开云体育

ctrl + shift + ? for shortcuts
© 2025 开云体育

Test fixture for dielectric measurements


 

Has anyone made test fixtures for measuring liquids or solids (essentially a capacitor with the DUT as dielectric)
I know they are available for their weight in gold but I would like to make some.
The solids are a problem as there are boundary effects where electrons in the dielectric don't like getting close to the conductive plate.
Amorphous solids (think talcum powder) are a problem for plate design and need compression forces. Other solids like sand or salt crystals are also a challenge.
Transducers like Barium Titanate probably have metal evaporated on the surface.
Anyone out there with experience?
Anyone have something for sale? (contact offline)


Dale H. Cook
 

At 05:21 PM 2/8/2019, Peter Bunge wrote:

Has anyone made test fixtures for measuring liquids or solids (essentially a capacitor with the DUT as dielectric)
The closest I have come is using a GR Type 544-B Megohm Bridge to measure wire insulation and coaxial dielectric leakage and breakdown, but that approach is only good to 500 VDC.

Dale H. Cook, GR/HP/Tek Collector, Roanoke/Lynchburg, VA


Chris Trask - N7ZWY/WDX3HLB
 

Somewhere around here, in a deep, dark, forgotten storage cabinet, I have a General Radio 1690-A "Dielectric Sample Holder". I picked it up at a junk yard decades ago. Used it a few times, then carefully put it away. Carefully enough that I couldn't tell you where it is even if my life depended on it. And if I could find it, I would never part with it.

You might want to download the PDF version of the manual at:



and see if it inspires any ideas on how to make such a fixture.


Has anyone made test fixtures for measuring liquids or solids
(essentially a capacitor with the DUT as dielectric)
I know they are available for their weight in gold but I would
like to make some.
The solids are a problem as there are boundary effects where
electrons in the dielectric don't like getting close to the
conductive plate.
Amorphous solids (think talcum powder) are a problem for plate
design and need compression forces. Other solids like sand or
salt crystals are also a challenge.
Transducers like Barium Titanate probably have metal evaporated
on the surface.
Anyone out there with experience?
Chris

When the going gets weird, the weird turn pro
- Hunter S. Thompson


 

On 2/8/19 6:31 PM, Chris Trask - N7ZWY/WDX3HLB wrote:
Somewhere around here, in a deep, dark, forgotten storage cabinet, I have a General Radio 1690-A "Dielectric Sample Holder". I picked it up at a junk yard decades ago. Used it a few times, then carefully put it away. Carefully enough that I couldn't tell you where it is even if my life depended on it. And if I could find it, I would never part with it.

You might want to download the PDF version of the manual at:

I have one of these as well. It's a marvel of amazing fine machinist
work. Unfortunately I didn't get mine from a junkyard; I paid Rather
Serious Money for it. (you lucky dog)

-Dave

--
Dave McGuire, AK4HZ
New Kensington, PA


Chris Trask - N7ZWY/WDX3HLB
 

I picked up quite a bit of GR stuff at that junk yard, including a pair of those nice variable trombone lines. I think I paid $40 each for them. They had been in storage somewhere for a very long time and were frozen, but I managed to get them loose. Should have bought the third one. Also got a variable length short.


Somewhere around here, in a deep, dark, forgotten storage
cabinet, I have a General Radio 1690-A "Dielectric Sample Holder".
I picked it up at a junk yard decades ago. Used it a few times,
then carefully put it away. Carefully enough that I couldn't tell
you where it is even if my life depended on it. And if I could find
it, I would never part with it.

