CHECK OUT THE WIKI The purpose of the wiki is mainly to allow you to find information on instruments, either from either
- The model number
- The function(s) listed below. Some instruments have multiple functions - for example, the 4195A is a VNA, spectrum analyzer and an impedance analyzer. Therefore the 4195A is listed in multiple categories
Please also check out HPWiki available here:
- Accessory kits - various types
- AC power analyzers - PA2201A and PA2203A
- AC power supplies 6811C, 6812C, 6813C
- Airlines
- Arbitrary waveform generators M8194A
- Amplifiers?493A, 495A?
- Attenuators (optical) 8156A, 8157A, 8158B, 81566A, 81576A,?
- Attenuators (RF) 8494A
- Attenuator set (500 Ω) 350C
- Attenuator set (600 Ω) 350D
- Attenuator switch driver
- Audio analyzers? 8903A, 8903B, 8903E,? ?
- Base station test sets
- Bit error rate testers (BERTs)
- Cables
- Capacitance meters U1701A, U1701B, 4272A, 4278A, 4279A
- Capacitor Bridge 4270A,
- Capacitor standards 16380A, 16380C,?
- Carrier noise test setsi
- Cesium frequency standards
- Clamp ammeters
- Close field probes
- Crystal Impedance E4915A, E4916A
- Data Acquisition Systems (DAQs)
- DC power analyzers
- DC power supplies 6030A , 6031A , 6032A, 6033A, 6035A, 6131C, 6621A, 6622A, 6623A, 6624A, 6627A, 6255A, 6645A, 6671A, 6672A, 6673A, 6674A, 6675A, 62003A, 62003C, 62003E, 62004A, 62004B, 62004E, 62005A, 62005B, 62005E, 62006A, 62006B, 62006E, 62010A, 62010C, 62010E, 62012A, 62012C, 62012E, 62015A, 62015C, 62015E, 62018A, 62018C, 62018E, 62024A, 62024C, 62024E, 62028A, 62028C, 62028E, 62048A, 62048C, 62048E
- Delay lines
- Detectors
- Device current waveform analyzers
- Digital communications analyzers
- Directional couplers
- Distortion analyzers 330B, 330C, 330D, 331A, 332A, 333A, 334A, 339A, 8903A, 8903B, 8903E,???
- Dynamic measurement DC source
- Electrometers
- Fading simulators
- Femto ammeters
- Filters
- Frequency counters 522B, 5342A 5343A 5352B
- Frequency standards?
- Function Generators ? 3310A,? 8165A,
- GPIB controllers, extenders, cables etc.
- GPS frequency standards
- Harmonic mixers
- High resistance meters 4339B
- High resistance meter fixtures 16008B
- HEV EV Grid Emulators and Test Systems
- In-circuit test systems
- Impedance analyzers 4195A, 4291A, 4291B, 4395A, 4396A, 4396B, 4294A, E4990A, E4991A
- Impedance Analyzer Accessories
- Impedance / Gain Phase analyzer 4194A
- Impedance Meter 4193A,
- Isolators
- LCR meters? U1701A, U1701B, U1731A,? U1731B, U1731C, U1732A, U1732B, U1732C, U1733C, 4191A , 4192A, 4194A, 4195A, E4196A,? 4216A, 4260A, 4261A, 4262A? 4263A, 4263B, 4271B, 4274A, 4275A, 4276A , 4277A, 4284A, 4285A, 4286A, 4287A, 4291A, 4291B, 4294A, 4332A, 4342A, 4395A, 4396A, 4396B, E4980A and E4980AL
- LCR meter calibration devices? 16380A 42030A? 42090A, 42091A and 42100A
- LCR meter accessories
- 2-Terminal BNCs.
- 4-Terminal Pair (BNC connectors)
- Cable extension 16048A, 16048D, 16048E, 16048G, 16048H
- DC current bias accessories 42841A, 42842A, 42842B, 42842C, 42843A
- DC voltage bias accessories 16065A, 16065C,
- Kelvin clips 16089A, 16089B, 16089C,16089E
- Lead Components 16047A,16047B, 16047D, 16047E
- Material 16451B, 16452A
- Probes 42941A
- SMD 16034E, 16034G, 16034H
- 2-port 16096A
- 7 mm (APC7)
- 2-Terminal BNCs.
- LCZ meters? 4276A, 4277A,
- Lightwave clock / data receivers
- Lightwave converter
- Lightwave component analyzer
- Lightwave measurement system mainframes
- Lightwave polarization analyzers 8509B
- Logic analyzers
- Nemo wireless network solutions.
