Re: Identify TDS640A 'hackable' options via serial number?
Hi Dave, You can unlock the options via GPIB. They are locked in EEPROM. See https://www.eevblog.com/forum/testgear/enabling-options-in-tektronix-tds500-tds600-and-tds700-series-scopes/ for example.
By walkerru3 · #208743 ·
Re: Identify TDS640A 'hackable' options via serial number?
You might try asking in the Test Equipment section of the eevblog forums <https://www.eevblog.com/forum/index.php>
By David C. Partridge · #208742 ·
Re: Salvage yard rescue: Tektronix TU-50. Some questions from a non-tech
Go to the photos section </g/TekScopes/photos> and create a folder called (e.g.) TU-50, and upload your pix there. David
By David C. Partridge · #208741 ·
Salvage yard rescue: Tektronix TU-50. Some questions from a non-tech
Hey everyone, I was pointed in the direction of this group from the Tektronix subreddit. Like the title says I don't have a formal background in EE or repairs, just some modest restorations of tape
By ricky covell · #208740 ·
Identify TDS640A 'hackable' options via serial number?
Hi Teksperts, Is it possible to identify which options are locked within a Tektronix TDS640A from the serial number? I understand that many options are present but 'locked' from access on lower spec
By Dave B · #208739 ·
Re: 5B44 Dual Timebase parts - knobs wanted!
Thank you all that responded, I have now contacted VintageTek Museum to inquire. QService where not listing these particular parts Thanks and kind regards, Daniel
By Daniel Wainwright · #208738 ·
Re: P6610 physical measurement amplifier
The Novasensor product line is still around. Jim Williams's AN61 (https://www.analog.com/media/en/technical-documentation/application-notes/an61fa.pdf) includes several application circuits that might
By mcbaneg@... · #208737 ·
Tek Manuals
While looking for some manuals I found the bitsavers.computerhistory.org site. It has some very high quality manuals including TEK ones. Regards, Brian
By Brian Symons · #208736 ·
Re: P6610 physical measurement amplifier
Thanks. The ebay listing is for the pressure transducer part of the probe set I have. Definitely labeled as Tektronix gear. The probe assembly has a Lemo power cable and a P6137 style BNC connector
By bill K7WXW · #208735 ·
Re: 2430A no display
Hi David, and thanks for the suggestion. That could well be the case. It seems that I need to figure out which part of the PSU board supplies the horizontal control circuit? Whatever is
By James55 · #208734 ·
Re: TDS744A flicking
Thank you, Jared, for the gesture. I really appreciate it. 73!
By rafael suarez · #208733 ·
Re: 2430A no display
A glow off to the right might suggest that the left horizontal plate signal was disconnected ...
By David C. Partridge · #208732 ·
Re: 2430A no display
Further testing.. Swapping the known good PSU and top boards into the original 2430A reveals that the display problem is indeed created by the original PSU. The original scope starts up and has
By James55 · #208731 ·
Re: 2430A no display
Shall keep this short, however... swapping the PSU from the 'original' (no display) scope into the working 2430A, brings back the 'BEEP' and the glow off of the right hand side of the CRT (no
By James55 · #208730 ·
Re: TDS3034 upgrade
Ok - thank you. I'll try it as soon as I got my 3034C
By C.W. · #208729 ·
Re: TDS744A flicking
These schematics are for the TDS540, but the power pins will be the same: https://w140.com/tekwiki/images/c/c9/TDS540_Schematics.pdf Jared
By Jared Cabot · #208728 ·
Re: Tek 2236 Control Button Error
Thanks for the reply... After having spent some time going through the CPU/CTM schematics I am inclined to agree with you, it's more probably a mechanical connection problem somewhere. The un-read
By George Werl · #208727 ·
Re: P6610 physical measurement amplifier
See https://www.ebay.com/itm/234280743238 It has an alternative part number (Not my auction just a Google result) Sent via mobile annoyance thingy, please pardon any typos. [email protected]>
By David Holland · #208726 ·
Re: P6610 physical measurement amplifier
If your probes don't seem to be Tek or related to engine analyzers, then also look at HP's foray into laser interferometer systems, which had some transducers for barometric pressure, temperature, and
By ed breya · #208725 ·
Re: P6610 physical measurement amplifier
Here's the rest. https://w140.com/tekwiki/wiki/Engine_Analyzer_System Ed