Do all the 140-series scopes have extremely noisy fans? My 140B is excessively noisy and I'm wondering if a different fan could be substituted. I don't mind the wind noise all that much, the whine that's associated with it is pretty obnoxious. I haven't lubricated it yet but I suspect that won't really knock down that whine very much (if at all). Looking for suggestions.
Thanks, Barry - N4BUQ
|
On my 141T, I replaced the Rotron "Muffin" with a "Whisper".
Dave Wise
toggle quoted message
Show quoted text
Do all the 140-series scopes have extremely noisy fans?? My 140B is excessively noisy and I'm wondering if a different fan could be substituted.? I don't mind the wind noise all that much, the whine that's associated with it is pretty
obnoxious.? I haven't lubricated it yet but I suspect that won't really knock down that whine very much (if at all).? Looking for suggestions.
Thanks,
Barry - N4BUQ
|
You must have an Opt 101X (aka "strangling hamster"). If lubrication doesn't take care of it, swap out the fan.
-- Cheers, Tom
-- Prof. Thomas H. Lee Allen Ctr., Rm. 205 420 Via Palou Mall Stanford University Stanford, CA 94305-4070
toggle quoted message
Show quoted text
On 4/2/2024 12:42 PM, n4buq wrote: Do all the 140-series scopes have extremely noisy fans? My 140B is excessively noisy and I'm wondering if a different fan could be substituted. I don't mind the wind noise all that much, the whine that's associated with it is pretty obnoxious. I haven't lubricated it yet but I suspect that won't really knock down that whine very much (if at all). Looking for suggestions.
Thanks, Barry - N4BUQ
|
My recently acquired HP-140B (yesterday!) has a fairly noisy fan. If it becomes bothersome, I¡¯ll replace it, as I did the fans in my 5245 and 5248 counters.?
I can¡¯t find a temperature spec for the 140B in either the manual or my HP catalogs, but typically the high temperature limit is 50C. If you have a ¡°laboratory¡± environment with reasonable temperatures, the ¡®scope can get by with a quieter fan that blows less air.?
toggle quoted message
Show quoted text
On Tue, Apr 2, 2024 at 12:51?PM Tom Lee < tomlee@...> wrote: You must have an Opt 101X (aka "strangling hamster"). If lubrication
doesn't take care of it, swap out the fan.
-- Cheers,
Tom
--
Prof. Thomas H. Lee
Allen Ctr., Rm. 205
Mall
Stanford University
Stanford, CA 94305-4070
On 4/2/2024 12:42 PM, n4buq wrote:
> Do all the 140-series scopes have extremely noisy fans?? My 140B is excessively noisy and I'm wondering if a different fan could be substituted.? I don't mind the wind noise all that much, the whine that's associated with it is pretty obnoxious.? I haven't lubricated it yet but I suspect that won't really knock down that whine very much (if at all).? Looking for suggestions.
>
> Thanks,
> Barry - N4BUQ
>
>
>
>
>
|
I have a HP5335 which has a verry niosy fan what should I replace it with?
|
I don't know what kind of fan is in mine.? It appears to have had a large round sticker attached to the back of it but that's gone.? All I know is it's a line-voltage fan.
Barry - N4BUQ
toggle quoted message
Show quoted text
From: "David Wise" <d44617665@...> To: "HP-Agilent-Keysight-equipment" <[email protected]> Sent: Tuesday, April 2, 2024 2:46:42 PM Subject: Re: [HP-Agilent-Keysight-equipment] HP 140 Series Fan Noise
On my 141T, I replaced the Rotron "Muffin" with a "Whisper".
Dave Wise
Do all the 140-series scopes have extremely noisy fans?? My 140B is excessively noisy and I'm wondering if a different fan could be substituted.? I don't mind the wind noise all that much, the whine that's associated with it is pretty
obnoxious.? I haven't lubricated it yet but I suspect that won't really knock down that whine very much (if at all).? Looking for suggestions.
Thanks,
Barry - N4BUQ
|
I don't know what that option is. The manual states the 140B doesn't have options and anything aside from the standard configuration is installed at the factory. I'll see if I can get some oil in at the proper lube point and see if that helps.
I do notice that if I restrict the air-flow with my hand, it gets a bit quieter. That's a bit impractical, though, as a workaround.
Thanks, Barry - N4BUQ
toggle quoted message
Show quoted text
You must have an Opt 101X (aka "strangling hamster"). If lubrication doesn't take care of it, swap out the fan.
