¿ªÔÆÌåÓý

Date
Re: Manual request
Looks like Artek Manuals has it. See < http://artekmanuals.com/manuals/hp-manuals/>. [hp_agilent_equipment] <hp_agilent_equipment@...> wrote:
By Sigur?ur ?sgeirsson · #83154 ·
Repair rights
Forwarded from Tekscopes.
By David C. Partridge · #83153 ·
Re: HP6034 PSU Q1, 2 fet
That is typical of a problem where a full wave doubler is working half wave. This is one of those jobs where an isolation transformer is a good idea. You can get by without by using two probes, and
By Chuck Harris · #83152 ·
Re: HP 8648B and an off topic request.
George, Some of the boards in the 8648 are very similar to those in the 8920. The Component Level Information Packet Manual Part Number: 08920-90107, has schematics. It is (or was) on Keysight.com.
By John Gord · #83151 ·
Re: HP 8648B and an off topic request.
Hi Don I now understand your email, having looked at the available HP manual, I will probe around the display connector and the RPG connector. However I suspect that without component level
By george edmonds · #83150 ·
Re: HP8970A automatic Noise Gain meter
Hi Chuck, Thanks for your comments on this. I will seek for a suitable battery to replace it. Studying the RAM chip data sheet learns that the data retention threshold voltage is 2V . Therefore 2,8V
By Herman van Rees · #83149 ·
Re: HP6034 PSU Q1, 2 fet
Hi The DC voltage across the AC to DC stage storage capacitors will give a good indication of the health of the storage capacitors as you vary the output load. These capacitors must be high ripple
By george edmonds · #83148 ·
Re: HP 8648B and an off topic request.
Hi Thank you for your reply, I still do have WinXP sp3 machines, however, the 82357B adapter is not easily available in the UK. The only ones listed on a famous action site are from China and vary in
By george edmonds · #83147 ·
Re: HP6034 PSU Q1, 2 fet
Known good HP 6038A #1 Vout = 23.5Vdc Iout = 5.5Amps Input Filter out: Vdc = 317.5Vdc Vac = 1.48Vac (HP6034A = 12.2Vac!) #2 Vout = 23.5Vdc Iout = 9.5Amps Input Filter out: Vdc = 310.9Vdc Vac = 2.58Vac
By KI6EZ · #83146 ·
Re: HP 8648B and an off topic request.
Hi Don Thank you for replying to me, it is much appreciated. First I have never seen an extension card that would work with this generator, the three plug in modules use a 9 pin D for the power and
By george edmonds · #83145 ·
Re: HP 8648B and an off topic request.
I've successfully run the 8648 calibration and motherboard utilities on WinXP SP3 w/Agilent IO v16.x and 82357B USB GPIB adapter. No special hacks needed other than it only worked on a physical
By Colby Burkett · #83144 ·
Re: HP6034 PSU Q1, 2 fet
Further updates, with some surprises.. I've performed PSU load test while monitoring output of the 'Input filter'. DVM (fluke 70) was utilized since a single ended scope probe is a wrong tool for this
By KI6EZ · #83143 ·
Re: HP 8648B and an off topic request.
George G6HIG, Looks like a useful instrument, I found several manuals searching "agilent 8648b manual." Also, try looking at the uP and the memory circuits supported by the backup battery. An
By Don Doughty · #83142 ·
Re: HP ESP120 Server 52V SMPS load removal question
Hi Chris In my experience most SMPS that were designed for computer use require a minimum load to operate correctly. Some will not turn on at all until a minimum load is present. It also strikes me
By george edmonds · #83141 ·
Re: HP6034 PSU Q1, 2 fet
Chuck Thank you for feedback. The voltage doubler filter (aka input filter) consist of 4 filter caps C2-C5 590uF 200VDC electrolytics (0180-2968) and a pair of inductors L1 and L2. As I have mentioned
By KI6EZ · #83140 ·
Re: HP 8660x series plug-in (8660 synthesizer) - rotary attenuation/
Andy, Thank you very much for the suggestion. I of course failed to add an important detail in my original post: the mainframe is known good, as another plug-in on this same mainframe works normally,
By David Feldman · #83139 ·
HP ESP120 Server 52V SMPS load removal question
29/04/2017 15:44 I am curious as to whether uncontrolled voltage regulation on the drains of big FETs used in a push pull low frequency (136kHz) RF amp run in quasi Class D may be caused by the 52V at
By Chris Wilson · #83138 ·
Re: 8592B to 8593A conversion (A25 counterlock retrofit)
Hi folks, just found my problem, there was another counter lock board, it's 08591-60027 and this one divides the sampling oscillator output by 40/41 and uses this signal as pll input. If I use my
By Thomas Gulden · #83137 ·
Re: HP8970A automatic Noise Gain meter
The Nominal 3.3V lithium cells are a really a 3.6V lithium thionyl chloride cells. If at low loads, they putter along at 3.65V for years and years, and then start a slow decline down to a couple of
By Chuck Harris · #83136 ·
Re: HP6034 PSU Q1, 2 fet
Typically there are two basic kinds of switchers: 1) full wave bridge/voltage doubler input, with a big capacitor filter. 2) power factor correcting preregulator, that drives main switcher, no big
By Chuck Harris · #83135 ·