Hi David,

We have one of these at work.
What info would you like to have? I can, when I get the time, take some pictures and do some measurements.
But as this is a costly piece, and we still use it, I will not take it apart more than normal use.

You might find some info here:

Regards,
Askild



On Wed, Oct 20, 2021 at 10:32 PM Dr. David Kirkby, Kirkby Microwave Ltd <drkirkby@...> wrote:
I would be interested in see exactl;y how this device


is made, with a view to making something similar.




 

On Fri, 22 Oct 2021 at 20:00, Askild <megafluffy@...> wrote:
Hi David,

We have one of these at work.
What info would you like to have? I can, when I get the time, take some pictures and do some measurements.
But as this is a costly piece, and we still use it, I will not take it apart more than normal use.

You might find some info here:

Regards,
Askild

Hi,
Thank you very much for the very? kind offer.?

Any photos and physical measurements you can take would be very helpful

There are 4 SMA connectors, which connect to Hcur, Lcur, Hpot & Lpot. Can you measure with a multimeter and determine which pieces of the fixture connect to each of the 4 connectors? It looks as though Lpot and Lcur connect to one side, and Hpot and Hcur connect to the other. But where do each go?

What pieces are electrically isolated from all 4 connectors, but electrically connected to the chassis?

The manual says that there are various values of capacitance, depending on what size plates are connected. Would I be right in assuming that if you assemble this, but don’t fully tighten the thumb screws, that those capacitance values decrease, since you are increase the spacing between the plates?

Am I right in assuming that the gap the manual speaks of, is the gap between the electrodes when fully assembled?
I assume that the ceramic is fixed to each side, and can not easily (if at all) be removed. Can you see how it is fixed to the body? I assume that ceramic sits a certain distance below the top surfaces of each side. What is that distance?

Is the ceramic tapped, so you screw into that when you insert the electrodes? If so, how far approximately do the screws go in? Can you measure the depth of where it's tapped?

I am obviously seriously confused about some aspects of the assembly.? I can't understand why increasing the thickness of the electrodes increases the gap. I would assume if you increase the thickness of the plates, that would decrease the gap between the electrodes, not increase it.

Let's assume hypothetically the ceramic is 4 mm below the surface on each side, so the space between the two pieces of ceramic are 8 mm. Then if you insert 3 mm thick electrodes each side, you take up 2 x 3 = 6 mm with metal, leaving a gap between the electrodes of 8-6=2 mm. That would match one of the data pints given in the manual.

3.0 mm thickness, 2 mm gap

Now, if you use thinner electrodes (say 1.3 mm) I would expect the gap to increase to 8 - 2 x 1.3 = 8 - 2.6 = 5.4 mm. But instead the? manual says

1.3 mm thickness, 0.3 mm gap.

How can using thinner electrodes result in decreasing the gap? I would have expected the dead opposite! I'm obviously totally mis-understanding the construction there.

There appears to me to be 3 connectors for liquid.
1) One of the side, at an angle, where liquid is inserted.
2) One at the bottom, where liquid is drained.
4) MAYBE one at the top, where I have put a question make on this drawing,

I can understand that trying to put liquid into this structure, with the bottom sealed, would be impossible unless the air could get out. So I assume at the top is an air-vent, where I have put a question mark on the attached diagram. Is that true?

There appears to be a lip on the electrodes. Am I right in assuming that the lip is placed in such a way that it goes over the ceramic? I assume that it perhaps helps centre the electrodes on the ceramric.

As you can see, I am a bit confused about the construction of this!

Dave


 

On Fri, 22 Oct 2021 at 20:00, Askild <megafluffy@...> wrote:
Hi David,

We have one of these at work.
What info would you like to have?

One other thing.

All the other fixtures I have come across that connect to the LCR meter do so with BNC connectors, as that's what the LCR meter uses. Yet this fixture is different. Given the LCR meters uses BNC connectors, why do you think HP used SMA connectors on the unit, then supplied adapters to BNC? Why did they not just fit BNCs? Could it be that the connectors need to be hermetically sealed, and that it's impossible to buy hermetically sealed BNCs, but it is SMAs? I know hermetically sealed Ns and SMAs are available, but I have never looked for BNCs, but it would not totally surprise me if they don't exist.

