Re: Cleaning Tek475A Attenuator Switches
Great job. I did the same thing on my 465 after several months of erratic behavior and several cleanups of contacts. In the end I decided to try the chance, took off modules attenuators, took off
By
unclebanjoman
·
#85573
·
|
Re: what is the differences between 465 and 465B?
There is a nice article in one of the Tekscopes flyers which covers the updates. If nobody else gets it in the next few hours, I'll see if I can find it and attach it to a new message. I have a
By
Bert Haskins
·
#85572
·
|
Re: Which would you pick? 454 or 475?
It is funny you should mention that. I told someone on the funwithtubes list that the warm up drift he complained about in the 454 was expected because its Nuvistor based vertical amplifiers are not
By
David Hess
·
#85570
·
|
Re: 2246 Mod A main board overloading the SMPS
Hi Priya, Resistances were taken on a nice working 2246 with a Fluke 110 which does not have enough voltage to forward bias the diodes. I tried this Fluke meter on some standard diodes and some
By
Tom Jobe <tomjobe@...>
·
#85571
·
|
Re: Need help choosing a scope and some other RF capable equipment.
Only Kansas and my high-school girlfriends are flat. :) 73 Jim N6OTQ
By
Jim <n6otq@...>
·
#85569
·
|
Re: Rear feet for 475
Could be. Admittedly all the functionality I have observed is at room temperature here in southern Louisiana. I can but assume they would/may turn quite a bit more brittle at subzero
By
Rob <rgwood@...>
·
#85568
·
|
Re: High Voltage Probe to work on Scopes.
05/01/2013 00:15 So they would be OK to look for ripple on a 5kV plate supply I have made? -- Best Regards, Chris Wilson.
By
Chris Wilson
·
#85567
·
|
Re: Which would you pick? 454 or 475?
The 475 is definitely a better scope, but... I really enjoy using the 454. The sharp trace is very, very nice. The thing I like least about it is the DC drift/unstable attenuator balance. You end up
By
David DiGiacomo
·
#85566
·
|
Re: High Voltage Probe to work on Scopes.
Yes, Craig is right. I only use it on a scope for < 100Hz stuff and if I'm *really* desperate since I don't have a better performing high voltage probe. Neil www.gruending.net
By
Neil Gruending
·
#85565
·
|
Re: what is the differences between 465 and 465B?
There are major differences between the models.? The service manual won't work for both. I have both (although the 465B is a parts unit) and the differences are so much that it's not possible to
By
Bob Albert <bob91343@...>
·
#85564
·
|
Re: High Voltage Probe to work on Scopes. [1 Attachment]
====================== I guess this is the graph, found by Google search <http://www.ko4bb.com/Manuals/Fluke/Bandwidth_Measurement_80K40.pdf>. It's valuable to know what's really happening. Looking at
By
Craig Sawyers <c.sawyers@...>
·
#85563
·
|
Re: Need help choosing a scope and some other RF capable equipment.
That is not resonance. A bias tee is not a big deal.
By
David DiGiacomo
·
#85562
·
|
what is the differences between 465 and 465B?
Another newbie question and let me say thanks for the groups patience beforehand!!? I have been trying to research the differences between the Tek 465 and 465B scopes.? Can I use the 465 Service
By
Matt Jodz <matt.inchina@...>
·
#85561
·
|
Re: 2246 Mod A main board overloading the SMPS
Completely agree. I ran the SMPS with just the fan for a whole afternoon before I was willing to go further. With the 40W series bulb removed :), the SMPS starts up correctly with just the SMPS and
By
Priya Jakatdar
·
#85560
·
|
Re: Which would you pick? 454 or 475? -->the 475!!!
realistically, the 475 is just a better scope in every way, and significantly newer and better desgined. and you have to love the 200+MHz bandwidth, and the ability to stop it down for lower frequency
By
wshawlee2 <walter2@...>
·
#85559
·
|
Re: Need help choosing a scope and some other RF capable equipment.
Hi jerry, I'm in Australia, shipping costs here would be much more than the scope. It looks like a nice package, I'm sure it will sell. I always enjoy your posts to the group, usually learn something
By
Don Black <donald_black@...>
·
#85558
·
|
Re: Need help choosing a scope and some other RF capable equipment.
Thanks Don, Yes it is 200MHZ. I will see if I can fix the listing. Where are you? Jerry Massengale
By
jerry massengale
·
#85557
·
|
Re: New old scope owner
Yes by logging into yahoogroups.com
By
John Griessen <john@...>
·
#85556
·
|
Re: Need help choosing a scope and some other RF capable equipment.
You mean it really isn't flat??? Don Black.
By
Don Black <donald_black@...>
·
#85555
·
|
New old scope owner
Hi to everyone, I've been reading the posts and researching the old analog Tek scopes. It's very exciting to see the interest shown even after all these years. My 453 has arrived. I haven't been able
By
Steve Purcell <steve.purcell.k@...>
·
#85554
·
|