CHECK OUT THE WIKI The purpose of the wiki is mainly to allow you to find information on instruments, either from either
- The model number
- The function(s) listed below. Some instruments have multiple functions - for example, the 4195A is a VNA, spectrum analyzer and an impedance analyzer. Therefore the 4195A is listed in multiple categories
Please also check out HPWiki available here:
- Accessory kits - various types
- AC power analyzers - PA2201A and PA2203A
- AC power supplies 6811C, 6812C, 6813C
- Airlines
- Arbitrary waveform generators M8194A
- Amplifiers?493A, 495A?
- Attenuators (optical) 8156A, 8157A, 8158B, 81566A, 81576A,?
- Attenuators (RF) 8494A
- Attenuator set (500 Ω) 350C
- Attenuator set (600 Ω) 350D
- Attenuator switch driver
- Audio analyzers? 8903A, 8903B, 8903E,? ?
- Base station test sets
- Bit error rate testers (BERTs)
- Cables
- Capacitance meters U1701A, U1701B, 4272A, 4278A, 4279A
- Capacitor Bridge 4270A,
- Capacitor standards 16380A, 16380C,?
- Carrier noise test setsi
- Cesium frequency standards
- Clamp ammeters
- Close field probes
- Crystal Impedance E4915A, E4916A
- Data Acquisition Systems (DAQs)
- DC power analyzers
- DC power supplies 6030A , 6031A , 6032A, 6033A, 6035A, 6131C, 6621A, 6622A, 6623A, 6624A, 6627A, 6255A, 6645A, 6671A, 6672A, 6673A, 6674A, 6675A, 62003A, 62003C, 62003E, 62004A, 62004B, 62004E, 62005A, 62005B, 62005E, 62006A, 62006B, 62006E, 62010A, 62010C, 62010E, 62012A, 62012C, 62012E, 62015A, 62015C, 62015E, 62018A, 62018C, 62018E, 62024A, 62024C, 62024E, 62028A, 62028C, 62028E, 62048A, 62048C, 62048E
- Delay lines
- Detectors
- Device current waveform analyzers
- Digital communications analyzers
- Directional couplers
- Distortion analyzers 330B, 330C, 330D, 331A, 332A, 333A, 334A, 339A, 8903A, 8903B, 8903E,???
- Dynamic measurement DC source
- Electrometers
- Fading simulators
- Femto ammeters
- Filters
- Frequency counters 522B, 5342A 5343A 5352B
- Frequency standards?
- Function Generators ? 3310A,? 8165A,
- GPIB controllers, extenders, cables etc.
- GPS frequency standards
- Harmonic mixers
- High resistance meters 4339B
- High resistance meter fixtures 16008B
- HEV EV Grid Emulators and Test Systems
- In-circuit test systems
- Impedance analyzers 4195A, 4291A, 4291B, 4395A, 4396A, 4396B, 4294A, E4990A, E4991A
- Impedance Analyzer Accessories
- Impedance / Gain Phase analyzer 4194A
- Impedance Meter 4193A,
- Isolators
- LCR meters? U1701A, U1701B, U1731A,? U1731B, U1731C, U1732A, U1732B, U1732C, U1733C, 4191A , 4192A, 4194A, 4195A, E4196A,? 4216A, 4260A, 4261A, 4262A? 4263A, 4263B, 4271B, 4274A, 4275A, 4276A , 4277A, 4284A, 4285A, 4286A, 4287A, 4291A, 4291B, 4294A, 4332A, 4342A, 4395A, 4396A, 4396B, E4980A and E4980AL
- LCR meter calibration devices? 16380A 42030A? 42090A, 42091A and 42100A
- LCR meter accessories
- 2-Terminal BNCs.
- 4-Terminal Pair (BNC connectors)
- Cable extension 16048A, 16048D, 16048E, 16048G, 16048H
- DC current bias accessories 42841A, 42842A, 42842B, 42842C, 42843A
- DC voltage bias accessories 16065A, 16065C,
- Kelvin clips 16089A, 16089B, 16089C,16089E
- Lead Components 16047A,16047B, 16047D, 16047E
- Material 16451B, 16452A
- Probes 42941A
- SMD 16034E, 16034G, 16034H
- 2-port 16096A
- 7 mm (APC7)
- 2-Terminal BNCs.
- LCZ meters? 4276A, 4277A,
- Lightwave clock / data receivers
- Lightwave converter
- Lightwave component analyzer
- Lightwave measurement system mainframes
- Lightwave polarization analyzers 8509B
- Logic analyzers
- Nemo wireless network solutions.