Dave



 

Hi,

I think you might be overthinking this test fixture.
There is no changing the electrodes. Each side of the fixture has its own fixed electrode. Its flush with the insulator.
If the insulator is ceramic or Teflon, I can't say now, but when I get the time I will look closer.
What you can change, is the spacer plate, that determine the distance between the electrodes.
It just go between the two half's of the fixture.
So a thicker spacing plate, longer distance between the electrodes, and lower capacitance.
But using this, the absolute capacitance is normally not of interest.
What you use is the difference between air filled, and liquid filled fixture.

This is just a guess, but I think thy used SMA because they are smaller, else the parts of the fixture had to be made thicker to use BNC.
In the set, there is SMA to BNC adapters.

And what I can see from the manuals, is that center pin on each side is connected together at the plates.
So its like kelvin clips, just for liquids.
It will be important to not change the tightening of the screws between air and liquid measurement.

When it comes to the three liquid ports, the bottom is closed for measurement, and used for draining.
The top one is to let the air out when filling in the liquid.
When I use it, I fill it from the side port, and fill until I see liquid in the top port.
At this point it is a bit overfilled, so removing the syringe from the side port, a bit will come out there.

This is really a simple overpriced piece of fixture.
But I'm sure that the total number of these produced, is not very high. Guess its not a stock item, but produced by order.
But the engineering going into this, and low volume, sort of justify high price.
(last part is just me guessing)

Regards,
Askild


On Sat, Oct 23, 2021 at 1:00 AM Dr. David Kirkby, Kirkby Microwave Ltd <drkirkby@...> wrote:
On Fri, 22 Oct 2021 at 20:00, Askild <megafluffy@...> wrote:
Hi David,

We have one of these at work.
What info would you like to have?

One other thing.

All the other fixtures I have come across that connect to the LCR meter do so with BNC connectors, as that's what the LCR meter uses. Yet this fixture is different. Given the LCR meters uses BNC connectors, why do you think HP used SMA connectors on the unit, then supplied adapters to BNC? Why did they not just fit BNCs? Could it be that the connectors need to be hermetically sealed, and that it's impossible to buy hermetically sealed BNCs, but it is SMAs? I know hermetically sealed Ns and SMAs are available, but I have never looked for BNCs, but it would not totally surprise me if they don't exist.

Dave



 

Looks relatively simple to me.? Knowing the area of the plates and the spacing between them, you can easily calculate the dielectric constant of the material between the plates.? You have a fairly good picture of the disassembled fixture, and more details can be found here:



That paper tells us that the electrodes are nickel plated cobalt (Fe 54%, Co 17%, Ni 29%), the insulator is Alumina (Al2O3), and the O-ring is
Viton (Fluro rubber).

There are a number of papers available, both on Keysight's website and other places, on measuring dielectric constant of water and other liquids, including papers from NBS (now NIST).? There are published tables of dielectric constants of dozens of materials.? I would probably start by reading all the available literature (there's a fair bit on the Keysight website, and likely more available in old HP literature and elsewhere), and then I would build a fixture and start making measurements.? That, of course, is where you would start finding the "unknown unknowns" (a term used by engineers long before anybody ever heard of Donald Rumsfeld).? As problems show up, you simply attack them, one by one.? An engineer I once worked with called this, appropriately, "peeling the onion."? Just poking around the web, I see that the dielectric constant of some liquids varies with temperature and pressure, so you'd want to factor that into your measurements.? (That would be one layer of the onion.)

I was once involved in the development of a soil moisture meter.? Most soils have a dielectric constant around 4, while water has a dielectric constant of 80 at 20° C, as I recall.? The dielectric constant of the soil/water matrix is therefore mildly dependent upon the type of soil, and highly dependent upon the concentration of water in the matrix.? Fascinating stuff.

Do let us know if you find anything interesting.


