Hi all. I'm back with another amateur mistake in my audio lab. My 5L4N Spectrum Analyzer died from a short in the 5440 scope. I was going to start a calibration procedure on it using cable extenders. The extender must have slid down a little in the mainframe scope. When I turned it on it smoked, hissed and blew the mainframe fuse, and fried two tants next to the plugin sockets. Once i got the scope going again I realized the 5L4N was no longer displaying any specific frequencies. Even the comb output wouldn't display anymore. Nothing looks burnt inside which surprised me because I probably sent 200v into it! I do have a display but just the baseline noise like when there no input signal connected. The noise display responds to the attenuator and span adjustments.
It's taken a few weeks to get hold of the final version service manual and make new cable extenders with keys to keep it from sliding down in the slot. I verified the comb output was good on my HP spectrum analyzer. This is a really complicated unit. So not exactly sure where to go after verifying the internal power rails. +15, -15, and -5v are all good inside the unit and power is intact at least through the input stage.
I was thinking that the problem was at or after the mixer because the comb signal enters at the mixer. But as I'm writing this realize that can't be true because you have to jumper the comb output to the input bnc. So the problem could be as early as in the input stage.
Is there are method I could use to determine what is an appropriate signal between the main stages on the block diagram? How would you guys approach this.
Thanks, Rick