-- Cheers, Tom
-- Prof. Thomas H. Lee Allen Ctr., Rm. 205 420 Via Palou Mall Stanford University Stanford, CA 94305-4070
On 4/2/2024 12:42 PM, n4buq wrote:
Do all the 140-series scopes have extremely noisy fans? My 140B is excessively noisy and I'm wondering if a different fan could be substituted. I don't mind the wind noise all that much, the whine that's associated with it is pretty obnoxious. I haven't lubricated it yet but I suspect that won't really knock down that whine very much (if at all). Looking for suggestions.
Thanks, Barry - N4BUQ
|
Shop for a physically similar fan (for ease of installation) with a lower current spec. It will blow less air and, hence, make less noise. Make sure the 5335 stays in a cool environment, say 80F or less, and it will be OK with the lower airflow.?
toggle quoted message
Show quoted text
On Tue, Apr 2, 2024 at 1:55?PM joe price < joeprice@...> wrote: I have a HP5335 which has a verry niosy fan what should I replace it with?
|
There are basically fans with two types of bearings out there ¨C ball bearing & bronze bushing.? Ball bearing fans start out with more noise and will simply get noisier over time as they wear.? End-of-life ball bearing fans can drive you out of the room with their noise.
?
Bronze bushing fans stay relatively silent over time.? They will get noisy if they lose their lubrication and the wear on the bushing gets excessive.? Then there is the other side of the coin where a fan will freeze up if the lubrication dries out and gets hard or gummy.
?
Greg
|
The whine is the ball bearings rolling on a not so smooth race. The problem could be the metal of the race has been brinelled by the balls smacking on it... kind of like little ball peen hammers. Or it could be chunks of dried grease making noise as they get into stuff. Most people that try to lubricate these fans don't realize that there are two ball bearings in the motor, one you can easily see if you remove the label, and the other spring loaded to the other side of the fan. You have to lube them both. The motors come apart pretty easy. Usually, there is a "C" ring, or a plastic washer like piece that is clipped over a groove in the bearing shaft. Remove it, and the fan, and both bearings will pop right out. Then remove one seal from each bearing, and wash them out with naptha, or varsol, or gasoline... It only takes a little, and blow them try. Don't let them spin very fast, as it will Brinel the bearing races. Put a little dab of lubriplate in the bearing, and you are good to go. Don't pack them full, that only turns the bearing into a grease pump, and makes everything turn slow, and get hot. Assembly is the reverse of disassembly. -Chuck Harris On Tue, 2 Apr 2024 15:42:51 -0400 (EDT) "n4buq" <n4buq@...> wrote: Do all the 140-series scopes have extremely noisy fans? My 140B is excessively noisy and I'm wondering if a different fan could be substituted. I don't mind the wind noise all that much, the whine that's associated with it is pretty obnoxious. I haven't lubricated it yet but I suspect that won't really knock down that whine very much (if at all). Looking for suggestions.
Thanks, Barry - N4BUQ
|
On Tue, Apr 2, 2024 at 08:43 PM, Jeremy Nichols wrote:
My recently acquired HP-140B (yesterday!) has a fairly noisy fan. If it becomes bothersome, I¡¯ll replace it, as I did the fans in my 5245 and 5248 counters.?
?
I can¡¯t find a temperature spec for the 140B in either the manual or my HP catalogs, but typically the high temperature limit is 50C. If you have a ¡°laboratory¡± environment with reasonable temperatures, the ¡®scope can get by with a quieter fan that blows less air.?
?
On Tue, Apr 2, 2024 at 12:51?PM Tom Lee < tomlee@...> wrote:
You must have an Opt 101X (aka "strangling hamster"). If lubrication doesn't take care of it, swap out the fan.
-- Cheers, Tom
-- Prof. Thomas H. Lee Allen Ctr., Rm. 205 Mall Stanford University Stanford, CA 94305-4070
On 4/2/2024 12:42 PM, n4buq wrote: > Do all the 140-series scopes have extremely noisy fans?? My 140B is excessively noisy and I'm wondering if a different fan could be substituted.? I don't mind the wind noise all that much, the whine that's associated with it is pretty obnoxious.? I haven't lubricated it yet but I suspect that won't really knock down that whine very much (if at all).? Looking for suggestions. > > Thanks, > Barry - N4BUQ > > > > >
Don't forgot that most of the plug-ins have a mix of glassware generating lots of heat and Ge transistors that don't like heat, this will depend on the plug-ins installed. Most of my 140x & 141x mainframes have a 115Vac ball bearing "Boxer" branded fan from IMC, I can't quite read the part number from the only picture I took with the filter removed. David
|
Hi, I also had a horrible noise from the fan in my hp5335A counter. What I did was to place 15 ohm, ? W resistors and 10 of them on a circuit board, in series with the fan, a total of 150 ohm. I placed the board on the inside of the counter on the right gable. This reduced the fan speed and also the noise substantially. I don¡¯t recall the temperature rise in steady state condition but it was uncritical.
