3586A Error 3.1 repaired..thanks to Didier


J.Byers
 

Hi Hank
HP Journal below has 3586A/B/C introductory article

PL Thomas "A programmable selective level meter (wave analyzer)
with synthesized tuning, autoranging, and automatic
calibration",
HEWLETT PACKARD JOURNAL, May, 1980 Vol 31, Number 5
Pages 3 thro 8

Thats from my catalog index.
I cant find all my issues at present (shifted house 2 years ago and much of
my stuff packed to the roof in a garden shed, I cant even open the door
without stuff cascading out!).
If I come across it I will email the pages to you

Best regards
JBEE (John Byers)

----- Original Message -----
From: Henry Kolesnik <kolesnik@...>
To: <hp_agilent_equipment@...>
Sent: Monday, August 08, 2005 10:34 PM
Subject: Re: [hp_agilent_equipment] 3586A Error 3.1 repaired..thanks to
Didier


John
Thanks for the tip, I'll file it away just in case I ever need to do it on
my 3586C. I'm still trying to find out if HP published an article in
their
Journal about the 3586 series?
tnx

73
Hank WD5JFR

----- Original Message -----
From: "hpxref" <face1941@...>
To: <hp_agilent_equipment@...>
Sent: Monday, August 08, 2005 1:47 AM
Subject: [hp_agilent_equipment] 3586A Error 3.1 repaired..thanks to Didier



Yahoo! Groups Links









Yahoo! Groups Links






hpxref
 

My 3586A E3.1 repair:
Many thanks to Didier (#1759)for his spot on suggestion as to the fix
for the 3.1 error on my 3586A. One difference: My VCXO oscillator
was oscillating amplitude wise OK, but had drifted low in frequency.
Others may benefit from my particular fix, different from Didiers, so
here goes.
With 3586A and measuring instruments warmed up 4 hours from cold at
20 deg C room temp:-
Instrumentation used:
HP34401A DMM, HP5316A Freq Cntr (with opt 001,but external frequency
in at rear fed from HP3586A HI STAB TCXO 10MHz out).
Systron Donner DL40 dual PSU fed in turn to 50k 10 turn Helipot,
bridge connected with center arm providing a variable external + / -
Vco via screened leads, +/- 8.6v, adjusted in 1.000v increments
as monitored on DMM.
A small pi section RF filter with a 100Hz LC `trap' section was used
in series with Vco leads into A11 to reduce any outside
interference. (50Hz line frequency over here).
The above instruments were in turn fed from an Advance CVN250A AC
regulator, waveform corrected

3586 Adjustments (A11 board):

Adjusting the slug outwards in L22 did not quite pull the XTAL
enough to lock, but unsoldering the XTAL case to its ground post and
lifting it up clear of the ground plane beneath did the trick here.
The reduction in XTAL case capacitance to deck raised its freq enough
in this case and no other changes to components were needed.
Before this fix my VCXO had resistor R27 lifted and an external
voltage fed to the cold end to check for frequency swing:
At 0.000 v Y20 (XTAL) frequency was 49.982775 Mhz, some 1600Hz lower
than it should have been
I measured a total deviation in Y20 frequency of from 2775Hz change
(+8.600 v in) to 3989Hz change for – 8.600 v in.)

After Xtal case lifted off ground and slug at minimum insertion, (the
now correctly locked case) total deviation for the same external Vco
inputs was:
1339 Hz change for +8.600 v in and
1772 Hz change for – 8.600 v in
or a total excursion of 3,111 Hz from the XTALs center frequency

With R27 re connected and external Vco removed, test point TP1 now
indicated a correctly locked frequency of 49.984375 MHz as indicated
on my 5316A counter.

Note that my total frequency excursion is near the same as the 3kHz
Didier experienced after adding extra C to Q22 on his unit. So this
value may well be close to HP's original specification as the only
circuit modification to mine was lifting the ground of the XTAL case.
As this is a relatively simple mod to make, others may find the same
in their case, avoiding changes or additions at the component level.
Again, many thanks to Didier for pointing out the area to check,
allowing my 3586A to work as it should (a lot less annoying whistles
and a normal dB readout of signals received at last!)… and for the
much appreciated downloaded details of the 3586A, a truly
magnificent piece of instrumentation and a tribute to HP's designers !

