Re: Tek 2230 - Storage does not work.
wrote: Here is something else to try. Swap P2111 and P2112 so that channel 1 and channel 2 are swapped and measure these voltages again.
By
David Hess
·
#70376
·
|
Re: Tek 2230 - Storage does not work.
wrote: 13 is good but 12 is off and being limited by the internal back-to-back diodes on the ouput. Yep. It is being pulled low by constant current sink Q2104 and not being balanced by Q2103 for some
By
David Hess
·
#70375
·
|
Re: Tek 2230 - Storage does not work.
-5% is Good. The Resistors all measure good and are connected. Both the chan 1 and 2 emitter networks measure the same values. 12 5.00 13 5.83 The base of Q2105 is -3.5 - which seems way off What
By
david_s5y
·
#70374
·
|
Re: TG501 option 1 connections
Hi, I do not have a spectrum analyzer but I am not surprised that a multiplier would have noisy outputs. Jerry Massengale jmassen418a@...
By
Jerry Massengale <jmassen418a@...>
·
#70373
·
|
Tektronix vacuum tube vintage oscilloscope - $75 (Seattle) 535
Say this on Craigslist http://seattle.craigslist.org/see/ele/2810475953.html I hope someone buys it to keep it from being torn apart. Bill Higdon here's the ad. " Tektronix 535 vacuum tube
By
Bill Higdon
·
#70372
·
|
Re: PG508 - Using 10-turn pots
It's a real circus in there. The coarse shaft has a gap, in which the fine shaft drives the readout via a bevel gear and several more gears. The slightest bind or friction will drive a prospective
By
Dave Wise
·
#70371
·
|
Re: PG508 - Using 10-turn pots
Those wire wound 0.1% resistors were not cheap and the mechanics of the display are complicated. The DVM should be more accurate. I like the multi-turn potentiometer but just got a pair of the
By
David Hess
·
#70370
·
|
Re: Tek 2230 - Storage does not work.
wrote: Good. These are the IC power pins. They are connected through a couple of diodes to ground and the 8.6 volt supply as shown on schematic 11. If you can dig up the Tek-Made book, the data sheet
By
David Hess
·
#70369
·
|
Re: PG508 - Using 10-turn pots
Kinda like the mechanical knob/switch version of the 7A13 over the later digital one with DVM. IMO, the switch is nicer to use than the multi turn pot. The switch has a group of 0.1% precision
By
Bernice Loui <rupunzels_window@...>
·
#70368
·
|
Re: Power supply problem with Tek 2220
Hi Tan, One more thing for today. On this 2230 I have, the voltage at the gate of Q935 is 0.211 volts more positive than the Kathode of the SCR, so it must be very close to firing the SCR! The 0.054
By
tom jobe <tomjobe@...>
·
#70367
·
|
Re: Tek 2230 - Storage does not work.
3 to gnd 1.38 - Where is this tied? 5 to gnd 6.8 4 Ch 1 or Both 1.68 Ch 2 4.0 14 Both Add 6.16 14 Both Alt/Chop 1.44 14 ch 1 Alt/Chop/add 1.44 14 ch 2 Alt/Chop/add 1.44 DC bias 15 1.8 16 1.04 1 1.16 2
By
david_s5y
·
#70366
·
|
Re: 2465A Options
Hi Carl, Option 10: GPIB (my 2465A has that, by the way) Option 11: Power connectors for active powered probes I do not know the meaning of the other options. Patrick Wong AK6C
By
honolulusnowwhite
·
#70365
·
|
Re: Intermittent triggering on 2465A?
Congratulations on your *two* scope purchases, Chip! I suggest that you fix the 475 because sooner or later the 2465A will die and you will need a scope to fix it. The main reason to own a scope is to
By
honolulusnowwhite
·
#70364
·
|
Re: TG501 option 1 connections
Congratulations, Jerry! I had a problem with intermittent output at the 10 ns (100 MHz) and 20 ns (50 MHz) positions and the problem was a flaky 10102 chip, U300. So I call that to your attention in
By
honolulusnowwhite
·
#70363
·
|
Re: AFG5101 questions...
This is part # 150-1206-00 but I can't find out any information about it. The pinout is identical to a generic Hitachi HD44780 type module - Anyone know if Tek used these elsewhere and if it really is
By
Nick de Smith
·
#70362
·
|
Re: Power supply problem with Tek 2220
Hi Tan, Thank you for the update on your progress. You have provided us with many things to think about! I wonder where you will go from here with your 2230? The voltage you have at R949 is now about
By
tom jobe <tomjobe@...>
·
#70361
·
|
Re: TEK parts, a reliable supplier
Confirm, I've purchased several items from them, all of good quality (including two VERY rare TDs for Tek!) Very very recommended. Max
By
unclebanjoman
·
#70360
·
|
Re: TEK parts, a reliable supplier
I have to agree that Qservice provides excellent service and order fulfillment and I am very glad that Dinos is supporting the Tek community. Patrick Wong AK6C
By
honolulusnowwhite
·
#70359
·
|
Re: Tek 2230 - Storage does not work.
Have you checked the control input levels at pin 4 and pin 14 and the power supply pin voltages at pin 3 and pin 5? What about the DC bias levels at all of the input and output pins? They are marked
By
David Hess
·
#70358
·
|
Re: PG508 - Using 10-turn pots
Think HPOS on lots of things, Volts on PS503A, both POS on 577, VPOS on 1502/3.... -ls- David Wise <david_wise@...> wrote:
By
teamlarryohio
·
#70357
·
|