G?ran
|
Hi, ???I also had a horrible noise from the fan in my hp5335A counter. What I did was to place 15 ohm, ? W resistors and 10 of them on a circuit board, in series with the fan, a total of 150 ohm. I placed the board on the inside of the counter on the right gable. This reduced the fan speed and also the noise substantially. I don¡¯t recall the temperature rise in steady state condition but it was uncritical.
G?ran
|
Hi.
There are now three bearing types available:- Ball bearing. Sleeve bearing (often sintered bronze* sometimes a plastic material.) And Magnetic.
(* With care, these can last a very long time.? Some are possible to extract and "refill" with lubrication, in a similar way to how we used to re-lubricate motorcycle chain's!)
The last type (magnetic) are (mechanically) virtually silent, but stray DC magnetic fields may be an issue in some cases.? Often found on smaller higher speed fans, at a price.? They can still fail, if dirt gets into the bearing area though, or they overheat.
The other source of "whine" or "turbine like" noise, is caused by something being too close to the rotating blades, often on the "air outlet" side.? That can be the motor strut's themselves (later types are profiled to reduce such noise) or a nearby cable, coax, or even other metalwork.
(Large items too close to the inlet side can also cause noise level to increase, as well as restricting airflow, depending on how the obstruction allows air to pass into the fan itself...)
Sometimes, changing the fan for one with very similar airflow characteristics, but with the inlet over the struts rather than the outlet (if that is the case) can help with the noise level.
Take care not to reverse the airflow through the instrument.? Many "wont like it" over the long term!? Most fans are marked showing the rotation and airflow direction, but it is not difficult to tell just by looking at the blade shape.? (The concave side is "usually" facing the outlet.)
Instruments in a rack, can have conflicting airflow requirements too.? It's not that uncommon to find issues caused by two or more exchanging each others warm air within a rack, rather than working together to pull ambient air "through" the rack.? In such cases, the rack is likely to need it's own fans too!
It is sadly too, not uncommon for a rack to be "over populated", so that instruments that pull air in from the underside, have little to no space for that as the unit below it is too close.? There is a reason for those 1 or 2'U' blank panels often seen, not just to make the front look neat.? And again, the type of support for the instrument.? Shelves and "L" shaped supports front to back, can sometimes obstruct airflow to/from the sides as well.
Over the years since many older instruments and equipments were developed, there have been vast improvements in fan technology, in particular blade and other airflow profile design aspects.? Some to benefit the ability to push or pull air through a restrictive passage, others to greatly reduce fan noise, without compromising (too much) on the airflow characteristics.? Most will use less electrical power for the same "duty" these days than the older designs, as motor tech' too has advanced.
The last thing you want, is for a fan to run, with the blades aerodynamically "stalled", often a hint is a "fluttering" sound to the airflow, and/or air exiting a fan being spat out radially, more than axially.? The resulting overall airflow is usually very low in those cases, regardless to if the fan is pushing air in, or pulling it out of a casing.
Old catalogues to identify the original fan type and performance (assuming you have the "originally fitted at manufacture" fan, and also new catalogues (web pages) so as to choose a modern unit with the needed airflow, and possibly reduced noise is what you need.? It does take more time than you might expect though, as often the choices are either vast, or limited.? :-(
Keeping air filters clean too, can help with noise levels, as well as keeping the instrument internals cooled correctly.
Hope something helps.? I hate to see otherwise great instruments compromised (or even damaged as a result of overheating) just because someone didn't like the noise, so put a "quieter" fan in.
HP (and other manufacturers) designers will have done the math, and specified the part used originally for a good reason.? Not just to prevent overheating, but also to produce a sensible warm up time, that results in relatively stable operating temperatures, for a given range of ambient temperatures.?? Compare the air outlet(s) temperatures once they have stabilised, for the same ambient and operating conditions, before and after replacing the fan, to be sure.? Much higher with the new fan, and it's probably not doing the job needed of it.? "How Much" is a moot point though.
Cheers All.
Dave 'KBV
(Who's populated and retrofitted customer racks with various "high dissipation things" in the past than I care to think about.? Not unusual to end up either fitting large fans to the top to pull air out, install ducting if a unit has the outlet hardware for it, or just remove back and sides, to allow free airflow, when in use.)
Noise?? You haven't lived until you work near a host of things with 400Hz fans in!? Some units we worked with had no less than three of those noisy things!? They do move heck of a lot of air for the size though...
toggle quoted message
Show quoted text
-- Sent from a Micro$oft free zone:
|
Thank you so much for this. Lots of very good information.