 

On Thu, 21 Oct 2021 at 09:22, Jim Allyn - N7JA <jim@...> wrote:

?Just poking around the web, I see that the dielectric constant of some liquids varies with temperature and pressure, so you'd want to factor that into your measurements.? (That would be one layer of the onion.)

There’s a document from NPL that discourages the use of the term “dielectric constant”, since it is not a constant, but as you say, varies with temperature, pressure etc. You can’t have a variable constant! The two words are mutually exclusive.?

The correct term to use is permittivity, or relative permittivity. Note it’s a complex number, with a real and imaginary part, with the imaginary part being the loss.

Er = Er’ - j Er”

((Note the dash and double dash).

Yes, I understand the basic way they Keysight fixture works, but I don’t understand how the thickness of the electrodes changes the spacing the way it does. I can’t understand how thicker electrodes results in a wider gap. ?

Dave
--
Dr. David Kirkby,
Kirkby Microwave Ltd,
drkirkby@...

Telephone 01621-680100./ +44 1621 680100

Registered in England & Wales, company number 08914892.
Registered office:
Stokes Hall Lodge, Burnham Rd, Althorne, Chelmsford, Essex, CM3 6DT, United Kingdom


 

It may look simple but making one is not. If the electrodes connecting to the impedance analyzer are not handled properly they can easily make a battery when immersed in a conductive solution. There are other issues as well that you will find out about if you attempt to build one, or even use one. Learn about the?guard ring on a solid dielectric test set which is not just two plates either side of a gap.
Try making a 4 terminal measurement on a resistor?or capacitor immersed in vinegar. Try measuring a 1pF capacitor accurately. All measurements have to deal with stray resistance, inductance, and capacitance. Try measuring a 270 pF dipped mica capacitor at 60 MHz.

On Sat, Oct 23, 2021 at 1:11 PM Jim Allyn - N7JA <jim@...> wrote:
Looks relatively simple to me.? Knowing the area of the plates and the spacing between them, you can easily calculate the dielectric constant of the material between the plates.? You have a fairly good picture of the disassembled fixture, and more details can be found here:



That paper tells us that the electrodes are nickel plated cobalt (Fe 54%, Co 17%, Ni 29%), the insulator is Alumina (Al2O3), and the O-ring is
Viton (Fluro rubber).

There are a number of papers available, both on Keysight's website and other places, on measuring dielectric constant of water and other liquids, including papers from NBS (now NIST).? There are published tables of dielectric constants of dozens of materials.? I would probably start by reading all the available literature (there's a fair bit on the Keysight website, and likely more available in old HP literature and elsewhere), and then I would build a fixture and start making measurements.? That, of course, is where you would start finding the "unknown unknowns" (a term used by engineers long before anybody ever heard of Donald Rumsfeld).? As problems show up, you simply attack them, one by one.? An engineer I once worked with called this, appropriately, "peeling the onion."? Just poking around the web, I see that the dielectric constant of some liquids varies with temperature and pressure, so you'd want to factor that into your measurements.? (That would be one layer of the onion.)

I was once involved in the development of a soil moisture meter.? Most soils have a dielectric constant around 4, while water has a dielectric constant of 80 at 20° C, as I recall.? The dielectric constant of the soil/water matrix is therefore mildly dependent upon the type of soil, and highly dependent upon the concentration of water in the matrix.? Fascinating stuff.

Do let us know if you find anything interesting.


 

Why the Cobalt alloy?? It may as well be solid nickel or nickel plated copper.?
Perhaps it has to do with the temperature coefficient. Another variable.
Does anyone know?


On Sat, Oct 23, 2021 at 1:11 PM Jim Allyn - N7JA <jim@...> wrote:
Looks relatively simple to me.? Knowing the area of the plates and the spacing between them, you can easily calculate the dielectric constant of the material between the plates.? You have a fairly good picture of the disassembled fixture, and more details can be found here:



That paper tells us that the electrodes are nickel plated cobalt (Fe 54%, Co 17%, Ni 29%), the insulator is Alumina (Al2O3), and the O-ring is
Viton (Fluro rubber).