- Noise and interference test set
- Noise figure analyzers
- Noise sources 346A, 346B. 346C ,
- Matching pads (50 ohm to 75 ohm or similar)
- Materials test equipment
- Microwave repeaters
- Microwave downconverters 70427A
- Microwave / THz sources
- Milliammeter 428B
- Milliohm meter
- Mobile communications DC source
- Modular instruments
- AXIe
- Data acquisition (DAQ)
- USB
- PXIe
- Modulation analyzers
- Multimeters 427A, 970A
- Optical attenuators
- Optical heads
- Optical sources
- Optical spectrum analyzers
- Oscilloscopes 120A, 120AR, 120B, 122A, 130A, 130B, 130BR, 130C, 140A, 140B, 141A, 150A, 150AR, 160B, 180A, 180AR, 180CD, 181A, 181AR, 181T, 181TR, 182C, 182T, 183A, 183B, 184A, 184B, 185A, 185B, 1200A, 1200B, 1220A, 1221A, 1703A, 1707A, 1707B, 1710A, 1710B, 1715A, 1722A, 1725A, 1726A, 1740A, 1741A, 1742A, 1743A, 1744A, 1746A, 1980A, 1980B, 5403A, 6000A, 6000L, 16533A, 16534A, 54100A, 5410B, 54100C, 5100D, 54111D, 54120A, 54120B, 54200A, 54501A, 54502A, 54503A, 54504A, 54520A, 54520C, 54540A, 54540C, 54542A, 54542C, 54600B, 54601A, 54601B, 54602B, 54603B,? 54645A, 54654N, 54710A, 54720A, 54750A, 54825N, E1428,?
- Oven controlled crystal oscillators (OCXOs)
- Pattern generators
- PCM terminal test set
- Phase noise measurement
- Pico ammeters
- Printers 2225
- Plotters 7470A, 7475A?
- Probes
- Protocol analyzers and exercisers.
- Power booster test sets
- Power meters 431A, 431B, 431C, 432A, 435A, 435B, 437B, 438A
- Power splitters
- Power supplies
- Pulse generators
- Q-meters 4342A?
- Q-meter calibration inductors 16470A
- Reflection transmission test set
- Return loss module (optical)
- Relays / switches / switch matrices (optical)
- Relays / switches / switch matrices (RF)
- Resistor standards 42030A?and 42100A
- S-parameter test sets
- Scalar network analyzers
- SCSI bus preprocessor interface E2324A
- Selective level meters 3746A
- Semiconductors
- Semiconductor parameter analyzers 4145A, 4155B, 4156B,
- Signal analyzers
- Signal generators / sweep generators / signal sources / oscillators 200CD, 201B, 209A, 204D,? 608A,? 8165A
- Software
- Source measure units
- Spectrum analyzers 4195A,???
- Switch control units
- SWR meter 415E?
- Time interval? counters
- Time mark generator 226A
- Timing and data state modules
- Torque wrenches
- Transmitter testers
- Trigger modules
- Ultrasound transducers
- Universal bridge? 4260A, 4265A, 4265B?
- Vacuum tube voltmeter 410C
- Vector Impedance Meter 4193A, 4800A, 4815A
- Vector Network Analyzers (VNAs) 4195A,? 8510A, 8510B, 8510C, 8753A, 8753B, 8753C, 8753D, 8753E, 8753ES, 8752ET, 8719A, 8719B, 8719C, 8719D, 8720A, 8720B, 8720C, 8720D, 8720ES, 8722A, 8722B, 8722C, 8722D, 8722ES,
- Vector Network Analyzers (VNA) calibration kits 85032B, 85032E, 85033C, 85033D, 85033E, 85050B, 85050C, 85050D, 85052B, 85052C, 85052D, 85054A, 85054B, 85054D, 85056A
- Vector Network Analyzer (VNA) verification kits
- Vector Signal Analyzer 89650S, 89600S
- Vector voltmeters 8405A, 8508A,
- VXI mainframes 70000B, 70000C
- Waveform and function generators
- Waveguide to waveguide and waveguide to coaxial transitions.
- Wireless 58 OTA chambers
- Wireless channel emulators
- Wireless network emulators
- Wireless communication test sets
?
Re: Schematics for HP 8558B SA
Thanks for the quick replies. Today I did find a complete military version of the HP 8558B manual with schematics. I'll start off with this and begin work on the SA, first checking and securing the
By
Paul
·
#86145
·
|
Re: Frequency calibration?
Right there with ya..
-Dave
[hp_agilent_equipment] wrote:
--
Dave McGuire, AK4HZ
New Kensington, PA
By
Dave McGuire
·
#86144
·
|
some 89410A/89441A notes
Here are some random notes on the expansion options for the
89410A/89441A, for the sake of the archives.
