Re: 425 Mil vertical module needed....seeking
Hi thanks Harvey....yes that's about it.....time saving with a good spares inventory ....replace the faulty aspect of equipment and then send through the system to component level repairs. Repair was
By
Jack
·
#152096
·
|
Re: No contribution from David Hess since July 18th
Hi Raymond, My records show he has never posted on TekScopes since we switched to Groups.io in Dec 2017. There were no posts from him on July 18 of this year either. Dennis Tillman W7PF
By
Dennis Tillman W7pF
·
#152095
·
|
Re: Tek 465 "B" Sweep
ok, all I recall is I was told ( by the counter tech at tektronix in sunnyvale when my 475 went in for cal service eons ago... ) the case had to be on for something due to sensitivity/shielding and
By
Ren¨¦e
·
#152094
·
|
Re: No contribution from David Hess since July 18th
He's active on EEVBlog forum, last post was yesterday. http://www.eevblog.com/forum/profile/?area=showposts;u=93452
By
Tom Gardner
·
#152093
·
|
No contribution from David Hess since July 18th
FYI: I'm missing contributions from our long time, very active, helpful and knowledgeable member David Hess. As far as I can see, his last message on this forum is from July 18th. On October 7th, I
By
Raymond Domp Frank
·
#152092
·
|
Re: N channel fet 151-1121-00 -- V10206
Well Dennis - I learned something there! I have a 7CT1N (no surprise there!) and have only occasionally used it for low power bipolars. I would have missed the oddball electrode mapping between
By
Craig Sawyers <c.sawyers@...>
·
#152091
·
|
Re: N channel fet 151-1121-00 -- V10206
Hi Raymond, That was the easy part, the photo shows he was using a 7633 scope (lucky guy to own one of those) so it had to be a 7CT1N. Slightly harder was verifying that the traces on the screen are
By
Dennis Tillman W7pF
·
#152090
·
|
Re: N channel fet 151-1121-00 -- V10206
On Fri, Nov 2, 2018 at 06:37 PM, Dennis Tillman W7PF wrote: Hi Dennis, > > COMMON MISTAKE: Your photo shows you did not switch the Drain and the Gate > leads when you were using the 7CT1N. I believe
By
Raymond Domp Frank
·
#152089
·
|
Re: Tek 465 "B" Sweep
I've never had a need to do any of these adjustments externally, "in the field" so to say, using 464, 465(B), 466 and 475(A) scopes for decades. The "construction" allows unguided poking at least 1.5"
By
Raymond Domp Frank
·
#152088
·
|
Re: N channel fet 151-1121-00 -- V10206
Hi, COMMON MISTAKE: Your photo shows you did not switch the Drain and the Gate leads when you were using the 7CT1N. I believe when you do switch them that you will see there is nothing wrong with the
By
Dennis Tillman W7pF
·
#152087
·
|
Re: N channel fet 151-1121-00 -- V10206
Yes that is the fet in the curve tracer. I hope I didn't blow it up trying to curve trace it.
By
Brendan
·
#152086
·
|
Re: N channel fet 151-1121-00 -- V10206
Not precisely sure what I'm looking at. Is that the said FET on a 7CT1N in your 'scope? Because if it is, I would say that was an ex-FET. It has gone to join the choir eternal.
By
Craig Sawyers <c.sawyers@...>
·
#152085
·
|
Re: Tek 465 "B" Sweep
I don't remember it that way. The entire 475 can be calibrated sitting out of its case on the bench. The entire 465 family can be calibrated sitting out of its case on the bench. The adjustment holes
By
Chuck Harris <cfharris@...>
·
#152084
·
|
Re: Tek 465 "B" Sweep
These holes provide access to e.g. vertical position, attenuator balance and DC gain settings. They are there as a convenience to allow some adjustments "in the field" without opening the case. These
By
Raymond Domp Frank
·
#152083
·
|
Re: Tek 465 "B" Sweep
check the side of the case there may be "holes" for some of the sensitive adjustments. there are on a 475 and the case needs to be installed for those adjustments. Ren¨¦e
By
Ren¨¦e
·
#152082
·
|
Re: N channel fet 151-1121-00 -- V10206
After messing around this is the trace I get. /g/TekScopes/photo/77313/0?p=Name,,,20,1,0,0
By
Brendan
·
#152081
·
|
N channel fet 151-1121-00 -- V10206
I just put this fet (151-1121-00) on the curve tracer and get nothing. should I follow the same procedure to trace this fet as a 151-1042-00 --- S2089? Or is something different? The 151-1121-00 in
By
Brendan
·
#152080
·
|
Re: Tek 465 "B" Sweep
That's a very good point. When I was using CFL illumination in a desk lamp, at my bench, I used to see a lot of excess noise signals in hi gain, hi-Z circuitry. In addition, carrying a smart phone
By
Chuck Harris <cfharris@...>
·
#152079
·
|
Re: Tek 465 "B" Sweep
No, it is not normal although the area is sensitive to external signals, like hum; you're covering and grounding a high-impedance part of the vertical input amp, directly after the attenuator block
By
Raymond Domp Frank
·
#152078
·
|
Re: Tek 465 "B" Sweep
Definitely not normal! If tektronix needed you to put a special shield there, there would be a shield with a part number specified in the list of things required for calibration. What you have is
By
Chuck Harris <cfharris@...>
·
#152077
·
|