You might want to download the PDF version of the manual at:

I have one of these as well. It's a marvel of amazing fine machinist
work. Unfortunately I didn't get mine from a junkyard; I paid Rather
Serious Money for it. (you lucky dog)

Chris

When the going gets weird, the weird turn pro
- Hunter S. Thompson


Dr. David Kirkby from Kirkby Microwave Ltd
 

On Fri, 8 Feb 2019, 22:21 peter bunge <bunge.pjp@... wrote:
Has anyone made test fixtures for measuring liquids or solids (essentially a capacitor with the DUT as dielectric)
I know they are available for their weight in gold but I would like to make some.
The solids are a problem as there are boundary effects where electrons in the dielectric don't like getting close to the conductive plate.
Amorphous solids (think talcum powder) are a problem for plate design and need compression forces. Other solids like sand or salt crystals are also a challenge.
Transducers like Barium Titanate probably have metal evaporated on the surface.
Anyone out there with experience?
Anyone have something for sale? (contact offline)

I have played with liquids.?

The frequency you want to make the measurement at is important, as is the loss, as they set what methods work and what will not.

Also is the liquid polar??

I have a document from NPL that lists about 50 ways of making such measurements!?

Andrew Gregory at NPL is very helpful



I managed to get a tour of the lab on the last NPL open day. The lab was not open to the public, but I arranged it in advance.?

I have used an air spaced Jackson capacitor and measured impedance on an LCR meter. It is not perfect as you get fringing fields with air, which are reduced with a dielectric.?

Polar liquids are a problem as they set up a boundary layer between the electrodes.?

I made an air-spaced 50 ohn transmission line, then filled the air with liquid. Unfortunately my design has no analytical solution, so I resorted to EM simulation, but a variation with stripline does?

I am told for low loss liquids like cyclohexane, the only method that works is a resonance method, looking for a reduction in Q with the dielectric.?

Dave


 

Glorious. Nice scores!

-Dave

On 2/8/19 7:14 PM, Chris Trask - N7ZWY/WDX3HLB wrote:
I picked up quite a bit of GR stuff at that junk yard, including a pair of those nice variable trombone lines. I think I paid $40 each for them. They had been in storage somewhere for a very long time and were frozen, but I managed to get them loose. Should have bought the third one. Also got a variable length short.


Somewhere around here, in a deep, dark, forgotten storage
cabinet, I have a General Radio 1690-A "Dielectric Sample Holder".
I picked it up at a junk yard decades ago. Used it a few times,
then carefully put it away. Carefully enough that I couldn't tell
you where it is even if my life depended on it. And if I could find
it, I would never part with it.

You might want to download the PDF version of the manual at:

I have one of these as well. It's a marvel of amazing fine machinist
work. Unfortunately I didn't get mine from a junkyard; I paid Rather
Serious Money for it. (you lucky dog)

Chris

When the going gets weird, the weird turn pro
- Hunter S. Thompson


--
Dave McGuire, AK4HZ
New Kensington, PA


 

Hmmm?

That is basically what I did my MS thesis on:

"An Automated Microwave Network Analyzer
and its Application in the Measurement
of the Dielectric Properties of Materials"

I put together a 8410 system out of manually controlled
8410 network analyzer bits, and some this and that,
automated it, and then made tons of measurements on the
dielectric properties of water at microwave frequencies.

Probably a bit more than you were looking to do.

The test cells were gold plated stainless steel sections
of coaxial "air line", that I filled with the liquids being
sampled. Distilled water, normal saline, and normal saline
with DNA.... Oh, and air.

I then wrote a function that computed the dielectric constant
of water according to the Debye equation to compare my results
to.

-Chuck Harris

peter bunge wrote:

Has anyone made test fixtures for measuring liquids or solids (essentially a capacitor with the DUT as dielectric)
I know they are available for their weight in gold but I would like to make some.
The solids are a problem as there are boundary effects where electrons in the dielectric don't like getting close to the conductive plate.
Amorphous solids (think talcum powder) are a problem for plate design and need compression forces. Other solids like sand or salt crystals are also a challenge.
Transducers like Barium Titanate probably have metal evaporated on the surface.
Anyone out there with experience?
Anyone have something for sale? (contact offline)




 

开云体育

This is a really interesting topic. I also wanted to do several measurements of (complex) permittivity and permeability. But some of the math is a bit cumbersome.
I also wonder what kind of test fixture could be home made (eg. using a ordinary lathe and/or mill).