Interesting that you mention airflow direction. That's one thing that has puzzled me about this scope. There has been a lot of discussion lately on other lists regarding whether fans should should push air into the instrument or pull it out and the consensus is that pulling it out is best but now I find that this scope pushes air in. I presume part of that is to maximize the air flow directly across the pass transistors but that's just a guess. I had considered reversing the airflow but now I don't think I will.
Thanks again, Barry - N4BUQ
toggle quoted message
Show quoted text
Hi.
There are now three bearing types available:- Ball bearing. Sleeve bearing (often sintered bronze* sometimes a plastic material.) And Magnetic.
(* With care, these can last a very long time.? Some are possible to extract and "refill" with lubrication, in a similar way to how we used to re-lubricate motorcycle chain's!)
The last type (magnetic) are (mechanically) virtually silent, but stray DC magnetic fields may be an issue in some cases.? Often found on smaller higher speed fans, at a price.? They can still fail, if dirt gets into the bearing area though, or they overheat.
The other source of "whine" or "turbine like" noise, is caused by something being too close to the rotating blades, often on the "air outlet" side.? That can be the motor strut's themselves (later types are profiled to reduce such noise) or a nearby cable, coax, or even other metalwork.
(Large items too close to the inlet side can also cause noise level to increase, as well as restricting airflow, depending on how the obstruction allows air to pass into the fan itself...)
Sometimes, changing the fan for one with very similar airflow characteristics, but with the inlet over the struts rather than the outlet (if that is the case) can help with the noise level.
Take care not to reverse the airflow through the instrument.? Many "wont like it" over the long term!? Most fans are marked showing the rotation and airflow direction, but it is not difficult to tell just by looking at the blade shape.? (The concave side is "usually" facing the outlet.)
Instruments in a rack, can have conflicting airflow requirements too. It's not that uncommon to find issues caused by two or more exchanging each others warm air within a rack, rather than working together to pull ambient air "through" the rack.? In such cases, the rack is likely to need it's own fans too!
It is sadly too, not uncommon for a rack to be "over populated", so that instruments that pull air in from the underside, have little to no space for that as the unit below it is too close.? There is a reason for those 1 or 2'U' blank panels often seen, not just to make the front look neat.? And again, the type of support for the instrument.? Shelves and "L" shaped supports front to back, can sometimes obstruct airflow to/from the sides as well.
Over the years since many older instruments and equipments were developed, there have been vast improvements in fan technology, in particular blade and other airflow profile design aspects.? Some to benefit the ability to push or pull air through a restrictive passage, others to greatly reduce fan noise, without compromising (too much) on the airflow characteristics.? Most will use less electrical power for the same "duty" these days than the older designs, as motor tech' too has advanced.
The last thing you want, is for a fan to run, with the blades aerodynamically "stalled", often a hint is a "fluttering" sound to the airflow, and/or air exiting a fan being spat out radially, more than axially.? The resulting overall airflow is usually very low in those cases, regardless to if the fan is pushing air in, or pulling it out of a casing.
Old catalogues to identify the original fan type and performance (assuming you have the "originally fitted at manufacture" fan, and also new catalogues (web pages) so as to choose a modern unit with the needed airflow, and possibly reduced noise is what you need.? It does take more time than you might expect though, as often the choices are either vast, or limited.? :-(
Keeping air filters clean too, can help with noise levels, as well as keeping the instrument internals cooled correctly.
Hope something helps.? I hate to see otherwise great instruments compromised (or even damaged as a result of overheating) just because someone didn't like the noise, so put a "quieter" fan in.
HP (and other manufacturers) designers will have done the math, and specified the part used originally for a good reason.? Not just to prevent overheating, but also to produce a sensible warm up time, that results in relatively stable operating temperatures, for a given range of ambient temperatures.?? Compare the air outlet(s) temperatures once they have stabilised, for the same ambient and operating conditions, before and after replacing the fan, to be sure.? Much higher with the new fan, and it's probably not doing the job needed of it.? "How Much" is a moot point though.
Cheers All.
Dave 'KBV
(Who's populated and retrofitted customer racks with various "high dissipation things" in the past than I care to think about.? Not unusual to end up either fitting large fans to the top to pull air out, install ducting if a unit has the outlet hardware for it, or just remove back and sides, to allow free airflow, when in use.)
Noise?? You haven't lived until you work near a host of things with 400Hz fans in!? Some units we worked with had no less than three of those noisy things!? They do move heck of a lot of air for the size though...
< -- Sent from a Micro$oft free zone:
|
Thanks, Chuck. The manual gives a procedure for lubricating the motor but, so far, I cannot find the places where it states to apply the oil. Apparently there should be holes to do that but I don't see them. I think I will go ahead and remove the fan (it needs a bit of cleaning anyway) and see if I can disassemble, clean, lube, and reassemble it. Hopefully that will help/cure most of the whine.