There are a number of papers available, both on Keysight's website and other places, on measuring dielectric constant of water and other liquids, including papers from NBS (now NIST).? There are published tables of dielectric constants of dozens of materials.? I would probably start by reading all the available literature (there's a fair bit on the Keysight website, and likely more available in old HP literature and elsewhere), and then I would build a fixture and start making measurements.? That, of course, is where you would start finding the "unknown unknowns" (a term used by engineers long before anybody ever heard of Donald Rumsfeld).? As problems show up, you simply attack them, one by one.? An engineer I once worked with called this, appropriately, "peeling the onion."? Just poking around the web, I see that the dielectric constant of some liquids varies with temperature and pressure, so you'd want to factor that into your measurements.? (That would be one layer of the onion.)

I was once involved in the development of a soil moisture meter.? Most soils have a dielectric constant around 4, while water has a dielectric constant of 80 at 20° C, as I recall.? The dielectric constant of the soil/water matrix is therefore mildly dependent upon the type of soil, and highly dependent upon the concentration of water in the matrix.? Fascinating stuff.

Do let us know if you find anything interesting.


 

Hi David,
Maybe you missed Akild's expalannation of the capacitnde issue.? The electrodes are fixed to the side plates. The spacing is changed by adding spacer rings. Thicker spacer? bigger gap and les capacitance. The document Jim provided a link to?
Has a mechanical drawing on page 27 which shows the connection details. You can calculate dimension S (electrode diameter) from the stated air capacitance with different spacers.


 

Hi David,?
As I tried to explain in my previous mail, you can not replace the electrodes.?
They are fixed in place.
The only thing you can change, is the spacer plate between the two halves of the fixture.?
So with thicker spacer, the greater spacing between electrodes,? and lower capacitance.?


Regards,?
Askild?

On Sun, 24 Oct 2021, 02:04 Dr. David Kirkby, Kirkby Microwave Ltd, <drkirkby@...> wrote:
On Thu, 21 Oct 2021 at 09:22, Jim Allyn - N7JA <jim@...> wrote:

?Just poking around the web, I see that the dielectric constant of some liquids varies with temperature and pressure, so you'd want to factor that into your measurements.? (That would be one layer of the onion.)

There’s a document from NPL that discourages the use of the term “dielectric constant”, since it is not a constant, but as you say, varies with temperature, pressure etc. You can’t have a variable constant! The two words are mutually exclusive.?

The correct term to use is permittivity, or relative permittivity. Note it’s a complex number, with a real and imaginary part, with the imaginary part being the loss.

Er = Er’ - j Er”

((Note the dash and double dash).

Yes, I understand the basic way they Keysight fixture works, but I don’t understand how the thickness of the electrodes changes the spacing the way it does. I can’t understand how thicker electrodes results in a wider gap. ?

Dave
--
Dr. David Kirkby,
Kirkby Microwave Ltd,
drkirkby@...

Telephone 01621-680100./ +44 1621 680100

Registered in England & Wales, company number 08914892.
Registered office:
Stokes Hall Lodge, Burnham Rd, Althorne, Chelmsford, Essex, CM3 6DT, United Kingdom


 

On Sun, 24 Oct 2021 at 18:46, Askild <megafluffy@...> wrote:
Hi David,?
As I tried to explain in my previous mail, you can not replace the electrodes.?
They are fixed in place.
The only thing you can change, is the spacer plate between the two halves of the fixture.?
So with thicker spacer, the greater spacing between electrodes,? and lower capacitance.?


Regards,?
Askild?
Yes, I understand that now. I assumed that this would have been made from stainless steel for maximum corrosion resistance. But I see it is 100% nickel, which is blinkin expensive! I could not actually find anyone with suitable material online, but I did not look that hard. I think I would probably get it made from a grade of stainless steel with a high nickel content, as that's much more affordable. I don't have any specific use cases for it - the one time I did want something similar for oil, I found out from someone at NPL that you can't measure the loss of low-loss liquids using an LCR meter.

If making one, I would bury a couple of pt100 platinum resistance thermistors in it, so the temperature can be measured accurately.

