Re: scanned: 545B and 561 Instrument Reference Books
A big THANX for all the effort !
By
tek_547
·
#150415
·
|
Re: Scope
On Sat, Aug 11, 2018 at 8:27 PM Thomas Dodge <tdodge2404@...> wrote: > Hi, > My name is Tom, and I haven't sent any emails in a while. I just bought a > Tektronix TDS 648A scope. I plugged it in
By
Thomas Dodge
·
#150414
·
|
Re: No trace on 556 :(
Only to be careful with the high voltage and the complex reassembly process. My replacement transformers were layer wound but the originals are pie (universal) wound. Chuck Harris on this group has
By
Morris Odell
·
#150413
·
|
Re: SG 504 leveled sine wave generator.
Took the chance to recal. the unit using better precision gear than specified in the manual. Improved the 6Mhz reference accuracy to -0.06dBm vs -0.12dBm with the R270 & R280 amplitude adjustments.
By
Ancel
·
#150412
·
|
Re: 2465B Replace Aluminum Electrolytics with Tantalum
And me 108 years.? That is if my electrolyics don't leak out.
By
machineguy59
·
#150411
·
|
SG 504 leveled sine wave generator.
Hi y'all: Was doing some calibrations on some precision, stub tuned (+/- 0.10dBm @ 4Vpp, full range) leveling heads over the past couple of weeks for the EBAY sale listing that I put up occasionally
By
Ancel
·
#150410
·
|
Re: 2465B Replace Aluminum Electrolytics with Tantalum
Lets see that would make me 107 Years Young. Sounds like you did a good job.?Leon Robinson ?? K5JLR Political Correctness is a Political Disease. To: [email protected] Sent: Saturday, August 11,
By
Leon Robinson
·
#150409
·
|
2465B Replace Aluminum Electrolytics with Tantalum
See this photo album /g/TekScopes/album?id=66462 for my recent repairs on a 2465B. This 2465B had modest corrosion on the A5 board from leaking aluminum electrolytic capacitors. On
By
machineguy59
·
#150408
·
|
Re: 2215A LVPS repair
On Sat, Aug 11, 2018 at 05:29 PM, tom jobe wrote: > > Hi Leo, > Yesterday I found the Fluke thermocouple and did some tests on a 2215A. > First I disconnected the optional fan, then let the scope run
By
satbeginner
·
#150407
·
|
Re: 2215A LVPS repair
Hi Leo, Yesterday I found the Fluke thermocouple and did some tests on a 2215A. First I disconnected the optional fan, then let the scope run with two traces like you did. The temperature of the
By
tom jobe <tomjobe@...>
·
#150406
·
|
Re: No trace on 556 :(
Thanks for the info morrios , do you have anything points that you think will help in this process ??? Things to pay attention to.? On Sat, Aug 11, 2018 at 5:22 AM, Morris
By
oliver johnson
·
#150405
·
|
Re: scanned: 545B and 561 Instrument Reference Books
I don't currently have Instrument Reference Books for any instruments later than the early 1960s. If I find them, I will scan them and share the scans.
By
Kurt Rosenfeld
·
#150404
·
|
Re: 464
Are all the low voltages correct?
By
Brendan
·
#150403
·
|
Re: No trace on 556 :(
I have uploaded scans of the hv transformer drawings to a folder in the Photos area called (not surprisingly) "556 high voltage transformers". Good luck! Morrios
By
Morris Odell
·
#150402
·
|
Re: No trace on 556 :(
I have the Tek specification drawings for both the 556 HV transformers - I'm not sure if they're in the group files but if not I'll upload them. When I mentioned a jig, what i was referring to was a
By
Morris Odell
·
#150401
·
|
464
hey all,i have a 464 with no crt display.front pannel lights work. i have the service man..wow what a book ! Any comon problems causing crt not to light?
By
kickingypsy9@...
·
#150400
·
|
Re: scanned: 545B and 561 Instrument Reference Books
I see you are very busy with the Reference Books Kurt so thanx for that. Are these books mainly about the "older" stuff from about the 1950 - 1960 era or made Tek them also for say the 547, 647a and
By
tek_547
·
#150399
·
|
Re: Tek 549 storage board anomaly
Could somebody forward this to the OP e-mail address, Tim Phillips (UK), please? Can't see it in the archives, as it's obfuscated by groups.io. Hopefully, at least the group moderators can see it, and
By
Sebastian Garcia
·
#150398
·
|
Re: 7704A Z-Axis Problem
(continued) typo, C4399 is at the +15V rail.
By
Albert Otten
·
#150397
·
|
Re: 7704A Z-Axis Problem
Hi Tomas, Since I was just curious what my 7704A would show I took off the upper side panels, attached probes and switched on the 7704A. A big very bright dot illuminated the center of the screen. At