. Does not change with no probe attached so it is getting an intermitant signal.

KT

--- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
>
> Maybe start by looking at the #1 "Vertical Inputs" schematic to see the simple circuit which starts at the junction of R101, C101 and R105 for channel 1.
> That junction point is connected to the readout ring behind the BNC connector, so you could test for continuity and shorting to ground, etc.
> If you compared both ends of R101 and R105 (Chan 1) with those of R102 and R106 (Chan 2) you might see if the readout voltage is getting that far.
> If that looked okay, you could go to U506 and see if it arrives at pins 12 and 15 as you would expect.
> With the scope on its back, you have excellent access to all of these resistors and U506 (16 pin DIP), with the blue outer cover slid back the width of your hand.
> From reading the service manual, it appears that the readout signal gets changed to a digital signal at U506, but is a simple voltage up to U506.
> tom jobe...
>
>
>
>
> ----- Original Message -----
> From: lldskjfoi
> To: TekScopes@...
> Sent: Wednesday, September 19, 2012 9:08 PM
> Subject: [TekScopes] Re: 2247A Channel 1 issue
>
>
>
> Some more checking shows channel 1 is not getting the code signal from the probe. Any idea on this circuit?
>
> KT
>
> --- In TekScopes@..., "lldskjfoi" <k.templin@> wrote:
> >
> > Hello everyone.
> >
> > Initial test show an issue with channel 1.
> >
> > Any direction on where to start?
> >
> > Channel 2 displays inputs correctly from the FG.
> >
> > Channel 1 displays 10 times the amplitude for the same signal.
> >
> > Example a 1 volt sine wave on channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions at .5v / div.
> >
> > The same signal on channel one reads as ~10v. When set to 5v / div it fills two div.
> >
> > Thanks, KT
> >
>


Tom Jobe
 

开云体育

I just checked R101 and R102 on a bare 2246 mainboard and they both measured 8.2k ohms in circuit.
If the resistance is correct on your scope, then something is drawing enough current to drop 0.7 volts across R101.
Nothing in that simple circuit looks like a suspect to me, no nasty 10uF to 47uF tantalums, etc.
tom jobe...
?
?
?

----- Original Message -----
From: lldskjfoi
Sent: Thursday, September 20, 2012 1:20 PM
Subject: [TekScopes] Re: 2247A Channel 1 issue

?


Thanks again for the responce. I think we are getting there.

Using cahannel 2 as the "working reference"

pin 12 (channel 2) on u506 goes from the +2.5 to +1.4 when a 10x probe is connected.

pin (15 channel 1) on U506 goes from +.9 to +.6 with the probe connected.

Checking R102 for channel 2 it has +2.5

R101, (channel 1), shows +2.5 in but only +1.8 out.

What are your thoughts? Is it simply a bad resistor or would one of the cap, C111 cause the voltage drop?

Thanks again, KT
--- In TekScopes@..., "Tom Jobe" wrote:
>
> For the record, I slid the cover back on a very dead 2246 and checked pin 12 and pin 15 of U506, they both had +2.49 volts with no probes attached.
> This scope does not have any lights on the front, and none of the buttons work, so I could not see the reading I would get at ground, or with a 1X or a 10X probe.
> tom jobe...
>
>
>
>
>
> ----- Original Message -----
> From: lldskjfoi
> To: TekScopes@...
> Sent: Thursday, September 20, 2012 8:28 AM
> Subject: [TekScopes] Re: 2247A Channel 1 issue
>
>
>
> Thanks, thanks the info I was looking for. I started getting into the schematic last night I saw the path to U506. Will start checking it out.
>
> Also notice just letting the scope run with the probe attached the attenuation factor jumps back and forth- correct / incorrect. Does not change with no probe attached so it is getting an intermitant signal.
>
> KT
>
> --- In TekScopes@..., "Tom Jobe" wrote:
> >
> > Maybe start by looking at the #1 "Vertical Inputs" schematic to see the simple circuit which starts at the junction of R101, C101 and R105 for channel 1.
> > That junction point is connected to the readout ring behind the BNC connector, so you could test for continuity and shorting to ground, etc.
> > If you compared both ends of R101 and R105 (Chan 1) with those of R102 and R106 (Chan 2) you might see if the readout voltage is getting that far.
> > If that looked okay, you could go to U506 and see if it arrives at pins 12 and 15 as you would expect.
> > With the scope on its back, you have excellent access to all of these resistors and U506 (16 pin DIP), with the blue outer cover slid back the width of your hand.
> > From reading the service manual, it appears that the readout signal gets changed to a digital signal at U506, but is a simple voltage up to U506.
> > tom jobe...
> >
> >
> >
> >
> > ----- Original Message -----
> > From: lldskjfoi
> > To: TekScopes@...
> > Sent: Wednesday, September 19, 2012 9:08 PM
> > Subject: [TekScopes] Re: 2247A Channel 1 issue
> >
> >
> >
> > Some more checking shows channel 1 is not getting the code signal from the probe. Any idea on this circuit?
> >
> > KT
> >
> > --- In TekScopes@..., "lldskjfoi" wrote:
> > >
> > > Hello everyone.
> > >
> > > Initial test show an issue with channel 1.
> > >
> > > Any direction on where to start?
> > >
> > > Channel 2 displays inputs correctly from the FG.
> > >
> > > Channel 1 displays 10 times the amplitude for the same signal.
> > >
> > > Example a 1 volt sine wave on channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions at .5v / div.
> > >
> > > The same signal on channel one reads as ~10v. When set to 5v / div it fills two div.
> > >
> > > Thanks, KT
> > >
> >
>


 

I get 8.2 and 8.2 as well.

KT

--- In TekScopes@..., "Tom Jobe" <tomjobe@...> wrote:

I just checked R101 and R102 on a bare 2246 mainboard and they both measured 8.2k ohms in circuit.
If the resistance is correct on your scope, then something is drawing enough current to drop 0.7 volts across R101.
Nothing in that simple circuit looks like a suspect to me, no nasty 10uF to 47uF tantalums, etc.
tom jobe...



----- Original Message -----
From: lldskjfoi
To: TekScopes@...
Sent: Thursday, September 20, 2012 1:20 PM
Subject: [TekScopes] Re: 2247A Channel 1 issue




Thanks again for the responce. I think we are getting there.

Using cahannel 2 as the "working reference"

pin 12 (channel 2) on u506 goes from the +2.5 to +1.4 when a 10x probe is connected.

pin (15 channel 1) on U506 goes from +.9 to +.6 with the probe connected.

Checking R102 for channel 2 it has +2.5

R101, (channel 1), shows +2.5 in but only +1.8 out.

What are your thoughts? Is it simply a bad resistor or would one of the cap, C111 cause the voltage drop?

