8594E Cal Signal not Found ?
After replacing battery I entered the Time Base Correction, Frequency Response Correction & Step Attenuator Corrections.
Now when I go to perform Cal Freq or Cal Amptd I’m getting a Cal Signal Not Found.
I checked the CAL OUT and I’m seeing 300.000824MHz -20dBm on my Spectrum Analyzer.
Input using an HP E4432B Signal Generator set 300MHz -20dBm I’m seeing 300.8MHz at -28dBm.
As I change the Span I get different frequency’s showing?
?
Andrew
?
? |
The workaround for the "Cal Signal Not Found" error is covered in the Assembly Level Repair manual.
?
If you had to reload all the correction constants, I would start with the "Analyzer Initialization" section in that manual (namely: setting DEFAULT CONFIG) and go from there, if you haven't already done so.
?
-mark
?
?
On Fri, Apr 4, 2025 at 02:51 PM, KB2LMN wrote:
|
Thanks Mark, ya I did all that but did just see this so I will try next.
?
Instrument firmware expects the cal output signal to be 300 MHz ± 2 MHz.Sometimestheinstrument default data is not able to tune the cal signal within this range and a “cal signal not found” message may appear on screen. Perform a cal output bypass check by pressing Frequency,-37, Hz, Cal, Cal Freq. This will bypass the cal check and start by calibrating the sweep ramp.
?
Andrew |
Hi Andrew,
?
It's ok to have 2 or 4 EEPROMs for version G (94.03.15).? You can read about the various versions and EEPROM part numbers here (original doc #08590-90324):
?
?
The latest version according to that doc is M (991130), but the latest out there in the various archives that I'm aware of is L (980615).
?
I don't recommend messing with firmware upgrades until everything is resolved with the self-cal.? One thing at a time.
?
-mark
?
?
On Fri, Apr 4, 2025 at 07:35 PM, KB2LMN wrote:
|
开云体育? I repaired an 8591A for a friend two weeks ago, and he had the same problem. In his case, it had lost its calibration constants due to a low battery. ? What caused the signal not found or LOCK OFF SRQ110 problem was that the units after a PRESET weren't dBm but mV, and also that the flatness data was empty...no value. For flatness, I had entered a value of -1 for the entire band. ? After fixing these two issues, I was able to run CAL FRE and CAL AMP without any problems. ? I'm not saying this is the problem with your 8594E, but out of curiosity, do either of these two cases apply to you regarding flatness and units after a PRESET? ? Yves ,_ |
Mark I did get everything working last night thank to your help I appreciate it! I ran it through some tests and it passed them all so the firmware was my next leap. Fortunately I have done some chip programming so this one should be an easy fix/ upgrade for me. I looked around and can’t find the G 980615 file for upgrade purposes.
?
Again thanks for sharing your expertise it would have taken me much, much longer to figure this all out on my own.
?
Andrew |