I think a open end semirigid cable could be used for dielectric measurements of liquids. At least the photographs in some Keysight application note about this topic look like this.

Tobias


-------- Original message --------
From: Chuck Harris <cfharris@...>
Date: 2/9/19 02:11 (GMT+01:00)
Subject: Re: [HP-Agilent-Keysight-equipment] Test fixture for dielectric measurements

Hmmm?

That is basically what I did my MS thesis on:

"An Automated Microwave Network Analyzer
?and its Application in the Measurement
?of the Dielectric Properties of Materials"

I put together a 8410 system out of manually controlled
8410 network analyzer bits, and some this and that,
automated it, and then made tons of measurements on the
dielectric properties of water at microwave frequencies.

Probably a bit more than you were looking to do.

The test cells were gold plated stainless steel sections
of coaxial "air line", that I filled with the liquids being
sampled.? Distilled water, normal saline, and normal saline
with DNA.... Oh, and air.

I then wrote a function that computed the dielectric constant
of water according to the Debye equation to compare my results
to.

-Chuck Harris

peter bunge wrote:
> Has anyone made test fixtures for measuring liquids or solids (essentially a capacitor with the DUT as dielectric)
> I know they are available for their weight in gold but I would like to make some.
> The solids are a problem as there are boundary effects where electrons in the dielectric don't like getting close to the conductive plate.
> Amorphous solids (think talcum powder) are a problem for plate design and need compression forces. Other solids like sand or salt crystals are also a challenge.
> Transducers like Barium Titanate probably have metal evaporated on the surface.
> Anyone out there with experience?
> Anyone have something for sale? (contact offline)
>
>
>
>




Dr. David Kirkby from Kirkby Microwave Ltd
 

On Sat, 9 Feb 2019, 11:14 Tobias Pluess <tobias.pluess@... wrote:
This is a really interesting topic. I also wanted to do several measurements of (complex) permittivity and permeability. But some of the math is a bit cumbersome.
I also wonder what kind of test fixture could be home made (eg. using a ordinary lathe and/or mill).

I think a open end semirigid cable could be used for dielectric measurements of liquids. At least the photographs in some Keysight application note about this topic look like this.

Tobias
To get both permittivity & permeability you need two independent measurements, such as reflection & transmission.?

A lot depends on the frequency. Putting the material under test inside a waveguide is very practical at microwave frequencies, whereas it is not in the MHz range, where a parallel plate capacitor is much more practical.?

Dave.?


 

Thanks Chuck; probably exactly what my friend Ron is looking for.
I hope you don't mind me forwarding this to him:

You might find his site interesting.?
Peter


On Fri, Feb 8, 2019 at 8:08 PM Chuck Harris <cfharris@...> wrote:
Hmmm?

That is basically what I did my MS thesis on:

"An Automated Microwave Network Analyzer
?and its Application in the Measurement
?of the Dielectric Properties of Materials"

I put together a 8410 system out of manually controlled
8410 network analyzer bits, and some this and that,
automated it, and then made tons of measurements on the
dielectric properties of water at microwave frequencies.

Probably a bit more than you were looking to do.

The test cells were gold plated stainless steel sections
of coaxial "air line", that I filled with the liquids being
sampled.? Distilled water, normal saline, and normal saline
with DNA.... Oh, and air.

I then wrote a function that computed the dielectric constant
of water according to the Debye equation to compare my results
to.

-Chuck Harris

peter bunge wrote:
> Has anyone made test fixtures for measuring liquids or solids (essentially a capacitor with the DUT as dielectric)
> I know they are available for their weight in gold but I would like to make some.
> The solids are a problem as there are boundary effects where electrons in the dielectric don't like getting close to the conductive plate.
> Amorphous solids (think talcum powder) are a problem for plate design and need compression forces. Other solids like sand or salt crystals are also a challenge.
> Transducers like Barium Titanate probably have metal evaporated on the surface.
> Anyone out there with experience?
> Anyone have something for sale? (contact offline)
>
>
>
>




 

