开云体育

ctrl + shift + ? for shortcuts
© 2025 开云体育
Re: HP 5245L counting/gate problems
OK Scott, it has never been recapped as far as I know. I’ll look into how/where the supply voltages are tested and will probably just go ahead and recap to avoid any future supply
By Tom Smith · #150382 ·
Re: HP 5245L counting/gate problems
Hi Pete, Thanks for the reply. The neon bulb is working on the start and stop test function. When placed on the start position, the bulb comes on but the counter doesn’t display any counting. All
By Tom Smith · #150381 ·
Re: HP 5245L counting/gate problems
I just resurrected one. If the power supply hasn't been re-capped, check the supply voltages first.? In my case every d*** capacitor was bad.? Check back then.
By Scott Townley · #150380 ·
Re: HP 5245L counting/gate problems
Just an idea: In the case of the gate indicator (a neon bulb): try shining a bright light on it to see if that "fixes" it. As for "not counting", I don't know an alternative other than to probe
By saipan59 (Pete) · #150379 ·
Re: HP 3403C True RMS Voltmeter display fault
There are days when Murphy's Law just seems to run in overdrive mode, but the fun will resume in a few day and I hope to have more news soon ! Joel
By Joel Setton · #150378 ·
Re: HP 3403C True RMS Voltmeter display fault
Well, It did not go well lol. This is actually not that badly packed. It is somewhat dense but this is really well designed. The boards can be freed from their metal case easily, and A22 (AD board) is
By AArnaud · #150377 ·
HP 5245L counting/gate problems
I have several of the HP 5245L counters and until recently, had at least one working. Now it seems that the last “working” counter I own has gone belly up. Two of them have the same problem, (the
By Tom Smith · #150376 ·
Re: 54003A success
Thank you for that detail! Jeremy [email protected]> wrote:
By Jeremy Nichols · #150375 ·
Re: 54003A success
Indeed, but with a 1000:1 probe that becomes a +/- 2000 volt input range which is sufficient. That was a work requirement and their solution covers more, although this would be sufficient for them
By Peter Gottlieb · #150374 ·
Re: 54003A success
Peter, didn’t you have some high-voltage requirements for your application? Looks like the 54003A is limited to +/- 2 Volts, after whatever reduction the probe provides. (Just curious. I missed the
By Jeremy Nichols · #150373 ·
54003A success
Thank to those who provided information I was able to successfully use 54003A input modules as 1 meg probe converter/extenders. For my company the current COTS device was used to allow remotely using
By Peter Gottlieb · #150372 ·
Re: Test Set Program Development Tool DEV_PL on 3.5" floppy
Not quite.? A USB solution was addressed earlier which is the defacto standard nowadays for communicating with old floppy deives under a modern OS.
By Matt Harris · #150371 ·
Re: Stuck APC-7 connector
Thanks, guys. Yes, what I really need to do above all is avoid over-torquing/over-stressing the connector at the point where it goes into the bulkhead. If I can get them apart without doing that, I
By Jinxie · #150370 ·
Re: Keysight U1242B Schematic
Hi Ozan. Thank you for that tip. Unfortunately the jack tests open. But the resistance on the jack contacts on the board reads 6.02 Mohm. So the fault is somewhere on the board. I have found other
By Mark Creighton · #150369 ·
Re: Service manual scan post processing
Which Minolta do you have? I am in the process of trying to get a 2400DSV-E to talk to a PC myself
By Peter Brown · #150368 ·
Re: Stuck APC-7 connector
Can you determine which end has the threads out - the N adapter or the port? Then using a padded jaw pliers and the hex wrench, turn the nut that doesn’t have the threads exposed. I would guess the
By Don Bitters · #150367 ·
Re: HP 3403C True RMS Voltmeter display fault
My guess is that the clock source must be OK, otherwise the blinking function would not work or the display would blink at a "wrong" rate. I think I would probe all pins around both the LD110 and
By Joel Setton · #150366 ·
Re: Stuck APC-7 connector
Put a few layers of masking tape around the knurled nut closest the panel and apply clockwise torque gently with channel-lock pliers. If it doesn't come loose easily, put a wrench on the flats of the
By ed breya · #150365 ·
Re: VNA Reference Impedance Renormalization
Good point. I use a DC block permanently installed at the RF input of any DC-sensitive piece of test equipment. It only takes a couple of hundred millivolts of DC to trash the front end of an
By Jinxie · #150364 ·
Re: VNA Reference Impedance Renormalization
Cheers Jim. Look forward to your findings. I've done similar things in the past, some even work with a Nano VNA, within reason... The secret I found in the immdiate past work life, is to
By Dave_G0WBX · #150363 ·