¿ªÔÆÌåÓý

ctrl + shift + ? for shortcuts
© 2025 Groups.io
Date
Re: TM 504 Power Module Questions
I think you got that backwards Pins 1 thru 13 (bottom) are all the same Pins 14 thru 28 (upper) are mostly custom select as per units allow. Don't forget side A and B looking from the front A is on
By [email protected] · #184850 ·
Re: SG503 question
Mine is an early serial number B0102xx an still well within specs. Jeff
By [email protected] · #184849 ·
Re: Paper Used in Older Tek Manuals
Dave Henderson of Artek manuals posted awhile back showing how to replace a comb binding. DaveD
By Dave Daniel · #184848 ·
Re: Paper Used in Older Tek Manuals
I bought a bunch of black and white combs and a small-capacity comb punch a while back. I used it to replace many crumbling Tektronix manual combs, as well as binding various other documents (such as
By Dave Daniel · #184847 ·
Re: Paper Used in Older Tek Manuals
Years ago (2006), I found a comb binder on ebay. It was about $112. Did not include supplies, which are easy to get. the trade name was Ibimatic, which is a brand of Ibico. It's handy to have. I don't
By rick · #184846 ·
Re: New (old) 2247a enquiry
Thanks, that fixed it. Bit embarassing, should have read the user manual before looking at the service manual.
By Richard Merifield · #184845 ·
Re: SG503 question
I can check the 2 of mine if that would help. I would suspect that they are close to spec if not in spec. I am curious if the 2 where the output is falling off are early serial numbers and the one you
By Zentronics42@... · #184844 ·
Re: SG503 question
Zen, Thanks for your reply. Table 3-3 of the manual suggests typical LF limits of 4.70 and 9.09 MHz on the two ranges in question. My units get very close to, or past, these values before the output
By romeo987 · #184843 ·
Re: Ballantine Calibrator manual
Thank you! I have scanned and uploaded the Ballantine 6126M manual to KO4bb and BAMA. I have an early and late 6125B/C manual that is in the que to be scanned and uploaded...
By Ken Eckert · #184842 ·
Ballantine Calibrator manual
Fellow Tekkies, You find the darndest things when you're cleaning up. In this case, it was a clean reproduction of the operating and service manuals for the Ballantine Labs 6127B O-Scope calibrator.
By Bruce Lane · #184841 ·
Re: Advice on fixing 2465B slow horiz sweep
Sorry that I haven¡¯t been following this thread closely, so forgive me if you¡¯ve already explicitly verified that the vertical calibration has not been affected. But if the vertical gain is off by
By Tom Lee · #184840 ·
Re: Paper Used in Older Tek Manuals
Thank you Toby and Ken for making those measurements. I have several reams of paper on order (28 lb and 32 lb because, as you say, heavier feels better). At the moment I'm using a blue 1/2 inch 3-ring
By Jeff Dutky · #184839 ·
Re: Paper Used in Older Tek Manuals
FWIW, I weighed 10 pages of a Tek 515A manual, I got approx 107 gsm -- approx 405 gsm for the blue covers. Of course heavier always feels better (I would try at least 120 gsm equivalent). --Toby
By toby@... · #184838 ·
Re: Advice on fixing 2465B slow horiz sweep
------- There are good suggestions in the other posts too. Given t
By Ozan · #184837 ·
Re: Advice on fixing 2465B slow horiz sweep
Hey Grayson, It'd be helpful if you could describe what you've done and seen in some detail. Detailed observations are vastly more likely to produce useful advice than writing up your conclusions.
By Sigur?ur ?sgeirsson · #184836 ·
Re: Advice on fixing 2465B slow horiz sweep
I suspect horizontal gain drift since that fits the symptoms and I don't know of any other problem. I replace all the electrolytics in the power supply and everywhere else I could find one. Just not
By Grayson Evans · #184835 ·
Re: DM 501A Mode Select Switches Issue
Jared, Thanks, very helpful. -- Michael Lynch Dardanelle, AR
By Mlynch001 · #184834 ·
Re: New (old) 2247a enquiry
If nothing else, sounds like you need to perform the trace rotation which is a screw trim adjustment which can be accessed from the front panel between the Focus and Readout control knobs just under
By Alex · #184833 ·
Re: AA 501A Parts, DM 501A Probes
Amphenol If they have the cable you need. Tend to be a little on the pricy side. $20 for a longish BNC but one of the top shelf brands before you get in to the exotic high frequency cables and
By Zentronics42@... · #184832 ·
Re: SG503 question
When the output falls is the frequency outside the selected range? I have one that will let me set the frequency quite a bit low and high on the selected range but the leveling is not specified
By Zentronics42@... · #184831 ·