I hope you don't mind that I forwarded this to my friend Ron.
You might find his site interesting:

Regards, Peter

On Fri, Feb 8, 2019 at 7:16 PM Dr. David Kirkby from Kirkby Microwave Ltd <drkirkby@...> wrote:
On Fri, 8 Feb 2019, 22:21 peter bunge <bunge.pjp@... wrote:
Has anyone made test fixtures for measuring liquids or solids (essentially a capacitor with the DUT as dielectric)
I know they are available for their weight in gold but I would like to make some.
The solids are a problem as there are boundary effects where electrons in the dielectric don't like getting close to the conductive plate.
Amorphous solids (think talcum powder) are a problem for plate design and need compression forces. Other solids like sand or salt crystals are also a challenge.
Transducers like Barium Titanate probably have metal evaporated on the surface.
Anyone out there with experience?
Anyone have something for sale? (contact offline)

I have played with liquids.?

The frequency you want to make the measurement at is important, as is the loss, as they set what methods work and what will not.

Also is the liquid polar??

I have a document from NPL that lists about 50 ways of making such measurements!?

Andrew Gregory at NPL is very helpful



I managed to get a tour of the lab on the last NPL open day. The lab was not open to the public, but I arranged it in advance.?

I have used an air spaced Jackson capacitor and measured impedance on an LCR meter. It is not perfect as you get fringing fields with air, which are reduced with a dielectric.?

Polar liquids are a problem as they set up a boundary layer between the electrodes.?

I made an air-spaced 50 ohn transmission line, then filled the air with liquid. Unfortunately my design has no analytical solution, so I resorted to EM simulation, but a variation with stripline does?

I am told for low loss liquids like cyclohexane, the only method that works is a resonance method, looking for a reduction in Q with the dielectric.?

Dave


 

At work we have the Agilent liquid test fixture, use together with a LCR meter to measure the permittivity of different oil's
I remember that my company didn't want to buy? more than one, due to price....

Keysight also has one for solids as well

The price is way to high for these, but if you study the manuals, you will see that the construction isn't that advanced, so making one should not be a problem if you have access to some proper machining tools.

And the dimensions isn't that critical as you compare the dialectic of what you want to measure against the dialectic of air.

But of course you will? need a instrument to connect to the test fixture.

BR,
Askild


On Sat, Feb 9, 2019 at 12:47 PM Dr. David Kirkby from Kirkby Microwave Ltd <drkirkby@...> wrote:
On Sat, 9 Feb 2019, 11:14 Tobias Pluess <tobias.pluess@... wrote:
This is a really interesting topic. I also wanted to do several measurements of (complex) permittivity and permeability. But some of the math is a bit cumbersome.
I also wonder what kind of test fixture could be home made (eg. using a ordinary lathe and/or mill).

I think a open end semirigid cable could be used for dielectric measurements of liquids. At least the photographs in some Keysight application note about this topic look like this.

Tobias
To get both permittivity & permeability you need two independent measurements, such as reflection & transmission.?

A lot depends on the frequency. Putting the material under test inside a waveguide is very practical at microwave frequencies, whereas it is not in the MHz range, where a parallel plate capacitor is much more practical.?

Dave.?


 

Thanks BR; the reason we will probably have to build our own Test Fixture is that, even with unlimited funds, the commercial fixtures are not suitable for the samples or to be adapted to heating to over 1000 deg C. Just a minor issue.
We have my 4192A that covers from 5 Hz to 13 MHz and my 4193A that covers from 400 kHz to 110 MHz.
My friend has a small business you may be interested in:



I do this for fun. I am retired and am not looking for a 'job' or to go back to 'work'.
Peter


On Sat, Feb 9, 2019 at 11:03 AM Askild <megafluffy@...> wrote:
At work we have the Agilent liquid test fixture, use together with a LCR meter to measure the permittivity of different oil's
I remember that my company didn't want to buy? more than one, due to price....

Keysight also has one for solids as well

The price is way to high for these, but if you study the manuals, you will see that the construction isn't that advanced, so making one should not be a problem if you have access to some proper machining tools.

