Re: 7704A Repair and modules. 7B53 in faster frames
bandwidth 7B53A's And Seconded - before I acquired 7B87, later -92As via a package deal: a 7B53 in my 7854 (400 MHz nom) scope worked fine. At fastest sweeps, with a known ~700 pS risetime pulse
By ashtonb@... · #263 ·
Tek 2230 CAL/Var Pot Broken
The pot/switch for the CAL/VAR volts on Ch1 of my 2230 is disfunctional and there is no detent. Manual says 311-2218-00 Tek part number, but it describes it as an Allen bradley 10K pot with
By vincent coppola · #262 ·
Re: 7704A Repair and modules.
There's not a darned thing wrong with using a 7B53A in a 7704 or 7704A. It's the 7B92/A that you can't easily use on lower bandwidth scopes only because they can't handle the faster sweeps. And
By dhuster@... · #261 ·
Re: 7704A Repair and modules.
The 7B53/7B53A is not recommended for use in the 7704/7704A. I don't remember why, but I think it was a good reason. The original poster needs to find a 7B70+7B71, 7B80+7B85, or 7B92/7B92A.
By dd@... · #260 ·
Re: 7704A Repair and modules.
The Often the Tantals for Supply Decoupling having a short. The Chokes which are lo
By Jorgen Lund-Nielsen · #259 ·
7704A Repair and modules.
Good Evening I have a 7704A mainframe that is in need of repair. I turned it on this morning after not having been powered up for about 2 years. The trace was moving around erratically. Then I
By Bob H. <fisher_c_body@...> · #258 ·
Re: Tektronix 466 Storage Scope Knobs Needed
Bill, if you don't mind knobs that aren't perfect, I have enough used ones to set up. They may be cracked, have a soldering iron ding, have some paint rubbed off, but they're serviceable. When I
By dhuster@... · #257 ·
7844 mainframe part needed
Hello, Would anyone have a vertical output hybrid for a 7844? Tek part 155-0064 I believe... If anyone has a 7844, I can email them the necessary pictorial info where the part is and how to remove it
By ilimzn · #256 ·
Re: 3C66 carrier amplifier
This allows as Interesting... thanks! So it's the 2/3 series equivalent of the type Q strain gauge then, I assume. George
By George <vee420@...> · #255 ·
Tektronix 466 Storage Scope Knobs Needed
Hello, if anyone has access to or knows a source for the following 466 parts, I would be grateful: - (2) A&B trigger level knobs #366-1280-00 - (1) B trigger slope knob #366-1278-00 - (1) vertical
By Bill Wollam · #254 ·
Re: Looking for 465B parts - air filters?
What's REALLY bad about leaving the filter out of the 465/475 scopes is when they're used "IBM" style -- rear on the floor while working on equipment. That fan "vacuums" up every dust moozie in
By dhuster@... · #253 ·
Re: Looking for 465B parts - air filters?
the is (and Hello Dean, I remember the Filter-Kote (milky water-soluble oil?) days for the metal filters in the 5xx scopes (!) I've been cutting up the
By ashtonb@... · #251 ·
Re: 3C66 carrier amplifier
Hi George, The 3C66 interfaces to a wide variety of mechanical transducers. This allows display of all kinds of mechanical measurements on the CRT face such as vibration, mechanical shock, shaft
By Stan or Patricia Griffiths <w7ni@...> · #252 ·
3C66 carrier amplifier
What does a 3C66 carrier amplifier do, or what do you use it for? Thanks! George
By George <vee420@...> · #250 ·
Re: Info Needed on 1M1 Test Unit
manual That can be some handy information. But be forewarned about U.S. Army calibration T.O.s (Technical Orders) which the USAF also uses. I have seen some that leave a lot to be desired. For
By dhuster@... · #249 ·
Re: Info Needed on 1M1 Test Unit
I'm not sure how helpful this is, but the government calibration manual for the TU-7 / 1M1 / 067-0521-00 / 067-0521-01 is available online: https://www.logsa.army.mil/etms/data/A/009074.pdf
By dd@... · #248 ·
Re: Info Needed on 1M1 Test Unit
Hi George, I think you can get by just fine with the 067-0521-00 manual for maintaining your 067-0521-01. Those units require almost zero maintenance anyway. You will probably never have occasion
By Stan or Patricia Griffiths <w7ni@...> · #247 ·
Re: Info Needed on 1M1 Test Unit
545A. later. I have a 067-0521-01, how much different is this and would a 067-0521-00 manual be sufficient for maintenance and usage of this unit? Thanks!
By George <vee420@...> · #246 ·
Re: Info Needed on 1M1 Test Unit
Hi Marc, Yes, the 1M1 is a very good tool to use in adjusting and checking a 545A. the 1M1 was also known as the TU-7 earlier and then the 067-0521-00 later. It is a newer instrument than the 545A
By Stan or Patricia Griffiths <w7ni@...> · #245 ·
Re: Info Needed on 1M1 Test Unit
Sure. This plugin is also known as the 067-0521-00. Manuals are available from: Bob Howard howardr@... http://www.magma.ca/~howardr/manuals.htm Ron Ford W.J. Ford Surplus