JBEE (John Byers)


 

John
Thanks for the tip, I'll file it away just in case I ever need to do it on my 3586C. I'm still trying to find out if HP published an article in their Journal about the 3586 series?
tnx

73
Hank WD5JFR

----- Original Message -----
From: "hpxref" <face1941@...>
To: <hp_agilent_equipment@...>
Sent: Monday, August 08, 2005 1:47 AM
Subject: [hp_agilent_equipment] 3586A Error 3.1 repaired..thanks to Didier


My 3586A E3.1 repair:
Many thanks to Didier (#1759)for his spot on suggestion as to the fix
for the 3.1 error on my 3586A. One difference: My VCXO oscillator
was oscillating amplitude wise OK, but had drifted low in frequency.
Others may benefit from my particular fix, different from Didiers, so
here goes.
With 3586A and measuring instruments warmed up 4 hours from cold at
20 deg C room temp:-
Instrumentation used:
HP34401A DMM, HP5316A Freq Cntr (with opt 001,but external frequency
in at rear fed from HP3586A HI STAB TCXO 10MHz out).
Systron Donner DL40 dual PSU fed in turn to 50k 10 turn Helipot,
bridge connected with center arm providing a variable external + / -
Vco via screened leads, +/- 8.6v, adjusted in 1.000v increments
as monitored on DMM.
A small pi section RF filter with a 100Hz LC `trap' section was used
in series with Vco leads into A11 to reduce any outside
interference. (50Hz line frequency over here).
The above instruments were in turn fed from an Advance CVN250A AC
regulator, waveform corrected

3586 Adjustments (A11 board):

Adjusting the slug outwards in L22 did not quite pull the XTAL
enough to lock, but unsoldering the XTAL case to its ground post and
lifting it up clear of the ground plane beneath did the trick here.
The reduction in XTAL case capacitance to deck raised its freq enough
in this case and no other changes to components were needed.
Before this fix my VCXO had resistor R27 lifted and an external
voltage fed to the cold end to check for frequency swing:
At 0.000 v Y20 (XTAL) frequency was 49.982775 Mhz, some 1600Hz lower
than it should have been
I measured a total deviation in Y20 frequency of from 2775Hz change
(+8.600 v in) to 3989Hz change for - 8.600 v in.)

After Xtal case lifted off ground and slug at minimum insertion, (the
now correctly locked case) total deviation for the same external Vco
inputs was:
1339 Hz change for +8.600 v in and
1772 Hz change for - 8.600 v in
or a total excursion of 3,111 Hz from the XTALs center frequency

With R27 re connected and external Vco removed, test point TP1 now
indicated a correctly locked frequency of 49.984375 MHz as indicated
on my 5316A counter.

Note that my total frequency excursion is near the same as the 3kHz
Didier experienced after adding extra C to Q22 on his unit. So this
value may well be close to HP's original specification as the only
circuit modification to mine was lifting the ground of the XTAL case.
As this is a relatively simple mod to make, others may find the same
in their case, avoiding changes or additions at the component level.
Again, many thanks to Didier for pointing out the area to check,
allowing my 3586A to work as it should (a lot less annoying whistles
and a normal dB readout of signals received at last!). and for the
much appreciated downloaded details of the 3586A, a truly
magnificent piece of instrumentation and a tribute to HP's designers !

JBEE (John Byers)









Yahoo! Groups Links


 

Thanks to John Byers there's a HP Journal article on the 3586.

Does anyone have the issue that they can scan?
HP Journal below has 3586A/B/C introductory article
HEWLETT PACKARD JOURNAL, May, 1980 Vol 31, Number 5 Pages 3 thro 8
PL Thomas "A programmable selective level meter (wave analyzer) with synthesized tuning, autoranging, and automatic calibration",

I'd sure like to read it in any form.

73
Hank WD5JFR

----- Original Message -----
From: "J.Byers" <face1941@...>
To: <hp_agilent_equipment@...>
Sent: Saturday, January 04, 1997 2:14 AM
Subject: Re: [hp_agilent_equipment] 3586A Error 3.1 repaired..thanks to Didier


Hi Hank
HP Journal below has 3586A/B/C introductory article

PL Thomas "A programmable selective level meter (wave analyzer)
with synthesized tuning, autoranging, and automatic
calibration",
HEWLETT PACKARD JOURNAL, May, 1980 Vol 31, Number 5
Pages 3 thro 8

Thats from my catalog index.
I cant find all my issues at present (shifted house 2 years ago and much of
my stuff packed to the roof in a garden shed, I cant even open the door
without stuff cascading out!).
If I come across it I will email the pages to you

Best regards
JBEE (John Byers)

----- Original Message -----
From: Henry Kolesnik <kolesnik@...>
To: <hp_agilent_equipment@...>
Sent: Monday, August 08, 2005 10:34 PM
Subject: Re: [hp_agilent_equipment] 3586A Error 3.1 repaired..thanks to
Didier


John
Thanks for the tip, I'll file it away just in case I ever need to do it on
my 3586C. I'm still trying to find out if HP published an article in
their
Journal about the 3586 series?
tnx

73
Hank WD5JFR

----- Original Message -----
From: "hpxref" <face1941@...>
To: <hp_agilent_equipment@...>
Sent: Monday, August 08, 2005 1:47 AM
Subject: [hp_agilent_equipment] 3586A Error 3.1 repaired..thanks to Didier



Yahoo! Groups Links









Yahoo! Groups Links








Yahoo! Groups Links






 

Hi all,
?
The HEWLETT PACKARD JOURNAL, ?May, 1980 Vol 31 can be downloaded from the hp archive, it's about 30Mb. There is also an article about the companion 3336? signal generator.
?
?
Regards,
?
Waz