Thanks again, KT
--- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
>
> For the record, I slid the cover back on a very dead 2246 and checked pin 12 and pin 15 of U506, they both had +2.49 volts with no probes attached.
> This scope does not have any lights on the front, and none of the buttons work, so I could not see the reading I would get at ground, or with a 1X or a 10X probe.
> tom jobe...
>
>
>
>
>
> ----- Original Message -----
> From: lldskjfoi
> To: TekScopes@...
> Sent: Thursday, September 20, 2012 8:28 AM
> Subject: [TekScopes] Re: 2247A Channel 1 issue
>
>
>
> Thanks, thanks the info I was looking for. I started getting into the schematic last night I saw the path to U506. Will start checking it out.
>
> Also notice just letting the scope run with the probe attached the attenuation factor jumps back and forth- correct / incorrect. Does not change with no probe attached so it is getting an intermitant signal.
>
> KT
>
> --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
> >
> > Maybe start by looking at the #1 "Vertical Inputs" schematic to see the simple circuit which starts at the junction of R101, C101 and R105 for channel 1.
> > That junction point is connected to the readout ring behind the BNC connector, so you could test for continuity and shorting to ground, etc.
> > If you compared both ends of R101 and R105 (Chan 1) with those of R102 and R106 (Chan 2) you might see if the readout voltage is getting that far.
> > If that looked okay, you could go to U506 and see if it arrives at pins 12 and 15 as you would expect.
> > With the scope on its back, you have excellent access to all of these resistors and U506 (16 pin DIP), with the blue outer cover slid back the width of your hand.
> > From reading the service manual, it appears that the readout signal gets changed to a digital signal at U506, but is a simple voltage up to U506.
> > tom jobe...
> >
> >
> >
> >
> > ----- Original Message -----
> > From: lldskjfoi
> > To: TekScopes@...
> > Sent: Wednesday, September 19, 2012 9:08 PM
> > Subject: [TekScopes] Re: 2247A Channel 1 issue
> >
> >
> >
> > Some more checking shows channel 1 is not getting the code signal from the probe. Any idea on this circuit?
> >
> > KT
> >
> > --- In TekScopes@..., "lldskjfoi" <k.templin@> wrote:
> > >
> > > Hello everyone.
> > >
> > > Initial test show an issue with channel 1.
> > >
> > > Any direction on where to start?
> > >
> > > Channel 2 displays inputs correctly from the FG.
> > >
> > > Channel 1 displays 10 times the amplitude for the same signal.
> > >
> > > Example a 1 volt sine wave on channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions at .5v / div.
> > >
> > > The same signal on channel one reads as ~10v. When set to 5v / div it fills two div.
> > >
> > > Thanks, KT
> > >
> >
>


Tom Jobe
 

开云体育

What are the resistance readings across R105 and R106?
That might help narrow the search down.
tom jobe...
?
?
?

----- Original Message -----
From: lldskjfoi
Sent: Thursday, September 20, 2012 5:19 PM
Subject: [TekScopes] Re: 2247A Channel 1 issue

?


I get 8.2 and 8.2 as well.

KT
--- In TekScopes@..., "Tom Jobe" wrote:
>
> I just checked R101 and R102 on a bare 2246 mainboard and they both measured 8.2k ohms in circuit.
> If the resistance is correct on your scope, then something is drawing enough current to drop 0.7 volts across R101.
> Nothing in that simple circuit looks like a suspect to me, no nasty 10uF to 47uF tantalums, etc.
> tom jobe...
>
>
>
> ----- Original Message -----
> From: lldskjfoi
> To: TekScopes@...
> Sent: Thursday, September 20, 2012 1:20 PM
> Subject: [TekScopes] Re: 2247A Channel 1 issue
>
>
>
>
> Thanks again for the responce. I think we are getting there.
>
> Using cahannel 2 as the "working reference"
>
> pin 12 (channel 2) on u506 goes from the +2.5 to +1.4 when a 10x probe is connected.
>
> pin (15 channel 1) on U506 goes from +.9 to +.6 with the probe connected.
>
> Checking R102 for channel 2 it has +2.5
>
> R101, (channel 1), shows +2.5 in but only +1.8 out.
>
> What are your thoughts? Is it simply a bad resistor or would one of the cap, C111 cause the voltage drop?
>
> Thanks again, KT
> --- In TekScopes@..., "Tom Jobe" wrote:
> >
> > For the record, I slid the cover back on a very dead 2246 and checked pin 12 and pin 15 of U506, they both had +2.49 volts with no probes attached.
> > This scope does not have any lights on the front, and none of the buttons work, so I could not see the reading I would get at ground, or with a 1X or a 10X probe.
> > tom jobe...
> >
> >
> >
> >
> >
> > ----- Original Message -----
> > From: lldskjfoi
> > To: TekScopes@...
> > Sent: Thursday, September 20, 2012 8:28 AM
> > Subject: [TekScopes] Re: 2247A Channel 1 issue
> >
> >
> >
> > Thanks, thanks the info I was looking for. I started getting into the schematic last night I saw the path to U506. Will start checking it out.
> >
> > Also notice just letting the scope run with the probe attached the attenuation factor jumps back and forth- correct / incorrect. Does not change with no probe attached so it is getting an intermitant signal.
> >
> > KT
> >
> > --- In TekScopes@..., "Tom Jobe" wrote:
> > >
> > > Maybe start by looking at the #1 "Vertical Inputs" schematic to see the simple circuit which starts at the junction of R101, C101 and R105 for channel 1.
> > > That junction point is connected to the readout ring behind the BNC connector, so you could test for continuity and shorting to ground, etc.
> > > If you compared both ends of R101 and R105 (Chan 1) with those of R102 and R106 (Chan 2) you might see if the readout voltage is getting that far.
> > > If that looked okay, you could go to U506 and see if it arrives at pins 12 and 15 as you would expect.
> > > With the scope on its back, you have excellent access to all of these resistors and U506 (16 pin DIP), with the blue outer cover slid back the width of your hand.
> > > From reading the service manual, it appears that the readout signal gets changed to a digital signal at U506, but is a simple voltage up to U506.
> > > tom jobe...
> > >
> > >
> > >
> > >
> > > ----- Original Message -----
> > > From: lldskjfoi
> > > To: TekScopes@...
> > > Sent: Wednesday, September 19, 2012 9:08 PM
> > > Subject: [TekScopes] Re: 2247A Channel 1 issue
> > >
> > >
> > >
> > > Some more checking shows channel 1 is not getting the code signal from the probe. Any idea on this circuit?
> > >
> > > KT
> > >
> > > --- In TekScopes@..., "lldskjfoi" wrote:
> > > >
> > > > Hello everyone.
> > > >
> > > > Initial test show an issue with channel 1.
> > > >
> > > > Any direction on where to start?
> > > >
> > > > Channel 2 displays inputs correctly from the FG.
> > > >
> > > > Channel 1 displays 10 times the amplitude for the same signal.
> > > >
> > > > Example a 1 volt sine wave on channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions at .5v / div.
> > > >
> > > > The same signal on channel one reads as ~10v. When set to 5v / div it fills two div.
> > > >
> > > > Thanks, KT
> > > >
> > >
> >
>


 

10k and 10k. What do you thin? the only thing left are the caps or the U506 itself.

KT

--- In TekScopes@..., "Tom Jobe" <tomjobe@...> wrote:

What are the resistance readings across R105 and R106?
That might help narrow the search down.
tom jobe...



----- Original Message -----
From: lldskjfoi
To: TekScopes@...
Sent: Thursday, September 20, 2012 5:19 PM
Subject: [TekScopes] Re: 2247A Channel 1 issue




I get 8.2 and 8.2 as well.