Dave .


 

On Sat, Oct 23, 2021 at 06:35 PM, peter bunge wrote:
Why the Cobalt alloy?? It may as well be solid nickel or nickel plated copper.?
Perhaps it has to do with the temperature coefficient.

I see there are at least a few hams here, so undoubtedly you will know about radio repeaters, and the duplexers (filters) they use to allow simultaneous transmission and reception with only one antenna.? The tuning rod in duplexer cavities is made of invar, which is 36% nickel and 64% iron.? Invar has a low thermal coefficient of expansion, and is used so the frequency of the cavity won't drift as the temperature changes.? I know of mountaintop repeater sites locally where the ambient temperature can vary at least 50° C. with the seasons.? There are other similar alloys such as Inovco and FeNiCo that also have low coefficient of expansion.? HP says the electrode is nickel plated cobalt (Fe 54%, Co 17%, Ni 29%).? I'd guess thermal coefficient of expansion is one of the reasons for this choice, perhaps combined with corrosion resistance, thermal conductivity, electrical resistivity, and HP-only-knows what else.


 

When I worked for E. I. DuPont Co. 30 years ago, we had a coaxial fixture made from brass, that had a connection on the bottom (banana pins?) that plugged into the top of a Hp 4342A Q meter. With a standard coil to establish resonance, we could measure low loss dielectric fluids quite accurately.


 

On Mon, 25 Oct 2021 at 06:00, John Lyles <jtml@...> wrote:
When I worked for E. I. DuPont Co. 30 years ago, we had a coaxial fixture made from brass, that had a connection on the bottom (banana pins?) that plugged into the top of a Hp 4342A Q meter. With a standard coil to establish resonance, we could measure low loss dielectric fluids quite accurately.
?
I spoke to someone at NPL who said that to measure low-loss liquids, you do indeed need a resonance technique, and look at the Q with and without the liquid. The Q would be calculated from the 3 dB bandwidth. The basic point I was trying to make is that you can't measure the low-loss liquids by measuring the impedance with an LCR meter. There's a large document from NPL I have cited before on this topic.

Dave


 

On Sat, Oct 23, 2021 at 05:03 PM, Dr. David Kirkby, Kirkby Microwave Ltd wrote:
There’s a document from NPL that discourages the use of the term “dielectric constant”, since it is not a constant, but as you say, varies with temperature, pressure etc. You can’t have a variable constant! The two words are mutually exclusive.?
The correct term to use is permittivity, or relative permittivity.


A search of the web site of the National Institute of Standards And Technology turns up 34718 documents that use the term "dielectric constant", but only 26531 documents that use "relative permittivity", and only 1016 instances of "permittivity."

"The United States and Great Britain are two countries separated by a common language."

"These chips Are fries, This queue A line, This tap A faucet, Wardrobe A closet, Vacation Holiday, Underground Subway, Chemist A drug store, Autumn Is fall, A garden Is yard, You see now It's not hard"

From "English to American" video:

https://www.youtube.com/watch?v=5Na7b5MiVak


 

On Tue, 26 Oct 2021 at 02:21, Jim Allyn - N7JA <jim@...> wrote:
On Sat, Oct 23, 2021 at 05:03 PM, Dr. David Kirkby, Kirkby Microwave Ltd wrote:
There’s a document from NPL that discourages the use of the term “dielectric constant”, since it is not a constant, but as you say, varies with temperature, pressure etc. You can’t have a variable constant! The two words are mutually exclusive.?
The correct term to use is permittivity, or relative permittivity.


A search of the web site of the National Institute of Standards And Technology turns up 34718 documents that use the term "dielectric constant", but only 26531 documents that use "relative permittivity", and only 1016 instances of "permittivity."

"The United States and Great Britain are two countries separated by a common language."

The term "dielectric constant" is widely used in the UK, so it's not just the USA which uses it. However, if you think about it logically

* Pi is a mathematical constant (3.14159..).
* c is a physical constant (299792458? m/s, which is approximately 3 x 10^8 m/s)
* something that changes can not be a constant.

