Re: HP 8920B Spares

 

Thank you for the replies on this subject of spares.
I suppose that what I would figure to be a part that does fail, the multi-function knob, is not high on the list of failure, which is surprising given its extensive use in navigating.
Side note; Terry Gillard, its nice to see that besides the Kenwood NX-5000 series we also have the HP 892x series in common. :-)
Thanks again,
Alex


Re: An RF relay becoming unstuck could be the explanation

 

Good day, and welcome to this group.

Over the years of working with relays of all kinds, including some very small signal relay like those used in RF test equipment, I've run into several occasions where a relay would develop a "perm."? "Perm" is a term that when associated with relays refers to some residual, permanent magnetism in either the pole piece or the armature.? It doesn't take much in the way of direct contact of the armature to the pole piece to cause a relay to not release when a perm develops.

In the past, and if it was possible to remove the relay, I would pass it slowly through an AC field, such as the type that was used to bulk erase magnetic recording tape.? This would usually cure the problem... for a while.

High quality relays will often have a piece of non-ferrous material in the armature where it contacts the pole piece.? This is to prevent the ferrous portion of the armature from directly contacting the pole piece.? When I first learned about this issue, I was able to put a piece of very thin piece of tape, such as the type of tape used in the olden days of splicing tape used in magnetic tape recording editing, across the pole to prevent the direct contact.? Obviously this can't be done in sealed relays.

The purpose of this epistle is to explain a common problem of why a relay might hang up.

Burt, K6OQK


Re: HP 8920B Spares

 

Hi Alex,

I have heard that the AC / DC Switch on the PSU can be swapped if you only use the set on AC. This forum has excellent Manuals, which you should download and memorize .. or perhaps Print out.

Be very careful when disconnecting the Power Switch from the Rod from the front, I use long nose pliers to hold the switch mechanism and jack the rod from the switch with a 1/4" blade screwdriver. DO NOT just pull on the rod.

I am still working slowly thru My 8921A, I have found Rick at Amtronix is very helpful. He often gives advice on this forum.

Regards

Bill M.??? VK6IC


Re: HP 8920B Spares

 

Alex,
?
I've heard that the power button mechanism is troublesome for the 8920 & 8924c machines.
The button is up front of course - but the actual device is at the rear of the unit. and has a long bar that reaches to it.
I've seen this on my 8924c
but have not confirmed it's the same on the 8920A. apparently the replacement power switch is either non existent or?
very rare to locate.
?
Respectfully,
?
Terry Gillard - NX7R


Re: HP 8920B Spares

 

I've worked on quite a few E8285A, 8935 (E6380 & E6381) & 8924E/C over the years. PSU / Regulator board CAPS are the most prolific failures. CRT/Plasma displays next. Several have had failed measurement units (A19 in 8920). In the same one, Sig Gen Synth A14 had a tantalum cap pop on one of the DC rails. So I replaced all tants in its A19 but it still failed to calibrate. Swapped it out in the end. I've got around 4 dud A19s in the parts box. Another common error is out of lock OO LEDs illuminating which normally is a sure sign something is up with the 40V PSU rail. they are a great analyzer for general analogue radio. The only other I'd ever get would be IFR1900. From same era, does the same things but goes about it quite differently..


HP 8920B Spares

 

Hello to the group,
I have always maintained spare parts for any test equipment that I own, i.e. IFR 1200 SS, Fluke, and the like.
Since I am new to the HP 8920B and have already done the LCD conversion, rebuilt the power supply, backed up the Cal Data, and renewed the memory battery are there any spares to keep an eye out for; such as the multi function dial, cards, or any parts (odd or rare IC's/transistors) that tend to fail with age?
I kept my spare CRT but if anyone needs one I am happy to part ways with it.
Thank you,
Alex, KD6VPH


Re: An RF relay becoming unstuck could be the explanation

 
Edited

Welcome.

If the 8920B has the full frequency option (0.4-1000MHz) and thus relay RF module that is possible. Another possibility is a poor coaxial connection.

Robert.


An RF relay becoming unstuck could be the explanation

 

Good morning friends,
?
?Forgive me for the lack of experience to write here in the group, I have an HP8920B and it stayed for 6 months measuring 4db above in spectrum analyzer mode.?
?
Suddenly one day i turned on the 8920B and the measurements are back to normal. I believe that some RF relay had the contacts stuck and after 6 months the contacts returned to work normally.?
?
It's been almost 2 years and the HP9020B is still working perfectly.
?
Forgive me for the bad english.


Re: 10MHz reference input - can a square wave with 10MHz frequency be used?

 

Your distribution amp will probably be happier with a sine-wave signal.


Re: 10MHz reference input - can a square wave with 10MHz frequency be used?

 
Edited

I realized that it is not the FS730 I have but rather the smaller one FS710:


Re: Amtronix is the best. I'm amazed at their service.

 

Thanks Rick. ?

?

My 8920A works perfectly fine and runs through all the diags for the various modules without errors and all measurements I have checked are within specs.

It is only the SA module that shows approx 3 dB to low (if I remember it correctly) which is sliiiightly lower than the specs +/- 2.5dB.


Re: Bessel Function on HP8920A...

 

Global user key..