KT
--- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
>
> I just checked R101 and R102 on a bare 2246 mainboard and they both measured 8.2k ohms in circuit.
> If the resistance is correct on your scope, then something is drawing enough current to drop 0.7 volts across R101.
> Nothing in that simple circuit looks like a suspect to me, no nasty 10uF to 47uF tantalums, etc.
> tom jobe...
>
>
>
> ----- Original Message -----
> From: lldskjfoi
> To: TekScopes@...
> Sent: Thursday, September 20, 2012 1:20 PM
> Subject: [TekScopes] Re: 2247A Channel 1 issue
>
>
>
>
> Thanks again for the responce. I think we are getting there.
>
> Using cahannel 2 as the "working reference"
>
> pin 12 (channel 2) on u506 goes from the +2.5 to +1.4 when a 10x probe is connected.
>
> pin (15 channel 1) on U506 goes from +.9 to +.6 with the probe connected.
>
> Checking R102 for channel 2 it has +2.5
>
> R101, (channel 1), shows +2.5 in but only +1.8 out.
>
> What are your thoughts? Is it simply a bad resistor or would one of the cap, C111 cause the voltage drop?
>
> Thanks again, KT
> --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
> >
> > For the record, I slid the cover back on a very dead 2246 and checked pin 12 and pin 15 of U506, they both had +2.49 volts with no probes attached.
> > This scope does not have any lights on the front, and none of the buttons work, so I could not see the reading I would get at ground, or with a 1X or a 10X probe.
> > tom jobe...
> >
> >
> >
> >
> >
> > ----- Original Message -----
> > From: lldskjfoi
> > To: TekScopes@...
> > Sent: Thursday, September 20, 2012 8:28 AM
> > Subject: [TekScopes] Re: 2247A Channel 1 issue
> >
> >
> >
> > Thanks, thanks the info I was looking for. I started getting into the schematic last night I saw the path to U506. Will start checking it out.
> >
> > Also notice just letting the scope run with the probe attached the attenuation factor jumps back and forth- correct / incorrect. Does not change with no probe attached so it is getting an intermitant signal.
> >
> > KT
> >
> > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
> > >
> > > Maybe start by looking at the #1 "Vertical Inputs" schematic to see the simple circuit which starts at the junction of R101, C101 and R105 for channel 1.
> > > That junction point is connected to the readout ring behind the BNC connector, so you could test for continuity and shorting to ground, etc.
> > > If you compared both ends of R101 and R105 (Chan 1) with those of R102 and R106 (Chan 2) you might see if the readout voltage is getting that far.
> > > If that looked okay, you could go to U506 and see if it arrives at pins 12 and 15 as you would expect.
> > > With the scope on its back, you have excellent access to all of these resistors and U506 (16 pin DIP), with the blue outer cover slid back the width of your hand.
> > > From reading the service manual, it appears that the readout signal gets changed to a digital signal at U506, but is a simple voltage up to U506.
> > > tom jobe...
> > >
> > >
> > >
> > >
> > > ----- Original Message -----
> > > From: lldskjfoi
> > > To: TekScopes@...
> > > Sent: Wednesday, September 19, 2012 9:08 PM
> > > Subject: [TekScopes] Re: 2247A Channel 1 issue
> > >
> > >
> > >
> > > Some more checking shows channel 1 is not getting the code signal from the probe. Any idea on this circuit?
> > >
> > > KT
> > >
> > > --- In TekScopes@..., "lldskjfoi" <k.templin@> wrote:
> > > >
> > > > Hello everyone.
> > > >
> > > > Initial test show an issue with channel 1.
> > > >
> > > > Any direction on where to start?
> > > >
> > > > Channel 2 displays inputs correctly from the FG.
> > > >
> > > > Channel 1 displays 10 times the amplitude for the same signal.
> > > >
> > > > Example a 1 volt sine wave on channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions at .5v / div.
> > > >
> > > > The same signal on channel one reads as ~10v. When set to 5v / div it fills two div.
> > > >
> > > > Thanks, KT
> > > >
> > >
> >
>


Tom Jobe
 

开云体育

What is the voltages?on either side of?R105 and R106?
It seems like C111 and U506 are the only possibilities if the voltage leaving R105 more or less agrees with what you see arriving at U506 pin 15.
tom jobe...
?
?
?

----- Original Message -----
From: lldskjfoi
Sent: Thursday, September 20, 2012 6:04 PM
Subject: [TekScopes] Re: 2247A Channel 1 issue

?


10k and 10k. What do you thin? the only thing left are the caps or the U506 itself.

KT
--- In TekScopes@..., "Tom Jobe" wrote:
>
> What are the resistance readings across R105 and R106?
> That might help narrow the search down.
> tom jobe...
>
>
>
> ----- Original Message -----
> From: lldskjfoi
> To: TekScopes@...
> Sent: Thursday, September 20, 2012 5:19 PM
> Subject: [TekScopes] Re: 2247A Channel 1 issue
>
>
>
>
> I get 8.2 and 8.2 as well.
>
> KT
> --- In TekScopes@..., "Tom Jobe" wrote:
> >
> > I just checked R101 and R102 on a bare 2246 mainboard and they both measured 8.2k ohms in circuit.
> > If the resistance is correct on your scope, then something is drawing enough current to drop 0.7 volts across R101.
> > Nothing in that simple circuit looks like a suspect to me, no nasty 10uF to 47uF tantalums, etc.
> > tom jobe...
> >
> >
> >
> > ----- Original Message -----
> > From: lldskjfoi
> > To: TekScopes@...
> > Sent: Thursday, September 20, 2012 1:20 PM
> > Subject: [TekScopes] Re: 2247A Channel 1 issue
> >
> >
> >
> >
> > Thanks again for the responce. I think we are getting there.
> >
> > Using cahannel 2 as the "working reference"
> >
> > pin 12 (channel 2) on u506 goes from the +2.5 to +1.4 when a 10x probe is connected.
> >
> > pin (15 channel 1) on U506 goes from +.9 to +.6 with the probe connected.
> >
> > Checking R102 for channel 2 it has +2.5
> >
> > R101, (channel 1), shows +2.5 in but only +1.8 out.
> >
> > What are your thoughts? Is it simply a bad resistor or would one of the cap, C111 cause the voltage drop?
> >
> > Thanks again, KT
> > --- In TekScopes@..., "Tom Jobe" wrote:
> > >
> > > For the record, I slid the cover back on a very dead 2246 and checked pin 12 and pin 15 of U506, they both had +2.49 volts with no probes attached.
> > > This scope does not have any lights on the front, and none of the buttons work, so I could not see the reading I would get at ground, or with a 1X or a 10X probe.
> > > tom jobe...
> > >
> > >
> > >
> > >
> > >
> > > ----- Original Message -----
> > > From: lldskjfoi
> > > To: TekScopes@...
> > > Sent: Thursday, September 20, 2012 8:28 AM
> > > Subject: [TekScopes] Re: 2247A Channel 1 issue
> > >
> > >
> > >
> > > Thanks, thanks the info I was looking for. I started getting into the schematic last night I saw the path to U506. Will start checking it out.
> > >
> > > Also notice just letting the scope run with the probe attached the attenuation factor jumps back and forth- correct / incorrect. Does not change with no probe attached so it is getting an intermitant signal.
> > >
> > > KT
> > >
> > > --- In TekScopes@..., "Tom Jobe" wrote:
> > > >
> > > > Maybe start by looking at the #1 "Vertical Inputs" schematic to see the simple circuit which starts at the junction of R101, C101 and R105 for channel 1.
> > > > That junction point is connected to the readout ring behind the BNC connector, so you could test for continuity and shorting to ground, etc.
> > > > If you compared both ends of R101 and R105 (Chan 1) with those of R102 and R106 (Chan 2) you might see if the readout voltage is getting that far.
> > > > If that looked okay, you could go to U506 and see if it arrives at pins 12 and 15 as you would expect.
> > > > With the scope on its back, you have excellent access to all of these resistors and U506 (16 pin DIP), with the blue outer cover slid back the width of your hand.
> > > > From reading the service manual, it appears that the readout signal gets changed to a digital signal at U506, but is a simple voltage up to U506.
> > > > tom jobe...
> > > >
> > > >
> > > >
> > > >
> > > > ----- Original Message -----
> > > > From: lldskjfoi
> > > > To: TekScopes@...
> > > > Sent: Wednesday, September 19, 2012 9:08 PM
> > > > Subject: [TekScopes] Re: 2247A Channel 1 issue
> > > >
> > > >
> > > >
> > > > Some more checking shows channel 1 is not getting the code signal from the probe. Any idea on this circuit?
> > > >
> > > > KT
> > > >
> > > > --- In TekScopes@..., "lldskjfoi" wrote:
> > > > >
> > > > > Hello everyone.
> > > > >
> > > > > Initial test show an issue with channel 1.
> > > > >
> > > > > Any direction on where to start?
> > > > >
> > > > > Channel 2 displays inputs correctly from the FG.
> > > > >
> > > > > Channel 1 displays 10 times the amplitude for the same signal.
> > > > >
> > > > > Example a 1 volt sine wave on channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions at .5v / div.
> > > > >
> > > > > The same signal on channel one reads as ~10v. When set to 5v / div it fills two div.
> > > > >
> > > > > Thanks, KT
> > > > >
> > > >
> > >
> >
>


