My 3586A E3.1 repair:
Many thanks to Didier (#1759)for his spot on suggestion as to the fix
for the 3.1 error on my 3586A. One difference: My VCXO oscillator
was oscillating amplitude wise OK, but had drifted low in frequency.
Others may benefit from my particular fix, different from Didiers, so
here goes.
With 3586A and measuring instruments warmed up 4 hours from cold at
20 deg C room temp:-
Instrumentation used:
HP34401A DMM, HP5316A Freq Cntr (with opt 001,but external frequency
in at rear fed from HP3586A HI STAB TCXO 10MHz out).
Systron Donner DL40 dual PSU fed in turn to 50k 10 turn Helipot,
bridge connected with center arm providing a variable external + / -
Vco via screened leads, +/- 8.6v, adjusted in 1.000v increments
as monitored on DMM.
A small pi section RF filter with a 100Hz LC `trap' section was used
in series with Vco leads into A11 to reduce any outside
interference. (50Hz line frequency over here).
The above instruments were in turn fed from an Advance CVN250A AC
regulator, waveform corrected
3586 Adjustments (A11 board):
Adjusting the slug outwards in L22 did not quite pull the XTAL
enough to lock, but unsoldering the XTAL case to its ground post and
lifting it up clear of the ground plane beneath did the trick here.
The reduction in XTAL case capacitance to deck raised its freq enough
in this case and no other changes to components were needed.
Before this fix my VCXO had resistor R27 lifted and an external
voltage fed to the cold end to check for frequency swing:
At 0.000 v Y20 (XTAL) frequency was 49.982775 Mhz, some 1600Hz lower
than it should have been
I measured a total deviation in Y20 frequency of from 2775Hz change
(+8.600 v in) to 3989Hz change for – 8.600 v in.)
After Xtal case lifted off ground and slug at minimum insertion, (the
now correctly locked case) total deviation for the same external Vco
inputs was:
1339 Hz change for +8.600 v in and
1772 Hz change for – 8.600 v in
or a total excursion of 3,111 Hz from the XTALs center frequency
With R27 re connected and external Vco removed, test point TP1 now
indicated a correctly locked frequency of 49.984375 MHz as indicated
on my 5316A counter.
Note that my total frequency excursion is near the same as the 3kHz
Didier experienced after adding extra C to Q22 on his unit. So this
value may well be close to HP's original specification as the only
circuit modification to mine was lifting the ground of the XTAL case.
As this is a relatively simple mod to make, others may find the same
in their case, avoiding changes or additions at the component level.
Again, many thanks to Didier for pointing out the area to check,
allowing my 3586A to work as it should (a lot less annoying whistles
and a normal dB readout of signals received at last!)… and for the
much appreciated downloaded details of the 3586A, a truly
magnificent piece of instrumentation and a tribute to HP's designers !
JBEE (John Byers)