Thanks again, Barry - N4BUQ
toggle quoted message
Show quoted text
The whine is the ball bearings rolling on a not so smooth race. The problem could be the metal of the race has been brinelled by the balls smacking on it... kind of like little ball peen hammers. Or it could be chunks of dried grease making noise as they get into stuff.
Most people that try to lubricate these fans don't realize that there are two ball bearings in the motor, one you can easily see if you remove the label, and the other spring loaded to the other side of the fan. You have to lube them both.
The motors come apart pretty easy. Usually, there is a "C" ring, or a plastic washer like piece that is clipped over a groove in the bearing shaft. Remove it, and the fan, and both bearings will pop right out.
Then remove one seal from each bearing, and wash them out with naptha, or varsol, or gasoline... It only takes a little, and blow them try. Don't let them spin very fast, as it will Brinel the bearing races.
Put a little dab of lubriplate in the bearing, and you are good to go. Don't pack them full, that only turns the bearing into a grease pump, and makes everything turn slow, and get hot.
Assembly is the reverse of disassembly.
-Chuck Harris
On Tue, 2 Apr 2024 15:42:51 -0400 (EDT) "n4buq" <n4buq@...> wrote:
Do all the 140-series scopes have extremely noisy fans? My 140B is excessively noisy and I'm wondering if a different fan could be substituted. I don't mind the wind noise all that much, the whine that's associated with it is pretty obnoxious. I haven't lubricated it yet but I suspect that won't really knock down that whine very much (if at all). Looking for suggestions.
Thanks, Barry - N4BUQ
|
It is easy to think that since HP equipment is designed for worst case environments such as overcrowded racks, high ambient temperatures and infrequent filter cleaning that it would be no problem to downsize the specifications of a replacement fan in the unit on your nice, clean workbench.? In fact, this is quite often the case - but, of course, it is the exceptions that will leave you with a very nice doorstop.? IMO the most overlooked of these exceptions is to think solely in terms of the fan's CFM and neglect the possibly critical static pressure requirements - often equipment, especially RF gear, requires a minimum static pressure from it's fan to get the air not just through the easy paths but also through the fine mesh air paths in it's heavily shielded cards.? This is not to say that a fan can't be replaced with a less noisy version but do your research first; if a piece of gear has what is effectively a small jet turbine hanging on back, there is probably a reason...
Hal
|
Pulling air out of an instrument means sucking air and dirt into the instrument, all around the edges. Not a good thing.?
toggle quoted message
Show quoted text
On Wed, Apr 3, 2024 at 7:38?AM n4buq < n4buq@...> wrote: Thank you so much for this.? Lots of very good information.
Interesting that you mention airflow direction.? That's one thing that has puzzled me about this scope.? There has been a lot of discussion lately on other lists regarding whether fans should should push air into the instrument or pull it out and the consensus is that pulling it out is best but now I find that this scope pushes air in.? I presume part of that is to maximize the air flow directly across the pass transistors but that's just a guess. I had considered reversing the airflow but now I don't think I will.
Thanks again,
Barry - N4BUQ
> Hi.
>
> There are now three bearing types available:-
> Ball bearing.
> Sleeve bearing (often sintered bronze* sometimes a plastic material.)
> And Magnetic.
>
> (* With care, these can last a very long time.? Some are possible to
> extract and "refill" with lubrication, in a similar way to how we used
> to re-lubricate motorcycle chain's!)
>
> The last type (magnetic) are (mechanically) virtually silent, but stray
> DC magnetic fields may be an issue in some cases.? Often found on
> smaller higher speed fans, at a price.? They can still fail, if dirt
> gets into the bearing area though, or they overheat.
>
> The other source of "whine" or "turbine like" noise, is caused by
> something being too close to the rotating blades, often on the "air
> outlet" side.? That can be the motor strut's themselves (later types are
> profiled to reduce such noise) or a nearby cable, coax, or even other
> metalwork.
>
> (Large items too close to the inlet side can also cause noise level to
> increase, as well as restricting airflow, depending on how the
> obstruction allows air to pass into the fan itself...)
>
> Sometimes, changing the fan for one with very similar airflow
> characteristics, but with the inlet over the struts rather than the
> outlet (if that is the case) can help with the noise level.
>
> Take care not to reverse the airflow through the instrument.? Many "wont
> like it" over the long term!? Most fans are marked showing the rotation
> and airflow direction, but it is not difficult to tell just by looking
> at the blade shape.? (The concave side is "usually" facing the outlet.)
>
> Instruments in a rack, can have conflicting airflow requirements too.
