7633/7b53a in XY
I have a 7633 loaded with a 7b53a and a 7a26. When the 7b53a is put into XY mode i get a fine horizontal adjustment from it, but my vertical adjustment moves diagonally. When I remove the 7b53a and put in my 7ct1n curve tracer and use it with the 7a26 I get normal vertical adjustment. Anyone experience this before? Thanks for reading!
|
On Wed, May 2, 2018 at 05:47 pm, lop pol wrote:
I just tested the 7b53a with the curve tracer and everything is normal. |
Hi Lop Pol,
toggle quoted message
Show quoted text
Not really. You can, of course, set the 7B53A to EXT, but you will be limited to the very few options of sensitivity and/or attenuation. You will basically have only 10mV / div ( with 10x activated and EXT with no attenuation), or 100mV/div (with x10 off), or yet, 1V/div, if you choose EXT¡Â10. But it also works with 7B53A source select to INT, in such case it will receive the X-axis signal routed through the 7633's triggering selection subsystem. When you do this, you use the 7A26's channel select to, say, CH1, and the 7A26's triggering selection to CH2, In the above example, CH1 will become the Y signal and CH2 will become the X signal, and you will have all the 7A26's sensitivity and attenuation choices for the X axis. Rgrds, Fabio On Wed, May 2, 2018 at 06:52 pm, lop pol wrote:
|