 

I have 2.5 at both ends of R106.

1.9 and 1 at R105.

--- In TekScopes@..., "Tom Jobe" <tomjobe@...> wrote:

What is the voltages on either side of R105 and R106?
It seems like C111 and U506 are the only possibilities if the voltage leaving R105 more or less agrees with what you see arriving at U506 pin 15.
tom jobe...



----- Original Message -----
From: lldskjfoi
To: TekScopes@...
Sent: Thursday, September 20, 2012 6:04 PM
Subject: [TekScopes] Re: 2247A Channel 1 issue




10k and 10k. What do you thin? the only thing left are the caps or the U506 itself.

KT
--- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
>
> What are the resistance readings across R105 and R106?
> That might help narrow the search down.
> tom jobe...
>
>
>
> ----- Original Message -----
> From: lldskjfoi
> To: TekScopes@...
> Sent: Thursday, September 20, 2012 5:19 PM
> Subject: [TekScopes] Re: 2247A Channel 1 issue
>
>
>
>
> I get 8.2 and 8.2 as well.
>
> KT
> --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
> >
> > I just checked R101 and R102 on a bare 2246 mainboard and they both measured 8.2k ohms in circuit.
> > If the resistance is correct on your scope, then something is drawing enough current to drop 0.7 volts across R101.
> > Nothing in that simple circuit looks like a suspect to me, no nasty 10uF to 47uF tantalums, etc.
> > tom jobe...
> >
> >
> >
> > ----- Original Message -----
> > From: lldskjfoi
> > To: TekScopes@...
> > Sent: Thursday, September 20, 2012 1:20 PM
> > Subject: [TekScopes] Re: 2247A Channel 1 issue
> >
> >
> >
> >
> > Thanks again for the responce. I think we are getting there.
> >
> > Using cahannel 2 as the "working reference"
> >
> > pin 12 (channel 2) on u506 goes from the +2.5 to +1.4 when a 10x probe is connected.
> >
> > pin (15 channel 1) on U506 goes from +.9 to +.6 with the probe connected.
> >
> > Checking R102 for channel 2 it has +2.5
> >
> > R101, (channel 1), shows +2.5 in but only +1.8 out.
> >
> > What are your thoughts? Is it simply a bad resistor or would one of the cap, C111 cause the voltage drop?
> >
> > Thanks again, KT
> > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
> > >
> > > For the record, I slid the cover back on a very dead 2246 and checked pin 12 and pin 15 of U506, they both had +2.49 volts with no probes attached.
> > > This scope does not have any lights on the front, and none of the buttons work, so I could not see the reading I would get at ground, or with a 1X or a 10X probe.
> > > tom jobe...
> > >
> > >
> > >
> > >
> > >
> > > ----- Original Message -----
> > > From: lldskjfoi
> > > To: TekScopes@...
> > > Sent: Thursday, September 20, 2012 8:28 AM
> > > Subject: [TekScopes] Re: 2247A Channel 1 issue
> > >
> > >
> > >
> > > Thanks, thanks the info I was looking for. I started getting into the schematic last night I saw the path to U506. Will start checking it out.
> > >
> > > Also notice just letting the scope run with the probe attached the attenuation factor jumps back and forth- correct / incorrect. Does not change with no probe attached so it is getting an intermitant signal.
> > >
> > > KT
> > >
> > > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
> > > >
> > > > Maybe start by looking at the #1 "Vertical Inputs" schematic to see the simple circuit which starts at the junction of R101, C101 and R105 for channel 1.
> > > > That junction point is connected to the readout ring behind the BNC connector, so you could test for continuity and shorting to ground, etc.
> > > > If you compared both ends of R101 and R105 (Chan 1) with those of R102 and R106 (Chan 2) you might see if the readout voltage is getting that far.
> > > > If that looked okay, you could go to U506 and see if it arrives at pins 12 and 15 as you would expect.
> > > > With the scope on its back, you have excellent access to all of these resistors and U506 (16 pin DIP), with the blue outer cover slid back the width of your hand.
> > > > From reading the service manual, it appears that the readout signal gets changed to a digital signal at U506, but is a simple voltage up to U506.
> > > > tom jobe...
> > > >
> > > >
> > > >
> > > >
> > > > ----- Original Message -----
> > > > From: lldskjfoi
> > > > To: TekScopes@...
> > > > Sent: Wednesday, September 19, 2012 9:08 PM
> > > > Subject: [TekScopes] Re: 2247A Channel 1 issue
> > > >
> > > >
> > > >
> > > > Some more checking shows channel 1 is not getting the code signal from the probe. Any idea on this circuit?
> > > >
> > > > KT
> > > >
> > > > --- In TekScopes@..., "lldskjfoi" <k.templin@> wrote:
> > > > >
> > > > > Hello everyone.
> > > > >
> > > > > Initial test show an issue with channel 1.
> > > > >
> > > > > Any direction on where to start?
> > > > >
> > > > > Channel 2 displays inputs correctly from the FG.
> > > > >
> > > > > Channel 1 displays 10 times the amplitude for the same signal.
> > > > >
> > > > > Example a 1 volt sine wave on channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions at .5v / div.
> > > > >
> > > > > The same signal on channel one reads as ~10v. When set to 5v / div it fills two div.
> > > > >
> > > > > Thanks, KT
> > > > >
> > > >
> > >
> >
>


Tom Jobe
 

开云体育

Well, unless I missed something you are down to C111 or U506 as being the culprit.
I would carefully lift one leg of C111.
It is near U506 pin 14, slightly further from the front of the scope.
It might also be orange in color.
tom jobe...
?

----- Original Message -----
From: lldskjfoi
Sent: Thursday, September 20, 2012 6:27 PM
Subject: [TekScopes] Re: 2247A Channel 1 issue

?