> It's not that uncommon to find issues caused by two or more exchanging
> each others warm air within a rack, rather than working together to pull
> ambient air "through" the rack.? In such cases, the rack is likely to
> need it's own fans too!
>
> It is sadly too, not uncommon for a rack to be "over populated", so that
> instruments that pull air in from the underside, have little to no space
> for that as the unit below it is too close.? There is a reason for those
> 1 or 2'U' blank panels often seen, not just to make the front look
> neat.? And again, the type of support for the instrument.? Shelves and
> "L" shaped supports front to back, can sometimes obstruct airflow
> to/from the sides as well.
>
> Over the years since many older instruments and equipments were
> developed, there have been vast improvements in fan technology, in
> particular blade and other airflow profile design aspects.? Some to
> benefit the ability to push or pull air through a restrictive passage,
> others to greatly reduce fan noise, without compromising (too much) on
> the airflow characteristics.? Most will use less electrical power for
> the same "duty" these days than the older designs, as motor tech' too
> has advanced.
>
> The last thing you want, is for a fan to run, with the blades
> aerodynamically "stalled", often a hint is a "fluttering" sound to the
> airflow, and/or air exiting a fan being spat out radially, more than
> axially.? The resulting overall airflow is usually very low in those
> cases, regardless to if the fan is pushing air in, or pulling it out of
> a casing.
>
> Old catalogues to identify the original fan type and performance
> (assuming you have the "originally fitted at manufacture" fan, and also
> new catalogues (web pages) so as to choose a modern unit with the needed
> airflow, and possibly reduced noise is what you need.? It does take more
> time than you might expect though, as often the choices are either vast,
> or limited.? :-(
>
> Keeping air filters clean too, can help with noise levels, as well as
> keeping the instrument internals cooled correctly.
>
> Hope something helps.? I hate to see otherwise great instruments
> compromised (or even damaged as a result of overheating) just because
> someone didn't like the noise, so put a "quieter" fan in.
>
> HP (and other manufacturers) designers will have done the math, and
> specified the part used originally for a good reason.? Not just to
> prevent overheating, but also to produce a sensible warm up time, that
> results in relatively stable operating temperatures, for a given range
> of ambient temperatures.?? Compare the air outlet(s) temperatures once
> they have stabilised, for the same ambient and operating conditions,
> before and after replacing the fan, to be sure.? Much higher with the
> new fan, and it's probably not doing the job needed of it.? "How Much"
> is a moot point though.
>
> Cheers All.
>
> Dave 'KBV
>
> (Who's populated and retrofitted customer racks with various "high
> dissipation things" in the past than I care to think about.? Not unusual
> to end up either fitting large fans to the top to pull air out, install
> ducting if a unit has the outlet hardware for it, or just remove back
> and sides, to allow free airflow, when in use.)
>
> Noise?? You haven't lived until you work near a host of things with
> 400Hz fans in!? Some units we worked with had no less than three of
> those noisy things!? They do move heck of a lot of air for the size
> though...
>
> ><
>
>
> --
> Sent from a Micro$oft free zone:
>
>
>
>
|
Is there supposed to be a dust filter in addition to the metal screen "filter"?? The screen is good for keeping fingers and larger animals out but for dust, not so much.? The manual states to wash and allow the filter to dry which makes me think there might have been something else behind the mesh.
Thanks, Barry - N4BUQ
toggle quoted message
Show quoted text
Pulling air out of an instrument means sucking air and dirt into the instrument, all around the edges. Not a good thing.?
On Wed, Apr 3, 2024 at 7:38?AM n4buq < n4buq@...> wrote: Thank you so much for this.? Lots of very good information.
Interesting that you mention airflow direction.? That's one thing that has puzzled me about this scope.? There has been a lot of discussion lately on other lists regarding whether fans should should push air into the instrument or pull it out and the consensus is that pulling it out is best but now I find that this scope pushes air in.? I presume part of that is to maximize the air flow directly across the pass transistors but that's just a guess. I had considered reversing the airflow but now I don't think I will.
Thanks again,
Barry - N4BUQ
> Hi.
>
> There are now three bearing types available:-
> Ball bearing.
> Sleeve bearing (often sintered bronze* sometimes a plastic material.)
> And Magnetic.
>
> (* With care, these can last a very long time.? Some are possible to
> extract and "refill" with lubrication, in a similar way to how we used
> to re-lubricate motorcycle chain's!)
>
> The last type (magnetic) are (mechanically) virtually silent, but stray
> DC magnetic fields may be an issue in some cases.? Often found on
> smaller higher speed fans, at a price.? They can still fail, if dirt
> gets into the bearing area though, or they overheat.