- Noise and interference test set
- Noise figure analyzers
- Noise sources 346A, 346B. 346C ,
- Matching pads (50 ohm to 75 ohm or similar)
- Materials test equipment
- Microwave repeaters
- Microwave downconverters 70427A
- Microwave / THz sources
- Milliammeter 428B
- Milliohm meter
- Mobile communications DC source
- Modular instruments
- AXIe
- Data acquisition (DAQ)
- USB
- PXIe
- Modulation analyzers
- Multimeters 427A, 970A
- Optical attenuators
- Optical heads
- Optical sources
- Optical spectrum analyzers
- Oscilloscopes 120A, 120AR, 120B, 122A, 130A, 130B, 130BR, 130C, 140A, 140B, 141A, 150A, 150AR, 160B, 180A, 180AR, 180CD, 181A, 181AR, 181T, 181TR, 182C, 182T, 183A, 183B, 184A, 184B, 185A, 185B, 1200A, 1200B, 1220A, 1221A, 1703A, 1707A, 1707B, 1710A, 1710B, 1715A, 1722A, 1725A, 1726A, 1740A, 1741A, 1742A, 1743A, 1744A, 1746A, 1980A, 1980B, 5403A, 6000A, 6000L, 16533A, 16534A, 54100A, 5410B, 54100C, 5100D, 54111D, 54120A, 54120B, 54200A, 54501A, 54502A, 54503A, 54504A, 54520A, 54520C, 54540A, 54540C, 54542A, 54542C, 54600B, 54601A, 54601B, 54602B, 54603B,? 54645A, 54654N, 54710A, 54720A, 54750A, 54825N, E1428,?
- Oven controlled crystal oscillators (OCXOs)
- Pattern generators
- PCM terminal test set
- Phase noise measurement
- Pico ammeters
- Printers 2225
- Plotters 7470A, 7475A?
- Probes
- Protocol analyzers and exercisers.
- Power booster test sets
- Power meters 431A, 431B, 431C, 432A, 435A, 435B, 437B, 438A
- Power splitters
- Power supplies
- Pulse generators
- Q-meters 4342A?
- Q-meter calibration inductors 16470A
- Reflection transmission test set
- Return loss module (optical)
- Relays / switches / switch matrices (optical)
- Relays / switches / switch matrices (RF)
- Resistor standards 42030A?and 42100A
- S-parameter test sets
- Scalar network analyzers
- SCSI bus preprocessor interface E2324A
- Selective level meters 3746A
- Semiconductors
- Semiconductor parameter analyzers 4145A, 4155B, 4156B,
- Signal analyzers
- Signal generators / sweep generators / signal sources / oscillators 200CD, 201B, 209A, 204D,? 608A,? 8165A
- Software
- Source measure units
- Spectrum analyzers 4195A,???
- Switch control units
- SWR meter 415E?
- Time interval? counters
- Time mark generator 226A
- Timing and data state modules
- Torque wrenches
- Transmitter testers
- Trigger modules
- Ultrasound transducers
- Universal bridge? 4260A, 4265A, 4265B?
- Vacuum tube voltmeter 410C
- Vector Impedance Meter 4193A, 4800A, 4815A
- Vector Network Analyzers (VNAs) 4195A,? 8510A, 8510B, 8510C, 8753A, 8753B, 8753C, 8753D, 8753E, 8753ES, 8752ET, 8719A, 8719B, 8719C, 8719D, 8720A, 8720B, 8720C, 8720D, 8720ES, 8722A, 8722B, 8722C, 8722D, 8722ES,
- Vector Network Analyzers (VNA) calibration kits 85032B, 85032E, 85033C, 85033D, 85033E, 85050B, 85050C, 85050D, 85052B, 85052C, 85052D, 85054A, 85054B, 85054D, 85056A
- Vector Network Analyzer (VNA) verification kits
- Vector Signal Analyzer 89650S, 89600S
- Vector voltmeters 8405A, 8508A,
- VXI mainframes 70000B, 70000C
- Waveform and function generators
- Waveguide to waveguide and waveguide to coaxial transitions.
- Wireless 58 OTA chambers
- Wireless channel emulators
- Wireless network emulators
- Wireless communication test sets
?
Re: Looking for the manual for the HP H04-3460B or 3460B opt H04
The H04 manual for the 3460 *A* is just the normal manual with some change sheets for the H04 modifications, if there is a 3460 *B* H04 version you should be able to use the 3460 *B* manual at the web
By
factory
·
#151083
·
|
Re: Looking for the manual for the HP H04-3460B or 3460B opt H04
I have some info for the 3460 *A* -H04 but not the B
Looks like specs / service information.