I have 2.5 at both ends of R106.

1.9 and 1 at R105.

--- In TekScopes@..., "Tom Jobe" wrote:
>
> What is the voltages on either side of R105 and R106?
> It seems like C111 and U506 are the only possibilities if the voltage leaving R105 more or less agrees with what you see arriving at U506 pin 15.
> tom jobe...
>
>
>
> ----- Original Message -----
> From: lldskjfoi
> To: TekScopes@...
> Sent: Thursday, September 20, 2012 6:04 PM
> Subject: [TekScopes] Re: 2247A Channel 1 issue
>
>
>
>
> 10k and 10k. What do you thin? the only thing left are the caps or the U506 itself.
>
> KT
> --- In TekScopes@..., "Tom Jobe" wrote:
> >
> > What are the resistance readings across R105 and R106?
> > That might help narrow the search down.
> > tom jobe...
> >
> >
> >
> > ----- Original Message -----
> > From: lldskjfoi
> > To: TekScopes@...
> > Sent: Thursday, September 20, 2012 5:19 PM
> > Subject: [TekScopes] Re: 2247A Channel 1 issue
> >
> >
> >
> >
> > I get 8.2 and 8.2 as well.
> >
> > KT
> > --- In TekScopes@..., "Tom Jobe" wrote:
> > >
> > > I just checked R101 and R102 on a bare 2246 mainboard and they both measured 8.2k ohms in circuit.
> > > If the resistance is correct on your scope, then something is drawing enough current to drop 0.7 volts across R101.
> > > Nothing in that simple circuit looks like a suspect to me, no nasty 10uF to 47uF tantalums, etc.
> > > tom jobe...
> > >
> > >
> > >
> > > ----- Original Message -----
> > > From: lldskjfoi
> > > To: TekScopes@...
> > > Sent: Thursday, September 20, 2012 1:20 PM
> > > Subject: [TekScopes] Re: 2247A Channel 1 issue
> > >
> > >
> > >
> > >
> > > Thanks again for the responce. I think we are getting there.
> > >
> > > Using cahannel 2 as the "working reference"
> > >
> > > pin 12 (channel 2) on u506 goes from the +2.5 to +1.4 when a 10x probe is connected.
> > >
> > > pin (15 channel 1) on U506 goes from +.9 to +.6 with the probe connected.
> > >
> > > Checking R102 for channel 2 it has +2.5
> > >
> > > R101, (channel 1), shows +2.5 in but only +1.8 out.
> > >
> > > What are your thoughts? Is it simply a bad resistor or would one of the cap, C111 cause the voltage drop?
> > >
> > > Thanks again, KT
> > > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
> > > >
> > > > For the record, I slid the cover back on a very dead 2246 and checked pin 12 and pin 15 of U506, they both had +2.49 volts with no probes attached.
> > > > This scope does not have any lights on the front, and none of the buttons work, so I could not see the reading I would get at ground, or with a 1X or a 10X probe.
> > > > tom jobe...
> > > >
> > > >
> > > >
> > > >
> > > >
> > > > ----- Original Message -----
> > > > From: lldskjfoi
> > > > To: TekScopes@...
> > > > Sent: Thursday, September 20, 2012 8:28 AM
> > > > Subject: [TekScopes] Re: 2247A Channel 1 issue
> > > >
> > > >
> > > >
> > > > Thanks, thanks the info I was looking for. I started getting into the schematic last night I saw the path to U506. Will start checking it out.
> > > >
> > > > Also notice just letting the scope run with the probe attached the attenuation factor jumps back and forth- correct / incorrect. Does not change with no probe attached so it is getting an intermitant signal.
> > > >
> > > > KT
> > > >
> > > > --- In TekScopes@..., "Tom Jobe" wrote:
> > > > >
> > > > > Maybe start by looking at the #1 "Vertical Inputs" schematic to see the simple circuit which starts at the junction of R101, C101 and R105 for channel 1.
> > > > > That junction point is connected to the readout ring behind the BNC connector, so you could test for continuity and shorting to ground, etc.
> > > > > If you compared both ends of R101 and R105 (Chan 1) with those of R102 and R106 (Chan 2) you might see if the readout voltage is getting that far.
> > > > > If that looked okay, you could go to U506 and see if it arrives at pins 12 and 15 as you would expect.
> > > > > With the scope on its back, you have excellent access to all of these resistors and U506 (16 pin DIP), with the blue outer cover slid back the width of your hand.
> > > > > From reading the service manual, it appears that the readout signal gets changed to a digital signal at U506, but is a simple voltage up to U506.
> > > > > tom jobe...
> > > > >
> > > > >
> > > > >
> > > > >
> > > > > ----- Original Message -----
> > > > > From: lldskjfoi
> > > > > To: TekScopes@...
> > > > > Sent: Wednesday, September 19, 2012 9:08 PM
> > > > > Subject: [TekScopes] Re: 2247A Channel 1 issue
> > > > >
> > > > >
> > > > >
> > > > > Some more checking shows channel 1 is not getting the code signal from the probe. Any idea on this circuit?
> > > > >
> > > > > KT
> > > > >
> > > > > --- In TekScopes@..., "lldskjfoi" <k.templin@> wrote:
> > > > > >
> > > > > > Hello everyone.
> > > > > >
> > > > > > Initial test show an issue with channel 1.
> > > > > >
> > > > > > Any direction on where to start?
> > > > > >
> > > > > > Channel 2 displays inputs correctly from the FG.
> > > > > >
> > > > > > Channel 1 displays 10 times the amplitude for the same signal.
> > > > > >
> > > > > > Example a 1 volt sine wave on channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions at .5v / div.
> > > > > >
> > > > > > The same signal on channel one reads as ~10v. When set to 5v / div it fills two div.
> > > > > >
> > > > > > Thanks, KT
> > > > > >
> > > > >
> > > >
> > >
> >
>


 

KT,

I money is on U506 has a bad input at pin 15. Lots of current's flowing into it, but it shouldn't be. Probably got zapped with a big dose of static electricity at the ring.

Jimmy

--- In TekScopes@..., "lldskjfoi" <k.templin@...> wrote:


I have 2.5 at both ends of R106.

1.9 and 1 at R105.


--- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:

What is the voltages on either side of R105 and R106?
It seems like C111 and U506 are the only possibilities if the voltage leaving R105 more or less agrees with what you see arriving at U506 pin 15.
tom jobe...



----- Original Message -----
From: lldskjfoi
To: TekScopes@...
Sent: Thursday, September 20, 2012 6:04 PM
Subject: [TekScopes] Re: 2247A Channel 1 issue




10k and 10k. What do you thin? the only thing left are the caps or the U506 itself.

