Re: 2215A LVPS repair
On Sat, Aug 11, 2018 at 05:29 PM, tom jobe wrote: > > Hi Leo, > Yesterday I found the Fluke thermocouple and did some tests on a 2215A. > First I disconnected the optional fan, then let the scope run
By satbeginner · #150407 ·
Re: 2215A LVPS repair
Hi Leo, Yesterday I found the Fluke thermocouple and did some tests on a 2215A. First I disconnected the optional fan, then let the scope run with two traces like you did. The temperature of the
By tom jobe <tomjobe@...> · #150406 ·
Re: No trace on 556 :(
Thanks for the info morrios , do you have anything points that you think will help in this process ??? Things to pay attention to.? On Sat, Aug 11, 2018 at 5:22 AM, Morris
By oliver johnson · #150405 ·
Re: scanned: 545B and 561 Instrument Reference Books
I don't currently have Instrument Reference Books for any instruments later than the early 1960s. If I find them, I will scan them and share the scans.
By Kurt Rosenfeld · #150404 ·
Re: 464
Are all the low voltages correct?
By Brendan · #150403 ·
Re: No trace on 556 :(
I have uploaded scans of the hv transformer drawings to a folder in the Photos area called (not surprisingly) "556 high voltage transformers". Good luck! Morrios
By Morris Odell · #150402 ·
Re: No trace on 556 :(
I have the Tek specification drawings for both the 556 HV transformers - I'm not sure if they're in the group files but if not I'll upload them. When I mentioned a jig, what i was referring to was a
By Morris Odell · #150401 ·
464
hey all,i have a 464 with no crt display.front pannel lights work. i have the service man..wow what a book ! Any comon problems causing crt not to light?
By kickingypsy9@... · #150400 ·
Re: scanned: 545B and 561 Instrument Reference Books
I see you are very busy with the Reference Books Kurt so thanx for that. Are these books mainly about the "older" stuff from about the 1950 - 1960 era or made Tek them also for say the 547, 647a and
By tek_547 · #150399 ·
Re: Tek 549 storage board anomaly
Could somebody forward this to the OP e-mail address, Tim Phillips (UK), please? Can't see it in the archives, as it's obfuscated by groups.io. Hopefully, at least the group moderators can see it, and
By Sebastian Garcia · #150398 ·
Re: 7704A Z-Axis Problem
(continued) typo, C4399 is at the +15V rail.
By Albert Otten · #150397 ·
Re: 7704A Z-Axis Problem
Hi Tomas, Since I was just curious what my 7704A would show I took off the upper side panels, attached probes and switched on the 7704A. A big very bright dot illuminated the center of the screen. At
By Albert Otten · #150396 ·
Re: 2215A LVPS repair
Hi all, Hi Tom, I tested for temperature this afternoon again, after I performed a full calibration on the scope. So, the scope had been running for several hours when I did this measurement. I used a
By satbeginner · #150395 ·
SENDING CLASSIC SCOPES AROUND THE WORLD
For as long as I have been a member of TekScopes there are two things I have heard regularly from our members outside the US: It costs too much to ship classic Tek scopes outside the US and they ae
By Dennis Tillman W7pF · #150394 ·
Re: 2215A LVPS repair
Be careful when making changes to the operating characteristics of the switching parts of a SMPS. Faster tr and tf can improve efficiency, but can aggravate PS noise and cause EMC issues. Ed
By Ed Breya · #150393 ·
Re: 2215A LVPS repair
Hi Leo, Could you give me the conditions under which you tested your Q9070 FET for the temperature, so I can get you an accurate temperature number to compare to a working 2215A here? The recent
By tom jobe <tomjobe@...> · #150392 ·
Re: scanned: 545B and 561 Instrument Reference Books
Here is the Type 60 / 2A60 Instrument Reference Book: http://w140.com/tekwiki/images/0/0e/Tek_type_60_irb.pdf
By Kurt Rosenfeld · #150391 ·
Re: 2215A LVPS repair
Hi Leo, Thanks for your feedback / experiments :-) About the fan, easiest is to use a simple fan without any electronics in it to begin with ;-) A standard two wire fan, or a 3 wire fan where you just
By Vince · #150390 ·
Re: 2215A LVPS repair
Hi all, here is my follow up on the FET replacement in my 2215A, and the adding of an optional fan. Using a P9NK70ZFP (the working one I started with) gave a FET housing temperature of 68 degree
By satbeginner · #150389 ·
Re: Calpak VS AFG
Timothy, The AFG 3252, for square waves, is spec'd for a time equal or less than 2.5ns, which is good enough to calibrate a 25MHz scope (which has a risetime of 14ns). But it wouldn't cut it for a
By Fabio Trevisan · #150388 ·