And the dimensions isn't that critical as you compare the dialectic of what you want to measure against the dialectic of air.

But of course you will? need a instrument to connect to the test fixture.

BR,
Askild

On Sat, Feb 9, 2019 at 12:47 PM Dr. David Kirkby from Kirkby Microwave Ltd <drkirkby@...> wrote:
On Sat, 9 Feb 2019, 11:14 Tobias Pluess <tobias.pluess@... wrote:
This is a really interesting topic. I also wanted to do several measurements of (complex) permittivity and permeability. But some of the math is a bit cumbersome.
I also wonder what kind of test fixture could be home made (eg. using a ordinary lathe and/or mill).

I think a open end semirigid cable could be used for dielectric measurements of liquids. At least the photographs in some Keysight application note about this topic look like this.

Tobias
To get both permittivity & permeability you need two independent measurements, such as reflection & transmission.?

A lot depends on the frequency. Putting the material under test inside a waveguide is very practical at microwave frequencies, whereas it is not in the MHz range, where a parallel plate capacitor is much more practical.?

Dave.?


 

In the late 1980s I worked at E. I. DuPont Co at their Engineering Physics Lab in Delaware. We did a lot of dielectric measurements for the company, and also had an 8510 plus the 1970s 8409 automatic network analyzer, a two rack wide series of instruments controlled with a 9836 computer. The basis was the 8410 network analyzer.

One of the typical measurements we made was liquid dielectric measurements. We had a brass cell, with O rings, that looked like a vertical coaxial cylinder. It was filled a certain height with the liquid, and it fitted on top of the HP 4342A Q meter. We made empty and full measurements and plugged the dial setttings into a program that computers K' and K"". I think we had a heat tape around it, and could measure vs temperatur. Later I automated it with a swept measurement system and had the dielectric properties vs temperature.?

Keysight and an aftermarket Swiss? company make a surface probe that can measure flat materials and liquids. Damaskos in Pennsylvania also makes some devices.


 

In the late 1980s I worked at E. I. DuPont Co at their Engineering Physics Lab in Delaware. We did a lot of dielectric measurements for the company, and also had an 8510 plus the 1970s 8409 automatic network analyzer, a two rack wide series of instruments controlled with a 9836 computer. The basis was the 8410 network analyzer.

One of the typical measurements we made was liquid dielectric measurements. We had a brass cell, with O rings, that looked like a vertical coaxial cylinder. It was filled a certain height with the liquid, and it fitted on top of the HP 4342A Q meter. We made empty and full measurements and plugged the dial setttings into a program that computers K' and K"". I think we had a heat tape around it, and could measure vs temperatur. Later I automated it with a swept measurement system and had the dielectric properties vs temperature.?

Keysight and an aftermarket Swiss? company make a surface probe that can measure flat materials and liquids. Damaskos in Pennsylvania also makes some devices.


Dr. David Kirkby from Kirkby Microwave Ltd
 

On Sat, 9 Feb 2019, 16:02 Askild <megafluffy@... wrote:
At work we have the Agilent liquid test fixture, use together with a LCR meter to measure the permittivity of different oil's

Is there any chance you could take some detailed photographs and detailed dimensions? I have looked at the manual, but some more details would be nice.?

What of the BNCs connect to the guard ring?

Whilst I take your point one is making comparisons to air. There is some software from Keysight that supports that fixture, which might be tailored to it.?

I believe the electrodes are plated in platinum. I don't know if that's the best possible material to reduce the barrier layer problem. I know getting things gold plated is easy, but I am unsure about platinum.


I remember that my company didn't want to buy? more than one, due to price....

Yes, I can understand that!


Keysight also has one for solids as well

The price is way to high for these, but if you study the manuals, you will see that the construction isn't that advanced, so making one should not be a problem if you have access to some proper machining tools.

I managed to pick up one of those. I have often wondered if it could be used for liquids too, if one made a small tank to fit inside - like a small fish tank! Although it would make the maths tricky, if the properties of the tanks material were first measured, an analytical solution (excuse the pun) could be derived.?