Science and engineering advances, and just because we have always done it one way, does not mean we should continue to do it that way. NPL's argument for discouraging the use of the term dielectric constant is to discourage people from thinking they are measuring a constant.

If you accept that relative permittivity is a better term, you can then, when appropriate, mention it is actually a complex number. The relative permittivity Er of a material is a complex number, normally written with a dash and double-dash like

Er = Er' - j Er''

where the Er'' is the loss term
?
Just because

* you Americans have used the term "dielectric constant" for years
* I myself have used the term "dielectric constant" for use
* many Brits have used the term for years

does not mean we should continue to use such a term, when someone points out that it is not a constant but a variable.

Feel free to continue to call this variable a constant if you wish - many Brits will no doubt do it too. But this particular British person is not a stick in the mud.

Dave G8WRB


 

On Tue, Oct 26, 2021 at 12:33 AM, Dr. David Kirkby, Kirkby Microwave Ltd wrote:
If you accept that relative permittivity is a better term
Oh, I agree that relative permittivity is probably a better term, and it is even possible that I will get used to it in the 20 or 25 years I may have left in this life.? I did, after all, get used to saying "Hertz" instead of cycles per second (or cps), "megahertz" instead of megacycles, and gigahertz instead of kmc.? But it's tough to stop using a term you have used for decades, especially when everybody around you still uses the old term. I try to think in metric sometimes, and it's getting a little more comfortable - very slowly.? Perhaps some time we will happen to be in the same country at the same time, and we can discuss it over a pint or two, which I understand is still the only legal measure used for beer in the UK.? In the meantime, remember:

“Science advances one funeral at a time.” - Max Planck

The switch to "relative permittivity" in the United States will probably go much as Mr. Planck suggests.


 

On Fri, 22 Oct 2021 at 20:00, Askild <megafluffy@...> wrote:
Hi David,

We have one of these at work.
What info would you like to have? I can, when I get the time, take some pictures and do some measurements.
But as this is a costly piece, and we still use it, I will not take it apart more than normal use.
Hi,
Can you determine how the electrodes are fixed into the outer case? I understand that there's one electrode on each side. The electrodes must be electrically isolated from the main body, so I assume they sit on? the ceramic insulator. But what holds them to the ceramic? What holds the ceramic in the case? I'm trying to work out all 3 of these conditions can be simultaneously satisfied.

1) Wires connect the electrode to the SMA connectors.
2) Electrode attached on an insulator
3) Insulator attached to the body.

Apart from using an adhesive, which I doubt is used, I can't see how everything can be fixed in place.

One other thing. There's an O-ring. Why not two, given there's a spacer? I would have thought there was a potential for leaks both sides of the spacer, so two o-rings would be needed.

Dave


 

Hi,

Sorry for slow reply, but I'm really busy at work.
I put some pictures in the photo album:

I can't see any way to disassemble it any more than this.

Let me know if there is anything else you would like to see.

Regards,
Askild


On Wed, Oct 27, 2021 at 2:47 PM Dr. David Kirkby, Kirkby Microwave Ltd <drkirkby@...> wrote:
On Fri, 22 Oct 2021 at 20:00, Askild <megafluffy@...> wrote:
Hi David,

We have one of these at work.
What info would you like to have? I can, when I get the time, take some pictures and do some measurements.
But as this is a costly piece, and we still use it, I will not take it apart more than normal use.
Hi,
Can you determine how the electrodes are fixed into the outer case? I understand that there's one electrode on each side. The electrodes must be electrically isolated from the main body, so I assume they sit on? the ceramic insulator. But what holds them to the ceramic? What holds the ceramic in the case? I'm trying to work out all 3 of these conditions can be simultaneously satisfied.

1) Wires connect the electrode to the SMA connectors.
2) Electrode attached on an insulator
3) Insulator attached to the body.

Apart from using an adhesive, which I doubt is used, I can't see how everything can be fixed in place.

One other thing. There's an O-ring. Why not two, given there's a spacer? I would have thought there was a potential for leaks both sides of the spacer, so two o-rings would be needed.