KT
--- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
>
> What are the resistance readings across R105 and R106?
> That might help narrow the search down.
> tom jobe...
>
>
>
> ----- Original Message -----
> From: lldskjfoi
> To: TekScopes@...
> Sent: Thursday, September 20, 2012 5:19 PM
> Subject: [TekScopes] Re: 2247A Channel 1 issue
>
>
>
>
> I get 8.2 and 8.2 as well.
>
> KT
> --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
> >
> > I just checked R101 and R102 on a bare 2246 mainboard and they both measured 8.2k ohms in circuit.
> > If the resistance is correct on your scope, then something is drawing enough current to drop 0.7 volts across R101.
> > Nothing in that simple circuit looks like a suspect to me, no nasty 10uF to 47uF tantalums, etc.
> > tom jobe...
> >
> >
> >
> > ----- Original Message -----
> > From: lldskjfoi
> > To: TekScopes@...
> > Sent: Thursday, September 20, 2012 1:20 PM
> > Subject: [TekScopes] Re: 2247A Channel 1 issue
> >
> >
> >
> >
> > Thanks again for the responce. I think we are getting there.
> >
> > Using cahannel 2 as the "working reference"
> >
> > pin 12 (channel 2) on u506 goes from the +2.5 to +1.4 when a 10x probe is connected.
> >
> > pin (15 channel 1) on U506 goes from +.9 to +.6 with the probe connected.
> >
> > Checking R102 for channel 2 it has +2.5
> >
> > R101, (channel 1), shows +2.5 in but only +1.8 out.
> >
> > What are your thoughts? Is it simply a bad resistor or would one of the cap, C111 cause the voltage drop?
> >
> > Thanks again, KT
> > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
> > >
> > > For the record, I slid the cover back on a very dead 2246 and checked pin 12 and pin 15 of U506, they both had +2.49 volts with no probes attached.
> > > This scope does not have any lights on the front, and none of the buttons work, so I could not see the reading I would get at ground, or with a 1X or a 10X probe.
> > > tom jobe...
> > >
> > >
> > >
> > >
> > >
> > > ----- Original Message -----
> > > From: lldskjfoi
> > > To: TekScopes@...
> > > Sent: Thursday, September 20, 2012 8:28 AM
> > > Subject: [TekScopes] Re: 2247A Channel 1 issue
> > >
> > >
> > >
> > > Thanks, thanks the info I was looking for. I started getting into the schematic last night I saw the path to U506. Will start checking it out.
> > >
> > > Also notice just letting the scope run with the probe attached the attenuation factor jumps back and forth- correct / incorrect. Does not change with no probe attached so it is getting an intermitant signal.
> > >
> > > KT
> > >
> > > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
> > > >
> > > > Maybe start by looking at the #1 "Vertical Inputs" schematic to see the simple circuit which starts at the junction of R101, C101 and R105 for channel 1.
> > > > That junction point is connected to the readout ring behind the BNC connector, so you could test for continuity and shorting to ground, etc.
> > > > If you compared both ends of R101 and R105 (Chan 1) with those of R102 and R106 (Chan 2) you might see if the readout voltage is getting that far.
> > > > If that looked okay, you could go to U506 and see if it arrives at pins 12 and 15 as you would expect.
> > > > With the scope on its back, you have excellent access to all of these resistors and U506 (16 pin DIP), with the blue outer cover slid back the width of your hand.
> > > > From reading the service manual, it appears that the readout signal gets changed to a digital signal at U506, but is a simple voltage up to U506.
> > > > tom jobe...
> > > >
> > > >
> > > >
> > > >
> > > > ----- Original Message -----
> > > > From: lldskjfoi
> > > > To: TekScopes@...
> > > > Sent: Wednesday, September 19, 2012 9:08 PM
> > > > Subject: [TekScopes] Re: 2247A Channel 1 issue
> > > >
> > > >
> > > >
> > > > Some more checking shows channel 1 is not getting the code signal from the probe. Any idea on this circuit?
> > > >
> > > > KT
> > > >
> > > > --- In TekScopes@..., "lldskjfoi" <k.templin@> wrote:
> > > > >
> > > > > Hello everyone.
> > > > >
> > > > > Initial test show an issue with channel 1.
> > > > >
> > > > > Any direction on where to start?
> > > > >
> > > > > Channel 2 displays inputs correctly from the FG.
> > > > >
> > > > > Channel 1 displays 10 times the amplitude for the same signal.
> > > > >
> > > > > Example a 1 volt sine wave on channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions at .5v / div.
> > > > >
> > > > > The same signal on channel one reads as ~10v. When set to 5v / div it fills two div.
> > > > >
> > > > > Thanks, KT
> > > > >
> > > >
> > >
> >
>


 

I think so too.

I have a "parts" 2246 that has the same chip.

I will swap out. It will be next week before I can get into the A10's.

I will let you know.

Thanks for help, KT

--- In TekScopes@..., "jtjewell83" <jtjewell83@...> wrote:

KT,

I money is on U506 has a bad input at pin 15. Lots of current's flowing into it, but it shouldn't be. Probably got zapped with a big dose of static electricity at the ring.

Jimmy

--- In TekScopes@..., "lldskjfoi" <k.templin@> wrote:


I have 2.5 at both ends of R106.

1.9 and 1 at R105.


--- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:

What is the voltages on either side of R105 and R106?
It seems like C111 and U506 are the only possibilities if the voltage leaving R105 more or less agrees with what you see arriving at U506 pin 15.
tom jobe...



----- Original Message -----
From: lldskjfoi
To: TekScopes@...
Sent: Thursday, September 20, 2012 6:04 PM
Subject: [TekScopes] Re: 2247A Channel 1 issue




10k and 10k. What do you thin? the only thing left are the caps or the U506 itself.