File /1Separate miscellaneous files/HP AN200 Fundamentals of the Electronic Counters.pdf uploaded #file-notice

Group Notification
 

The following files and folders have been uploaded to the Files area of the [email protected] group.

By: Hank Riley <n1ltv@...>

Description:
The basic principles and some advanced techniques for getting the best results from a frequency counter. 44 pages


Re: 10MHz reference input - can a square wave with 10MHz frequency be used?

 

For the reference input it does not state the need for a sine wave because there is no need.? Just like single purpose frequency counters do not require a sinewave.

The reference out is a sine wave because that is the industry practice for high grade sources.? A sine wave is ready to distribute in a lab to multiple destinations without causing all kinds of RF pollution that could be generated by a non-sinusoidal signal.

So unless your Bodnar signal will be distributed it will be fine as is.


On Wednesday, August 2, 2023 at 03:34:28 AM EDT, CrazyFin wrote:

[several snips made for brevity]

I have read through the user manual and on page 531 (in the version I have of the user manual) I find the paragraph "10 MHz REF INPUT" which does not state sine wave only.

However, under the paragraph?10 MHz REF OUTPUT it clearly states
"Waveform = sinewave"

So the question is if I can use a 10MHz square wave on the 10MHz ref input on the 8920A?


Re: Amtronix is the best. I'm amazed at their service.

 

¿ªÔÆÌåÓý

Wow,? $1400 for a calibration is outlandish. If you run all your diagnostics, chances are your 8920 is working okay but may be off frequency, etc.. If your doing commercial work requiring a cal, you may need it. Just to let all know, the calibration on the 8920 units involve running the internal diagnostics then the "percal" program to adjust the 8920 internal frequency standard, voltage standard and audio levels and a few other minor things. Then the power meter and DC offset are zeroed from the front panel. After that, we run the performance evaluation. This checks most all of the AM, FM, rf in/out levels, audio, power meter, receive and generate and spectrum levels, spurious levels, etc. This is an automated test that produces some 20 pages of performance data. The performance evaluation insures that the 8920 is accurate and tests much more than the internal diagnostic program. Sometimes, the performance evaluation reveals faulty modules or hardware failures that the diagnostics don't show. I hope this helps....Rick

On 8/2/2023 3:46 AM, CrazyFin wrote:

?

Fully agree even though I have not used Ricks services for actual repairs or mods but only for advice as well as Rick's excellent posts here in the group.

I just wish we would have a similar workshop here in Sweden (or within Europe) that has the same experience and knowledge that Rick and his Amtronix has.
I am located in Sweden so shipping my 8920A back and forth to USA is just to expensive.

I would like to get my spectrum analyzer module calibrated and Agilent in Sweden would charge me 1400 USD to calibrate my 8920A (which would be a full calibration and not just the SA module).
It would also be nice to upgrade from 60W input module to the high power option 100W.


Re: 10MHz reference input - can a square wave with 10MHz frequency be used?

 

?

Great! Thx for all input.??

Will build my own 10MHz low pass filter as per the article Robert linked to or just take the easy way out and purchase the one on eBay.
I'll dig through my storage boxes first though, I might actually have a 10MHz low pass filter from Mini Circuits that would work in that case.



Re: 10MHz reference input - can a square wave with 10MHz frequency be used?

 

I have used this one from ebay - about 3 dB insertion loss, about $10.50 - eBay item number:385415398319


Re: 10MHz reference input - can a square wave with 10MHz frequency be used?

 

Yes.
I'd put it through a low pass filter though. The 8920 input is sensitive enough. Here is one example:


looks like same design with stripline and SMD components.


Re: Amtronix is the best. I'm amazed at their service.

 

?

Fully agree even though I have not used Ricks services for actual repairs or mods but only for advice as well as Rick's excellent posts here in the group.

I just wish we would have a similar workshop here in Sweden (or within Europe) that has the same experience and knowledge that Rick and his Amtronix has.
I am located in Sweden so shipping my 8920A back and forth to USA is just to expensive.

I would like to get my spectrum analyzer module calibrated and Agilent in Sweden would charge me 1400 USD to calibrate my 8920A (which would be a full calibration and not just the SA module).
It would also be nice to upgrade from 60W input module to the high power option 100W.


10MHz reference input - can a square wave with 10MHz frequency be used?

 

I have a Spectracom GPS clock 8183A as my 10MHz frequence reference in my lab to my various instruments.

I am interested in testing the Leo Bodnar Electronics?Precision GPS Reference Clock:


However it seems that it only produces a square wave with 10MHz so my question is if the reference input on the HP 8920A accepts a square wave or does it only work with a 10MHz sine wave?

I have read through the user manual and on page 531 (in the version I have of the user manual) I find the paragraph "10 MHz REF INPUT" which does not state sinus wave only.
It just says:

=========

10 MHz REF INPUT??

Operating Considerations
Input frequency = 1, 2, 5, or 10 MHz
Input level >0.15 V rms
Input impedance approximately 50?.
When a valid signal is applied to the 10 MHz REF INPUT, the Test Set
automatically switches from internal to external reference.?
=========

However, under the paragraph?10 MHz REF OUTPUT it clearly states
"Waveform = sinewave"

So the question is if I can use a 10MHz square wave on the 10MHz ref input on the 8920A?