Hi Tim, Nice. Interesting that it is that accurate at 4GHz after the conversion. I have more than several of these generators and I think they are great. I noticed on mine that if you do a hard reset
By Tom B · #94330 ·
Re: HP34702A Ohms converter issues
That's weird - I thought I saw another response to this thread but when I refreshed the page, it disappeared!? :-(
By gregdunn@... · #94328 ·
Re: HP34702A Ohms converter issues
I put a pair of 2N3904 in for Q13-14 - no change unfortunately.? :-( Yes, it's hard to figure out what Q11 needs to be - HP lists 'FN2960' as an equivalent but that doesn't come back with a match for
By gregdunn@... · #94327 ·
Re: HP34702A Ohms converter issues
I was just looking back at the schematic and OP - I see that you reported Q11 was bad and replaced, but not with the same type. This could be the problem, since JFETs all work more or less the same,
By Ed Breya · #94325 ·
Re: HP34702A Ohms converter issues
If U2 is railed negative, then Q11 should cut off. Are the parts in sockets? If so, try pulling Q13, which should drop the gain and the current to a very low value. If there's no effect, then try Q14.
By Ed Breya · #94324 ·
Prologix has updated the EZGPIB file and is working on the C++ example.
The title says it all. You will find many examples in the folder when you download it. Just click on EZGPIB.exe to run. Investigate the?DF6JB Plotter as well. No affiliation, I just like this
By peter bunge · #94323 ·
Re: HP34702A Ohms converter issues
I hope so.? ;-) In this setup (Q12 / CR13 out of circuit), the gates of Q15 are offset just a bit - A: -5.0V B: -5.8V? The drains are of course then far apart - A: 2.8V B: 8.46V? and the sources
By gregdunn@... · #94322 ·
Re: HP34702A Ohms converter issues
The problem is obviously revolving around the op-amp which is made up of Q15 and U2. The two gates of Q15 should be balanced at all times, otherwise the output of U2 would be stuck at the +12 or the
By Joel Setton · #94321 ·
Re: VNA advice wanted
Pat, Thanks. I was pretty sure David had the A/B designations swapped, but confirmation greatly appreciated. And it would be very easy for someone reading the thread later to get things backwards with
By Reginald Beardsley · #94320 ·
Re: Interested in helping start/run a Wiltron/Anritsu group on groups.io?
Might take a few hours to be approved and propagate. [email protected]> wrote:
By bownes · #94319 ·
Re: Interested in helping start/run a Wiltron/Anritsu group on groups.io?
*[email protected] <[email protected]> *has been created. Have fun! [email protected]> wrote:
By bownes · #94318 ·
Re: HP34702A Ohms converter issues
With the protection circuitry disabled, the 1K resistor reads 1.87K. U3's inputs are both at -6.4V.? Its output is at -7.4V. Q15 seems to depend on what I'm measuring; for this test, its inputs are
By gregdunn@... · #94317 ·
Interested in helping start/run a Wiltron/Anritsu group on groups.io?
Looking for person (one or more) interested in starting and running a Wiltron/Anritsu equipment group (possibly modeled after this HP-Agilent-Keysight-equipment group) on groups.io. If you are
By David Feldman · #94316 ·
Re: VNA advice wanted
I bought mine from Globaltest a few years ago, I think I paid about $40 for it. -Dave -- Dave McGuire, AK4HZ New Kensington, PA
By Dave McGuire · #94315 ·
Re: VNA advice wanted
Dave, In regard to the collet extractor for APC-7, where did you find one cheap? I've been looking high and low for an inexpensive collet extractor without success. -Eric
By Eric Garner · #94314 ·
Re: Test fixture for dielectric measurements
Hmmm... If solid, you could hold the material in air away from the electrodes Chris Trask N7ZWY / WDX3HLB Senior Member IEEE http://www.home.earthlink.net/~christrask/
By Chris Trask - N7ZWY/WDX3HLB <christrask@...> · #94313 ·
Re: HP8714ET Impedance
Just going through my old emails. You may find the attached useful. Don't even think about getting into the bridge to change the impedance. HP uses stripline on a saphire substrate (on the 85046A
By peter bunge · #94312 ·
Re: VNA advice wanted
With all this great talk about VNA's it reminds me¡­ I have an extra 8753ES, 50ohm, internal s-para, working and looking beautiful¡­ 3ghz unit. $3k and also an n8357a 6ghz pna for sale¡­ same
By Pete Manfre <pmanfre@...> · #94311 ·
Re: HP34702A Ohms converter issues
Hello Greg, I think we're getting close to the root cause... but I thhink I need some more data. First, temporarily unsolder the collector lead of Q12 and one lead of CR13, which will disable the
By Joel Setton · #94310 ·
Re: VNA advice wanted
Yes, my error. I got the 85046A and 85046B transposed! The A is the 50 ohm version with APC7 connectors. I doubt the RF source, A, B or R receiver N connectors would get damaged as they are rarely