KT
--- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
>
> What are the resistance readings across R105 and R106?
> That might help narrow the search down.
> tom jobe...
>
>
>
> ----- Original Message -----
> From: lldskjfoi
> To: TekScopes@...
> Sent: Thursday, September 20, 2012 5:19 PM
> Subject: [TekScopes] Re: 2247A Channel 1 issue
>
>
>
>
> I get 8.2 and 8.2 as well.
>
> KT
> --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
> >
> > I just checked R101 and R102 on a bare 2246 mainboard and they both measured 8.2k ohms in circuit.
> > If the resistance is correct on your scope, then something is drawing enough current to drop 0.7 volts across R101.
> > Nothing in that simple circuit looks like a suspect to me, no nasty 10uF to 47uF tantalums, etc.
> > tom jobe...
> >
> >
> >
> > ----- Original Message -----
> > From: lldskjfoi
> > To: TekScopes@...
> > Sent: Thursday, September 20, 2012 1:20 PM
> > Subject: [TekScopes] Re: 2247A Channel 1 issue
> >
> >
> >
> >
> > Thanks again for the responce. I think we are getting there.
> >
> > Using cahannel 2 as the "working reference"
> >
> > pin 12 (channel 2) on u506 goes from the +2.5 to +1.4 when a 10x probe is connected.
> >
> > pin (15 channel 1) on U506 goes from +.9 to +.6 with the probe connected.
> >
> > Checking R102 for channel 2 it has +2.5
> >
> > R101, (channel 1), shows +2.5 in but only +1.8 out.
> >
> > What are your thoughts? Is it simply a bad resistor or would one of the cap, C111 cause the voltage drop?
> >
> > Thanks again, KT
> > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
> > >
> > > For the record, I slid the cover back on a very dead 2246 and checked pin 12 and pin 15 of U506, they both had +2.49 volts with no probes attached.
> > > This scope does not have any lights on the front, and none of the buttons work, so I could not see the reading I would get at ground, or with a 1X or a 10X probe.
> > > tom jobe...
> > >
> > >
> > >
> > >
> > >
> > > ----- Original Message -----
> > > From: lldskjfoi
> > > To: TekScopes@...
> > > Sent: Thursday, September 20, 2012 8:28 AM
> > > Subject: [TekScopes] Re: 2247A Channel 1 issue
> > >
> > >
> > >
> > > Thanks, thanks the info I was looking for. I started getting into the schematic last night I saw the path to U506. Will start checking it out.
> > >
> > > Also notice just letting the scope run with the probe attached the attenuation factor jumps back and forth- correct / incorrect. Does not change with no probe attached so it is getting an intermitant signal.
> > >
> > > KT
> > >
> > > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
> > > >
> > > > Maybe start by looking at the #1 "Vertical Inputs" schematic to see the simple circuit which starts at the junction of R101, C101 and R105 for channel 1.
> > > > That junction point is connected to the readout ring behind the BNC connector, so you could test for continuity and shorting to ground, etc.
> > > > If you compared both ends of R101 and R105 (Chan 1) with those of R102 and R106 (Chan 2) you might see if the readout voltage is getting that far.
> > > > If that looked okay, you could go to U506 and see if it arrives at pins 12 and 15 as you would expect.
> > > > With the scope on its back, you have excellent access to all of these resistors and U506 (16 pin DIP), with the blue outer cover slid back the width of your hand.
> > > > From reading the service manual, it appears that the readout signal gets changed to a digital signal at U506, but is a simple voltage up to U506.
> > > > tom jobe...
> > > >
> > > >
> > > >
> > > >
> > > > ----- Original Message -----
> > > > From: lldskjfoi
> > > > To: TekScopes@...
> > > > Sent: Wednesday, September 19, 2012 9:08 PM
> > > > Subject: [TekScopes] Re: 2247A Channel 1 issue
> > > >
> > > >
> > > >
> > > > Some more checking shows channel 1 is not getting the code signal from the probe. Any idea on this circuit?
> > > >
> > > > KT
> > > >
> > > > --- In TekScopes@..., "lldskjfoi" <k.templin@> wrote:
> > > > >
> > > > > Hello everyone.
> > > > >
> > > > > Initial test show an issue with channel 1.
> > > > >
> > > > > Any direction on where to start?
> > > > >
> > > > > Channel 2 displays inputs correctly from the FG.
> > > > >
> > > > > Channel 1 displays 10 times the amplitude for the same signal.
> > > > >
> > > > > Example a 1 volt sine wave on channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions at .5v / div.
> > > > >
> > > > > The same signal on channel one reads as ~10v. When set to 5v / div it fills two div.
> > > > >
> > > > > Thanks, KT
> > > > >
> > > >
> > >
> >
>


 

The 4051 CMOS multiplexor is one of those very common "jelly bean" ICs
with many alternate sources and even improved versions.

On Fri, 21 Sep 2012 01:51:39 -0000, "lldskjfoi"
<k.templin@...> wrote:

I think so too.

I have a "parts" 2246 that has the same chip.

I will swap out. It will be next week before I can get into the A10's.

I will let you know.

Thanks for help, KT

--- In TekScopes@..., "jtjewell83" <jtjewell83@...> wrote:

KT,

I money is on U506 has a bad input at pin 15. Lots of current's flowing into it, but it shouldn't be. Probably got zapped with a big dose of static electricity at the ring.

Jimmy

--- In TekScopes@..., "lldskjfoi" <k.templin@> wrote:


I have 2.5 at both ends of R106.

1.9 and 1 at R105.


--- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:

What is the voltages on either side of R105 and R106?
It seems like C111 and U506 are the only possibilities if the voltage leaving R105 more or less agrees with what you see arriving at U506 pin 15.
tom jobe...



----- Original Message -----
From: lldskjfoi
To: TekScopes@...
Sent: Thursday, September 20, 2012 6:04 PM
Subject: [TekScopes] Re: 2247A Channel 1 issue




10k and 10k. What do you thin? the only thing left are the caps or the U506 itself.

KT
--- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
>
> What are the resistance readings across R105 and R106?
> That might help narrow the search down.
> tom jobe...
>
>
>
> ----- Original Message -----
> From: lldskjfoi
> To: TekScopes@...
> Sent: Thursday, September 20, 2012 5:19 PM
> Subject: [TekScopes] Re: 2247A Channel 1 issue
>
>
>
>
> I get 8.2 and 8.2 as well.
>
> KT
> --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
> >
> > I just checked R101 and R102 on a bare 2246 mainboard and they both measured 8.2k ohms in circuit.
> > If the resistance is correct on your scope, then something is drawing enough current to drop 0.7 volts across R101.
> > Nothing in that simple circuit looks like a suspect to me, no nasty 10uF to 47uF tantalums, etc.
> > tom jobe...
> >
> >
> >
> > ----- Original Message -----
> > From: lldskjfoi
> > To: TekScopes@...
> > Sent: Thursday, September 20, 2012 1:20 PM
> > Subject: [TekScopes] Re: 2247A Channel 1 issue
> >
> >
> >
> >
> > Thanks again for the responce. I think we are getting there.
> >
> > Using cahannel 2 as the "working reference"
> >
> > pin 12 (channel 2) on u506 goes from the +2.5 to +1.4 when a 10x probe is connected.
> >
> > pin (15 channel 1) on U506 goes from +.9 to +.6 with the probe connected.
> >
> > Checking R102 for channel 2 it has +2.5
> >
> > R101, (channel 1), shows +2.5 in but only +1.8 out.
> >
> > What are your thoughts? Is it simply a bad resistor or would one of the cap, C111 cause the voltage drop?
> >
> > Thanks again, KT
> > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
> > >
> > > For the record, I slid the cover back on a very dead 2246 and checked pin 12 and pin 15 of U506, they both had +2.49 volts with no probes attached.
> > > This scope does not have any lights on the front, and none of the buttons work, so I could not see the reading I would get at ground, or with a 1X or a 10X probe.
> > > tom jobe...
> > >
> > >
> > >
> > >
> > >
> > > ----- Original Message -----
> > > From: lldskjfoi
> > > To: TekScopes@...
> > > Sent: Thursday, September 20, 2012 8:28 AM
> > > Subject: [TekScopes] Re: 2247A Channel 1 issue
> > >
> > >
> > >
> > > Thanks, thanks the info I was looking for. I started getting into the schematic last night I saw the path to U506. Will start checking it out.
> > >
> > > Also notice just letting the scope run with the probe attached the attenuation factor jumps back and forth- correct / incorrect. Does not change with no probe attached so it is getting an intermitant signal.
> > >
> > > KT
> > >
> > > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
> > > >
> > > > Maybe start by looking at the #1 "Vertical Inputs" schematic to see the simple circuit which starts at the junction of R101, C101 and R105 for channel 1.
> > > > That junction point is connected to the readout ring behind the BNC connector, so you could test for continuity and shorting to ground, etc.
> > > > If you compared both ends of R101 and R105 (Chan 1) with those of R102 and R106 (Chan 2) you might see if the readout voltage is getting that far.
> > > > If that looked okay, you could go to U506 and see if it arrives at pins 12 and 15 as you would expect.
> > > > With the scope on its back, you have excellent access to all of these resistors and U506 (16 pin DIP), with the blue outer cover slid back the width of your hand.
> > > > From reading the service manual, it appears that the readout signal gets changed to a digital signal at U506, but is a simple voltage up to U506.
> > > > tom jobe...
> > > >
> > > >
> > > >
> > > >
> > > > ----- Original Message -----
> > > > From: lldskjfoi
> > > > To: TekScopes@...
> > > > Sent: Wednesday, September 19, 2012 9:08 PM
> > > > Subject: [TekScopes] Re: 2247A Channel 1 issue
> > > >
> > > >
> > > >
> > > > Some more checking shows channel 1 is not getting the code signal from the probe. Any idea on this circuit?
> > > >
> > > > KT
> > > >
> > > > --- In TekScopes@..., "lldskjfoi" <k.templin@> wrote:
> > > > >
> > > > > Hello everyone.
> > > > >
> > > > > Initial test show an issue with channel 1.
> > > > >
> > > > > Any direction on where to start?
> > > > >
> > > > > Channel 2 displays inputs correctly from the FG.
> > > > >
> > > > > Channel 1 displays 10 times the amplitude for the same signal.
> > > > >
> > > > > Example a 1 volt sine wave on channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions at .5v / div.
> > > > >
> > > > > The same signal on channel one reads as ~10v. When set to 5v / div it fills two div.
> > > > >
> > > > > Thanks, KT
> > > > >
> > > >
> > >
> >
>


 

The culprit was C111. Sniped one end and bingo.