>
> The other source of "whine" or "turbine like" noise, is caused by
> something being too close to the rotating blades, often on the "air
> outlet" side.? That can be the motor strut's themselves (later types are
> profiled to reduce such noise) or a nearby cable, coax, or even other
> metalwork.
>
> (Large items too close to the inlet side can also cause noise level to
> increase, as well as restricting airflow, depending on how the
> obstruction allows air to pass into the fan itself...)
>
> Sometimes, changing the fan for one with very similar airflow
> characteristics, but with the inlet over the struts rather than the
> outlet (if that is the case) can help with the noise level.
>
> Take care not to reverse the airflow through the instrument.? Many "wont
> like it" over the long term!? Most fans are marked showing the rotation
> and airflow direction, but it is not difficult to tell just by looking
> at the blade shape.? (The concave side is "usually" facing the outlet.)
>
> Instruments in a rack, can have conflicting airflow requirements too.
> It's not that uncommon to find issues caused by two or more exchanging
> each others warm air within a rack, rather than working together to pull
> ambient air "through" the rack.? In such cases, the rack is likely to
> need it's own fans too!
>
> It is sadly too, not uncommon for a rack to be "over populated", so that
> instruments that pull air in from the underside, have little to no space
> for that as the unit below it is too close.? There is a reason for those
> 1 or 2'U' blank panels often seen, not just to make the front look
> neat.? And again, the type of support for the instrument.? Shelves and
> "L" shaped supports front to back, can sometimes obstruct airflow
> to/from the sides as well.
>
> Over the years since many older instruments and equipments were
> developed, there have been vast improvements in fan technology, in
> particular blade and other airflow profile design aspects.? Some to
> benefit the ability to push or pull air through a restrictive passage,
> others to greatly reduce fan noise, without compromising (too much) on
> the airflow characteristics.? Most will use less electrical power for
> the same "duty" these days than the older designs, as motor tech' too
> has advanced.
>
> The last thing you want, is for a fan to run, with the blades
> aerodynamically "stalled", often a hint is a "fluttering" sound to the
> airflow, and/or air exiting a fan being spat out radially, more than
> axially.? The resulting overall airflow is usually very low in those
> cases, regardless to if the fan is pushing air in, or pulling it out of
> a casing.
>
> Old catalogues to identify the original fan type and performance
> (assuming you have the "originally fitted at manufacture" fan, and also
> new catalogues (web pages) so as to choose a modern unit with the needed
> airflow, and possibly reduced noise is what you need.? It does take more
> time than you might expect though, as often the choices are either vast,
> or limited.? :-(
>
> Keeping air filters clean too, can help with noise levels, as well as
> keeping the instrument internals cooled correctly.
>
> Hope something helps.? I hate to see otherwise great instruments
> compromised (or even damaged as a result of overheating) just because
> someone didn't like the noise, so put a "quieter" fan in.
>
> HP (and other manufacturers) designers will have done the math, and
> specified the part used originally for a good reason.? Not just to
> prevent overheating, but also to produce a sensible warm up time, that
> results in relatively stable operating temperatures, for a given range
> of ambient temperatures.?? Compare the air outlet(s) temperatures once
> they have stabilised, for the same ambient and operating conditions,
> before and after replacing the fan, to be sure.? Much higher with the
> new fan, and it's probably not doing the job needed of it.? "How Much"
> is a moot point though.
>
> Cheers All.
>
> Dave 'KBV
>
> (Who's populated and retrofitted customer racks with various "high
> dissipation things" in the past than I care to think about.? Not unusual
> to end up either fitting large fans to the top to pull air out, install
> ducting if a unit has the outlet hardware for it, or just remove back
> and sides, to allow free airflow, when in use.)
>
> Noise?? You haven't lived until you work near a host of things with
> 400Hz fans in!? Some units we worked with had no less than three of
> those noisy things!? They do move heck of a lot of air for the size
> though...
>
> ><
>
>
> --
> Sent from a Micro$oft free zone:
>
>
>
>
|
It used to be that for the older tek equipment (500 series scopes), you washed the filter then sprayed it with something that acted like a dust magnet.
Harvey
------------------------------------------------------------------------
toggle quoted message
Show quoted text
On 4/3/2024 11:47 AM, n4buq wrote: Is there supposed to be a dust filter in addition to the metal screen "filter"?? The screen is good for keeping fingers and larger animals out but for dust, not so much.? The manual states to wash and allow the filter to dry which makes me think there might have been something else behind the mesh.
Thanks, Barry - N4BUQ
Pulling air out of an instrument means sucking air and dirt into the instrument, all around the edges. Not a good thing.
Jeremy
On Wed, Apr 3, 2024 at 7:38?AM n4buq <n4buq@...> wrote:
Thank you so much for this.? Lots of very good information.