It is part of a set of fiches that cover 3460A specials and is a little randomly organized and poorly
By
Peter Brown
·
#151082
·
|
Re: Repair HP3455A Multimeter
Ralf,
Unless I am very mucht mistaken (...) the 3325 uses the chassis as GND, and **all** screws fastening the PCB should better be nice and tight. The PCB has no overall GND itself. This can drive
By
Wilko Bulte
·
#151081
·
|
Looking for the manual for the HP H04-3460B or 3460B opt H04
I'm looking for the Service manual for the H04 version of the HP 3460B. H04 adds an additional digit to the display and has improved accuracy.
Richard
By
Richard Parrish
·
#151080
·
|
Re: HP6209B component location layout
The current through the 100k pot is normally 3.2mA.to change that to 50k
would require doubling the current to 6.4mA. This additional current will
show up as an offset on the current meter. The 6207
By
Hugh Gilbert
·
#151079
·
|
HP6209B component location layout
I recently acquired a HP6209B, and already have the same manual as the one here in the files section, but it does not include the component layout, as it seems that page was left out during the scan.
By
Alex
·
#151078
·
|
LCR Meter E4980A Power On Test Fail
Hi,
I am currently trying to repair an E4980A LCR Meter that is experiencing a power-on failure. When the unit is powered on, the self-test displays the following error message:
*"E1000: Power On
By
KC Chong
·
#151077
·
|
Repair HP3455A Multimeter
Hi,
my 3325A is driving me crazy and no one could help yet, so I went over to a faulty HP3455A multimeter. It worked many years, but now it comes up with a OL display in DC.
Cal test stops at
By
Ralf Rosche
·
#151076
·
|
Re: DIY: Repair of HP Step Attenuators
Here’s a YouTube video I found when I was repairing the attenuator in my 8902a. ?I followed his procedure and used the same watch repair o ring kit from Amazon (under $10 shipped).
By
pac1085
·
#151075
·
|
HP CRT "rough" equivalence info needed
I got a HP8754A recently that mostly works but seems low on available trace intensity. Maybe it just needs some internal adjustment, or maybe the CRT is weak on emission. Now that I have this one,
By
ed breya
·
#151074
·
|
Re: CAL TIMEBASE 8591A
I found the problem, I simply made a mistake in the procedure. Everything is
correct now.
Yves
By
Yves Tardif
·
#151073
·
|
Re: DIY: Repair of HP Step Attenuators
Hi Don I searched as you described but didn't find your repair procedure. What is the message number please? Thanks Jason NJ7K
By
fishtronics
·
#151072
·
|
Re: 89605 / 89606 cal file
It seems the cal files have one gain parameter starting at 50MHz in 50MHz steps. The smallest of the three files must’ve been intended for a system with an E2730 and the others for systems with
By
Ed Marciniak
·
#151071
·
|
CAL TIMEBASE 8591A
Hi,
I have two 8591As to check, and for both, I'd like to correct the timebase.
I followed the procedure in the service manual; it worked, and I correctly
executed the CAL STORE command when
By
Yves Tardif
·
#151070
·
|
Re: DIY: Repair of HP Step Attenuators
Thanks
Just ordered 50 ? orings.? I have 2 Attenuators to repair.
By
ROBERT LEVY
·
#151069
·
|
Re: DIY: Repair of HP Step Attenuators
How many of what size/type would be needed to replace all the ones used in
an 8566B stack? I'd be interested in buying a set to refurbish my spec an
when it becomes
By
Dave Daniel
·
#151068
·
|
Re: DIY: Repair of HP Step Attenuators
Is anyone going to be buying 0-rings now? I'd like to do a group order since I now have an old, mostly working 8566B SA.
These can be mailed around in letters for $.73...
By
John Griessen
·
#151067
·
|
HP 435B, Zero Carryover Test - How noisy should it be?
Hey All,
I noticed what looked to be a zero issue with my 435B, so I thought I'd run it through the zero carryover test (OSM 4-6) to see what was going on.
While the meter meets all the ranges given
By
Tony Goodhew
·
#151066
·
|
Re: DIY: Repair of HP Step Attenuators
I used 0.5mm CS X 1.0mm ID 70 durometer from theoringstore.com on an attenuator from my HP
By
Sam Reaves
·
#151065
·
|
Re: DIY: Repair of HP Step Attenuators
I have done over a hundred repairs of the 3332xA/B style attenuators. In the file listing in this site you will find a detailed step by step procedure with HPAK part nos. and Apple Rubber part nos.
By
Don Bitters
·
#151064
·
|