Second Channel
--------------
It's possible to add the second input channel option
By
Dave McGuire
·
#86143
·
|
Re: HP AN/USM 281A Oscilliscope needs repair (San Antonio TX)
One last question for the night. Is there a component failure that would cause the beam/dot not to show on the screen when I apply low voltage to the test leads (1.5v battery) I have to hold the TEST
By
Andrew Maich <andremajic@...>
·
#86142
·
|
Re: HP AN/USM 281A Oscilliscope needs repair (San Antonio TX)
Cool. Thank you for the tip on where to check.
I'm still trying to figure out if there are any settings I need to try without cracking the case open. I'll save this email for reference when I open it
By
Andrew Maich <andremajic@...>
·
#86141
·
|
Re: HP AN/USM 281A Oscilliscope needs repair (San Antonio TX)
Thanks for the help. What can I use for a signal? I don't have a signal generator. I tried testing it with a 3v 800ma wall wart using the EXT INPUT and didn't get a dot or anything.
I tried using a
By
Andrew Maich <andremajic@...>
·
#86140
·
|
Re: Frequency calibration?
Unless there is a reason for extreme precision in your frequency measuring, you don't need to spend too much money adjusting the oscillator to a reasonable precision. The old method using WWV with a
By
Stuart Landau
·
#86139
·
|
Re: [TekScopes] Fast Rise/Fall Time Pulser
Mine arrived today. Alas, it uses a USB B connector instead of a more common USB mini or micro; I gotta get a cable.
Mine has the comparator IC marked only, "AJK BAA." Yes, this is an Analog Devices
By
Larry McDavid
·
#86138
·
|
Re: Frequency calibration?
What's wrong with my Rubidium reference that would make me choose a GPS disciplined? Costed me $75 on EPay, but still needs a power supply, case and diastribution amplifiers, yet another project to
By
Ovidiu Popa [Business]
·
#86137
·
|
Re: Frequency calibration?
I have the same equipment and don't need insane accuracy (1Hz in 100MHz is more the sufficient - 10e-8). I will put the 1PPS signal into the A input, the 10MHz from the ref osc out on the back into
By
WA2SCL_Jeff
·
#86136
·
|
Re: Frequency calibration?
They say coffee helps you live longer, I don’t know about how true that is but I do know that it definitely makes me enjoy the time I do have much more!
Peter
By
Peter Gottlieb
·
#86135
·
|
Re: Frequency calibration?
?????? There are many of us who are slaves to coffee. Have few other bad habits but I can't even talk in the morning before drinking a cup.
--
Richard Knoppow
dickburk@...
WB6KBL
By
Richard Knoppow
·
#86134
·
|
Re: Fast Rise/Fall Time Pulser
Yes, works as advertised as far as I can tell. There is plenty of
discussion and many screenshots
By
Orin Eman
·
#86133
·
|
Re: Frequency calibration?
Two questions for the original poster: How accurate do you need (want) to be? and what are you using the counter for?
Time and frequency precision can become an obsession - Time Nuttery.
Robert
By
Robert G8RPI
·
#86132
·
|
Fast Rise/Fall Time Pulser
There has been discussion recently about generating fast rise/fall time pulses by various means. Here is a link to a Leo Bodnar device that will provide <40 ps rise/fall time, 50% duty cycle, 10 MHz
By
Larry McDavid
·
#86131
·
|
Re: Frequency calibration?
Correct! Need more caffeinated beverage. Thanks!
[hp_agilent_equipment] <hp_agilent_equipment@...> wrote:
By
bownes
·
#86130
·
|
Re: Frequency calibration?
?????? On a standard scope with the same signal on X and Y in phase is a line at 45degrees to the right, 180degrees out of phase is a straight line tilted 45 degrees to the left, a circle is 90
By
Richard Knoppow
·
#86128
·
|
Re: VNA design engineer on the list?
Dr. Joel Dunsmore hangs out in the Forums on HP/Agilent/Keysight.
[hp_agilent_equipment] <hp_agilent_equipment@...> wrote:
By
bownes
·
#86129
·
|
VNA design engineer on the list?
If anyone here designed commercial VNAs, I'd like to call attention to some recent posts I've made to the EEVblog metrology forum on the topic of using an SA and a computer to measure amplitude and
By
Reginald Beardsley
·
#86127
·
|
Re: Frequency calibration?
By far and away the best choice is going to be a GPS disciplined oscillator
such as an HP 3801A or a Thunderbolt. A good GPSDO should be good to at
least 1x10^-10
? (1ppb) over the course of a day.
By
bownes
·
#86126
·
|