Amazing what a 10 cent part can do.

Thanks Tom I really appreciate your help. You start overthinking the problem and going off in a wrong direction.

It's nice to have those "outside eyes" helping.

Snipping the cap was a great idea.

KT

--- In TekScopes@..., "Tom Jobe" <tomjobe@...> wrote:

Well, unless I missed something you are down to C111 or U506 as being the culprit.
I would carefully lift one leg of C111.
It is near U506 pin 14, slightly further from the front of the scope.
It might also be orange in color.
tom jobe...

----- Original Message -----
From: lldskjfoi
To: TekScopes@...
Sent: Thursday, September 20, 2012 6:27 PM
Subject: [TekScopes] Re: 2247A Channel 1 issue




I have 2.5 at both ends of R106.

1.9 and 1 at R105.

--- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
>
> What is the voltages on either side of R105 and R106?
> It seems like C111 and U506 are the only possibilities if the voltage leaving R105 more or less agrees with what you see arriving at U506 pin 15.
> tom jobe...
>
>
>
> ----- Original Message -----
> From: lldskjfoi
> To: TekScopes@...
> Sent: Thursday, September 20, 2012 6:04 PM
> Subject: [TekScopes] Re: 2247A Channel 1 issue
>
>
>
>
> 10k and 10k. What do you thin? the only thing left are the caps or the U506 itself.
>
> KT
> --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
> >
> > What are the resistance readings across R105 and R106?
> > That might help narrow the search down.
> > tom jobe...
> >
> >
> >
> > ----- Original Message -----
> > From: lldskjfoi
> > To: TekScopes@...
> > Sent: Thursday, September 20, 2012 5:19 PM
> > Subject: [TekScopes] Re: 2247A Channel 1 issue
> >
> >
> >
> >
> > I get 8.2 and 8.2 as well.
> >
> > KT
> > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
> > >
> > > I just checked R101 and R102 on a bare 2246 mainboard and they both measured 8.2k ohms in circuit.
> > > If the resistance is correct on your scope, then something is drawing enough current to drop 0.7 volts across R101.
> > > Nothing in that simple circuit looks like a suspect to me, no nasty 10uF to 47uF tantalums, etc.
> > > tom jobe...
> > >
> > >
> > >
> > > ----- Original Message -----
> > > From: lldskjfoi
> > > To: TekScopes@...
> > > Sent: Thursday, September 20, 2012 1:20 PM
> > > Subject: [TekScopes] Re: 2247A Channel 1 issue
> > >
> > >
> > >
> > >
> > > Thanks again for the responce. I think we are getting there.
> > >
> > > Using cahannel 2 as the "working reference"
> > >
> > > pin 12 (channel 2) on u506 goes from the +2.5 to +1.4 when a 10x probe is connected.
> > >
> > > pin (15 channel 1) on U506 goes from +.9 to +.6 with the probe connected.
> > >
> > > Checking R102 for channel 2 it has +2.5
> > >
> > > R101, (channel 1), shows +2.5 in but only +1.8 out.
> > >
> > > What are your thoughts? Is it simply a bad resistor or would one of the cap, C111 cause the voltage drop?
> > >
> > > Thanks again, KT
> > > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
> > > >
> > > > For the record, I slid the cover back on a very dead 2246 and checked pin 12 and pin 15 of U506, they both had +2.49 volts with no probes attached.
> > > > This scope does not have any lights on the front, and none of the buttons work, so I could not see the reading I would get at ground, or with a 1X or a 10X probe.
> > > > tom jobe...
> > > >
> > > >
> > > >
> > > >
> > > >
> > > > ----- Original Message -----
> > > > From: lldskjfoi
> > > > To: TekScopes@...
> > > > Sent: Thursday, September 20, 2012 8:28 AM
> > > > Subject: [TekScopes] Re: 2247A Channel 1 issue
> > > >
> > > >
> > > >
> > > > Thanks, thanks the info I was looking for. I started getting into the schematic last night I saw the path to U506. Will start checking it out.
> > > >
> > > > Also notice just letting the scope run with the probe attached the attenuation factor jumps back and forth- correct / incorrect. Does not change with no probe attached so it is getting an intermitant signal.
> > > >
> > > > KT
> > > >
> > > > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote:
> > > > >
> > > > > Maybe start by looking at the #1 "Vertical Inputs" schematic to see the simple circuit which starts at the junction of R101, C101 and R105 for channel 1.
> > > > > That junction point is connected to the readout ring behind the BNC connector, so you could test for continuity and shorting to ground, etc.
> > > > > If you compared both ends of R101 and R105 (Chan 1) with those of R102 and R106 (Chan 2) you might see if the readout voltage is getting that far.
> > > > > If that looked okay, you could go to U506 and see if it arrives at pins 12 and 15 as you would expect.
> > > > > With the scope on its back, you have excellent access to all of these resistors and U506 (16 pin DIP), with the blue outer cover slid back the width of your hand.
> > > > > From reading the service manual, it appears that the readout signal gets changed to a digital signal at U506, but is a simple voltage up to U506.
> > > > > tom jobe...
> > > > >
> > > > >
> > > > >
> > > > >
> > > > > ----- Original Message -----
> > > > > From: lldskjfoi
> > > > > To: TekScopes@...
> > > > > Sent: Wednesday, September 19, 2012 9:08 PM
> > > > > Subject: [TekScopes] Re: 2247A Channel 1 issue
> > > > >
> > > > >
> > > > >
> > > > > Some more checking shows channel 1 is not getting the code signal from the probe. Any idea on this circuit?
> > > > >
> > > > > KT
> > > > >
> > > > > --- In TekScopes@..., "lldskjfoi" <k.templin@> wrote:
> > > > > >
> > > > > > Hello everyone.
> > > > > >
> > > > > > Initial test show an issue with channel 1.
> > > > > >
> > > > > > Any direction on where to start?
> > > > > >
> > > > > > Channel 2 displays inputs correctly from the FG.
> > > > > >
> > > > > > Channel 1 displays 10 times the amplitude for the same signal.
> > > > > >
> > > > > > Example a 1 volt sine wave on channel 2 displays on the voltmeter function correctly at ~1v and 2 divisions at .5v / div.
> > > > > >
> > > > > > The same signal on channel one reads as ~10v. When set to 5v / div it fills two div.
> > > > > >
> > > > > > Thanks, KT
> > > > > >
> > > > >
> > > >
> > >
> >
>