As I write before I have reasonable results with using an air spaced capacitor, but the lack of a guard ring is going to reduce accuracy.?

And the dimensions isn't that critical as you compare the dialectic of what you want to measure against the dialectic of air.

But of course you will? need a instrument to connect to the test fixture.

I have a 4284A & 4285A, which were top of the line instruments in their day. Unfortunately the fixtures sell on the used market for more than the instruments do!

BR,
Askild

Dave



 



On Mon, Feb 11, 2019 at 1:49 PM Dr. David Kirkby from Kirkby Microwave Ltd <drkirkby@...> wrote:
On Sat, 9 Feb 2019, 16:02 Askild <megafluffy@... wrote:
At work we have the Agilent liquid test fixture, use together with a LCR meter to measure the permittivity of different oil's

Is there any chance you could take some detailed photographs and detailed dimensions? I have looked at the manual, but some more details would be nice.?

I can maybe do this a bit later when I get the time for it...
?

What of the BNCs connect to the guard ring?

The test jig has SMA connectors, the BNC's are just adapters.? All the SMA connectors are screw directly onto the test fixture that is all metal. This will probably get more clear when I get some pictures for you.



Whilst I take your point one is making comparisons to air. There is some software from Keysight that supports that fixture, which might be tailored to it.?

I believe the electrodes are plated in platinum. I don't know if that's the best possible material to reduce the barrier layer problem. I know getting things gold plated is easy, but I am unsure about platinum.


I remember that my company didn't want to buy? more than one, due to price....

Yes, I can understand that!


Keysight also has one for solids as well

The price is way to high for these, but if you study the manuals, you will see that the construction isn't that advanced, so making one should not be a problem if you have access to some proper machining tools.

I managed to pick up one of those. I have often wondered if it could be used for liquids too, if one made a small tank to fit inside - like a small fish tank! Although it would make the maths tricky, if the properties of the tanks material were first measured, an analytical solution (excuse the pun) could be derived.?

As I write before I have reasonable results with using an air spaced capacitor, but the lack of a guard ring is going to reduce accuracy.?

And the dimensions isn't that critical as you compare the dialectic of what you want to measure against the dialectic of air.

But of course you will? need a instrument to connect to the test fixture.

I have a 4284A & 4285A, which were top of the line instruments in their day. Unfortunately the fixtures sell on the used market for more than the instruments do!

BR,
Askild

Dave


Askild?



Dr. David Kirkby from Kirkby Microwave Ltd
 

On Fri, 8 Feb 2019, 22:21 peter bunge <bunge.pjp@... wrote:
Has anyone made test fixtures for measuring liquids or solids (essentially a capacitor with the DUT as dielectric)

Attached is a booklet from NPL on dielectric measurements.?

Is is noteworthy they advise against the term "dielectric constant" as no dielectric has constant properties.?

Dave.?




 

Thanks John, we appreciate any information we can get.
Even the old stuff which has a lot of information.
PeterB

On Mon, Feb 11, 2019 at 1:44 AM John Lyles <jtml@...> wrote:
In the late 1980s I worked at E. I. DuPont Co at their Engineering Physics Lab in Delaware. We did a lot of dielectric measurements for the company, and also had an 8510 plus the 1970s 8409 automatic network analyzer, a two rack wide series of instruments controlled with a 9836 computer. The basis was the 8410 network analyzer.

One of the typical measurements we made was liquid dielectric measurements. We had a brass cell, with O rings, that looked like a vertical coaxial cylinder. It was filled a certain height with the liquid, and it fitted on top of the HP 4342A Q meter. We made empty and full measurements and plugged the dial setttings into a program that computers K' and K"". I think we had a heat tape around it, and could measure vs temperatur. Later I automated it with a swept measurement system and had the dielectric properties vs temperature.?

Keysight and an aftermarket Swiss? company make a surface probe that can measure flat materials and liquids. Damaskos in Pennsylvania also makes some devices.