Interesting that you mention airflow direction. That's one thing that has puzzled me about this scope.? There has been a lot of discussion lately on other lists regarding whether fans should should push air into the instrument or pull it out and the consensus is that pulling it out is best but now I find that this scope pushes air in.? I presume part of that is to maximize the air flow directly across the pass transistors but that's just a guess. I had considered reversing the airflow but now I don't think I will.
Thanks again, Barry - N4BUQ
> Hi. > > There are now three bearing types available:- > Ball bearing. > Sleeve bearing (often sintered bronze* sometimes a plastic material.) > And Magnetic. > > (* With care, these can last a very long time. Some are possible to > extract and "refill" with lubrication, in a similar way to how we used > to re-lubricate motorcycle chain's!) > > The last type (magnetic) are (mechanically) virtually silent, but stray > DC magnetic fields may be an issue in some cases.? Often found on > smaller higher speed fans, at a price.? They can still fail, if dirt > gets into the bearing area though, or they overheat. > > The other source of "whine" or "turbine like" noise, is caused by > something being too close to the rotating blades, often on the "air > outlet" side.? That can be the motor strut's themselves (later types are > profiled to reduce such noise) or a nearby cable, coax, or even other > metalwork. > > (Large items too close to the inlet side can also cause noise level to > increase, as well as restricting airflow, depending on how the > obstruction allows air to pass into the fan itself...) > > Sometimes, changing the fan for one with very similar airflow > characteristics, but with the inlet over the struts rather than the > outlet (if that is the case) can help with the noise level. > > Take care not to reverse the airflow through the instrument.? Many "wont > like it" over the long term!? Most fans are marked showing the rotation > and airflow direction, but it is not difficult to tell just by looking > at the blade shape.? (The concave side is "usually" facing the outlet.) > > Instruments in a rack, can have conflicting airflow requirements too. > It's not that uncommon to find issues caused by two or more exchanging > each others warm air within a rack, rather than working together to pull > ambient air "through" the rack.? In such cases, the rack is likely to > need it's own fans too! > > It is sadly too, not uncommon for a rack to be "over populated", so that > instruments that pull air in from the underside, have little to no space > for that as the unit below it is too close. There is a reason for those > 1 or 2'U' blank panels often seen, not just to make the front look > neat.? And again, the type of support for the instrument.? Shelves and > "L" shaped supports front to back, can sometimes obstruct airflow > to/from the sides as well. > > Over the years since many older instruments and equipments were > developed, there have been vast improvements in fan technology, in > particular blade and other airflow profile design aspects.? Some to > benefit the ability to push or pull air through a restrictive passage, > others to greatly reduce fan noise, without compromising (too much) on > the airflow characteristics.? Most will use less electrical power for > the same "duty" these days than the older designs, as motor tech' too > has advanced. > > The last thing you want, is for a fan to run, with the blades > aerodynamically "stalled", often a hint is a "fluttering" sound to the > airflow, and/or air exiting a fan being spat out radially, more than > axially.? The resulting overall airflow is usually very low in those > cases, regardless to if the fan is pushing air in, or pulling it out of > a casing. > > Old catalogues to identify the original fan type and performance > (assuming you have the "originally fitted at manufacture" fan, and also > new catalogues (web pages) so as to choose a modern unit with the needed > airflow, and possibly reduced noise is what you need.? It does take more > time than you might expect though, as often the choices are either vast, > or limited.? :-( > > Keeping air filters clean too, can help with noise levels, as well as > keeping the instrument internals cooled correctly. > > Hope something helps.? I hate to see otherwise great instruments > compromised (or even damaged as a result of overheating) just because > someone didn't like the noise, so put a "quieter" fan in. > > HP (and other manufacturers) designers will have done the math, and > specified the part used originally for a good reason.? Not just to > prevent overheating, but also to produce a sensible warm up time, that > results in relatively stable operating temperatures, for a given range > of ambient temperatures.?? Compare the air outlet(s) temperatures once > they have stabilised, for the same ambient and operating conditions, > before and after replacing the fan, to be sure. Much higher with the > new fan, and it's probably not doing the job needed of it.? "How Much" > is a moot point though. > > Cheers All. > > Dave 'KBV > > (Who's populated and retrofitted customer racks with various "high > dissipation things" in the past than I care to think about.? Not unusual > to end up either fitting large fans to the top to pull air out, install > ducting if a unit has the outlet hardware for it, or just remove back > and sides, to allow free airflow, when in use.) > > Noise?? You haven't lived until you work near a host of things with > 400Hz fans in!? Some units we worked with had no less than three of > those noisy things!? They do move heck of a lot of air for the size > though... > > >< > > > -- > Sent from a Micro$oft free zone: > > > >
|